James,
 thank you for the updates.
 I will talk to Phillip about the lifting.
 Do you have data for flow test performed in 2015 ?
      Regards, Tsuneo Kageya.
________________________________
From: James Brock <jbrock at jlab.org>
Sent: Wednesday, August 6, 2025 9:40 AM
To: Pablo Campero Rojas <campero at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Xiangdong Wei <xwei at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Cc: Denny Insley <dinsley at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
Thanks Pablo. I believe Xiangdong is on vacation, Phillip and Tsuneo can get a power strip installed in the rack.  Beyond that Pablo, are all the controls working as intended for the experiment?
Tsuneo and Phillip, unless the target pressure issue has been resolved, the cell membrane, gas panel valve seals, and internal plumbing integrity need to be verified.   Please coordinate with Denny to pick up the Target from the top ISO Flange with the gantry in the ESB (there are three lifting eyes). Rigging should be one sling (down stream) and two come-alongs (up stream) to level the flange before lifting.  When it's lifted I can help with shield removal and the cell inspection.  Just let me know when.
After the inspection, the flow tests need to be made again. Hopefully it will reproduce the measurements made in 2015 and the modifications Bob planned can proceed.  The schedule he presented in June had the target modifications completed at the end of this month.  These changes can't proceed until the testing is complete.
Xiangdong, Phillip, and Tsuneo, PRad needs more attention because time is running out. I can help out but again, I have other priorities I'm responsible for.
-James
James Brock
Polarized Target Group
Thomas Jefferson National Accelerator Facility
12000 Jefferson Ave. Suite 12
Newport News, VA 23606
[cid:98943579-0e5c-4ef1-b93e-7feb9fabefe6]
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Tuesday, August 5, 2025 9:37 AM
To: James Brock <jbrock at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Xiangdong Wei <xwei at jlab.org>; dsg-hallb_prad_ii at jlab.org <dsg-hallb_prad_ii at jlab.org>
Subject: Re: PRad Controls Issues
Hello James and Xiangdong,
Yesterday, I went ahead and checked the connections for the NI-DAQ-USB-6001 module DI channels prior to consider your suggested Datexel Modbus DAQ modules and found that the power supply inside the EV Aux Box that provides the +5VDC to NI-DAQ-6001 module DI channels was not connected. I proceeded and installed a power cable to the power supply and also wired the +5VDC to the NI-DAQ-6001 module. All work was performed with no power.
Tsuneo and I tested the control and position readback of EV1, EV2 valves using LabVIEW PRad program, and noticed no issues. You might want to test the status readback for the Gate Valve and Vacuum Switch later when running the system. Fixing the readout issues of the NI-DAQ-6001 module DI channels was the quickest solution to have the target controls running.
Currently, the +5VDC power supply is connected to a temporary power extension but I suggest installing another power strips in the rack since there are no available outlets on the existing ones.
The estimated arrival time for the ordered NI-DAQ-USB-6421 module is around August 11th.
Regards,
Pablo
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Thursday, July 31, 2025 1:56 PM
To: James Brock <jbrock at jlab.org>; Pablo Campero Rojas <campero at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>
Cc: Christopher Keith <ckeith at jlab.org>; Denny Insley <dinsley at jlab.org>; Chris Carlin <carlin at jlab.org>
Subject: Re: PRad Controls Issues
Hi James,
Thank you very for the briefing, I appreciate it.
Phillip will work with you I'm on vacation.
Thanks,
Xiangdong
________________________________
From: James Brock <jbrock at jlab.org>
Sent: Thursday, July 31, 2025 1:14 PM
To: Xiangdong Wei <xwei at jlab.org>; Pablo Campero Rojas <campero at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>
Cc: Christopher Keith <ckeith at jlab.org>; Denny Insley <dinsley at jlab.org>; Chris Carlin <carlin at jlab.org>
Subject: Re: PRad Controls Issues
Hi All,
My apologies I intended to send an email out yesterday. From my perspective the schedule for PRad is slipping and there remains a lot of work to do. Many of the jobs scheduled cannot be performed in parallel. I'm happy to help with PRad but I have other priorities that need attention so my time will be limited.
I told Tsuneo his suspicion about the target window is worth checking. This would explain the cell pressure issue. Another thing to check is the integrity of EV1 and EV2, just to verify they seal completely.
Denny, let me know when you're planning on pulling the target and I can meet you there to help and inspect the target cell.
Pablo, I found two Datexel Modbus Daq modules that should cover everything that needs to be read out for PRad II. It might be faster to use these we have onhand.  Chris Carlin may have a quick solution or daq from NI that could work as well. The bottom line is the target controls needs to be sorted out asap. Here are the link to the datasheets.
https://www.datexel.com/0-10v-to-ethernet-dat8017-v.html
https://www.datexel.com/modbus-tcp-digital-io-dat8130.html
Next week I will be out Monday and Tuesday, August  4th & 5th .
-James
James Brock
Polarized Target Group
Thomas Jefferson National Accelerator Facility
12000 Jefferson Ave. Suite 12
Newport News, VA 23606
[cid:fb9f56c1-b398-46f2-a8da-842c158f2d5d]
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Thursday, July 31, 2025 11:43 AM
To: James Brock <jbrock at jlab.org>; Pablo Campero Rojas <campero at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Tsuneo Kageya <kageya at jlab.org>
Cc: Christopher Keith <ckeith at jlab.org>; Denny Insley <dinsley at jlab.org>
Subject: Re: PRad Controls Issues
Hi James,
Tsuneo told me that you were thinking to open the PRad-II target to inspect target cell windows and the gas feed line to find out why the cell pressure is low.
