    <div class="moz-cite-prefix">Hi All,</div>
    <div class="moz-cite-prefix">inside the rich e-panel there are two
      air manifolds but I do not remember if we have the possibility to
      adjust the flow separately.</div>
    <div class="moz-cite-prefix">What might be interesting to test is to
      change the flow in the two manifolds separately in order to find
      the best setup that minimizes the flow towards the detector
      keeping a good safety margin on the max temperatures on the
      electronic boards.</div>
    <div class="moz-cite-prefix">May be I am dreaming...</div>
    <div class="moz-cite-prefix">George is this possible in your
      opinion?</div>
    <div class="moz-cite-prefix">Hope to see you soon <br>
    </div>
    <div class="moz-cite-prefix">Many thanks<br>
    </div>
    <div class="moz-cite-prefix">Sandro<br>
    </div>
    <div class="moz-cite-prefix"><br>
    </div>
    <div class="moz-cite-prefix">Il 24/09/2019 14:23, Tyler Lemon ha
      scritto:<br>
    </div>
    <blockquote type="cite"
cite="mid:CH2PR09MB39766D74D2C14351262D8437BA840@CH2PR09MB3976.namprd09.prod.outlook.com">
      <meta http-equiv="Content-Type" content="text/html;
        charset=windows-1252">
      <style type="text/css" style="display:none;"> P {margin-top:0;margin-bottom:0;} </style>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        Hello,</div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        From Ben and Mateo's reply, it seems the goal of the test should
        be to keep the same temperatures in RICH as previous running
        periods.</div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        Ideally, the test would involve finding what airflow settings
        are required to maintain ~65 C in RICH FPGAs. It may be that 900
        slm is actually required to keep temperatures in the appropriate
        range.<br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        At a bare minimum, the tests should involve adjusting compressor
        settings with the goal of maintaining 900 slm total flow. For
        this, RICH electronics can be left off since the airflow
        measurements would be the indicator of test results.
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        It is as Marco said, we never tried to optimize the cooling
        system. These tests would be the only way to see what is
        actually needed to cool RICH and to see whether the current
        cooling system could even provide enough flow for two RICH
        sectors.</div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        - Tyler<br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <div style="font-family: Calibri, Arial, Helvetica, sans-serif;
        font-size: 12pt; color: rgb(0, 0, 0);">
        <br>
      </div>
      <hr style="display:inline-block;width:98%" tabindex="-1">
      <div id="divRplyFwdMsg" dir="ltr"><font style="font-size:11pt"
          face="Calibri, sans-serif" color="#000000"><b>From:</b> Marco
          Mirazita <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
          <b>Sent:</b> Tuesday, September 24, 2019 4:18 AM<br>
          <b>To:</b> Amrit Yegneswaran <a class="moz-txt-link-rfc2396E" href="mailto:yeg@jlab.org"><yeg@jlab.org></a><br>
          <b>Cc:</b> Benjamin Raydo <a class="moz-txt-link-rfc2396E" href="mailto:braydo@jlab.org"><braydo@jlab.org></a>; Tyler Lemon
          <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a>; dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
          <b>Subject:</b> Re: [Dsg-rich] [EXTERNAL] RICH electronics</font>
        <div> </div>
      </div>
      <div class="BodyFragment"><font size="2"><span
            style="font-size:11pt;">
            <div class="PlainText">Hi all,<br>
              let me make a couple of remarks.<br>
              <br>
              1) The temperature limits were set to ensure a safe
              operation of the <br>
              whole system, which, as Ben said, includes not just the
              FPGA, but <br>
              cables, MAROC, external temperature close to TOF, etc. I
              don't want to <br>
              change these limits unless strictly necessary and only if
              with no risk <br>
              for the detectors.<br>
              However, we know that the current running conditions are
              mostly driven <br>
              by one FPGA which is producing more heat (its temperature
              is about 2 or <br>
              3 degrees higher than the neighbors) and by chance it is
              also in the <br>
              hottest spot of the panel.<br>
              <br>
              2) The cooling system was originally designed to be able
              to feed two <br>
              RICH modules with one compressor, in case of failure of
              the other one. <br>
              We are now far from this situation. We run one RICH with
              900 slm with a <br>
              maximum of 1200 slm per compressor. But we never attempted
              a real <br>
              optimization of the air flow. For example, the lower line
              is flowing <br>
              less air than the upper one, while I would expect the
              opposite.<br>
              <br>
              For these reasons, I think there could be room to optimize
              the running <br>
