[dsg-svt] DSG work request
Marc McMullen
mcmullen at jlab.org
Tue Jul 17 11:37:55 EDT 2018
Yuri,
We have tested both flow meters, and noted inconsistencies in flow using
multiple configurations. During testing this morning the "in" meter
stopped working. We took the cover off and noted not only a moisture on
the display cover, but standing water in the bottom of the enclosure.
The boards appear to be coated for moisture protection, however, the
connector has evidence of corrosion which has eaten away the power
connection. After further testing the meter still is operational,
however the connector needs to be replaced in order for it to work. The
power connection is corroded and broken at the level of the base of the
connection, so repair will be difficult if at all possible.
That said, these meters are not designed to function where they can
accumulate more moisture than the internal desiccant can absorb. The
desiccant bags are completely saturated. While the pcbs are coated to
prevent damage, the power/data contacts were only covered by heat
shrink, _not potted to prevent actual moisture damage_. Additionally,
the connector is positioned at the bottom of the enclosure and recessed,
so it actually collects fluid.
When both meters were working the "in" meter measured "closer" to the
manual rotameter than the "out" meter. We hooked up each meter
individually, and in series. While in series, we noted that the "out"
meter only recorded flow when it was the first meter in the series.
Rotameter gph (Lpm)
Out (Lpm)
In (Lpm)
10 (0.631)
0.00
0.96
14 (0.88)
1.641
1.638
16 (1.01)
1.725
1.654
18 (1.13)
2.15
1.804
Hall B Engineering and you should evaluate the setup and operation of
the chiller to determine if this is the right equipment for that
particular setup. If not, you should consult with Hall B engineering to
select one that may be more appropriate. I will put the the meters back
in the hall before lunch today.
Regards,
Marc
On 7/16/2018 9:10 AM, Peter Bonneau wrote:
>
> FYI: The SVT flow meter was selected by Hall B engineering (see email
> below).
>
> DSG placed the order for the meters as requested by Hall B engineering.
>
> Hall B engineering should select a meter that meets SVT requirements.
>
>
>
>
> -------- Forwarded Message --------
> Subject: Re: Flow Meter for SVT
> Date: Thu, 31 Jul 2014 14:44:09 -0400
> From: Saptarshi Mandal <mandal at jlab.org>
> To: Peter Bonneau <bonneau at jlab.org>
>
>
>
> Hi Peter,
>
> This one has bigger connections, so might be lesser pressure drop. Will
> this work for you ? FLR6302D should work for us.
>
> http://www.omega.com/pptst/FLR-D.html
>
> Saptarshi
>
> On 7/31/2014 1:13 PM, Peter Bonneau wrote:
> > Hi Saptarshi,
> >
> >
> > I know we've talked about this before, but it needs to be ordered now.
>
> >
> > For an interface, we need a 0-5 V output.
> >
> >
> > Thanks,
> >
> > Peter
>
>
>
> On 7/13/2018 12:39 PM, Yuri Gotra wrote:
>>
>> Thank you Peter,
>>
>> A comment on your statement on the flow meters. The models and the
>> number of the flow meters purchased (no spares) were defined by the
>> HBI group. I reported several times that this model is not designed
>> to work in magnetic field. The flow meters failed the very first time
>> the solenoid was turned ON. Please select a model which would be
>> compatible with requirements of the detector and let me and Bob
>> Miller know.
>>
>>
>> On 7/13/18 11:56 AM, Peter Bonneau wrote:
>>>
>>> DSG comments regarding these work requests are in *blue.*
>>>
>>> On 7/13/2018 9:55 AM, Yuri Gotra wrote:
>>>> For testing the cooling system in preparation of the SVT for beam
>>>> operations it is critical to have measurements of the coolant flow.
>>>> Both flow meters observed issues with correct reading of the flow.
>>>> The flow meters have been disconnected from the cooling lines
>>>> (control lines are still connected). Please test the flow meters asap.
>>> DSG will schedule in this request in with all the other work in
>>> Halls B, C, and D we are doing in preparation for the upcoming run.
>>> Hall B SVT does not have cooling system repair components (external
>>> to the chillers) which was first mentioned in our email of 1/31/2017
>>> (listed below).
>>
>
>
>
> _______________________________________________
> dsg-svt mailing list
> dsg-svt at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-svt
--
Marc McMullen
Detector Support Group
757-269-7738
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180717/b77048b2/attachment-0002.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ffdnfgbahljakpen.jpg
Type: image/jpeg
Size: 76912 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180717/b77048b2/attachment-0008.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ecejfmklgikopcmj.jpg
Type: image/jpeg
Size: 90788 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180717/b77048b2/attachment-0009.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: jdbmnkfkoehcfbnp.jpg
Type: image/jpeg
Size: 66653 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180717/b77048b2/attachment-0010.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ijchdeinfhajbmkm.jpg
Type: image/jpeg
Size: 86981 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/dsg-hallb_svt/attachments/20180717/b77048b2/attachment-0011.jpg>
More information about the dsg-svt
mailing list