<div dir='auto'><div>Green light is on. I can also ping the address you gave, but still does not see any flow. The pressure gauge in front is showing around 7psi, so there is gas upto MFC. I set the flow at 100sccm, but dont see any flow.<br><div class="gmail_extra"><br><div class="gmail_quote">On Mar 19, 2020 18:46, Brian Eng <beng@jlab.org> wrote:<br type="attribution"><blockquote class="quote" style="margin:0 0 0 .8ex;border-left:1px #ccc solid;padding-left:1ex"><p dir="ltr">First make sure it has power (LEDs are on and/or blinking, but no red ones!).
<br>
Then see if it is online (ping hb-mfc-bonus-rtpc.jlab.org).
<br>
Lastly reinitialize the MFCs from the LabVIEW GUI (I'm not sure on the exact labeling of the button but that should be close).
<br>

<br>
_Both_ MFC 104 & 308 need to online in order for the cRIO to communicate with them. If you see the initialization counter get to 3 there's a problem.
<br>

<br>
________________________________________
<br>
From: Dsg-rtpc <dsg-rtpc-bounces@jlab.org> on behalf of Carlos Ayerbe Gayoso <gayoso@jlab.org>
<br>
Sent: Thursday, March 19, 2020 6:42 PM
<br>
To: dsg-rtpc@jlab.org
<br>
Cc: JIWAN POUDEL
<br>
Subject: [Dsg-rtpc] MFC308
<br>

<br>
Hi Guys,
<br>

<br>
Jiwan told me that he has problems to communicate with the MFC308 (I guess the number is correct, I have goldfish memory) the one for the CO2/He.
<br>

<br>
Can you give us a hint how to proceed?
<br>

<br>
Thank you
<br>

<br>
-Carlos