[dsg-hallb_gas] AHDC gas system
Marc Mcmullen
mcmullen at jlab.org
Mon Feb 3 10:44:38 EST 2025
Hi Gabriel,
I was able to log in to BTECH03 remotely. But it was a lot of permission changes on the computer to allow this. I believe Denny owns that computer and may be able to help with determining permissions and contacting the computer center if need be. He is the Hall B Work Coordinator so he may be able to help with that.
Again, I have put a controls GUI at this location: O:\hallb_eng_open\ALERT GUI, Hall B should be able to help with that as well.
Marc
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Monday, February 3, 2025 10:28 AM
To: Marc Mcmullen <mcmullen at jlab.org>; Nathan Baltzell <baltzell at jlab.org>
Subject: [EXTERNAL] [ALERT] AHDC gas system
Hi Marc and Nathan,
I would like to control the AHDC gas system. Right now, it is possible to monitor it via EPICS (thank you Nathan!) but not to control it.
A control via EPICS is not necessary but I know a computer in the counting room can control the gas system, right Marc?
Does any of you, know how I can connect to it and start the monitoring app?
Have a nice day,
--
Gabriel
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
À: "Marc Mcmullen" <mcmullen at jlab.org>
Envoyé: Vendredi 31 Janvier 2025 16:25:22
Objet: Re: Purge the gas line
Hi Marc,
Do you know if I know have remote access to the computer from which I can control the gas flow?
Have a nice day,
--
Gabriel
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
Cc: "Bob Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>, "Morgan Cook, IV" <mcookiv at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Jeudi 19 Décembre 2024 16:37:15
Objet: Re: Purge the gas line
Hi Gabriel,
I put in a request to the computer center to look at BTECH03, and passed the status over to Nathan with hall b. They are in the process of figuring out the computer issue.
Nathan has made a monitoring screen for EPICS that you can use to see how the system is doing. He can probably help you with that.
Regards,
Marc
Get Outlook for iOS<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_o0ukef&d=DwMFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=hQJSuIbzwxGOX7LZbVne0gHeC6U5hlDu99LTH_I02ZM&m=nX8aCyYM5jysCeIT408JUpMG-cWb5opn5tszrL-B7LyAAHMNQIC5sMf6j0AySmO1&s=OjgwD-RziXWnM1WiDtwDzL-nQBRAnv_GnsM553Jf084&e=>
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Wednesday, December 18, 2024 5:30:50 AM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; Morgan Cook IV <mcookiv at jlab.org>; Marc Mcmullen via dsg-hallb_gas <dsg-hallb_gas at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hello Marc,
Did you have time to see how I can get access to BTECH03 and start the GUI?
Have a nice day,
--
Gabriel
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
Cc: "Bob Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>, "Morgan Cook, IV" <mcookiv at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Vendredi 6 Décembre 2024 16:13:14
Objet: Re: Purge the gas line
Hi Gabriel,
I need to check with the computer center, there may be permissions issues with logging into BTECH03. They should be able to allow permission, so we will see.
>From the other aspect, I still need to send over a GUI to PV "map" to Nathan. I'll see what I can do to get that done by sometime next week.
I'll send an update when I have any news.
Marc
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Friday, December 6, 2024 9:52 AM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; Morgan Cook IV <mcookiv at jlab.org>; Marc Mcmullen via dsg-hallb_gas <dsg-hallb_gas at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hello,
Did you have time to get the GUI ready for me to be able to check it remotely?
If yes, we could try connect and start everything early next week. I just don't want to start the gas system without be able to check it from afar.
Have a nice day,
--
Gabriel
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
À: "Marc Mcmullen" <mcmullen at jlab.org>
Cc: "Bob Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>, "Morgan Cook, IV" <mcookiv at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Jeudi 28 Novembre 2024 15:58:38
Objet: Re: Purge the gas line
Hi Marc,
It is perfect if EPICs can read the numbers, display and save them as needed.
It is also perfect if EPIC does not allow the control of gas systems.
Then, I'll just to learn how to connect to BTECH03 and start the GUI on it.
Have a nice day,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>, "Bob Miller" <rmiller at jlab.org>
Cc: "Denny Insley" <dinsley at jlab.org>, "Morgan Cook, IV" <mcookiv at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Mercredi 27 Novembre 2024 19:19:30
Objet: Re: Purge the gas line
Hi Gabriel,
Brian and I are looking into what we feel is the easiest way to approach this. Hall B has a windows machine (BTECH03) with LabVIEW already installed, and it is on the correct subnet.
What this means is that we can put a GUI on the computer and you can just remote login using your CUE account and use the GUI to control the ALERT DC gas and the Target Straw purge screens.
So, we have EPICs PVs which our software updates already in the system. But this may be as simple as using remote desktop.
