[Dsg-dc] DC Mix 2 CO2 Flow
Brian Eng
beng at jlab.org
Sat Jul 28 19:41:26 EDT 2018
Since the tank pressure has dropped enough I've turned back on automatic mode.
The high limit for mix 2 has been lowered to 90 psi until the regulators are adjusted.
----- Original Message -----
> From: "Brian Eng" <beng at jlab.org>
> To: "dsg-dc" <dsg-dc at jlab.org>, "Hallb-eng" <hallb-engineering at jlab.org>
> Sent: Saturday, July 28, 2018 2:27:15 PM
> Subject: [Dsg-dc] DC Mix 2 CO2 Flow
> The regulator for the CO2 used for Mix 2 needs to be adjusted.
>
> The MFC valve was at 100% but not flowing anything, which is almost always due
> to the upstream pressure not being high enough. In the attached plot the scale
> for the MFC is -2 to 2 SLM.
>
> I put the mixing system in manual mode and set the flows for mix 2 to 0 so the
> tank won't be filled with just Argon.
>
> Once the regulator is adjusted or the pressure in the tank gets below ~95 it can
> be put back into automatic mode.
More information about the Dsg-dc
mailing list