Interesting. I have the other crate in the ESB -- I'll ask Chuck to
move it back to your area.
Any idea if the firmware on the crate you have is current?
-- Brad
On Tue, 30 Jul 2019, Marc Mcmullen wrote:
> Hi Simona and Brad,
> DSG is investigating the issue Pablo described in the previous email,
> where certain channel parameters of a CAEN HV module will arbitrarily
> change. The tests Pablo ran were on the A1535 modules, which were tested
> in slot 13. All of these modules were old, and it was not clear if the
> issue was local to that particular type of module or the slot.
> I am currently testing the brand new A7030TN modules on multiple slots.
> Module 0297 is located in slot 1, channel 28 failed to complete a full
> current test by ramping to ~1500V with a 1.5M ohm load, it stopped at
> 400V. I checked the parameters using the CAEN GECO interface and the SVMax
> was set to 400V.
> Prior to running the test I had set all slot 1 SVMax values to 3000V.
> Before rerunning the test this morning I set the SVMax back to 3000V
> (using the GECO interface) and ran the test. The test stopped at 600V, I
> checked the SVMax and it was set to 3000V. At that point I asked Pablo to
> check the PV using CSS to look directly into the server and the SVMax
> process variable was indeed set to 600V, the GECO interface was still
> displaying 3000V.
> I used the GECO interface and re-entered 3000V and noted that as I
> highlighted the value, the GECO display changed to 400V. I changed the
> value to 3000V and re-ran the test and it ramped to the expected voltage
> before and tripped as expected.
> This confirms a suspicion that the embedded EPICS server software in the
> sy4527 mainframe has an issue. The process variable changed multiple times
> to different values. We need to test this on another sy4527 crate to see
> if this is a systemic problem or just a bad crate. When it is convenient
> please send us the other crate and we will continue 7030TN testing, and
> see if the issue manifests itself on the second crate.
> *
> Note in the pics below that channel 01.028 SVMax is set to 3000 V in the
> top screen, which is the GECO interface, while below CSS is displaying 600
> V for the same variable.
> GECO
> CSS
> Regards,
> Marc
>
> --------------------------------------------------------------------------
>
> From: Dsg-hallc_controls <dsg-hallc_controls-bounces at jlab.org> on behalf
> of Pablo Campero Rojas <campero at jlab.org>
> Sent: Thursday, July 25, 2019 1:49 PM
> To: dsg-hallc controls <dsg-hallc_controls at jlab.org>
> Cc: Patrizia Rossi <rossi at jlab.org>
> Subject: [Dsg-hallc_controls] CAEN-HV boards test results
>
>
> Hello Simona,
>
> We developed an EPICS based program and CSS screens to test the three
> CAEN-HV boards that brought to the DSG-lab last week.
>
> From the test performed, we found that all three boards have CAEN software
> issues and one of them has a hardware issue as well.
>
> Software issues:
>
> 1. Some channels' parameters like "I[mon]", "V[set]", V[Rdwn]", "I[max]",
> "V[mon]" arbitrarily change from its pre-set values (set values at
> test time) to random values.
> 2. Discrepancies between GECO-2020 (CAEN Control software) and the PVs
> values (Generated by CAEN EPICS Server) displayed for some of the
> above mentioned variables.
>
> However, we were able to ramp all channels to the set voltage after we
> solved these CAEN software issues by re-entering set values, re-setting
> power On/Off channels, and sometimes resetting the built-in CAEN EPICS
> Server.
>
> Hardware issue:
>
> 1. Channel 22 of HV board (S/N: 556) does not to ramp up.
>
> We recommend sending HV board S/N: 556 back to CAEN.
>
> Attached below is a table with the summary of the findings for each board
> tested. To see detailed results from the test, you are most than welcome
> to come and visit us in the DSG room EEL-121.
