[Dsg-hallb_magnets] Monitoring Solenoid vacuum signals- initial checks

Amrit Yegneswaran yeg at jlab.org
Mon Jul 31 19:50:51 EDT 2017


gr8 job pablo and brian, 
you guys seem to be doing so many nice catches that next time i go deep see fishing, i'm taking you guys onboard with me. 
catch me the trades boys and we'll heave away to singapore. 


https://www.youtube.com/watch?v=vkYt-ToxfkM (waits) 





From: "George Biallas" <biallas at jlab.org> 
To: "Pablo Campero" <campero at jlab.org> 
Cc: "David Anderson" <dla at jlab.org>, "dsg-hallb magnets" <dsg-hallb_magnets at jlab.org> 
Sent: Monday, July 31, 2017 4:29:56 PM 
Subject: Re: [Dsg-hallb_magnets] Monitoring Solenoid vacuum signals- initial checks 

Pablo, 
Nice catch. 

Pfeiffer's interchanging the pin positions 1 to 2 an 2 to 1 seems nonsensical. Go figure! 




From: "Pablo Campero" <campero at jlab.org> 
To: "David Anderson" <dla at jlab.org>, "Scot Spiegel" <spiegel at jlab.org> 
Cc: "dsg-hallb magnets" <dsg-hallb_magnets at jlab.org> 
Sent: Monday, July 31, 2017 3:23:19 PM 
Subject: Re: [Dsg-hallb_magnets] Monitoring Solenoid vacuum signals- initial checks 




BQ_BEGIN

Hello David, 

With regards to the correction that you mention to the PLC vacuum signals readouts, I verified that the assignation for the PLC channels for each vacuum gauge are correct. However I found that the analog outputs wiring from the Vacuum controller perhaps is not correct. 


Scot, 

Looking to the DWG B00000-09-00-0656, sheet 2, I noticed that the wiring of the pin-outs for the analog outputs in the Vacuum controller Pfiffer TPG-362 are reversed for pin 1 and pin 2. 
According to the Pfiffer TPG-362 manual (see attached, page 15): pin 2 is associate with the gauge # 1 (CG8606) and pin 1 corresponds to gauge # 2 (CG8600TB). 
In the mentioned DGW above the pin 1 is used for the analog output signal (0-10 V) for gauge #1 (CG8606) and Pin 2 for gauge #2 (CG8600TB). 
I believe that this is why the vacuum signals are reversed and the wiring and drawing should to be modified. Can you please check and confirm this. 


Best Regards, 
Pablo 





From: "David Anderson" <dla at jlab.org> 
To: campero at jlab.org 
Cc: "George Biallas" <biallas at jlab.org>, "Ruben Fair" <rfair at jlab.org> 
Sent: Monday, July 31, 2017 11:10:16 AM 
Subject: Re: Monitoring Solenoid vacuum signals- initial checks 

The CG gauge signals are reversed in the PLC.... 
Please correct... 

Thanks, 
David 

----- Original Message ----- 
From: "Ruben Fair" <rfair at jlab.org> 
To: "George Biallas" <biallas at jlab.org>, "David Anderson" <dla at jlab.org> 
Sent: Friday, July 28, 2017 5:23:56 PM 
Subject: Fwd: Monitoring Solenoid vacuum signals- initial checks 

FYI 


From: "Pablo Campero Rojas" <campero at jlab.org> 
To: "Ruben Fair" <rfair at jlab.org> 
Cc: "dsg-hallb magnets" <dsg-hallb_magnets at jlab.org> 
Sent: Friday, July 28, 2017 5:12:19 PM 
Subject: Monitoring Solenoid vacuum signals- initial checks 

Hello Ruben, 

The following are task completed today with regards to the readout of the vacuum signals. 

1. Removed temporary PLC logic program that was used to monitor the temporary vacuum gauge on CG8606. 
2. Wiring for temporary gauge was removed from the terminal block in the controls rack to connect actual vacuum gauge signal CG8606. 
3.Verified correct assignation in the PLC analog input channels (0-10 V) for three vacuum signals (CG8606, CG8600TB, and TB8600) 
4. The connector hooked in Turbo pump TC-1200 remote was modified, pin 1 (+24VDC) was disable to avoid turning on of the turbo pump, This remote option not required by our control system; connector only used to monitor the speed in the turbo pump (pin 12: 0-10 V output). 
5. Verified wiring (continuity and sense) for the three signals. 
6. Measured voltage in the terminal block to make sure the PLC analog input operation range 0-10 V. 
7. Actual readouts for the vacuum signals were compared between hardware vacuum controllers and the readouts on the Solenoid control systems (PLC and EPICS). The initial readouts as follow: 
Vacuum signal 
Pfeiffer DCU-002 Unit - Turbo Pump 
Pfeiffer TPG-362 - Vacuum controller PLC/EPICS 
CG8606 
4.27 [torr] 4.191 / 4.18 [torr] 
CG8600TB 
1.51*10^-2 [torr] 1.512*10^-2 / 0.01 [torr] 
TB8600 0 % 
0.101 / 0.10 % 

Best Regards, 
Pablo 

_______________________________________________ 
Dsg-hallb_magnets mailing list 
Dsg-hallb_magnets at jlab.org 
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_magnets 

BQ_END


-- 
George 

_______________________________________________ 
Dsg-hallb_magnets mailing list 
Dsg-hallb_magnets at jlab.org 
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_magnets 

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20170731/aadfa48a/attachment-0001.html>


More information about the Dsg-hallb_magnets mailing list