Regards,
Marc
Get Outlook for iOS<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_o0ukef&d=DwMFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=hQJSuIbzwxGOX7LZbVne0gHeC6U5hlDu99LTH_I02ZM&m=l5NmyUsUb6B6a6DvWJbrRPggGyy9U6d0CV7ftF59vKyoYibSp2Ml4Zofr00dUAl1&s=V6rYOh2qnXtPVA4bCRTbs_58F9Scd_KwVvHcOpE_2Ls&e=>
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Tuesday, November 26, 2024 4:04:55 PM
To: Bob Miller <rmiller at jlab.org>
Cc: Denny Insley <dinsley at jlab.org>; Morgan Cook IV <mcookiv at jlab.org>; Marc Mcmullen <mcmullen at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hi Bob,
Thank you for your email and all your efforts.
I'm glad things are moving on smoothly so far.
Next week I'm busy teaching at a gaseous detector school. So if we can plane to complete all connexions to the detector on Monday 9th, it would be perfect.
Marc, are you in touch with Nathan to allow remote access to your system? Or at least, readout the main values of it? If everything on this side could also be ready Monday 9, it would be even perfect-er.
Have a nice day,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: Bob <rmiller at jlab.org>
à: Gabriel <gabriel.charles at ijclab.in2p3.fr>
Cc: Denny <dinsley at jlab.org>; Morgan <mcookiv at jlab.org>; Marc <mcmullen at jlab.org>
Envoyé: mardi 26 novembre 2024 19:20 CET
Sujet : Re: Purge the gas line
Hi Gabriel,
The ALERT gas system is fully functional in local control.
All relief valves have been tested and all leak checks are complete.
The detector supply line to the hall is set at 15 psi with all valves closed. The target supply line to the hall is set at 20 psi with all valves closed.
The detector is open to atmosphere through the supply and return lines and are not currently connected to the gas system.
The target straw is under pressure with a separate N2 supply regulated to 5 psi. The target is not currently connected to the gas system.
Please notify Morgan when you are ready to connect the detector and target to the gas system.
Thanks,
Bob Miller
Hall B Lead Engineer
(757) 822-9586
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Sunday, November 24, 2024 3:35 PM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; Marc Mcmullen via dsg-hallb_gas <dsg-hallb_gas at jlab.org>; Patrizia Rossi <rossi at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hi Marc,
Thank you for your email, work and detailled tests.
What I really appreciate is that despite the complexity of the gas system, you managed to keep the interface simple.
We will get in touch with Bob and maybe Nathan to see how I can access the control panel from afar.
Have a nice day,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Bob Miller" <rmiller at jlab.org>, "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
Cc: "Denny Insley" <dinsley at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>, "Patrizia Rossi" <rossi at jlab.org>
Envoyé: Vendredi 22 Novembre 2024 20:30:46
Objet: Re: Purge the gas line
Hi Gabriel and Bob,
The ALERT DC and BoNuS target controls are currently online and working.
The BoNuS target controls were used to pressurize the gas line which feeds the target straw (although the straw is not connected).
The ALERT DC controls are operating in auto-mode and holding pressure in the system to a set point of 47 mbar above atmosphere. I have run the following tests:
1.
Raise the DC pressure manually to exceed the low pressure fault (ensuring manual control works)
2.
Set the system to auto and observed it increasing pressure to a set point of 47 mbar
3.
Changing the set point to 55 mbar and observed the system increase pressure in the DC to 55 mbar and hold
4.
Change the set point back to 47 mbar and observed the system decrease pressure in the DC back to 47 mbar and hold
5.
Intentionally release the target straw gas pressure to cause a fault, observe the DC controls release the pressure in the system at the max of the exhaust MFC (500 sccm)
6.
Re-pressurized the target straw circuit to 40 psi, observed the DC controls clear the target straw low pressure fault and return to automated control and return to the last set point (47 mbar)
Bob can now leak check the system. I'll contact Bob on Monday about operating the system.
Regards,
Marc
________________________________
From: Marc Mcmullen <mcmullen at jlab.org>
Sent: Thursday, November 21, 2024 8:53 AM
To: Bob Miller <rmiller at jlab.org>; Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Cc: Denny Insley <dinsley at jlab.org>; Marc Mcmullen via dsg-hallb_gas <dsg-hallb_gas at jlab.org>
Subject: Re: Purge the gas line
Hi Gabriel,
We are currently integrating the ALERT DC gas controls used in the EEL to the cRIO controller in the hall. We have read back from the sensors. Right now we are working on the solenoid control for ALERT and BoNuS. Then we will start pressure control of the panel.
It may take a few days to get it up and running.
After that Bob will do some leak checks.
I’ll provide updates as the project progresses.
Regards,
Marc
Get Outlook for iOS<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_o0ukef&d=DwMFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=plE8x1Qj6-oBUv4DJULncvm2rlHV-MV7JFc6UEKBC1w&m=LLR9Tch5GHNiwYDY0__ZJHaMHNui2tc7Ewl_Oxu9UuB6D11rnIztNYd2EHe2Ontl&s=8p9qsvz0AqyIqvHm7ALUE5OzcF5fsqMOQXbVucItn6E&e=>
________________________________
From: Bob Miller <rmiller at jlab.org>
Sent: Thursday, November 21, 2024 8:36:24 AM
To: Marc Mcmullen <mcmullen at jlab.org>; Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Cc: Denny Insley <dinsley at jlab.org>
Subject: Re: Purge the gas line
Hi All,
I have another meeting this morning, so I will not be able to attend the ALERT meeting.
