[Dsg-htcc] HTCC gas flow rate for KPP
    Brian Eng 
    beng at jlab.org
       
    Mon Jan 23 16:33:18 EST 2017
    
    
  
Based on George's recommendation I had Marc set the flow to 6.5 slm remotely.
> On Jan 23, 2017, at 2:53 PM, George Jacobs <jacobsg at jlab.org> wrote:
> 
> Hi Youri,
> 
> I recommend you increase your CO2 flow rate. The current rate of 2 slm is too low. I recommend a flow of 6.5 slm until the KPP run  is completed.
> 
> The calculated flow rates for the HTCC are as follows;
> Purge Flow - 20 slm - 1.5 volume exchanges per day- the detector volume will be purged in 3.4 days
> Running Flow - 6.5 slm - 0.5 volume exchanges per day 
> Reduced Flow - 3.25 slm - 0.25 volume exchanges per day
> 
> George
    
    