[Dsg-hallb_magnets] Solenoid MPS water flow operations
Pablo Campero
campero at jlab.org
Wed Dec 5 09:58:46 EST 2018
Dear Dr. Rossi,
With regards to the safety operation concerns/questions with the Solenoid MPS, after investigations I can mention the following:
1. Recommended water supply flows and ΔP by Danfysik are:
1.1. Flow has to be < 180 [l/min] At ΔP (Sup-Ret) = 6 [Bar].
1.2. Flow has to be < 115 [l/min] at ΔP (Sup-Ret) = 3 [Bar].
2. Danfysik performed a pressure test for the Solenoid MPS water cooling system at 31 [Bar] for 1 [min], during factory acceptance test protocol on 14/04/2015.
3. The max absolute pressure rated by Danfysik = 12 [Bar].
4. The table below shows a summary of the values for MPS internal flows recommended and the actual flows as of now:
Flow Switch
Flow Transmitter Dynamic Range [l/min] Recommended Flows by Danfysik [l/min] Measured Flows [l/min]
FSW1 1-80 50 67
FSW2 1-15 8.22 8.5
FSW3 1-80 50 71.87
FSW4 1-15 1.45 1.65
FSW5 1-15 1.45 1.55
Total Flow -- 111.12 150.57
Since actual measured supply flow is 150.57 [l/min] at ΔP (Sup-Ret) = 5.17 [Bar] within the recommended operational values, we can say that the MPS water cooling pressure and flow are safe.
The MPS will not be damaged.
Best Regards,
Pablo