              conditions and get closer to the original design.<br>
              But let me stress again that this optimization must ensure
              to stay <br>
              within the safety limits.<br>
              <br>
              Best regards,<br>
              Marco<br>
              <br>
              <br>
              <br>
              <br>
              Il 2019-09-24 04:12 Amrit Yegneswaran ha scritto:<br>
              > given ben's last sentence<br>
              > <br>
              >  (highlighted; i guess "it" in the sentence refers to
              "the<br>
              > temperature" as does "that".<br>
              >  from which we have it = that => i = hat or i
              =tha, aah the pronouns<br>
              > they are precious!)<br>
              > <br>
              >  anyway i'm now very confused about the planned test
              and the intent<br>
              > thereof?<br>
              > <br>
              >  are we thinking reducing the airflow from ~950 slm
              will reduce the<br>
              > temperature below 65c?<br>
              >  pardon my obtuseness.<br>
              > <br>
              >  so what mods are you conjuring up tyler?<br>
              > <br>
              > -------------------------<br>
              > <br>
              > FROM: Benjamin Raydo <a class="moz-txt-link-rfc2396E" href="mailto:braydo@jlab.org"><braydo@jlab.org></a><br>
              >  SENT: Monday, September 23, 2019 5:33 PM<br>
              >  TO: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a>; Amrit
              Yegneswaran <a class="moz-txt-link-rfc2396E" href="mailto:yeg@jlab.org"><yeg@jlab.org></a>;<br>
              > Marco Mirazita <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
              >  SUBJECT: Re: [Dsg-rich] [EXTERNAL] RICH electronics<br>
              > <br>
              >  Hi Tyler,<br>
              > <br>
              >  Matteo is correct about the temperature limit of 85C
              to respect the<br>
              > FPGA timing constraints, but we also have fiber
              cables running close<br>
              > to those parts that are rated for 80C (I'd expect
              some deformation<br>
              > possible if we go above this for a long period).<br>
              > <br>
              >  Can we set a safe limit of 75C for any reported FPGA
              temperature for<br>
              > your tests (short term limit: <1hr)? If your tests
              need to run in such<br>
              > a way that you want a higher limit, let us discuss
              again if you don't<br>
              > mind. For long term it would be great to keep it as
              low as possible<br>
              > and for the moment I believe we've been running with
              the hottest FPGA<br>
              > temp to be around 65C so don't want to see that
              increase significantly<br>
              > if there are any changes to the cooling.<br>
              > <br>
              >  Hope that helps.<br>
              > <br>
              >  Thanks,<br>
              >  Ben<br>
              > <br>
              > -------------------------<br>
              > <br>
              > FROM: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a><br>
              >  SENT: Monday, September 23, 2019 4:59 PM<br>
              >  TO: Amrit Yegneswaran <a class="moz-txt-link-rfc2396E" href="mailto:yeg@jlab.org"><yeg@jlab.org></a>; Marco
              Mirazita<br>
              > <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a>; Benjamin Raydo
              <a class="moz-txt-link-rfc2396E" href="mailto:braydo@jlab.org"><braydo@jlab.org></a><br>
              >  CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
              >  SUBJECT: Re: [Dsg-rich] [EXTERNAL] RICH electronics<br>
              > <br>
              >  Hi Ben,<br>
              > <br>
              >  In October, we're planning on doing some testing to
              optimize RICH's<br>
              > cooling system and see what would needed to modify
              the system to cool<br>
              > two RICH sectors.<br>
              > <br>
              >  With respect to the RICH readout electronics, FPGAs,
              and their<br>
              > timing, are there any temperature limits we should
              take into<br>
              > consideration when testing the cooling system?<br>
              > <br>
              >  Thanks,<br>
              >  Tyler<br>
              > <br>
              > -------------------------<br>
              > <br>
              > FROM: Amrit Yegneswaran <a class="moz-txt-link-rfc2396E" href="mailto:yeg@jlab.org"><yeg@jlab.org></a><br>
              >  SENT: Monday, September 23, 2019 4:48 PM<br>
              >  TO: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a>; Marco
              Mirazita<br>
              > <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
              >  SUBJECT: Re: [Dsg-rich] [EXTERNAL] RICH electronics<br>
              > <br>
              >  tyler<br>
              >  address fpga temp issue with ben.<br>
              >  i believe 60c is right at the edge of timing issues,
              not sure.<br>
              > -------------------------<br>
              > <br>
              > FROM: Dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich-bounces@jlab.org"><dsg-rich-bounces@jlab.org></a> on