Before knowing you were going to open the PRad-II target, I had asked Denny to help me lifting the target with the chain hoist nearby. He checked the area and said OK to help. Could you coordinate with him with a lifting plan (that's Hall-B work area)?
Phillip will work with you on the PRad-II target.
I'm travelling (on vacation) between 8/3 and 8/10, so I won't be able to work on the PRad-II target in the next two weeks (I also need to prepare the fusion poster and rehearsals for the DOE head and governor's visit during and after vacation.
What was conclusion of meeting with Pablo? Have you discussed the LabView features you think should be added on (or remove)?
I'll check my email every night while on vacation.
Thanks,
Xiangdong
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, July 30, 2025 10:35 AM
To: James Brock <jbrock at jlab.org>; Pablo Campero Rojas <campero at jlab.org>
Cc: Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
My email did not refresh soon enough (until I was in ESB). I might not be able to join this afternoon.
Thanks,
Xiangdong
Get Outlook for iOS<https://aka.ms/o0ukef>
________________________________
From: James Brock <jbrock at jlab.org>
Sent: Wednesday, July 30, 2025 10:09:04 AM
To: Xiangdong Wei <xwei at jlab.org>; Pablo Campero Rojas <campero at jlab.org>
Cc: Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
It will be ~1, Pablo can't make 10. Sorry Wei.
-James
Get Outlook for Android<https://aka.ms/AAb9ysg>
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Wednesday, July 30, 2025 9:31:54 AM
To: James Brock <jbrock at jlab.org>; Pablo Campero Rojas <campero at jlab.org>
Cc: Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
Hi James and Pablo,
I'm tied up with preparing a poster for the DOE secretary's visit today but will stop by the ESB for a short time around 10.
Regards,
Xiangdong
________________________________
From: James Brock <jbrock at jlab.org>
Sent: Wednesday, July 30, 2025 8:55 AM
To: Pablo Campero Rojas <campero at jlab.org>
Cc: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
Hi Pablo,
I would like meet you in the ESB at 10 and quickly go over the problem(s). Once I'm more familiar with the issue(s), together we may be able to work out an easy solution today.
Thanks,
James
________________________________
From: Pablo Campero Rojas <campero at jlab.org>
Sent: Tuesday, July 29, 2025 3:54:24 PM
To: James Brock <jbrock at jlab.org>
Cc: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: Re: PRad Controls Issues
Hello James,
The AIO com error indicator that you show in the attached screenshot  is assigned to check the comms with the 6001 USB-DAQ device  which is used to read analog and digital signals, we are aware that  the 6001 module has issues with its DI channels, as you will notice the EV1 and EV2 indicators do not change to open even when we checked that the valves are physically opened. Also, the vacuum switch or turbo pump enable status don't work, only the AI channels to read the pressure for the Bottle and Cell/Chamber are working.
We agreed to replace the 6001-USB-DAQ module by a newer version NI USB-6421with more AI channels to read the turbo pumps' speeds. The replacement 6421module was ordered but not received yet.
If you want to meet, I will be available tomorrow around 10:00.
Regards,
Pablo
________________________________
From: James Brock <jbrock at jlab.org>
Sent: Tuesday, July 29, 2025 3:07 PM
To: Pablo Campero Rojas <campero at jlab.org>
Cc: Xiangdong Wei <xwei at jlab.org>; Tsuneo Kageya <kageya at jlab.org>; Phillip Dobrenz <dobrenz at jlab.org>; Christopher Keith <ckeith at jlab.org>
Subject: PRad Controls Issues
Hi Pablo,
I tried to help Wei with the PRad II ESB Tests but was unable to do so. The LabVIEW vi you wrote for PRad II is having communication errors causing it to stop working.  The COM ERROR labeled as AIO repeatedly and randomly presents an error. I'm not sure if it's an issue is with address ports assigned to the USB to RS232 box or some latency com issues with AIO but it needs to be addressed and corrected.  The PRad test needs to be concluded for other modification to take place and time is running out. I can meet you out there to perform a quick room temp flow measurement using your program so we can identify the problem and fix it. When are you avaiable?
-James
[cid:9f1622dc-8364-412c-b97e-adca950556a0]
James Brock
Polarized Target Group
Thomas Jefferson National Accelerator Facility
12000 Jefferson Ave. Suite 12
Newport News, VA 23606
[cid:4cbe942a-4c1c-4880-a87b-0434ce299dfe]
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250806/e82997f1/attachment-0001.htm>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image.png
Type: image/png
Size: 2137659 bytes
Desc: image.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250806/e82997f1/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-jdj11rg2.png
Type: image/png
Size: 4094 bytes
Desc: Outlook-jdj11rg2.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250806/e82997f1/attachment-0005.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-myeukcsa.png
Type: image/png
Size: 4094 bytes
Desc: Outlook-myeukcsa.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250806/e82997f1/attachment-0006.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Outlook-atgkeim5.png
Type: image/png
Size: 4094 bytes
Desc: Outlook-atgkeim5.png
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_prad_ii/attachments/20250806/e82997f1/attachment-0007.png>
    
    
More information about the Dsg-hallb_prad_ii
mailing list