Regards,
Bob Miller
Hall B Lead Engineer
(757) 822-9586
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Thursday, November 21, 2024 8:26 AM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hi Marc,
We will have an ALERT meeting today where I'd like to report on your progress with the gas system for ALERT.
Can you give me an update of what you've been able to complete this week, please?
Sometimes Bob attend the meeting too, so if you've already agreed that Bob will give the update, don't bother sending me an email.
Have a nice day,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
Cc: "Bob Miller" <rmiller at jlab.org>, "Denny Insley" <dinsley at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Vendredi 15 Novembre 2024 20:13:52
Objet: Re: Purge the gas line
Hi Gabriel,
The panel was moved to hall b earlier this week. DSG is working with hall b engineering and mechanical to connect the instrumentation and get the system running.
I expect this to be done in the early part of the coming week.
We will also have the pressure and flows (for the ALERT DC and Target straw) sent to archived process variables which you can monitor through Mya.
I’ll send an update next week as the system progresses.
Regards,
Marc
Get Outlook for iOS<https://urldefense.proofpoint.com/v2/url?u=https-3A__aka.ms_o0ukef&d=DwMFaQ&c=CJqEzB1piLOyyvZjb8YUQw&r=plE8x1Qj6-oBUv4DJULncvm2rlHV-MV7JFc6UEKBC1w&m=Ggy0n3SUo1eS1q0utKiSYSv0N6Km4c8zJ9iwVmoZ-8gWJCzwPf2uWg09Jzg0IK7w&s=iGweCepCsv9yp7WTnDLnBTAp0za6jWMnW1zNAkbI_YY&e=>
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Thursday, November 14, 2024 10:25:45 PM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; Marc Mcmullen via dsg-hallb_gas <dsg-hallb_gas at jlab.org>
Subject: [EXTERNAL] Re: Purge the gas line
Hello,
As you probably know the panel is now inside the Hall and ready to be set.
The temporary setup installed by Bob was very helpful but of course the lack of controls forces us to check that everything is fine everyday.
Even when I was on site, I did not do it properly: I did see a bottle was empty and then didn't see that the flow meter was closed.
It is anyway not possible to know immediately when a bottle is empty. With the new system this kind of problem will be avoided.
Hence I'm going to wait that the gas system is accessible via EPICS. So please, whevener you have time, can you work on it?
Bob, right now everything is opened as I was still running HV tests tonight. HV are now off.
Can you close the flow meter please? And then fill the Hall pipes with N2 at twice the atmospheric pressure. That will keep water away and speed up the restart process.
See you when the beam will be in Hall B,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
Pictures about detectors:
https://www.instagram.com/particledetectors/
________________________________
De: "Marc Mcmullen" <mcmullen at jlab.org>
À: "Bob Miller" <rmiller at jlab.org>, "Gabriel Charles" <gabriel.charles at ijclab.in2p3.fr>
Cc: "Denny Insley" <dinsley at jlab.org>, "Marc Mcmullen via dsg-hallb_gas" <dsg-hallb_gas at jlab.org>
Envoyé: Mercredi 6 Novembre 2024 16:39:43
Objet: Re: Purge the gas line
Hi Bob,
I am working on integrating the ALERT DC controls software to the cRIO in the hall. It will take a few days to get it up and running and tested.
Whenever the panel moves, Brian will asses the hardware connections needed and get with DSG staff so the pressure sensors, the solenoid, and the pump are connected as I am working on the software.
Please keep us updated on the panel movement.
Regards,
Marc
________________________________
From: Bob Miller <rmiller at jlab.org>
Sent: Wednesday, November 6, 2024 10:20 AM
To: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Cc: Denny Insley <dinsley at jlab.org>; Marc Mcmullen <mcmullen at jlab.org>
Subject: Re: Purge the gas line
Hi Denny,
When do you expect to install the gas panel next to ALERT?
Hi Marc,
When do you expect the gas panel to be functional in the Hall?
Thanks,
Bob Miller
Hall B Lead Engineer
(757) 822-9586
________________________________
From: Gabriel Charles <gabriel.charles at ijclab.in2p3.fr>
Sent: Wednesday, November 6, 2024 8:44 AM
To: Bob Miller <rmiller at jlab.org>
Subject: [EXTERNAL] [ALERT] Purge the gas line
Hi Bob,
As we discussed yesterday, would it be possible to start purging the Hall B gas line? As per the current schedule, ALERT detector will be at the beam level on Thursday. It's probably a good point to start purging then.
In case the installation is taking longer we could stop purging over the weekend and leave 2 atmospheric pressure in the tube.
What do you think?
Have a nice day,
--
Gabriel CHARLES
IJCLab/IN2P3/CNRS, University Paris-Saclay
+33 1 69 15 64 42
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_gas/attachments/20250203/4cdd1bcc/attachment-0001.html>
More information about the dsg-hallb_gas
mailing list