              behalf of Tyler Lemon<br>
              > <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a><br>
              >  SENT: Monday, September 23, 2019 3:11 PM<br>
              >  TO: Marco Mirazita
              <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
              >  SUBJECT: Re: [Dsg-rich] [EXTERNAL] RICH electronics<br>
              > <br>
              >  Hi Marco,<br>
              > <br>
              >  Looking back at EP temperatures during steady
              operations (I picked<br>
              > 3/10/2019 to 3/31/2019), Hardware Interlock RTDs were
              ~40 deg C and<br>
              > FPGA temperatures were ~60 deg C.<br>
              > <br>
              > Is there a temperature increase that would be
              considered acceptable?<br>
              > For example if the FPGA temperatures were ~70 deg C
              rather than ~60<br>
              > deg C?<br>
              > <br>
              >  The first test should be determining what flow rate
              is actually<br>
              > needed to maintain EP temperatures in that acceptable
              range with all<br>
              > RICH electronics on.<br>
              > <br>
              >  -Tyler<br>
              > <br>
              > -------------------------<br>
              > <br>
              > FROM: Marco Mirazita
              <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  SENT: Monday, September 23, 2019 9:01 AM<br>
              >  TO: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a><br>
              >  CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a><br>
              >  SUBJECT: Re: [Dsg-rich] [EXTERNAL] RICH electronics<br>
              > <br>
              > Hi Tyler,<br>
              >  sorry, I had in mind that the replacement of the PT
              was scheduled for<br>
              > <br>
              >  last Wednesday.<br>
              > <br>
              >  The tests we have in mind to optimize the cooling
              system is basically<br>
              > <br>
              >  what you said. But we still have to make a plan of
              the specific<br>
              >  measurements we need. We will do this plan in the
              next weeks.<br>
              >  Any suggestion from you and the DSG is welcome.<br>
              > <br>
              >  Best regards,<br>
              >  Marco<br>
              > <br>
              >  Il 2019-09-20 20:46 Tyler Lemon ha scritto:<br>
              >  > Hi Marco,<br>
              >  ><br>
              >  > The PT replacement has not been completed. It
              is scheduled for next<br>
              >  > Wednesday September 25.<br>
              >  ><br>
              >  > As for the visit to optimize RICH running
              conditions, I will not be<br>
              >  > at work for at least part of the week of
              October 21 to October 25,<br>
              >  > however there should be other people from DSG
              that would be able to<br>
              >  > help.<br>
              >  ><br>
              >  > What tests are being planned for the visit?
              Would it essentially be<br>
              >  > varying the settings of the cooling system
              (flow rate, buffer tank<br>
              >  > pressure, compressor output pressure) and
              seeing how the<br>
              > temperature<br>
              >  > within RICH is affected?<br>
              >  ><br>
              >  > Best regards,<br>
              >  > Tyler<br>
              >  ><br>
              >  > -------------------------<br>
              >  ><br>
              >  > FROM: Marco Mirazita
              <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  > SENT: Friday, September 20, 2019 12:21 PM<br>
              >  > TO: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a><br>
              >  > CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a>; Valery
              Kubarovsky <a class="moz-txt-link-rfc2396E" href="mailto:vpk@jlab.org"><vpk@jlab.org></a>;<br>
              >  > marco contalbrigo <a class="moz-txt-link-rfc2396E" href="mailto:mcontalb@fe.infn.it"><mcontalb@fe.infn.it></a><br>
              >  > SUBJECT: Re: [EXTERNAL] RICH electronics<br>
              >  ><br>
              >  > Hi Tyler,<br>
              >  > do you know if Bob has completed the
              replacement of the FT?<br>
              >  > If so, the compressor could be restarted on
              Monday morning, once we<br>
              >  > are<br>
              >  > sure everything is working properly I can turn
              on the RICH<br>
              >  > electronics<br>
              >  > and keep it on for the night and then make the
              tests on Tuesday<br>
              >  > morning.<br>
              >  ><br>
              >  > As you know, we didn't have much time to
              optimize the working point<br>
              >  > of<br>
              >  > the cooling system after the RICH installation,
              we simply choose<br>
              >  > running<br>
              >  > conditions that ensure a safe operation of the
              detector.<br>
              >  > But these running conditions might be
              problematic with two RICH<br>
