[Dsg-hallc_controls] CAEN-HV boards test results

Pablo Campero Rojas campero at jlab.org
Tue Jul 30 12:55:18 EDT 2019


Hello Brad,


Currently, the firmware running on the HVCAEN SY4527 crate mainframe  is the latest one available by CAEN (Rel. 2.0.2 - March 2019).



Best Regards,

Pablo

________________________________
From: Brad Sawatzky <brads at jlab.org>
Sent: Tuesday, July 30, 2019 12:46:44 PM
To: Marc Mcmullen <mcmullen at jlab.org>
Cc: Pablo Campero Rojas <campero at jlab.org>; dsg-hallc controls <dsg-hallc_controls at jlab.org>; Patrizia Rossi <rossi at jlab.org>
Subject: Re: [Dsg-hallc_controls] CAEN-HV boards test results

Interesting.  I have the other crate in the ESB -- I'll ask Chuck to
move it back to your area.

Any idea if the firmware on the crate you have is current?

-- Brad

On Tue, 30 Jul 2019, Marc Mcmullen wrote:

>    Hi Simona and Brad,
>    DSG is investigating the issue Pablo described in the previous email,
>    where certain channel parameters of a CAEN HV module will arbitrarily
>    change. The tests Pablo ran were on the A1535 modules, which were tested
>    in slot 13. All of these modules were old, and it was not clear if the
>    issue was local to that particular type of module or the slot.
>    I am currently testing the brand new A7030TN modules on multiple slots.
>    Module 0297 is located in slot 1, channel 28 failed to complete a full
>    current test by ramping to ~1500V with a 1.5M ohm load, it stopped at
>    400V. I checked the parameters using the CAEN GECO interface and the SVMax
>    was set to 400V.
>    Prior to running the test I had set all slot 1 SVMax values to 3000V.
>    Before rerunning the test this morning I set the SVMax back to 3000V
>    (using the GECO interface) and ran the test. The test stopped at 600V, I
>    checked the SVMax and it was set to 3000V. At that point I asked Pablo to
>    check the PV using CSS to look directly into the server and the SVMax
>    process variable was indeed set to 600V, the GECO interface was still
>    displaying 3000V.
>    I used the GECO interface and re-entered 3000V and noted that as I
>    highlighted the value, the GECO display changed to 400V. I changed the
>    value to 3000V and re-ran the test and it ramped to the expected voltage
>    before and tripped as expected.
>    This confirms a suspicion that the embedded EPICS server software in the
>    sy4527 mainframe has an issue. The process variable changed multiple times
>    to different values. We need to test this on another sy4527 crate to see
>    if this is a systemic problem or just a bad crate. When it is convenient
>    please send us the other crate and we will continue 7030TN testing, and
>    see if the issue manifests itself on the second crate.
>    *
>    Note in the pics below that channel 01.028 SVMax is set to 3000 V in the
>    top screen, which is the GECO interface, while below CSS is displaying 600
>    V for the same variable.
>    GECO
>    CSS
>    Regards,
>    Marc
>
>    --------------------------------------------------------------------------
>
>    From: Dsg-hallc_controls <dsg-hallc_controls-bounces at jlab.org> on behalf
>    of Pablo Campero Rojas <campero at jlab.org>
>    Sent: Thursday, July 25, 2019 1:49 PM
>    To: dsg-hallc controls <dsg-hallc_controls at jlab.org>
>    Cc: Patrizia Rossi <rossi at jlab.org>
>    Subject: [Dsg-hallc_controls] CAEN-HV boards test results
>
>
>    Hello Simona,
>
>    We developed an EPICS based program and CSS screens to test the three
>    CAEN-HV boards that brought to the DSG-lab last week.
>
>    From the test performed, we found that all three boards have CAEN software
>    issues and one of them has a hardware issue as well.
>
>      Software issues:
>
>     1. Some channels' parameters like "I[mon]", "V[set]", V[Rdwn]", "I[max]",
>        "V[mon]" arbitrarily change from its pre-set values (set values at
>        test time) to random values.
>     2. Discrepancies between GECO-2020 (CAEN Control software) and the PVs
>        values (Generated by CAEN EPICS Server) displayed for some of the
>        above mentioned variables.
>
>     However, we were able to ramp all channels to the set voltage after we
>    solved these CAEN software issues by re-entering set values, re-setting
>    power On/Off  channels, and sometimes resetting the built-in CAEN EPICS
>    Server.
>
>      Hardware issue:
>
>     1. Channel 22 of HV board (S/N: 556) does not to ramp up.
>
>    We recommend sending HV board S/N: 556 back to CAEN.
>
>    Attached below is a table with the summary of the findings for each board
>    tested. To see detailed results from the test, you are most than welcome
>    to come and visit us in the DSG room EEL-121.
>
>    Best Regards,
>
>    Pablo
>
>
>
>            +-------------------------------------------------------+
>            |Board|  Board   |Serial|Hardware|Software|  Comments   |
>            |  #  |  Model   |Number| Issue  |        |             |
>            |     |          |      |        | Issue  |             |
>            |-----+----------+------+--------+--------+-------------|
>            |     |          |      |        |        |For some     |
>            |     |          |      |        |        |channels,    |
>            |     |          |      |        |        |V[set] stuck |
>            |     |          |      |        |        |at a         |
>            |     |          |      |        |        |different    |
>            |     |          |      |        |        |value than   |
>            |     |          |      |        |        |the pre-set  |
>            |     |          |      |        |        |value during |
>            |     |          |      |        |        |the test.    |
>            |     |          |      |        |        | One of the  |
>            |     |          |      |        |        |channels     |
>            |     |          |      |        |        |tripped,     |
>            |     |          |      |        |        |because      |
>            |     |          |      |        |        |"Imax" value |
>            |     |          |      |        |        |changed from |
>            |     |          |      |        |        |its set value|
