    Hi Matt,<br>
    <br>
    The length of tubing from the rack on the H2 Target gas pad to the
    FT gas chassis is 115 ft. We will need 40 ft of NEW 1/4" SS line to
    complete the connection. The distance to the MVT gas chassis should
    be the same. I have 3 cylinders of 10% pre-mix left over from the
    2011 experiment.<br>
    <br>
    George<br>
    <br>
    <div class="moz-cite-prefix">On 9/14/2017 9:07 AM, George Jacobs
      wrote:<br>
    </div>
    <blockquote type="cite"
      cite="mid:90a604fe-fe53-8bd4-7df8-e8c72efa8561@jlab.org">
      <meta http-equiv="Content-Type" content="text/html; charset=utf-8">
      Hi Matt,<br>
      <br>
      The P&I only shows a single box as this was originally
      designed to test the FT tracker in the hall using pre-mix. That
      setup never happened. So, currently, the MVT group needs to test
      the MVT detector prior to magnet insertion. This can be
      accomplished using the system show on the P&I. The P&I
      would only require both the FT and MVT gas distribution chassis if
      both require testing or if this system were to be a temporary
      supply. There are only 2-3 cylinders of premix remaining, so it
      would be a short term supply.<br>
      <br>
      The exhaust line shown is the current one in the hall that is
      shared by many systems.<br>
      <br>
      This is the same system as in EEL rm 124, just with longer lines.
      That system was determined to be a pressure system.<br>
      <br>
      I will get a line length estimate for you today. The 1/4" tubing
      has 0.029 wall, so the ID is 0.192.<br>
      <br>
      thanks,<br>
      George<br>
      <br>
      <div class="moz-cite-prefix">On 9/13/2017 2:49 PM, Matthew
        Marchlik wrote:<br>
      </div>
      <blockquote type="cite"
        cite="mid:1078197907.7566486.1505328597170.JavaMail.zimbra@jlab.org">
        <pre wrap="">Hi Maxime,

The system may not even qualify as a "pressure system". I'll determine the stored energy and update you. If it is deemed a PS I estimate 1-1.5 weeks to complete all the paperwork. Maybe longer if I cannot find an inspector and/or DA to sign off the technical review form in a timely manner.

Even if the system is deemed to not be a pressure system, an analysis per "Flammable Gas for Physics Experiments" supplement and an OSP are still required.


George,

Your P&ID only shows one box to represent the FT(will change to MVT). Shouldn't it be two boxes, the first is the FT supply box, second would be the FT?

Is the exhaust line shown one that is currently available or does it represent the one that is being installed for the MVT and FT exhaust?

Orifices and relief valves will most likely be required. If you need help sizing these, I can help you even if its not a PS.

I will need more system parameters to determine if the mechanical and chemical stored energy is below 1,000 ft-lb. What is the ID and total length of the 1/4 tubing? The two 1/4 in lines should combine to 1/2 in since its a long run down to the MVT supply box. What is the ID and total length of this line?

As a reminder. Anyone proficient in performing the stored energy calcs and meeting applicable requirements of the PVSSS may make the determination whether or not a system is considered a "pressure system", it doesn't require a DA.

Matt

----- Original Message -----
From: "mdefurne" <a class="moz-txt-link-rfc2396E" href="mailto:mdefurne@jlab.org" moz-do-not-send="true"><mdefurne@jlab.org></a>
To: <a class="moz-txt-link-abbreviated" href="mailto:dsg-mvt@jlab.org" moz-do-not-send="true">dsg-mvt@jlab.org</a>, "Bob Miller" <a class="moz-txt-link-rfc2396E" href="mailto:rmiller@jlab.org" moz-do-not-send="true"><rmiller@jlab.org></a>, "Matthew Marchlik" <a class="moz-txt-link-rfc2396E" href="mailto:marchlik@jlab.org" moz-do-not-send="true"><marchlik@jlab.org></a>, "George Jacobs" <a class="moz-txt-link-rfc2396E" href="mailto:jacobsg@jlab.org" moz-do-not-send="true"><jacobsg@jlab.org></a>
Sent: Tuesday, September 12, 2017 11:25:55 AM
Subject: Re: [Dsg-mvt] Mixing system

Hi George, Hi Bob, Hi Matt,

1) Can you give an estimate of how much time it takes to do the Pressure 
system requirements?

2) Do we have all components in the house? If not, do we have to include 
a 6-week delivery time for the components? How much would it cost?

3) Bob, can you send me the latest schedule you have for the 
installation of MVT in the Hall? If the mixing system cannot be ready 
for October 15th, can you give me a date on which you are absolutely 
sure it will be ready to be used? November 1st?

I apologize for the many questions. But I want to be sure it is worth 
spending time and money on this premix system.

Thank you for you replies,
Have a good day,

Maxime

Le 11/09/2017 à 12:48, George Jacobs a écrit :
</pre>
        <blockquote type="cite">
          <pre wrap="">Hi Maxime,

Pre-mix supply system assembly can happen in 1-3 days after all 
pressure systems requirements have been completed, as long as all 
components are in house.

It's the pressure system requirements that require a long lead time. 
More time is also required if we need to order anything such as 
orifices or relief valves that have lead times.

George

On 9/10/2017 10:38 PM, mdefurne wrote:
</pre>
          <blockquote type="cite">
            <pre wrap="">Hello,

If this is the most "convenient" back-up system, then I would say go 
for it. Nevertheless mixing system must be the priority. If it can be 
ready by the time we move the MVT in the Hall, then let's forget 
about this pre-mix gas supply.

I do not know the time scale. George, how long do you need to have 
this pre-mix system ready? Matt, how long to you need to approve this 
system if we need it? If you tell me X days, then let's check the 
mixing system status X days before the MVT installation. If we think 
the mixing system will be ready, then there will be no need to spend 
time on the premix supply in the Hall.

