From: Anuruddha Rathnayake <adr4zs at virginia.edu>
Sent: Monday, December 21, 2020 9:18 AM
To: Chuck Long <mlong at jlab.org>
Cc: kagnanvo <kagnanvo at jlab.org>; Liyanage, Nilanga K (nl8n) <nl8n at virginia.edu>; Marc Mcmullen <mcmullen at jlab.org>
Subject: Re: [EXTERNAL] Re: Notification for connection of gas supply on UVa GEMs in Bigbite in TEDF
Hi Chuck,
Sounds good. I'll be there too but I'll be a little late like 10, 15 mins I think.
Best
Anu
On Mon, Dec 21, 2020 at 9:02 AM Chuck Long <mlong at jlab.org<mailto:mlong at jlab.org>> wrote:
Hi Anu,
As we determined Friday there is a leak in the gas panel due to a bad quick connect fitting - tubing was leaking, then popped out while performing a pull test, would not stay in under pressure. Seems to be the fitting is not grabbing onto the wall of the tubing.
I will replace the fitting this morning and we can proceed with running the lines to the chambers. I will be over there around 10am.
-Chuck
________________________________
From: Anuruddha Rathnayake <adr4zs at virginia.edu<mailto:adr4zs at virginia.edu>>
Sent: Friday, December 18, 2020 4:19 PM
To: kagnanvo <kagnanvo at jlab.org<mailto:kagnanvo at jlab.org>>
Cc: Chuck Long <mlong at jlab.org<mailto:mlong at jlab.org>>; Liyanage, Nilanga K (nl8n) <nl8n at virginia.edu<mailto:nl8n at virginia.edu>>; Marc Mcmullen <mcmullen at jlab.org<mailto:mcmullen at jlab.org>>
Subject: Re: [EXTERNAL] Re: Notification for connection of gas supply on UVa GEMs in Bigbite in TEDF
Hi Kondo,
Okay yeah this calculation makes sense to me. So I think for the 5 units flow rate, it's equivalent to (300 cc/min)/4 = 75 cc/min which is the flow that we have been flushing our modules on Nitrogen when we are keeping them idle.
We will try and set the flow rate to that value.
However, we will not be able to start flushing the modules in the TEDF with Nitrogen today as it seems like for a 4 x 100 cc/min flow rate, the gas bottle is being consumed way more quickly than it's supposed to. Chuck said he will have to troubleshoot about what's going on and we didn't want to hook up the modules in that situation. We'll give you an update on Monday.
Best
Anu
On Fri, Dec 18, 2020 at 3:07 PM kagnanvo <kagnanvo at jlab.org<mailto:kagnanvo at jlab.org>> wrote:
Hi Anu,
That's a good question. I remember we went through this many years ago but I don't remember the details
What I know is that when you are at 20 units, it is equivalent to 5 volume change / hour of our GEM chambers
The volume of the GEM is rougly 5 x 6 x 0.12 l = 3.6 l
so this will give you 5 x 3600 cc / 60 mm = 300 cc / min
I hope I did not make any mistake in my calculations. This should be pretty close to the actual numbers
Best regards
Kondo
On 12/18/2020 2:17 PM, Anuruddha Rathnayake wrote:
Hi Kondo,
Sounds good. This is something I should know but do you know what's the unit of flow meters we use in our clean room. It doesn't show any units and I always used numbers printed on them to say the flow rate.
The ones we have here in the TEDF are marked in cubic centimeters(cc)/min.
If you can jist give a value that you would like to flow in cc/min, I guess that would be fine too.
Best
Anu
On Fri, 18 Dec 2020, 12:59 kagnanvo, <kagnanvo at jlab.org<mailto:kagnanvo at jlab.org>> wrote:
Hi Anu,
Thanks for taking care of it.
Before starting flushing, if you are doing it today, please be very careful with the flow rate. Again we are not really familiar with the new system. But of course you have a lot of experience with this and also Chuck will be there to help
Best regards
Kondo
On 12/18/2020 12:43 PM, Anuruddha Rathnayake wrote:
Hi Kondo,
Sounds good!
