<html><body><div style="font-family: times new roman,new york,times,serif; font-size: 12pt; color: #000000"><div>Hello Ruben,<br data-mce-bogus="1"></div><div><br data-mce-bogus="1"></div><div>I have been monitoring the Torus Live Fast-Daq EPICS screen in preparation for testing the digital filters and have noticed odd signals in IDCCT1.<br data-mce-bogus="1"></div><div><br data-mce-bogus="1"></div><div>IDCCT1 Fast-Daq signal is changing in amplitude as I monitor it. So far I have seen it steady at several amplitudes including 4 [A], 0.025 [A], and 1 [A]. Attached are screen shots of the EPICS Fast-Daq Live screen showing these signals and also a screen shot of the signal in transition from one amplitude to another.<br></div><div><br data-mce-bogus="1"></div><div>01 Torus Fast-Daq IDCCT1- large 4 A amplitude signal- 2017-01-18 at 933AM -- shows 4 [A] amplitude signal seen at 9:33AM.<br data-mce-bogus="1"></div><div>02 Torus Fast-Daq IDCCT- transition to small amplitude- 2017-01-18 933AM-- shows transition from 4 [A] amplitude to 0.025 [A] amplitude.<br data-mce-bogus="1"></div><div>03 Torus Fast-Daq IDCCT1- 0.025 A amplitude signal- 2017-01-18 at 934AM -- shows 0.025 [A] amplitude signal seen at 9:35AM<br data-mce-bogus="1"></div><div>04 Torus Fast-Daq IDCCT1- 1 A amplitude signal- 2017-01-18 at 950AM -- shows 1 [A] amplitude signal seen at 9:50AM.<br data-mce-bogus="1"></div><div><br data-mce-bogus="1"></div><div>Is there any work occurring on the Torus system that would affect the IDCCT1 signal or Fast-Daq signals? If so, should I hold off on any testing of the digital filters until the work is done?<br data-mce-bogus="1"></div><div><br data-mce-bogus="1"></div><div>Best regards, <br data-mce-bogus="1"></div><div>Tyler<br data-mce-bogus="1"></div></div></body></html>