>            |     |          |      |        |        |during test. |
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |Discrepancies|
>            |     |          |      |        |        |between      |
>            |     |          |      |        |        |GECO-2020    |
>            |     |          |      |        |        |values and   |
>            |  1  |CAEN-A1535| 775  |        |   x    |EPICS PVs    |
>            |     |          |      |        |        |values for   |
>            |     |          |      |        |        |I[mon].      |
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |All affected |
>            |     |          |      |        |        |channels     |
>            |     |          |      |        |        |ramped to    |
>            |     |          |      |        |        |desired set  |
>            |     |          |      |        |        |value after  |
>            |     |          |      |        |        |the set      |
>            |     |          |      |        |        |values were  |
>            |     |          |      |        |        |re-entered   |
>            |     |          |      |        |        |and power    |
>            |     |          |      |        |        |On/Off reset.|
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |For the      |
>            |     |          |      |        |        |recovery of  |
>            |     |          |      |        |        |some         |
>            |     |          |      |        |        |channels,    |
>            |     |          |      |        |        |reset of the |
>            |     |          |      |        |        |built-in CAEN|
>            |     |          |      |        |        |EPICS Server |
>            |     |          |      |        |        |was required.|
>            |-----+----------+------+--------+--------+-------------|
>            |     |          |      |        |        |For some     |
>            |     |          |      |        |        |channels,    |
>            |     |          |      |        |        |V[set] value |
>            |     |          |      |        |        |changed from |
>            |     |          |      |        |        |its pre-set  |
>            |     |          |      |        |        |value.       |
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |For variable |
>            |     |          |      |        |        |V[mon], found|
>            |     |          |      |        |        |discrepancies|
>            |     |          |      |        |        |between      |
>            |     |          |      |        |        |GECO-2020    |
>            |  2  |CAEN-A1535| 776  |        |   x    |values and   |
>            |     |          |      |        |        |EPICS PVs    |
>            |     |          |      |        |        |values.      |
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |All affected |
>            |     |          |      |        |        |channels     |
>            |     |          |      |        |        |ramped to    |
>            |     |          |      |        |        |desired set  |
>            |     |          |      |        |        |value after  |
>            |     |          |      |        |        |the set      |
>            |     |          |      |        |        |values were  |
>            |     |          |      |        |        |re-entered   |
>            |     |          |      |        |        |and channels |
>            |     |          |      |        |        |were reset.  |
>            |-----+----------+------+--------+--------+-------------|
>            |     |          |      |        |        |Set values   |
>            |     |          |      |        |        |for V[set],  |
>            |     |          |      |        |        |V[Rdwn] and  |
>            |     |          |      |        |        |V[max]       |
>            |     |          |      |        |        |changed (PVs |
>            |     |          |      |        |        |and value in |
>            |     |          |      |        |        |GECO-2020)   |
>            |     |          |      |        |        |from its     |
>            |     |          |      |        |        |pre-set      |
>            |     |          |      |        |        |values during|
>            |     |          |      |        |        |the test.    |
>            |     |          |      |        |        |             |
>            |     |          |      |        |        |Re-entering  |
>            |  3  |CAEN-A1535| 556  |   x    |   x    |values for   |
>            |     |          |      |        |        |V[set],      |
>            |     |          |      |        |        |V[Rdwn] and  |
>            |     |          |      |        |        |V[max], then |
>            |     |          |      |        |        |reset power  |
>            |     |          |      |        |        |On/Off       |
>            |     |          |      |        |        |allowed ramp |
>            |     |          |      |        |        |up all       |
>            |     |          |      |        |        |channels to  |
>            |     |          |      |        |        |the desired  |
>            |     |          |      |        |        |set voltage  |
>            |     |          |      |        |        |except for   |
>            |     |          |      |        |        |channel 22,  |
>            |     |          |      |        |        |which did not|
>            |     |          |      |        |        |ramp up.     |
>            +-------------------------------------------------------+




--
Brad Sawatzky, PhD <brads at jlab.org>  -<>-  Jefferson Lab / Hall C / C111
Ph: 757-269-5947  -<>-  Fax: 757-269-5235  -<>- Pager: brads-page at jlab.org
The most exciting phrase to hear in science, the one that heralds new
  discoveries, is not "Eureka!" but "That's funny..."   -- Isaac Asimov
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallc_controls/attachments/20190730/cd43f176/attachment-0001.html>


More information about the Dsg-hallc_controls mailing list