[Dsg-hallb_magnets] Hall B Solenoid - temperature sensors
    Dave Kashy 
    kashy at jlab.org
       
    Wed Jul 19 20:21:52 EDT 2017
    
    
  
Hi Pablo,
To get below 1K takes very different refrigeration techniques.
Our polarized targets operate at~0.050K using dilution refrigerators 
The lowest temperature you should see on the solenoid is 4.2K if we are venting to atmosphere.
But at 1.2 ATM return pressure it can't go lower than 4.4K
All due to the properties on helium
Dave
Sent from my iPhone
> On Jul 17, 2017, at 5:23 PM, Pablo Campero <campero at jlab.org> wrote:
> 
> Hello Ruben,
> 
> I checked the sensors that is reading back 0.0 K 
> 
> TR8671
> TR8673
> TR8611
> TR8672
> TR8674
> TR8610
> 
> The mentioned sensors are read by the Cryocon Units, so I have to go and check the cryocon units readouts and also make sure that the mentioned temperature sensors are physically connected in the Solenoid Service Tower.  
> PLC is reading  0.0 K values because the Cryocon Unit is reading a real zero (which we know is not possible at this moment) or it has a ''clip'' readback on its channels where the temperature sensors is connected (Most probably for now).
> 
> 
> Best Regards,
> Pablo 
> 
> 
> From: "Ruben Fair" <rfair at jlab.org>
> To: "Pablo Campero Rojas" <campero at jlab.org>, "Probir Ghoshal" <ghoshal at jlab.org>
> Cc: "Scot Spiegel" <spiegel at jlab.org>
> Sent: Monday, July 17, 2017 4:35:26 PM
> Subject: Hall B Solenoid - temperature sensors
> 
> Pablo, Probir,
> 
> All the temperature sensors are now installed and the connectors are still plugged in.
> 
> Please review all the sensors on EPICS and see if they are all reading what you expect.
> 
> If not, please work with each other to assign correct calibration curves.
> 
> I shall disconnect the cables first thing tomorrow morning at 7am as welding will start shortly after.
> 
> Regards
> 
> Ruben
> 
> _______________________________________________
> Dsg-hallb_magnets mailing list
> Dsg-hallb_magnets at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-hallb_magnets
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20170719/fe3dda2b/attachment.html>
    
    
More information about the Dsg-hallb_magnets
mailing list