[Dsg-hallb_magnets] Hall B magnet interlock	and	instrumentation	checkouts complete
    David Kashy 
    kashy at jlab.org
       
    Wed Jul 15 10:33:35 EDT 2020
    
    
  
Hi Brian,
What type of limits are you talking about. Alarm limits or something else?
Thanks
Dave
________________________________
From: Brian Eng <beng at jlab.org>
Sent: Wednesday, July 15, 2020 10:31 AM
To: David Kashy <kashy at jlab.org>; Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Should I change the limits for that PT from something other than 1-1.5 then?
________________________________
From: Dsg-hallb_magnets <dsg-hallb_magnets-bounces at jlab.org> on behalf of David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 9:53 AM
To: Tyler Lemon <tlemon at jlab.org>; Ruben Fair <rfair at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Tyler,
1.5 is the setpoint for control so it will read that most of the time.
Dave
________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Wednesday, July 15, 2020 9:37 AM
To: Ruben Fair <rfair at jlab.org>; David Kashy <kashy at jlab.org>; dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Cc: Probir Ghoshal <ghoshal at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Ruben, Dave,
The "failure" for PT8554 is that the script expects the sensor's reading to be between 1 atma and 1.5 atma. Since the value the script saw for PT8554 was 1.5 atma, it put the status as "FAIL".
Tyler
________________________________
From: Ruben Fair <rfair at jlab.org>
Sent: Wednesday, July 15, 2020 9:27 AM
To: Tyler Lemon <tlemon at jlab.org>
Cc: Probir Ghoshal <ghoshal at jlab.org>
Subject: Fw: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi Tyler,
Could you comment on Dave's question below?
Cheers
Ruben
________________________________
From: David Kashy <kashy at jlab.org>
Sent: Wednesday, July 15, 2020 8:46 AM
To: Ruben Fair <rfair at jlab.org>; Probir Ghoshal <ghoshal at jlab.org>; Denny Insley <dinsley at jlab.org>
Cc: Nicholas Sandoval <sandoval at jlab.org>
Subject: Re: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hi All,
It is correct that TD8522TR has been very unreliable in the past, especially when below 10K. During warm up and cooldown it seems to work better. It is also true that it is not used for control so running without it is not a big deal.
As far as PT8554, that is another story. This sensor is the input parameter for the LN2 reservoir back pressure control. It seems to be working fine at least now when I looked.
What was the indication that caused it to be "failed" during the checkout?
Thanks
Dave
________________________________
From: Ruben Fair <rfair at jlab.org>
Sent: Tuesday, July 14, 2020 3:27 PM
To: David Kashy <kashy at jlab.org>; Probir Ghoshal <ghoshal at jlab.org>; Denny Insley <dinsley at jlab.org>
Cc: Nicholas Sandoval <sandoval at jlab.org>
Subject: Fw: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Dave,
See elog below from the magnet pre-check today - I believe neither the temp or press sensor is being used to control anything?
https://logbooks.jlab.org/entry/3814661
[cid:a085db29-3ee8-43e5-9b24-e8fabd360697]
[cid:9d782124-1718-4ea5-b08f-00251ddc11b5]
Probir,
I think we did at one time try swapping leads to activate the spare TD8522TR but saw the same behavior?
Ruben
________________________________
From: Dsg-hallb_magnets <dsg-hallb_magnets-bounces at jlab.org> on behalf of Tyler Lemon <tlemon at jlab.org>
Sent: Tuesday, July 14, 2020 2:59 PM
To: dsg-hallb magnets <dsg-hallb_magnets at jlab.org>
Subject: [Dsg-hallb_magnets] Hall B magnet interlock and instrumentation checkouts complete
Hello
Pablo, Probir, and I have completed the pre-power-up interlock and instrumentation checkouts for the Hall B Solenoid and Torus. The checkout results have been posted to the logbooks at the links below.
Torus interlock - https://logbooks.jlab.org/entry/3814708
Torus instrumentation - https://logbooks.jlab.org/entry/3814474 , https://logbooks.jlab.org/entry/3814475
Solenoid interlock - https://logbooks.jlab.org/entry/3814709
Solenoid instrumentation - https://logbooks.jlab.org/entry/3814661
Tyler
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/c8eb3340/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 6442 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/c8eb3340/attachment-0002.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 4359 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_magnets/attachments/20200715/c8eb3340/attachment-0003.png>
    
    
More information about the Dsg-hallb_magnets
mailing list