Chuck,
Please let me know a good time for you (either Monday or even today afternoon) and I can work with you to hook the GEMs up to the gas supply and start flushing them.
Best
Anu
On Fri, Dec 18, 2020 at 12:31 PM kagnanvo <kagnanvo at jlab.org<mailto:kagnanvo at jlab.org>> wrote:
Hi Anu,
We have been given the green light to connect the gas supply to our BB GEM layer in TEDF. So it will be great if you could see with Chuck to have the system hooked up by Monday afternoon, so that we could get support from Marc to run the monitoring software before the Xmas break
Per Marc indication, we should hooked our chambers to the flow-meter channels #1 to #4 with the system installed in the rack. At the beginning we will have to operate at very low flow rate to make sure that we really understand the flow going into the chamber.
I plan to be at JLab on Monday or Tuesday depending on the status of the Alu frames here at UVa but go ahead anyway with the gas connection
Best regards
Kondo
-------- Forwarded Message --------
Subject: Re: Notification for connection of gas supply on UVa GEMs in Bigbite in TEDF
Date: Fri, 18 Dec 2020 10:43:20 -0500
From: Jack Segal <segal at jlab.org><mailto:segal at jlab.org>
To: kagnanvo <kagnanvo at jlab.org><mailto:kagnanvo at jlab.org>, Marc Mcmullen <mcmullen at jlab.org><mailto:mcmullen at jlab.org>, dsg-halla_gas at jlab.org<mailto:dsg-halla_gas at jlab.org> <dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org>
CC: nilanga at virginia.edu<mailto:nilanga at virginia.edu> <nilanga at virginia.edu><mailto:nilanga at virginia.edu>, Chuck Long <mlong at jlab.org><mailto:mlong at jlab.org>
You should be good to go.
On 12/18/2020 10:36 AM, kagnanvo wrote:
Dear Jack,
The Marc Mullen from DSG group has just set up the SBS gas system prototype for us in the TEDF highbay area to try it out on the UVa GEM layer we just recently installed in Bigbite frame.
We would like to notify to that we are now ready to connect the gas supply to the GEM chambers and seek your approval to proceed.
We are planning to use a gas cylinder and regulator system identical to the one we have installed for the commissioning of the UVa GEM layers in EEL124 clean room and set up for us by Chuck. This system will be hooked up to the monitoring system installed by Marc to supply the BB GEM chambers.
Please let us know if there is any step we need to take before going ahead with connecting the system to our chambers. Ideally we would like to have N2 flowing into the chambers before the xmas break
Best regards
Kondo
On 12/16/2020 5:05 PM, Marc Mcmullen wrote:
Hi Kondo,
I have received clearance from DSG and Physics management to travel to Jlab to install the Raspberry Pi. I plan to make the trip tomorrow or Friday and will work alone per Jlab COVID procedures. I will send an update on the progress as it becomes available.
Regards,
Marc
---------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
*From:* Marc Mcmullen <mcmullen at jlab.org><mailto:mcmullen at jlab.org>
*Sent:* Wednesday, December 16, 2020 10:26 AM
*To:* kagnanvo <kagnanvo at jlab.org><mailto:kagnanvo at jlab.org>; Gnanvo, Kondo (kg6cq) <kg6cq at virginia.edu><mailto:kg6cq at virginia.edu>; dsg-halla_gas at jlab.org<mailto:dsg-halla_gas at jlab.org> <dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org>
*Cc:* nilanga at virginia.edu<mailto:nilanga at virginia.edu> <nilanga at virginia.edu><mailto:nilanga at virginia.edu>; Chuck Long <mlong at jlab.org><mailto:mlong at jlab.org>
*Subject:* Re: [Dsg-halla_gas] Prototype Gas Distribution System
Hi Kondo,
I will coordinate my trip to the lab to install the Raspberry Pi and start the software with the DSG manager as well as Physics division management.
I will update you on the progress.
