[Dsg-rtpc] RTPC gas system

Narbe Kalantarians narbe at jlab.org
Fri Mar 23 11:35:18 EDT 2018


Hi Sebastian,

I'm checking to see what I can find.

- Narbe

----- Original Message -----
From: "kuhn" <kuhn at jlab.org>
To: "George Jacobs" <jacobsg at jlab.org>
Cc: dsg-rtpc at jlab.org
Sent: Friday, March 23, 2018 11:29:46 AM
Subject: Re: [Dsg-rtpc] RTPC gas system

O.k., so I assume that the pressure sensor are (or will be) taken care of by you guys (with W&M and Narbe). 
About the temperature sensors: Again, it doesn’t have to be super-precise (up to 5 degrees F relative), and it doesn’t have to cover a huge range (50 - 100 F), but it MUST be compact and ideally need only 2 (or 3) leads to be brought out of the RTPC. I agree that it most likely will just measure ambient temperature, but it should be the temperature of the RTPC itself, not somewhere “roughly in the vicinity”. I could start digging for a sensor like that, but frankly I was hoping that the gas system experts in the BONuS collaboration can do the sleuthing.

- Sebastian

> On Mar 21, 2018, at 9:47 AM, George Jacobs <jacobsg at jlab.org> wrote:
> 
> Hi Sebastian,
> 
> - The pressure sensors, 1 each of an absolute pressure transducer and 1 each of a differential pressure transducer. These sensors connect to the gas volume via tubing and therefore do not take up space inside the detector volume. The tubing connection locations must be determined along with the method of the connection. The sensors themselves, should be located outside the magnetic field. These transducers are shown in the P&I diagram, PT1 and PT2. These are connected via the blue lines, tubing, to the RTPC and DMS.
> 
> - The temperature sensor, first we need to define what temperature to monitor. If it is the temperature of the gas, then we must define the location where the sensor measures the gas temp. With the flow rates intended, <100sccm, this will most likely be a measurement of ambient temperature. We must also determine the type of sensor to use, since the standard ones are too large, along with the location of the sensor and it's connection to the readout.
> 
> Cheers,
> George
> 
> On 3/20/2018 4:55 PM, kuhn wrote:
>> My only concern is about the sensors - everything else seems to be in good hands. To reiterate, here is the dream wish list:
>> 
>> 1 combination sensor (or small separate sensors) for ABSOLUTE PRESSURE and for Temperature. The required range for both is very small (50 - 100 F and 0.9 to 1.1 atm) but they have to be tiny to fit inside the RTPC. (We need two of each so we can also monitor the DMS, although there space is not quite as much at a premium. We MIGHT get around the pressure sensor for the RTPC if instead we have an extra (tight) gas line that attaches to one end of a differential pressure meter, with the other end attached to the DMS. We must know the absolute pressure in both to within 0.01 (0.14 psi) if at all possible, while the temperature is less sensitive (3 degrees Celsius or 5 F). We can accomplish this either by having an absolute pressure meter at at least one of these two places, or relative pressure sensors combined with a Hall barometer, or some combination thereof with a differential pressure sensor (probably most sensitive). 
>> I hope the gas group can look into possible sensors for this that are available on the market. The ones suggested by George look way too big, at least for the RTPC. (I think for the temperature sensor, a simple T-dependent resistor or diode should be good enough!)
>> 
>> Thanks - Sebastian
>> 
>> 
>>> On Mar 20, 2018, at 4:47 PM, George Jacobs <jacobsg at jlab.org>
>>>  wrote:
>>> 
>>> Hi Carlos,
>>> 
