[dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC

Eugene Pasyuk pasyuk at jlab.org
Tue Jul 28 15:28:47 EDT 2020


Thanks Marc!
Mohammad, is there anything else you need to do in the hall?

Thanks,

-Eugene

From: Marc McMullen <mcmullen at jlab.org>
Date: Tuesday, July 28, 2020 at 3:26 PM
To: Eugene Pasyuk <pasyuk at jlab.org>, Jiwan Poudel <jpoudel at jlab.org>, Sebastian Kuhn <skuhn at odu.edu>, Bob Miller <rmiller at jlab.org>, Denny Insley <dinsley at jlab.org>, "dsg-hallb_rtpc at jlab.org" <dsg-hallb_rtpc at jlab.org>
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC

Hi Eugene,

I have completed all the RTPC pressure transducer work in the hall.

Regards,
Marc
________________________________
From: Eugene Pasyuk <pasyuk at jlab.org>
Sent: Tuesday, July 28, 2020 1:55 PM
To: Marc Mcmullen <mcmullen at jlab.org>; Jiwan Poudel <jpoudel at jlab.org>; Kuhn, Sebastian E. <skuhn at odu.edu>; Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; dsg-hallb_rtpc at jlab.org <dsg-hallb_rtpc at jlab.org>
Subject: RE: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC


Just got a message from Bob. The beam line work is done. Please do whatever you need to do and let me know when you are done.



-Eugene



From: Marc Mcmullen <mcmullen at jlab.org>
Sent: Tuesday, July 28, 2020 12:51
To: Jiwan Poudel <jpoudel at jlab.org>; Eugene Pasyuk <pasyuk at jlab.org>; Kuhn, Sebastian E. <skuhn at odu.edu>; Bob Miller <rmiller at jlab.org>; Denny Insley <dinsley at jlab.org>; dsg-hallb_rtpc at jlab.org
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



Mohammad,



I will be at the lab later to confirm the wiring connection.



Marc

________________________________

From: dsg-hallb_rtpc <dsg-hallb_rtpc-bounces at jlab.org<mailto:dsg-hallb_rtpc-bounces at jlab.org>> on behalf of mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>>
Sent: Tuesday, July 28, 2020 12:48 PM
To: Jiwan Poudel <jpoudel at jlab.org<mailto:jpoudel at jlab.org>>; Eugene Pasyuk <pasyuk at jlab.org<mailto:pasyuk at jlab.org>>; Kuhn, Sebastian E. <skuhn at odu.edu<mailto:skuhn at odu.edu>>; Bob Miller <rmiller at jlab.org<mailto:rmiller at jlab.org>>; Denny Insley <dinsley at jlab.org<mailto:dinsley at jlab.org>>
Cc: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



Dear Marc and DSG members,

   We just heard an update from Marco Battaglieri that we may get the beam to the Hall by this weekend. Bob Miller will be putting the beamline components back. I would like to ask you kindly to help with fixing the pressure transducer as soon as you can.



Eugene, Please don't start sweeping the magnets  till I confirm that we are done with this fix.



Best regards,

Mohammad.



On Tue, Jul 28, 2020 at 10:51 AM mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>> wrote:

If you could not open the picture that Jiwan attached to his replay, here it is.



Mohammad.



On Tue, Jul 28, 2020 at 10:48 AM mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>> wrote:

Hello Marc,

 Following our phone call this morning, It has to be what you call PDT2.



Mohammad.



On Tue, Jul 28, 2020 at 10:45 AM Jiwan Poudel <jpoudel at jlab.org<mailto:jpoudel at jlab.org>> wrote:

Hello all,



We have two differential transducer (PDTs): RTPC-DMS (PDT332) and RTPC-Buffer(PDT208). So I am taking about RTPC-DMS pressure transducer. I have only this photo right now, where you can see one two wires are coming out from PDT.



Thank you,

Jiwan



[Image removed by sender.]



On Jul 28, 2020, at 10:19 AM, mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>> wrote:



Hello Jiwan,

   Sounds like the DSG group is not fully sure about the specification of this transducer as they used to name it PT2. Can you please send them some pics of the transducer as the ones I did for the other one.



