    OK, I will explain how the controls are working without the PID
    pressure control system installed. This info is also included in the
    operators manual and is the same as it was during CLAS 6 ops.<br>
    <br>
    First of all, N2 gas flows from the N2 manifold and supplies the
    MFC. The MFC will try to flow at the setpoint unless the gas supply
    solenoid safety valve is closed.<br>
    <br>
    The gas supply solenoid safety valve closes when pressure in the
    detector reaches 2.0 inches wc. This is an active protection to
    prevent window failure.  This is active until the pressure decreases
    to 1.9 inches water column ("wc), due to the programmed 0.100"wc
    deadband. So, gas will not flow if the pressure does not decrease
    enough to reset the relay and open the valve.<br>
    <br>
    So, how do you flow gas to leak check if the gas pressure is too
    high? You must first lower the pressure in the sector you are
    checking. The easiest and most direct method to do this is to remove
    the 1" cap from the exhaust manifold and open the exhaust valve to
    lower pressure for the sector you want to flow gas to. Another
    method is to remove one of the 1" lines from that sector on the LTCC
    valve panel. Once the pressure is lower, the solenoid relay will
    reset and you can once again flow gas. Please note that if you do
    not shut the valve, pressure will decrease to 0.000"wc.<br>
    <br>
    The pressure protection bubblers are a passive mode of protection.
    Currently, not all bubblers are set at the proper oil height to
    bubble at 2.5"wc. Some currently bubble at 2.0"wc . This task is on
    the to do list along with several others to be completed during the
    KPP setup.<br>
    <br>
    George<br>
    <br>
    <div class="moz-cite-prefix">On 12/7/2016 8:17 AM, David Anderson
      wrote:<br>
    </div>
    <blockquote
      cite="mid:1327155831.3919581.1481116653149.JavaMail.zimbra@jlab.org"
      type="cite">
      <pre wrap="">Gentlemen,
 For some reason I have lost all flow to sector 1, can I get some assistance with this?
I am still tracing down leaks in the system and can't troubleshoot with zero flow available.
Also, can we verify the flow rates/set points for the other sectors? 
There are several bubblers that are venting due to some variations in the set points and in the bubbler levels.

Thanks,
David  

----- Original Message -----
From: "David Anderson" <a class="moz-txt-link-rfc2396E" href="mailto:dla@jlab.org"><dla@jlab.org></a>
To: "Brian Eng" <a class="moz-txt-link-rfc2396E" href="mailto:beng@jlab.org"><beng@jlab.org></a>
Cc: "Douglas Tilles" <a class="moz-txt-link-rfc2396E" href="mailto:tilles@jlab.org"><tilles@jlab.org></a>, "Bob Miller" <a class="moz-txt-link-rfc2396E" href="mailto:rmiller@jlab.org"><rmiller@jlab.org></a>, "George Jacobs" <a class="moz-txt-link-rfc2396E" href="mailto:jacobsg@jlab.org"><jacobsg@jlab.org></a>, "Marc McMullen" <a class="moz-txt-link-rfc2396E" href="mailto:mcmullen@jlab.org"><mcmullen@jlab.org></a>
Sent: Monday, December 5, 2016 9:04:11 AM
Subject: Re: [Dsg-ltcc] LTCC gas system

OK guys I found some 1" line and tied the supply and return together at sector 6.
It appears the gas circuit is leak free and the #6 chamber is leaking.
I am planning to leak check that chamber by pressurizing and injecting 134a so I can leak check just the chamber as soon as I am able to. 
Due to drift chamber work ongoing getting manlift time to do so is difficult.
Also, as George pointed out the bubbler levels need to be tweaked as well.

Thanks,
David

----- Original Message -----
From: "David Anderson" <a class="moz-txt-link-rfc2396E" href="mailto:dla@jlab.org"><dla@jlab.org></a>
To: "Brian Eng" <a class="moz-txt-link-rfc2396E" href="mailto:beng@jlab.org"><beng@jlab.org></a>
Cc: "dsg-ltcc" <a class="moz-txt-link-rfc2396E" href="mailto:dsg-ltcc@jlab.org"><dsg-ltcc@jlab.org></a>, "Douglas Tilles" <a class="moz-txt-link-rfc2396E" href="mailto:tilles@jlab.org"><tilles@jlab.org></a>, "Bob Miller" <a class="moz-txt-link-rfc2396E" href="mailto:rmiller@jlab.org"><rmiller@jlab.org></a>
Sent: Monday, December 5, 2016 6:47:56 AM
Subject: Re: [Dsg-ltcc] LTCC gas system

Ok thanks Brian, leak checking will continue this morning.
George and I were discussing Friday about isolating the gas circuit from the chambers.
By connecting the supply and return together at the chamber I can quickly tell if the leak is in the gas circuit of the chamber itself.
That may save a lot of time....I'll need a 12' section of 1" tubing that terminates with a male fitting on each end.
Any help rounding that up would be helpful.

Thanks,
David 

----- Original Message -----
From: "Brian Eng" <a class="moz-txt-link-rfc2396E" href="mailto:beng@jlab.org"><beng@jlab.org></a>
To: <a class="moz-txt-link-abbreviated" href="mailto:dsg-ltcc@jlab.org">dsg-ltcc@jlab.org</a>
Cc: <a class="moz-txt-link-abbreviated" href="mailto:dla@jlab.org">dla@jlab.org</a>
Sent: Friday, December 2, 2016 4:59:33 PM
Subject: Re: [Dsg-ltcc] LTCC gas system

So I finally saw the S4 flow turn on for the first time; also S1 has turned on as well so it is still leaking, albeit much improved. I still haven't seen S5 turn on yet.

So currently the LTCC leak rates look like the following (from worst [biggest leak] to best):

S2 > S6 > S3 > S1 > S4 > S5

--
Brian Eng
Detector Support Group
757-269-6018

</pre>
      <blockquote type="cite">
        <pre wrap="">On Dec 2, 2016, at 2:51 PM, Brian Eng <a class="moz-txt-link-rfc2396E" href="mailto:beng@jlab.org"><beng@jlab.org></a> wrote:

Hi David,

All the flows were reduced to 0.5 LPM except for S2 & S6 (5 LPM & 3 LPM respectively). Those were set to ~1 LPM higher than when the flow as on 100% of the time, so figure the leak rate is somewhere between those numbers.

The leak on S1 is much improved if not completely fixed as I haven't seen the flow kick on recently.

S3 still has a leak as I see the flow (at 0.5 LPM) turn on periodically.
</pre>
      </blockquote>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a>

(website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a></pre>
  </body>
</html>