    Leak check of the splice and bubbler are complete. Result - no
    detectable leakage from these connections<br>
    <br>
    George<br>
    <br>
    <div class="moz-cite-prefix">On 1/29/2018 8:56 AM, George Jacobs
      wrote:<br>
    </div>
    <blockquote type="cite"
      cite="mid:814984234.4828014.1517234160106.JavaMail.zimbra@jlab.org">
      <div style="font-family: arial,helvetica,sans-serif; font-size:
        12pt; color: #000000">
        <div>LTCC S5 Fill Status - Almost there!<br>
        </div>
        <div><br data-mce-bogus="1">
        </div>
        <div>
          <div>LTCC S5 is currently filling. I have attached the
            pressure strip chart showing the 2 pressure transducer
            outputs.<br>
          </div>
          <br>
          <div>The blue line is the transducer located at the detector
            gas inlet, This is located at the lowest point of the
            detector volume. <br>
          </div>
          <br>
          <div>The brown line is the output of the pressure transducer
            connected to the pressure protection bubbler. (was the S4
            transducer)</div>
          <div><br data-mce-bogus="1">
          </div>
          <div>8:30an 29 Jan 2018 - Detector looks like it is nearly
            full. Detector pressure at the bottom corner is 2.1"wc. 
            Time to start reducing flow in order to determine a workable
            maintenance flow. This process will allow us to estimate how
            long the remaining gas will last. The scale shows 77kg
            remaining.<br data-mce-bogus="1">
          </div>
          <div><br data-mce-bogus="1">
          </div>
          <div>George<br data-mce-bogus="1">
          </div>
        </div>
        <div><br data-mce-bogus="1">
        </div>
        <div><br>
        </div>
        <hr id="zwchr" data-marker="__DIVIDER__">
        <div data-marker="__HEADERS__"><b>From: </b>"George Jacobs"
          <a class="moz-txt-link-rfc2396E" href="mailto:jacobsg@jlab.org"><jacobsg@jlab.org></a><br>
          <b>To: </b><a class="moz-txt-link-abbreviated" href="mailto:dsg-ltcc@jlab.org">dsg-ltcc@jlab.org</a>, "Bob Miller"
          <a class="moz-txt-link-rfc2396E" href="mailto:rmiller@jlab.org"><rmiller@jlab.org></a>, "Doug Tilles"
          <a class="moz-txt-link-rfc2396E" href="mailto:tilles@jlab.org"><tilles@jlab.org></a>, "Denny Insley"
          <a class="moz-txt-link-rfc2396E" href="mailto:dinsley@jlab.org"><dinsley@jlab.org></a>, "David Anderson"
          <a class="moz-txt-link-rfc2396E" href="mailto:dla@jlab.org"><dla@jlab.org></a>, "Morgan Cook" <a class="moz-txt-link-rfc2396E" href="mailto:mcookiv@jlab.org"><mcookiv@jlab.org></a>,
          "Todd Ewing" <a class="moz-txt-link-rfc2396E" href="mailto:jtewing@jlab.org"><jtewing@jlab.org></a>, "Calvin Mealer"
          <a class="moz-txt-link-rfc2396E" href="mailto:mealer@jlab.org"><mealer@jlab.org></a><br>
          <b>Cc: </b>"Amrit Yegneswaran" <a class="moz-txt-link-rfc2396E" href="mailto:yeg@jlab.org"><yeg@jlab.org></a><br>
          <b>Sent: </b>Tuesday, January 23, 2018 11:46:52 AM<br>
          <b>Subject: </b>LTCC S5 Fill Status and Further leak check
          required<br>
        </div>
        <div><br>
        </div>
        <div data-marker="__QUOTED_TEXT__">
          <div style="font-family: arial,helvetica,sans-serif;
            font-size: 12pt; color: #000000"
            data-mce-style="font-family: arial,helvetica,sans-serif;
            font-size: 12pt; color: #000000;">
            <div>LTCC S5 C4F10 Fill Status<br>
            </div>
            <br>
            <div>LTCC S5 is currently filling. I have attached the
              pressure strip chart showing the 2 pressure transducer
              outputs.<br>
            </div>
            <br>
            <div>The blue line is the transducer located at the detector
              gas inlet, This is located at the lowest point of the
              detector volume. <br>
            </div>
            <br>
            <div>The brown line is the output of the pressure transducer
              connected to the pressure protection bubbler. (was the S4
              transducer)<br>
            </div>
            <br>
            <div>9:30am 22 Jan 2018 - turned C4F10 flow on.<br>
            </div>
            <br>
            <div>11:20 am         Spliced the pressure protection
              bubbler into the S5 Exhaust/Vent line</div>
            <br>
            <div>1:40 pm            Disconnected the pressure protection
              bubbler from the gas lines in order to reduce oil level to
              2.5 inches<br>
            </div>
            <br>
            <div>2:15 pm            Reconnected piping to pressure
              protection bubbler<br>
            </div>
            <br>
            <div>18:10 pm          Pressure is limited by the Vent
              bubbler oil level. Fill continues, increasing the pressure
              at the gas inlet, blue line. Brown line indicated pressure
              at the protection bubbler is also increasing, but at a
              lesser rate, due to C4F10 carry over into the exhaust.</div>
            <br>
            <div>This is a good start. </div>
            <br>
            <div>Leak Check Required - Since I spliced in the S4 Bubbler
              and removed the bubbler piping in order to changed the oil
              levels in it, these connection should be leak checked
              ASAP. As can be seen on the pressure chart, detector
              pressure was very low when I reconnected the lines. Not
              enough pressure for a good leak check. An additional leak
              check should be performed using the sniffer, both at the
              bubbler connection splice Tee and the Bubbler piping
              connections.  <br>
            </div>
            <br>
            <div>Thanks,<br>
            </div>
            <div>George<br>
            </div>
            <div><br>
            </div>
            <br>
            <div>-- <br>
            </div>
            <div>George Jacobs<br>
              Physics Detector Support Group<br>
              Jefferson Lab<br>
              12000 Jefferson Ave.<br>
              STE 12<br>
              Newport News, VA 23606<br>
              <br>
              (office) 757-269-7115<br>
              <br>
              (cell)   757-876-0480<br>
              <br>
              (email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a><br>
              <br>
              (website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a></div>
          </div>
          <br>
        </div>
        <div><br>
        </div>
        <div data-marker="__SIG_POST__">-- <br>
        </div>
        <div>George Jacobs<br>
          Physics Detector Support Group<br>
          Jefferson Lab<br>
          12000 Jefferson Ave.<br>
          STE 12<br>
          Newport News, VA 23606<br>
          <br>
          (office) 757-269-7115<br>
          <br>
          (cell)   757-876-0480<br>
          <br>
          (email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a><br>
          <br>
          (website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a><br>
        </div>
      </div>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a>

(website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a></pre>
  </body>
</html>