>
> Best Regards,
>
> Pablo
>
>
>
> +-------------------------------------------------------+
> |Board| Board |Serial|Hardware|Software| Comments |
> | # | Model |Number| Issue | | |
> | | | | | Issue | |
> |-----+----------+------+--------+--------+-------------|
> | | | | | |For some |
> | | | | | |channels, |
> | | | | | |V[set] stuck |
> | | | | | |at a |
> | | | | | |different |
> | | | | | |value than |
> | | | | | |the pre-set |
> | | | | | |value during |
> | | | | | |the test. |
> | | | | | | One of the |
> | | | | | |channels |
> | | | | | |tripped, |
> | | | | | |because |
> | | | | | |"Imax" value |
> | | | | | |changed from |
> | | | | | |its set value|
> | | | | | |during test. |
> | | | | | | |
> | | | | | |Discrepancies|
> | | | | | |between |
> | | | | | |GECO-2020 |
> | | | | | |values and |
> | 1 |CAEN-A1535| 775 | | x |EPICS PVs |
> | | | | | |values for |
> | | | | | |I[mon]. |
> | | | | | | |
> | | | | | |All affected |
> | | | | | |channels |
> | | | | | |ramped to |
> | | | | | |desired set |
> | | | | | |value after |
> | | | | | |the set |
> | | | | | |values were |
> | | | | | |re-entered |
> | | | | | |and power |
> | | | | | |On/Off reset.|
> | | | | | | |
> | | | | | |For the |
> | | | | | |recovery of |
> | | | | | |some |
> | | | | | |channels, |
> | | | | | |reset of the |
> | | | | | |built-in CAEN|
> | | | | | |EPICS Server |
> | | | | | |was required.|
> |-----+----------+------+--------+--------+-------------|
> | | | | | |For some |
> | | | | | |channels, |
> | | | | | |V[set] value |
> | | | | | |changed from |
> | | | | | |its pre-set |
> | | | | | |value. |
> | | | | | | |
> | | | | | |For variable |
> | | | | | |V[mon], found|
> | | | | | |discrepancies|
> | | | | | |between |
> | | | | | |GECO-2020 |
> | 2 |CAEN-A1535| 776 | | x |values and |
> | | | | | |EPICS PVs |
> | | | | | |values. |
> | | | | | | |
> | | | | | |All affected |
> | | | | | |channels |
> | | | | | |ramped to |
> | | | | | |desired set |
> | | | | | |value after |
> | | | | | |the set |
> | | | | | |values were |
> | | | | | |re-entered |
> | | | | | |and channels |
> | | | | | |were reset. |
> |-----+----------+------+--------+--------+-------------|
> | | | | | |Set values |
> | | | | | |for V[set], |
> | | | | | |V[Rdwn] and |
> | | | | | |V[max] |
> | | | | | |changed (PVs |
> | | | | | |and value in |
> | | | | | |GECO-2020) |
> | | | | | |from its |
> | | | | | |pre-set |
> | | | | | |values during|
> | | | | | |the test. |
> | | | | | | |
> | | | | | |Re-entering |
> | 3 |CAEN-A1535| 556 | x | x |values for |
> | | | | | |V[set], |
> | | | | | |V[Rdwn] and |
> | | | | | |V[max], then |
> | | | | | |reset power |
> | | | | | |On/Off |
> | | | | | |allowed ramp |
> | | | | | |up all |
> | | | | | |channels to |
> | | | | | |the desired |
> | | | | | |set voltage |
> | | | | | |except for |
> | | | | | |channel 22, |
> | | | | | |which did not|
> | | | | | |ramp up. |
> +-------------------------------------------------------+
--
Brad Sawatzky, PhD <brads at jlab.org> -<>- Jefferson Lab / Hall C / C111
Ph: 757-269-5947 -<>- Fax: 757-269-5235 -<>- Pager: brads-page at jlab.org
The most exciting phrase to hear in science, the one that heralds new
discoveries, is not "Eureka!" but "That's funny..." -- Isaac Asimov
More information about the Dsg-hallc_controls
mailing list