Mohammad





On Tue, Jul 28, 2020 at 10:04 AM mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>> wrote:

Dear DSG members,

    Please send us the cables color code for PDT332 to connect it back.



Regards,

Mohammad.



On Mon, Jul 27, 2020 at 20:37 <jpoudel at jlab.org<mailto:jpoudel at jlab.org>> wrote:

I am sorry to write again, but I have to inform to you all correctly. I wanted to write "cable from transducer pt332 is not attached to chassis but GUI is showing a large negative value. How?" Previus email was confusing (isn't is changed to is) so have to write again.





On Jul 27, 2020 20:09, jpoudel at jlab.org<mailto:jpoudel at jlab.org> wrote:

PDT208 is working fine, the connection is good. I verified connection and connected it to chassis. Its showing quite reasonable reading also.



PDT332( RTPC-DMS diff) is showing some weird number in epics GUI even before its connected to system. Cable from transducer is attached at chassis. How? I have no idea.( see attached) I think it should show zero. I compare PDT208(rtpc-buffer) reading removing cable from chassis, and it shows zero.



In addition, Connecting wires at transducer PDT332 are coming out. So we need color code of cables before connecting it.



Thank you,

Jiwan







On Jul 27, 2020 19:18, jpoudel at jlab.org<mailto:jpoudel at jlab.org> wrote:

Hello Marc,



Can you please send me the connection of PDT 332 (color code of cables) also?



In PDT 208, white and green are together. Are they -15V?



Thank you,

Jiwan



On Jul 27, 2020 18:40, JIWAN POUDEL <jpoud001 at odu.edu<mailto:jpoud001 at odu.edu>> wrote:

I will check the connections and verify it is correct before plugging in to chassis.



Thank you,

Jiwan



On Jul 27, 2020, at 6:22 PM, mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>> wrote:



Thanks Marc,

   Maybe the beacon was on, but for sure the magnets are off. Unfortunately, I can’t get to the Lab till Wednesday. Jiwan can you connect it back?



Mohammad



On Mon, Jul 27, 2020 at 18:18 Marc Mcmullen <mcmullen at jlab.org<mailto:mcmullen at jlab.org>> wrote:

Hi Mohammad and Jiwan,



I have replaced the fuse in the chassis and checked the internal components for issues. I did not reconnect the connector for the PDT 208 (differential pressure DMS to RTPC) because I could not verify if the contacts are wired correctly since it was reconnected, because the magnet beacon is on. I have attached a pinout which can be used to verify connections prior to plugging it back into the interface chassis.



Regards,

Marc

________________________________

From: Marc Mcmullen <mcmullen at jlab.org<mailto:mcmullen at jlab.org>>
Sent: Friday, July 24, 2020 11:18 AM

To: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



Thanks Mohammad,



I'll look into the spares first. Then contact you about the repair.



Regards,

Marc

________________________________

From: dsg-hallb_rtpc <dsg-hallb_rtpc-bounces at jlab.org<mailto:dsg-hallb_rtpc-bounces at jlab.org>> on behalf of mhattawy at odu.edu<mailto:mhattawy at odu.edu> <mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>>
Sent: Friday, July 24, 2020 11:16 AM
To: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



Hello Marc,

    The magnets will be energized starting from 3:30 pm today to Monday morning. You won’t be able to get to the gas panel, that is within the fields of the magnets.





Mohammad



On Fri, Jul 24, 2020 at 11:11 Marc Mcmullen <mcmullen at jlab.org<mailto:mcmullen at jlab.org>> wrote:

Hi Jiwan,



I am going to see if we have some spare supplies for that chassis. When I have the swap out in  hand, I'll need to coordinate with your group to complete the repair.



The chassis will have to be de-energized and removed from the rack for repair. This means the MFC and the other connected sensors will have to be shut down while I make the repairs.



Additionally, I will be doing this work either on the weekend or after hours, to avoid coming into contact with others.



I'll let you know the status of the spares sometime today, so we can coordinate the system outage.