>>> The misc swagelock hardware includes valves, fittings, adapters, ferrules, etc. 
>>> 
>>> The mass flow controllers have the longest lead times and should be ordered ASAP. A minimum of 2 mass flow controllers should be ordered in order to keep our options between series and parallel DMS setup.
>>> 
>>> As far as the temp sensor, would one of the following work?
>>> 
>>> https://www.mcmaster.com/#6467t24/=1c2110g
>>> 
>>> or
>>> 
>>> https://www.mcmaster.com/#3866k43/=1c211mt
>>> 
>>> 
>>> Cheers,
>>> George
>>> 
>>> On 3/20/2018 3:05 PM, Carlos Ayerbe wrote:
>>> 
>>>> Hi George, Marc and Brian and all, 
>>>> 
>>>> I got noticed from Keith, that we can spend some money on the acquisition of several components for the gas system. 
>>>> 
>>>> I was checking and these attachments correspond to the latest configuration. I noticed that the excel file is not updated according to the scheme (just some discrepancy numbers, like how many MFC items are counted). 
>>>> 
>>>> There are a few items I would like to understand:
>>>> - the temperature probe is under interrogation mark, do we have any reference about it? In other words, do we know what do we want? 
>>>> - the item 'misc Swagelock hardware' is $2,500 but I don't know what are w talking about, valves, hoses, regulators, gauges? 
>>>> 
>>>> I would like to confirm with Sebastian (and the others) that the scheme is more or less the starting point to begin building the system, so we can discuss the exact number of components to buy as soon as possible. So, in accordance, update the excel file. 
>>>> 
>>>> Thank you 
>>>> 
>>>> Cheers,
>>>> 
>>>> -Carlos
>>>> 
>>>> 
>>>> ----- Original Message -----
>>>> From: "George Jacobs" 
>>>> 
>>>> <jacobsg at jlab.org>
>>>> 
>>>> 
>>>> To: 
>>>> 
>>>> dsg-rtpc at jlab.org, "kuhn" <kuhn at jlab.org>
>>>> 
>>>> 
>>>> Sent: Monday, February 12, 2018 9:25:58 AM
>>>> Subject: Re: [Dsg-rtpc] RTPC gas system
>>>> 
>>>> Hi Sebastian, 
>>>> 
>>>> Please find attached P&I diagram along with the components list for Series configuration. The list is also valid for the Parallel configuration, so there are options if needed. 
>>>> 
>>>> According to the markings on the deck, the magnetic field extends out slightly beyond the cart. Not sure if any part of it is free of the magnetic field. Maybe the field is low enough all the way outboard at floor level? This may require longer and larger ID gas lines. 
>>>> 
>>>> Cheers, 
>>>> George 
>>>> 
>>>> 
>>>> 
>>>> 
>>>> _______________________________________________
>>>> Dsg-rtpc mailing list
>>>> 
>>>> 
>>>> Dsg-rtpc at jlab.org
>>>> https://mailman.jlab.org/mailman/listinfo/dsg-rtpc
>>> -- 
>>> George Jacobs
>>> Jefferson Lab (TJNAF)
>>> STE 12
>>> 12000 Jefferson Ave.
>>> Newport News, VA 23606
>>> 
>>> (office) 757-269-7115
>>> 
>>> (cell)   757-876-0480
>>> 
>>> (email) 
>>> 
>>> jacobsg at jlab.org
>>> 
>>> 
>>> 
>>> (website) 
>>> 
>>> https://userweb.jlab.org/~jacobsg
>>> 
>>> _______________________________________________
>>> Dsg-rtpc mailing list
>>> 
>>> Dsg-rtpc at jlab.org
>>> https://mailman.jlab.org/mailman/listinfo/dsg-rtpc
>> 
>> _______________________________________________
>> Dsg-rtpc mailing list
>> 
>> Dsg-rtpc at jlab.org
>> https://mailman.jlab.org/mailman/listinfo/dsg-rtpc
> 
> -- 
> George Jacobs
> Jefferson Lab (TJNAF)
> STE 12
> 12000 Jefferson Ave.
> Newport News, VA 23606
> 
> (office) 757-269-7115
> 
> (cell)   757-876-0480
> 
> (email) 
> jacobsg at jlab.org
> 
> 
> (website) 
> https://userweb.jlab.org/~jacobsg


_______________________________________________
Dsg-rtpc mailing list
Dsg-rtpc at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-rtpc




More information about the Dsg-rtpc mailing list