    Hi Dave,<br>
    <br>
    Here is some info for the DC Gas System. These relief valves are
    critical path for operation of the gas mixing system for the drift
    chambers. Do you have a time estimate for this? We cannot test the
    drift chambers until we can mix gas.<br>
    <br>
    Thanks,<br>
    George<br>
    <div class="moz-forward-container"><br>
      <br>
      -------- Forwarded Message --------
      <table class="moz-email-headers-table" border="0" cellpadding="0"
        cellspacing="0">
        <tbody>
          <tr>
            <th align="RIGHT" nowrap="nowrap" valign="BASELINE">Subject:
            </th>
            <td>[Dsg-dc] ASME relief valves for the DC gas ASME storage
              tanks</td>
          </tr>
          <tr>
            <th align="RIGHT" nowrap="nowrap" valign="BASELINE">Date: </th>
            <td>Fri, 11 Nov 2016 14:41:59 -0500</td>
          </tr>
          <tr>
            <th align="RIGHT" nowrap="nowrap" valign="BASELINE">From: </th>
            <td>George Jacobs <a class="moz-txt-link-rfc2396E" href="mailto:jacobsg@jlab.org"><jacobsg@jlab.org></a></td>
          </tr>
          <tr>
            <th align="RIGHT" nowrap="nowrap" valign="BASELINE">To: </th>
            <td>Dave Meekins <a class="moz-txt-link-rfc2396E" href="mailto:meekins@jlab.org"><meekins@jlab.org></a></td>
          </tr>
          <tr>
            <th align="RIGHT" nowrap="nowrap" valign="BASELINE">CC: </th>
            <td><a class="moz-txt-link-abbreviated" href="mailto:dsg-dc@jlab.org">dsg-dc@jlab.org</a></td>
          </tr>
        </tbody>
      </table>
      <br>
      <br>
      <meta http-equiv="content-type" content="text/html; charset=utf-8">
      Hi Dave,<br>
      <br>
      Please find attached system flow diagram for the DC gas mixing
      system. Gas mixing supply pressure at the argon and CO2 pressure
      regulators is 200 psi. Gas supply pressure to the inlet of the
      argon and CO2 MFCs is 140 psi. Mixed gas pressure in the storage
      buffer tanks is maintained 80-100 psi by increasing or decreasing
      the gas mixing MFC set points. A pic of the tanks is attached. The
      lower pair of 5000gal tanks only have 3 connections, 2 on top and
      a drain at the bottom. Both of the upper connections are in use,
      so we will need to modify that connection for the relief
      connection. The upper pair of 2000 gal tanks has an extra
      connection on top that can be used for the relief. In all cases,
      green epoxy was used on the connections. I can show you these
      tanks and connections if you want.<br>
      <br>
      Max gas flow rates to the tanks.<br>
      With the MFC valve at 100% open max flow is 250 slm N2 equivalent,
      using the gas correction factors for argon and CO2, we have the
      following maximum flow rates;<br>
      For argon 1.39 x 250 = 347.5 slm<br>
      For CO2 0.7 X 250 = 175 slm<br>
      total of 522.5 slm per pair ( 2 independent mixing systems)<br>
      <br>
      The MFC failure mode for an inlet overpressure is that the valve
      cannot open against the high pressure.<br>
      The MFC failure mode for an excessive differential pressure across
      it is that the valve cannot open due to the excessive differential
      pressure.<br>
      <br>
      Detailed specifications can be found here,    <a
        moz-do-not-send="true" class="moz-txt-link-freetext"
        href="https://www.mksinst.com/product/Product.aspx?ProductID=1458">https://www.mksinst.com/product/Product.aspx?ProductID=1458</a><br>
      <br>
      The mixing buffers do not require relief valves. These are 250cc
      volume HP filter housings, not ASME, listing 3,000 max pressure on
      the body. I can show you one, I have a couple spares in the gas
      shed and I also use  them in the EEL for gas mixing setups. <br>
      <br>
      Thanks,<br>
      George<br>
      <br>
      <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a moz-do-not-send="true" class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a>

(website) <a moz-do-not-send="true" class="moz-txt-link-freetext" href="https://userweb.jlab.org/%7Ejacobsg">https://userweb.jlab.org/~jacobsg</a></pre>
    </div>
  </body>
</html>