              >  > modules<br>
              >  > and definitely not sustainable in case of
              failure of one of the two<br>
              >  > compressors.<br>
              >  > Therefore, also in preparation of the upcoming
              CLAS12 data taking,<br>
              >  > I'm<br>
              >  > planning a visit to JLab for Dario and Sandro
              to try to optimize<br>
              > the<br>
              >  > running conditions, and to be effective this
              work must be<br>
              > coordinated<br>
              >  ><br>
              >  > with you and the DSG.<br>
              >  > The proposed dates are from October 21 to
              October 31. Is this plan<br>
              > ok<br>
              >  ><br>
              >  > for you?<br>
              >  > Thank you,<br>
              >  > Marco<br>
              >  ><br>
              >  > Il 2019-09-18 14:12 Tyler Lemon ha scritto:<br>
              >  > > Hi Marco,<br>
              >  > ><br>
              >  > > Bob Miller has scheduled the replacement
              of the faulty pressure<br>
              >  > > transducer (PT) on the air-cooling buffer
              tank for next<br>
              > Wednesday,<br>
              >  > > September 25.<br>
              >  > ><br>
              >  > > Could the tests be scheduled to start
              Thursday September 26?<br>
              >  > ><br>
              >  > > Best regards,<br>
              >  > > Tyler<br>
              >  > ><br>
              >  > > -------------------------<br>
              >  > ><br>
              >  > > FROM: Marco Mirazita
              <a class="moz-txt-link-rfc2396E" href="mailto:Marco.Mirazita@lnf.infn.it"><Marco.Mirazita@lnf.infn.it></a><br>
              >  > > SENT: Wednesday, September 18, 2019 4:08
              AM<br>
              >  > > TO: Tyler Lemon <a class="moz-txt-link-rfc2396E" href="mailto:tlemon@jlab.org"><tlemon@jlab.org></a><br>
              >  > > CC: dsg-rich <a class="moz-txt-link-rfc2396E" href="mailto:dsg-rich@jlab.org"><dsg-rich@jlab.org></a>;
              Valery Kubarovsky<br>
              > <a class="moz-txt-link-rfc2396E" href="mailto:vpk@jlab.org"><vpk@jlab.org></a>;<br>
              >  > > marco contalbrigo
              <a class="moz-txt-link-rfc2396E" href="mailto:mcontalb@fe.infn.it"><mcontalb@fe.infn.it></a><br>
              >  > > SUBJECT: [EXTERNAL] RICH electronics<br>
              >  > ><br>
              >  > > Hi Tyler,<br>
              >  > > the RICH electronics is off since quite a
              lot of time and in<br>
              >  > > preparation<br>
              >  > > of the upcoming CLAS12 running I would
              like to turn it on to<br>
              > verify<br>
              >  > > that<br>
              >  > > everything is still working and eventually
              to take new scaler<br>
              >  > > calibration data.<br>
              >  > > I could plan this work for the beginning
              of the next week.<br>
              >  > > In preparation for that, we should
              coordinate to have all the<br>
              >  > systems<br>
              >  > ><br>
              >  > > ready. Would you be available to start the
              compressor for the<br>
              > test?<br>
              >  > > Thank you,<br>
              >  > > Marco<br>
              >  ><br>
              >  > _______________________________________________<br>
              >  > Dsg-rich mailing list<br>
              >  > <a class="moz-txt-link-abbreviated" href="mailto:Dsg-rich@jlab.org">Dsg-rich@jlab.org</a><br>
              >  > <a
                href="https://mailman.jlab.org/mailman/listinfo/dsg-rich"
                moz-do-not-send="true">https://mailman.jlab.org/mailman/listinfo/dsg-rich</a>
              [1]<br>
              > <br>
              > <br>
              > Links:<br>
              > ------<br>
              > [1] <a
                href="https://mailman.jlab.org/mailman/listinfo/dsg-rich"
                moz-do-not-send="true">https://mailman.jlab.org/mailman/listinfo/dsg-rich</a><br>
            </div>
          </span></font></div>
      <br>
      <fieldset class="mimeAttachmentHeader"></fieldset>
      <pre class="moz-quote-pre" wrap="">_______________________________________________
Dsg-rich mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Dsg-rich@jlab.org">Dsg-rich@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/dsg-rich">https://mailman.jlab.org/mailman/listinfo/dsg-rich</a>
</pre>
    </blockquote>
    <p><br>
    </p>
    <pre class="moz-signature" cols="72">-- 
-----------------------------------------
Dr. Eng. Sandro Tomassini
Istituto Nazionale di Fisica Nucleare
Laboratori Nazionali di Frascati
---Research Division---
P.O.Box 13 00044 Frascati (RM) Italy
Tel. office: +39 06 9403 2919/8089
Fax: +39 06 94032256
e-mail: <a class="moz-txt-link-abbreviated" href="mailto:sandro.tomassini@lnf.infn.it">sandro.tomassini@lnf.infn.it</a>
<a class="moz-txt-link-freetext" href="https://urldefense.proofpoint.com/v2/url?u=http-3A__www.lnf.infn.it_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=sLmN_kLNB8f_QBCYTgZaO1cK1zcpXSUHz0BQf_04bYk&m=Ks_koQB8zfdhAZpqWi1ulFQVWFGW3EuWLeBlPjTyTDs&s=sUK8i5XKeQizqnO1zVv4v1N6p7Mr7ouujo4N8w-3gRQ&e=">http://www.lnf.infn.it/</a>
-----------------------------------------