Regards,

Marc

________________________________

From: dsg-hallb_rtpc <dsg-hallb_rtpc-bounces at jlab.org<mailto:dsg-hallb_rtpc-bounces at jlab.org>> on behalf of Jiwan Poudel <jpoudel at jlab.org<mailto:jpoudel at jlab.org>>
Sent: Friday, July 24, 2020 9:33 AM
To: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: Re: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



Hello Brian,



Those power pins are not showing what it should show. Pin 6 and 5 showing -28.7V and pins 7 and 5 showing only 1.3V.



Thank you,

Jiwan



On Jul 23, 2020, at 2:29 PM, Brian Eng <beng at jlab.org<mailto:beng at jlab.org>> wrote:



Thanks Mohammad,



Fortunately it already has the pinout on it.



Unplug that connector and measure between pins 6 & 5 for -15 VDC, 7 & 5 for +15 VDC. If those are good then it isn't power.



Next open the backshell of that connector and connect probes to 2 & 12 to see if it is reading the same voltage as the cRIO (~4 VDC). If the reading at the sensor matches the cRIO it probably isn't the cabling.



Lastly I would disconnect the detector and open the valves so that sensor is exposed to ambient pressure and see what it reads, if it's still ~400 torr then it is most likely the sensor, but could be the cabling/chassis.



To eliminate that possibility the sensor would need to be removed and tested on a bench using the same pins as listed above.



I don't think we (DSG) have a direct replacement for that, so I'd try Hall B and if not you'd need to talk with Bob as that involves changing a pressure system component.



________________________________

From: dsg-hallb_rtpc <dsg-hallb_rtpc-bounces at jlab.org<mailto:dsg-hallb_rtpc-bounces at jlab.org>> on behalf of mhattawy at odu.edu<mailto:mhattawy at odu.edu><mohammad.hattawy at gmail.com<mailto:mohammad.hattawy at gmail.com>>
Sent: Thursday, July 23, 2020 1:21 PM
To: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: [dsg-hallb_rtpc] [EXTERNAL] Re: Pressure transducer of RTPC



<image_123923953.JPG>

<image_123923953.JPG>

<image_6483441.JPG>





On Thu, Jul 23, 2020 at 12:38 Brian Eng <beng at jlab.org<mailto:beng at jlab.org>> wrote:

Hi Jiwan,



The voltage being read back by the cRIO is ~4V which is being converted correctly to ~400 torr.



Can you get the full model number of PT334 so I can better give some suggestions on what you guys could measure to help troubleshoot?



Off-hand some of the things that could be wrong is: the sensor not getting proper supply voltage, a loose connection of the output to the cRIO as it is getting some voltage back, or a bad sensor (let's hope it isn't that as that was the only one we had in that range).



________________________________

From: Jiwan Poudel <jpoudel at jlab.org<mailto:jpoudel at jlab.org>>
Sent: Thursday, July 23, 2020 11:20 AM
To: Brian Eng <beng at jlab.org<mailto:beng at jlab.org>>
Cc: dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org> <dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>>
Subject: Pressure transducer of RTPC



Hello Brian,



RTPC abs. Pressure transducer seems not working properly. This should show around 760 torr (~ 1 atm), but showing about 400 torr only. What could be the reason?



Thank you,

Jiwan



[Image removed by sender.]



_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc

--

Best regards,
Mohammad Hattawy.

_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc



_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc

--

Best regards,
Mohammad Hattawy.

_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc

--

Best regards,
Mohammad Hattawy.

_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc









_______________________________________________
dsg-hallb_rtpc mailing list
dsg-hallb_rtpc at jlab.org<mailto:dsg-hallb_rtpc at jlab.org>
https://mailman.jlab.org/mailman/listinfo/dsg-hallb_rtpc

--

Best regards,
Mohammad Hattawy.


--

Best regards,
Mohammad Hattawy.




--

Best regards,
Mohammad Hattawy.


--

Best regards,
Mohammad Hattawy.


--

Best regards,
Mohammad Hattawy.
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_rtpc/attachments/20200728/b1b9b623/attachment-0001.html>


More information about the dsg-hallb_rtpc mailing list