Regards,
Marc
---------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
*From:* kagnanvo <kagnanvo at jlab.org><mailto:kagnanvo at jlab.org>
*Sent:* Tuesday, December 15, 2020 9:04 PM
*To:* Marc Mcmullen <mcmullen at jlab.org><mailto:mcmullen at jlab.org>; Gnanvo, Kondo (kg6cq) <kg6cq at virginia.edu><mailto:kg6cq at virginia.edu>; dsg-halla_gas at jlab.org<mailto:dsg-halla_gas at jlab.org> <dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org>
*Cc:* nilanga at virginia.edu<mailto:nilanga at virginia.edu> <nilanga at virginia.edu><mailto:nilanga at virginia.edu>; Chuck Long <mlong at jlab.org><mailto:mlong at jlab.org>
*Subject:* Re: [Dsg-halla_gas] Prototype Gas Distribution System
Hi Marc,
With Chuck help today, we moved the rack in the TEDF building where the Bigbite frame is located (where our GEM layer is installed). I don't really know the room number of the area but maybe Chuck could help me out here
Chuck also has prepared the gas cylinder to be used for the test and I believe everything is ready to go. So when you are available, we can go ahead with the installation.
I am not at JLab tomorrow so if you plan to work tomorrow, Chuck could help also. if not, I will be back at JLab on Thursday and we can plan accordingly
Please let me know
Best regards
Kondo
On 12/14/2020 12:09 PM, Marc Mcmullen wrote:
Kondo,
You are correct, the distribution rack is located in EEL 125, near Chuck's gas bottles. They do need to be moved to the test location in the TEDF.
Please let me know, where that location is in the TEDF (room number).
As far as calibrating, I believe you just want to connect your old flow meters in-line to check if the flow values match. You may need to get with Jack (system owner) on that for those kind of questions.
let me know if you have any questions or when the distribution panel is moved and I'll make arrangements for the software start up.
Marc
---------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------------
*From:* kagnanvo <kagnanvo at jlab.org><mailto:kagnanvo at jlab.org> <mailto:kagnanvo at jlab.org><mailto:kagnanvo at jlab.org>
*Sent:* Monday, December 14, 2020 12:01 PM
*To:* Marc Mcmullen <mcmullen at jlab.org><mailto:mcmullen at jlab.org> <mailto:mcmullen at jlab.org><mailto:mcmullen at jlab.org>; Gnanvo, Kondo (kg6cq) <kg6cq at virginia.edu><mailto:kg6cq at virginia.edu> <mailto:kg6cq at virginia.edu><mailto:kg6cq at virginia.edu>; dsg-halla_gas at jlab.org<mailto:dsg-halla_gas at jlab.org> <mailto:dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org> <dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org> <mailto:dsg-halla_gas at jlab.org><mailto:dsg-halla_gas at jlab.org>
*Cc:* nilanga at virginia.edu<mailto:nilanga at virginia.edu> <mailto:nilanga at virginia.edu><mailto:nilanga at virginia.edu> <nilanga at virginia.edu><mailto:nilanga at virginia.edu> <mailto:nilanga at virginia.edu><mailto:nilanga at virginia.edu>
*Subject:* Re: [Dsg-halla_gas] Prototype Gas Distribution System
Hi again Marc,
Just to clarify a few things once again. So right the rack is in EEL125 and you want us to move it to TEDF right?
I can see that with Chuck tomorrow then
Then after that I will have to get in touch with you for setting the raspberry pi and the associate software.
We will have to also set up a mechanism to calibrate the gas flow meter that you provide with the one we were using in EEL124 clean room.
Can we install temporarily one of our flow meter on the panel to perform this calibration?
Bets regards
Kondo
On 12/14/2020 10:44 AM, Marc Mcmullen wrote:
-->
--
John(Jack) Segal
Halls A and C Spectrometer Support Manager
Thomas Jefferson National Accelerator Facility
12000 Jefferson Avenue
Suite 12
Newport News, VA 23606-4486
Phone: (757) 269-7242
Cell: (757) 320-9977
-------------- next part --------------