Hi Carlos,

I’m wondering whether that diff pressure sensor is not already used - we changed the sensor for the 4He buffer volume to measure the differential pressure with the surrounding air, so there are already 2 diff press sensors “in service” in the panel as of now. If there is one more in the EEL, would you know where to look for it?

Alternatively, we could (with minimal rerouting) use THAT diff press. sensor to connect it to the RTPC witness line instead of leaving it open to air. That would minimize the amount of work required, no more additional sensors needed (lowering hardware and software cost), and we still would have the full information by comparing the absolute pressure in the RTPC with the absolute pressure measured in the Hall, while having the highest sensitivity where it counts - the differential pressure across the ground foil. How does that sound?

- Sebastian

On Jan 3, 2020, at 6:45 PM, Carlos Ayerbe Gayoso <gayoso at jlab.org<mailto:gayoso at jlab.org>> wrote:

Hello All and Happy New Year,

there is a differential pressure sensor available (the original one used in the panel), BUT I don't know if it is enough accurate. There are two Honeywell 19C015PA4K pressure transducer available, but they were discarded due to the accuracy.

The differential is still at EEL, the transducer are at W&M.

Cheers

-Carlos
________________________________
From: Sebastian Kuhn <kuhn at jlab.org<mailto:kuhn at jlab.org>>
Sent: Friday, January 3, 2020 2:34 PM
To: dsg-rtpc at jlab.org<mailto:dsg-rtpc at jlab.org> <dsg-rtpc at jlab.org<mailto:dsg-rtpc at jlab.org>>
Cc: JIWAN POUDEL <jpoud001 at odu.edu<mailto:jpoud001 at odu.edu>>; Bob Miller <rmiller at jlab.org<mailto:rmiller at jlab.org>>; Mohammad Hattawy <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>>; Nathan Baltzell <nbaltzel at odu.edu<mailto:nbaltzel at odu.edu>>; Carlos Ayerbe Gayoso <gayoso at jlab.org<mailto:gayoso at jlab.org>>; Narbe Kalantarians <narbe at jlab.org<mailto:narbe at jlab.org>>
Subject: Re: [Dsg-rtpc] [EXTERNAL] Gas fittings for the BONuS12 DMS

Dear Brian,

I realize that this is very late in the game. I agree that if we can’t get it done in time, we will (have to) live with the present arrangement. The reason that I would prefer a differential pressure meter (as opposed to dead-reckoning with absolute pressure measurements) is that the this pressure difference between RTPC and 4He buffer is directly relevant for the strain on the thin ground foil, as well as potentially the leak rate from the detector.

I am cc’ing Narbe and Carlos in case they are aware of any still existing differential pressure meters that we could use; in any case, to be useful, the meter should have the following parameters:

1) Sensitivity/accuracy: Ideally about 0.013*10^-3 atm or 0.005 iwc or 0.0002 psi or 0.01 Torr (can be a bit less accurate)
2) Range: At least +/-1.3*10^-3 atm or 0.5 iwc or 0.02 psi or 1 Torr (it doesn’t hurt if it is more, obviously, as long as it is sensitive enough).
I believe these specs are compatible with the differential pressure sensor we are presently using for the RTPC - DMS sensor PDT332 (MKS226A).

- Sebastian


> On Jan 3, 2020, at 1:58 PM, Brian Eng <beng at jlab.org<mailto:beng at jlab.org>> wrote:
>
> Hi Sebastian,
>
> Regarding this point, The Hall B Gas System already has a sensor/PV that is reading the absolute pressure in the Hall itself (i.e. ambient). This can be used to convert any absolute pressure to a relative one.
>
> So we already have enough sensors that we could calculate this value (DP between RTPC & He Buffer), but I don't know what kind of accuracy/speed you require.
>
> If you require something with higher accuracy we would need to know what pressure range you expected to see if there is any sensors available in-hand that would work. The lead-time on these is long enough that we wouldn't get it in time for the experiment.
>
>> On Jan 3, 2020, at 11:58 AM, Sebastian Kuhn <kuhn at jlab.org<mailto:kuhn at jlab.org>> wrote:
>>
>> 2) Is there a possibility to add one more item to the gas panel? (rather belated, I know - my apologies) I realized that we really would like to be able to measure the differential pressure between the 4He gas buffer volume  (orange in the gas  panel design drawings) and the RTPC (green). This would
>> a) close the loop so we have both absolute and relative pressure for all relevant parts of the system
>> b) allow us to monitor (and alarm on) any pressure differential between the RTPC and the buffer volume that exceeds the (rather limited) safety margin for the thin ground foil between the two.
>
> _______________________________________________
> Dsg-rtpc mailing list
> Dsg-rtpc at jlab.org<mailto:Dsg-rtpc at jlab.org>
> https://mailman.jlab.org/mailman/listinfo/dsg-rtpc

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rtpc/attachments/20200104/b752422a/attachment-0002.html>


More information about the Dsg-rtpc mailing list