[Dsg-ltcc] [New Logentry] S5 Flow Rate

Maurizio Ungaro ungaro at jlab.org
Wed Jan 31 09:13:30 EST 2018


Hi Brian, George, 

a 0.065 l / minute flow is still 93 liters / day, which is about 1 1KG / day. George please correct me if I'm wrong.

The expected losses may be less than that. I'm worried about forcing flow and losing gas this way.

Thanks,
Mauri


----- Original Message -----
From: beng at jlab.org
To: dsg-ltcc at jlab.org
Sent: Monday, January 29, 2018 1:30:04 PM
Subject: [Dsg-ltcc] [New Logentry] S5 Flow Rate

S5 Flow Rate 


Lognumber 3523300 . Submitted by beng on Mon, 01/29/2018 - 13:28. Logbooks: 	HBLOG LTCC 
References: 	3523143 - Follow-up Re: Follow-up Re: Began filling LTCC S5 with C4F10 gas @ 0.500 slm 


The flow has been reduced another 0.1 slm from 0.4 slm to 0.3 slm. 

The plan is to reduce the flow every hour or so by 0.1 slm until we get to 0.1 slm (or the pressure starts to drop) and hold it there for the night. 

If 0.1 slm holds the flow will be set to 0.065 slm, which is slightly above the minimum flow value of 0.063 slm that the MFC can control. 