[BTeam] Elevated chronic beam loss in NL
Jay Benesch
benesch at jlab.org
Tue Oct 19 07:26:01 EDT 2021
https://logbooks.jlab.org/entry/3927857 by Pavel
The level of the chronic beam loss in NL has become much higher (a
factor of ~3-5) around 8 pm as measured by the NDX detectors. Neutron
readout is above the readout saturation in NL23. The photon channels are
not saturated, showing the dose rates about 1 kR/h, and up to ~2kR/h in
spikes. I will decrease the integration time to 0.5 sec to allow for
higher dynamic range.
We should discuss this at B team.
If lem was a significant driver, efforts to understand what went wrong
with the beam-based gradient calibration data taken in August should be
elevated and a beam studies of just a few zones scheduled to see if the
problem repeats. https://opsweb.acc.jlab.org/CSUEApps/atlis/task/22480
The average scaling factor found in August was ~0.8 while the lem fudge
factor is 1.012 aka the gradient-weighted scaling factor should be
~1.01. Chris and Adam have reanalyzed the data taken in August and it
looks clean. We're at a loss. New eyes needed.
In addition, thought should be given to changing the NL optics so as to
increase beta function on lower passes and decrease it on the higher
passes.
What happened overnight:
https://logbooks.jlab.org/entry/3927791
https://logbooks.jlab.org/entry/3927845
https://logbooks.jlab.org/entry/3927849
https://logbooks.jlab.org/entry/3927946
https://logbooks.jlab.org/entry/3927975
Brian did a little steering 4R into 5L around the time things quieted
down in the NDXs.
More information about the BTeam
mailing list