[dsg-svt] FYI: pending SVT support work requests to the DSG
Peter Bonneau
bonneau at jlab.org
Mon Feb 5 15:43:21 EST 2018
See attached DSG response.
On 2/3/2018 10:34 AM, Yuri Gotra wrote:
> The CLAS engineering run is ending in 2 days. We are approaching the
> start of physics data taking.
> Here is the list of SVT support work requested from the DSG which is
> still pending.
>
> - Implementing the SVT hardware safety interlocks control interface in
> CSS, the chosen CLAS environment for slow controls and monitoring. The
> current LabView version of the interface requires a dedicated MS
> Windows pc to check the status, monitoring the system state, and
> resetting the latched interlocks. This pc is not on the CUE management
> system and requires manual installation of security patches and system
> updates.
> - Implementing the hardware interlock alarm logging. Lack of logging
> features limits debugging options of the primary reasons for interlock
> actions and delays resolving the alarm incidents.
> - Broken hardware interlock on ambient temperature T2. This interlock
> had to be disabled, a detector safety issue.
> - Failing reading of the inlet coolant temperature in the hardware
> interlock. This problem requires correct conversion of the raw sensor
> data at negative temperatures to be implemented. This interlock had to
> be disabled, a detector safety issue.
> - Failing reading of the outlet coolant flow meter. An attempt to fix
> it was made, the reading is still incorrect. This reading provides
> data to the SVT software and hardware interlocks, a detector safety
> issue.
> - Installing the leak detection system. The leak sensor is connected
> to the hardware interlock but is not yet monitoring the tube connected
> to the drain pan, a detector safety issue.
> - Fixing bad connection of the hybrid temperature reading for R3S11B.
> Several attempts to fix unreliable contact have been made, the reading
> is still bad. The corresponding module PS interlock had to be
> disabled, a module safety issue.
> - Installing the nitrogen cooling line temperature monitoring. The
> temperature sensor has been ordered.
>
> A new request. Anova A40 chiller has issues related to operating at
> negative temperatures and was removed from the hall this week. This
> chiller is being tested by the engineering group. The SVT is using one
> of the Lauda chillers connected to the cold plate. This chiller has a
> history of blowing the fuses when it was operated in the EEL building.
> The spare Lauda chiller has 208 V plug. In order to minimize the down
> time of the detector in case of failure of the main chiller, it is
> required to assemble a second relay box connected to the hardware
> interlock. The relay box currently in use was designed for 110 V
> outlet compatible with A40 and one of the Lauda chillers.
>
>
> _______________________________________________
> dsg-svt mailing list
> dsg-svt at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-svt
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180205/d6f70e28/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: DSG SVT email response of 2-5-2018.pdf
Type: application/pdf
Size: 271399 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180205/d6f70e28/attachment-0002.pdf>
More information about the dsg-svt
mailing list