[Dsg-hallb_magnets] Hall B Solenoid - Update on SOE relay string / dump switch operation check out

Pablo Campero campero at jlab.org
Fri May 5 15:51:54 EDT 2017


Hello Ruben, 
Attached find the updated spreadsheet with the results of the QD - SOE and MPS Dump Switch functionality Test. 

Nick, 
We did not complete with the Water Cooled Leads Flow Test, will continue the test on Monday (05//08/17). 
After that, you will be able to pull out the SOE module and go ahead with your point testing hardware for the SOE module display. 

Best Regards, 
Pablo 

From: "Probir Ghoshal" <ghoshal at jlab.org> 
To: "Brian Eng" <beng at jlab.org> 
Cc: "Sarin Philip" <philip at jlab.org>, "Krister Bruhwel" <bruhwel at jlab.org>, "Pablo Campero Rojas" <campero at jlab.org>, "Ruben Fair" <rfair at jlab.org> 
Sent: Friday, May 5, 2017 1:07:56 PM 
Subject: Re: Hall B Solenoid - Update on SOE relay string / dump switch operation check out 

Hi Brian, 

I will be in the hall with Pablo again this afternoon flow water flow test in the hall..! 
Please come over. 
We did the tests with our QD after the interlock issue that was from the Lead temp...! 
Pablo has taken care off and we have completed our tests on the QD successfully. 

Please come over to the hall and see if we can get hold of Mark. 

Regards 

Probir 


----- Original Message ----- 
From: "Ruben Fair" <rfair at jlab.org> 
To: "Probir Ghoshal" <ghoshal at jlab.org> 
Sent: Friday, May 5, 2017 12:58:08 PM 
Subject: Fwd: Hall B Solenoid - Update on SOE relay string / dump switch operation check out 

Begin forwarded message: 

> From: Brian Eng <beng at jlab.org> 
> Date: May 5, 2017 at 12:42:36 PM EDT 
> To: Ruben Fair <rfair at jlab.org> 
> Cc: Sarin Philip <philip at jlab.org>, Krister Bruhwel <bruhwel at jlab.org>, Pablo Campero Rojas <campero at jlab.org> 
> Subject: Re: Hall B Solenoid - Update on SOE relay string / dump switch operation check out 
> 
> Sarin, 
> 
> Here's the A to ppm at a few values after Pablo changed the code to reflect the 4000A capacity. 
> 
> 4 A - 1000 ppm 
> 25 A - 6250 ppm 
> 60 A - 15000 ppm 
> 100 A - 25000 ppm 
> 
> The Amps comes from the EPICS screen while the ppm was from the LCD display. 
> 
> Do you think Mark would be available after lunch to check to see if the safety OCP interlock is still happening? 
> 
> Pablo, 
> 
> I can verify that the lowest slew rate you can set is indeed 273 mA/s, this is due to DA 4 being set to have a min DAC value of 32. 
> 
>> On May 2, 2017, at 3:40 PM, Ruben Fair <rfair at jlab.org> wrote: 
>> 
>> Hi Sarin, 
>> 
>> Yes please. 
>> 
>> Pablo, Brian, Krister, 
>> 
>> Can you please liaise with Sarin to be able to work with Mark Todd to see why this interlock keeps tripping and also to confirm the A to ppm calibration? 
>> 
>> Many thanks 
>> 
>> Ruben 
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20170505/d2f57729/attachment.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Solenoid- SOE and Dump Switch MPS Test - V1.xlsx
Type: application/vnd.openxmlformats-officedocument.spreadsheetml.sheet
Size: 12401 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20170505/d2f57729/attachment.xlsx>


More information about the Dsg-hallb_magnets mailing list