Apparently premix bottles will not be ordered anymore. But we will 
keep 2-3 of them in case the mixing system is not ready (one bottle 
lasts 6 days if no leak between the detector and the bottle).

Have a good day,

Maxime


Le 08/09/2017 à 13:35, George Jacobs a écrit :
</pre>
            <blockquote type="cite">
              <pre wrap="">There was a plan earlier this year to test the FT with pre-mix, so I 
produced the attached diagram for testing the FT with pre-mix in the 
hall at that time. This setup uses the pre-existing lines that run 
from the target gas pad to the downstream end of the space frame. 
Replacing the FT chassis in the diagram with the MVT chassis is all 
that is required to make it the MVT pre-mix gas supply. This setup 
is quite similar to the one in EEL rm 124, but still requires a DA.

George

On 9/8/2017 12:14 PM, mdefurne wrote:
</pre>
              <blockquote type="cite">
                <pre wrap="">Hi Marc,

Thank you for the update. Before inserting the MVT in the solenoid, 
we will have to check that everything has been plugged back 
properly. To do so, we will need gas.

A back-up plan will be to have the gas line ready and plug a premix 
bottle on it. What is the procedure to follow to have this back up 
plan ready for the MVT installation in the Hall? Unfortunately it 
is necessary to prepare this back-up plan because we won t insert 
the CVT in the solenoid without checking that everything is OK.

Kind Regards,

Maxime


Le 08/09/2017 à 10:09, Marc McMullen a écrit :
</pre>
                <blockquote type="cite">
                  <pre wrap="">Hi Maxime,

Currently the MVT mixing system is being evaluated by the Pressure 
Systems authority and a location is being discussed by the fire 
safety authority. If the location is moved outdoors, consideration 
must be given for the mass flow controllers. The controllers 
purchased are not specified to operate at a temperature lower than 
50F.

DSG is currently looking into another option for an MFC, while 
Hall B is looking into locations and climate control possibilities 
for the MFCs.

With that said, Amrit and I have discussed the possibility for the 
mixing system to be completed by 9/15 (possibly not even 10/15) 
and we think it is not likely. George can provide further 
confirmation. If it is necessary to run a premix by then, that 
option needs to be expedited.

Regards,

Marc



On 9/6/2017 1:41 PM, mdefurne wrote:
</pre>
                  <blockquote type="cite">
                    <pre wrap="">Dear George,

Will the mixing system be ready for October 1st? If not when do 
you expect to have it ready?

If it is not ready before the 15th, we will absolutely need the 
gas line ready so that we can plug a premix bottle to run tests: 
We will not be able to enter the CVT in the solenoid if we do not 
check the electronics connexion and gas lines. Indeed, once the 
CVT inside the solenoid, the micromegas tiles and disks will not 
be accessible anymore.

If there is any administrative paperwork to have a premix bottle 
plugged on the gas line, please let me know.

Thank you,

Kind regards,

Maxime

_______________________________________________
Dsg-mvt mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Dsg-mvt@jlab.org" moz-do-not-send="true">Dsg-mvt@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-mvt" moz-do-not-send="true">https://mailman.jlab.org/mailman/listinfo/dsg-mvt</a>
</pre>
                  </blockquote>
                </blockquote>
              </blockquote>
              <pre wrap="">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(<a class="moz-txt-link-abbreviated" href="mailto:email%29jacobsg@jlab.org" moz-do-not-send="true">email)jacobsg@jlab.org</a>

(website)<a class="moz-txt-link-freetext" href="https://userweb.jlab.org/%7Ejacobsg" moz-do-not-send="true">https://userweb.jlab.org/~jacobsg</a>


_______________________________________________
Dsg-mvt mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Dsg-mvt@jlab.org" moz-do-not-send="true">Dsg-mvt@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-mvt" moz-do-not-send="true">https://mailman.jlab.org/mailman/listinfo/dsg-mvt</a>
</pre>
            </blockquote>
          </blockquote>
          <pre wrap="">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(<a class="moz-txt-link-abbreviated" href="mailto:email%29jacobsg@jlab.org" moz-do-not-send="true">email)jacobsg@jlab.org</a>

(website)<a class="moz-txt-link-freetext" href="https://userweb.jlab.org/%7Ejacobsg" moz-do-not-send="true">https://userweb.jlab.org/~jacobsg</a>


_______________________________________________
Dsg-mvt mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Dsg-mvt@jlab.org" moz-do-not-send="true">Dsg-mvt@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-mvt" moz-do-not-send="true">https://mailman.jlab.org/mailman/listinfo/dsg-mvt</a>
</pre>
        </blockquote>
      </blockquote>
      <br>
      <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org" moz-do-not-send="true">jacobsg@jlab.org</a>

(website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/%7Ejacobsg" moz-do-not-send="true">https://userweb.jlab.org/~jacobsg</a></pre>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <br>
      <pre wrap="">_______________________________________________
Dsg-mvt mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Dsg-mvt@jlab.org">Dsg-mvt@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-mvt">https://mailman.jlab.org/mailman/listinfo/dsg-mvt</a>
</pre>
    </blockquote>
    <br>
    <pre class="moz-signature" cols="72">-- 
George Jacobs
Jefferson Lab (TJNAF)
STE 12
12000 Jefferson Ave.
Newport News, VA 23606

(office) 757-269-7115

(cell)   757-876-0480

(email) <a class="moz-txt-link-abbreviated" href="mailto:jacobsg@jlab.org">jacobsg@jlab.org</a>

(website) <a class="moz-txt-link-freetext" href="https://userweb.jlab.org/~jacobsg">https://userweb.jlab.org/~jacobsg</a></pre>
  </body>
</html>