[Dsg-ft] Chiller digital module test
    Raffaella De Vita 
    Raffaella.Devita at ge.infn.it
       
    Wed Nov  1 06:24:18 EDT 2017
    
    
  
Hi Amanda,
thanks for the update. Please keep us informed of progresses. For now we 
will tell Sergey and Ben to wait until the interlock work is complete.
Best regards,
     Raffaella
Amanda Hoebel wrote:
> Hi Raffaella,
>
> Pablo, Pete, and I have been working on the interlocks today. We have connected the card to the interlocks and are testing the readback.
> We have recently noticed the calorimeter temperature sensors are not working. After we installed the chiller card to the interlocks we checked the signals and they were working fine. We will be working on the interlocks for the next few days as we investigate the temperature sensor problem and connect the condensation sensor cable to the interlocks cRIO.
>
> Regards,
> Amanda
>
> ----- Original Message -----
> From: "raffaella devita" <Raffaella.Devita at ge.infn.it>
> To: "Amanda Hoebel" <amandah at jlab.org>
> Sent: Tuesday, October 31, 2017 9:59:12 AM
> Subject: Re: [Dsg-ft] Chiller digital module test
>
> Hi Amanda,
> I wanted to ask you when do you plan to connect the card to the
> interlocks: Sergey Boiarinov and Ben Raydo would like to switch on the
> Forward Tagger during the next days to do DAQ and trigger tests and I
> wanted to make sure this would not happen while you will be testing the
> interlocks.
> Thanks,
>       Raffaella
>
>
> Raffaella De Vita wrote:
>> Hi Amanda,
>> this are very good news! I'm really glad this is working as expected.
>> Thanks for taking care of this.
>> Best regards,
>>      Raffaella
>>
>> Amanda Hoebel wrote:
>>> Hi Raffaella,
>>>
>>> Pablo and I installed the digital module in the chiller. The module turned the pump on and off with no heating problem.
>>>
>>> Attached is a picture of the controller screen. The first peak is when the chiller was turned off for the module installation. The following three peaks are three tests Pablo and I did where we put the chiller into standby mode through the digital module. The new card is not connected to the hardware interlocks yet but we will let you know when it is.
>>>
>>> Regards,
>>> Amanda
>>>
>>>
>>> _______________________________________________
>>> Dsg-ft mailing list
>>> Dsg-ft at jlab.org
>>> https://mailman.jlab.org/mailman/listinfo/dsg-ft
> _______________________________________________
> Dsg-ft mailing list
> Dsg-ft at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-ft
    
    
More information about the Dsg-ft
mailing list