[Dsg-rtpc] [EXTERNAL] Gas fittings for the BONuS12 DMS
Sebastian Kuhn
kuhn at jlab.org
Fri Jan 3 14:34:51 EST 2020
Dear Brian,
I realize that this is very late in the game. I agree that if we can’t get it done in time, we will (have to) live with the present arrangement. The reason that I would prefer a differential pressure meter (as opposed to dead-reckoning with absolute pressure measurements) is that the this pressure difference between RTPC and 4He buffer is directly relevant for the strain on the thin ground foil, as well as potentially the leak rate from the detector.
I am cc’ing Narbe and Carlos in case they are aware of any still existing differential pressure meters that we could use; in any case, to be useful, the meter should have the following parameters:
1) Sensitivity/accuracy: Ideally about 0.013*10^-3 atm or 0.005 iwc or 0.0002 psi or 0.01 Torr (can be a bit less accurate)
2) Range: At least +/-1.3*10^-3 atm or 0.5 iwc or 0.02 psi or 1 Torr (it doesn’t hurt if it is more, obviously, as long as it is sensitive enough).
I believe these specs are compatible with the differential pressure sensor we are presently using for the RTPC - DMS sensor PDT332 (MKS226A).
- Sebastian
> On Jan 3, 2020, at 1:58 PM, Brian Eng <beng at jlab.org> wrote:
>
> Hi Sebastian,
>
> Regarding this point, The Hall B Gas System already has a sensor/PV that is reading the absolute pressure in the Hall itself (i.e. ambient). This can be used to convert any absolute pressure to a relative one.
>
> So we already have enough sensors that we could calculate this value (DP between RTPC & He Buffer), but I don't know what kind of accuracy/speed you require.
>
> If you require something with higher accuracy we would need to know what pressure range you expected to see if there is any sensors available in-hand that would work. The lead-time on these is long enough that we wouldn't get it in time for the experiment.
>
>> On Jan 3, 2020, at 11:58 AM, Sebastian Kuhn <kuhn at jlab.org> wrote:
>>
>> 2) Is there a possibility to add one more item to the gas panel? (rather belated, I know - my apologies) I realized that we really would like to be able to measure the differential pressure between the 4He gas buffer volume (orange in the gas panel design drawings) and the RTPC (green). This would
>> a) close the loop so we have both absolute and relative pressure for all relevant parts of the system
>> b) allow us to monitor (and alarm on) any pressure differential between the RTPC and the buffer volume that exceeds the (rather limited) safety margin for the thin ground foil between the two.
>
> _______________________________________________
> Dsg-rtpc mailing list
> Dsg-rtpc at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-rtpc
More information about the Dsg-rtpc