[Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin Question
    Greg Kibilko 
    greg at setechsales.com
       
    Thu Feb 27 09:27:37 EST 2020
    
    
  
Hi Aaron - Thank you very much.
*Ciao Alberto and Gianni - If you are not currently duplicating this 
setup then perhaps this information may help you replicate and/or better 
understand the root cause of these disconnects.**
*
Kind regards,
Greg
On 2/27/2020 8:13 AM, Aaron Brown wrote:
> Hello Greg,
>
> So, let me try to explain what was going on when we noticed the random 
> disconnections.
> We were conducting a test with load where the voltage was ramped up 
> step-wise. There were a series of relays that would quickly cycle 
> through all of the channels and a digital multi-meter would measure 
> the voltage output of the channel.
> The disconnection occurred on the board attached to the load, and 
> sometimes to other random boards populated in the crate.
> I have attached a talk given by one of our group members. Pages 12-14 
> detail the test setup.
> Let me know if you have any other questions.
> Hopefully this will be helpful.
>
> Best regards,
> Aaron Brown
> ------------------------------------------------------------------------
> *From:* Greg Kibilko <greg at setechsales.com>
> *Sent:* Tuesday, February 25, 2020 9:21 AM
> *To:* Aaron Brown <ambrown at jlab.org>
> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
> Question
> Thank you sir.  Alberto's diagnosis on the source of the parameter 
> change problem is heartening, but we're just trying to understand why 
> the other problem (random disconnect) is still not being seen at CAEN.
>
> Cheers,
> Greg
>
> On 2/25/2020 9:19 AM, Aaron Brown wrote:
>> Good morning Greg,
>>
>> To my knowledge there is no correlation between the pin issue and the 
>> disconnects.
>> I will ask around if anyone has noticed a connection between these 
>> two issues, but it is my understanding that they are separate.
>>
>> Best Regards,
>> Aaron Brown
>> ------------------------------------------------------------------------
>> *From:* Greg Kibilko <greg at setechsales.com> <mailto:greg at setechsales.com>
>> *Sent:* Tuesday, February 25, 2020 8:28 AM
>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>> <mailto:dsg-hallc_controls at jlab.org>
>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>> Question
>> Hi Aaron,
>>
>> I'm curious if you've seen any correlation between the disconnects 
>> and the problems with the boards that have the SL pin pushing out?  
>> Could there be a connection?
>>
>> Best,
>> Greg
>>
>> On 2/24/2020 11:38 AM, Aaron Brown wrote:
>>> Hello Alberto,
>>>
>>> This is wonderful news!
>>> I look forward to receiving the new HiVocs version soon.
>>>
>>> Best Regards,
>>> Aaron Brown
>>> ------------------------------------------------------------------------
>>> *From:* Alberto Lucchesi <a.lucchesi at caen.it> 
>>> <mailto:a.lucchesi at caen.it>
>>> *Sent:* Monday, February 24, 2020 10:22 AM
>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>; Greg 
>>> Kibilko <greg at setechsales.com> <mailto:greg at setechsales.com>; Amrit 
>>> Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>> <mailto:dsg-hallc_controls at jlab.org>
>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>> Question
>>> Hi Aaron,
>>>
>>> I'was able to reproduce the random parameter changes issue.
>>>
>>> The issue was not visible with only one board because it is related 
>>> to the parallelism inside the IOC.
>>>
>>> What was happening is that with several boards the EPICS framework 
>>> (our IOC is based over it) calls parts of our code in parallel and 
>>> there was an internal structure that was not thread-safe. The 
>>> results were that some settings were not  applied to the right 
>>> parameter but to onother one (the aren't random parameter changes, 
>>> The values were right but the parameters were wrong! (i.e., a Trip = 
>>> 40 were codified as VSet = 40 or a PW = On where codified as RUp = 1).
>>>
>>> I fixed the issue and now I'm testing the results and the issues 
>>> seem to be resolved. I'll do other tests in these days then I'll 
>>> send you a new HiVocs version within a couple of days.
>>>
>>> I don't have news about the other issue (the board disconnection) in 
>>> all my test I never experienced a board disconnection.
>>>
>>>
>>> Best Regards,
>>> Albert
>>>
>>>
>>> Il 24/02/2020 14:14, Aaron Brown ha scritto:
>>>> Hello Gianni,
>>>>
>>>> Yes, April works for us. We will have all the boards in our 
>>>> possession on hand for any rework that can be done.
>>>>
>>>> Best regards,
>>>> Aaron Brown
>>>> ------------------------------------------------------------------------
>>>> *From:* Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>>>> *Sent:* Friday, February 21, 2020 5:13 PM
>>>> *To:* Greg Kibilko <greg at setechsales.com> 
>>>> <mailto:greg at setechsales.com>; Aaron Brown <ambrown at jlab.org> 
>>>> <mailto:ambrown at jlab.org>; Amrit Yegneswaran <yeg at jlab.org> 
>>>> <mailto:yeg at jlab.org>
>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>>> Question
>>>> Dear Aaron,
>>>>
>>>> as told by Greg April could be OK with us to visit you.
>>>> If possible I would like to have a look to all the boards 
>>>> connectors ( to rework partially the RADIALL connector and in 
>>>> particularly the safety loop pin).
>>>> Anyway for the moment just let me know if April is OK,  I will send 
>>>> later a list of rework I would like to do on the boards and cables.
>>>>
>>>> Have a nice week end to everybody
>>>> Gianni
>>>> Il 18/02/2020 21:03, Greg Kibilko ha scritto:
>>>>> Hi Aaron,
>>>>>
>>>>> Very good.  Gianni is home sick, but I did Skype with him and his 
>>>>> schedule for March is full. We're aiming for April.  I should be 
>>>>> able to propose some dates for you this week.
>>>>>
>>>>> Best,
>>>>> Greg
>>>>>
>>>>> On 2/18/2020 1:31 PM, Aaron Brown wrote:
>>>>>> Hi Greg,
>>>>>>
>>>>>> If April is the best time for you guys, it works for us.
>>>>>>
>>>>>> Best regards,
>>>>>> Aaron
>>>>>> ------------------------------------------------------------------------
>>>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>>>> <mailto:greg at setechsales.com>
>>>>>> *Sent:* Tuesday, February 18, 2020 10:38 AM
>>>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>>>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>; 
>>>>>> Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector Pin 
>>>>>> Question
>>>>>> Hi Aaron,
>>>>>>
>>>>>> Ok, fantastic.  This gives us plenty of time to work with.
>>>>>>
>>>>>> How about timeline?  Is April too late?
>>>>>>
>>>>>> Cheers,
>>>>>> Greg
>>>>>>
>>>>>> On 2/18/2020 10:36 AM, Aaron Brown wrote:
>>>>>>> Morning Greg,
>>>>>>>
>>>>>>> The only day/time I can think of that /might/ pose a problem 
>>>>>>> would be Wednesday afternoons between 1 and 3.
>>>>>>> Other than that, there is nothing that can't be worked around.
>>>>>>>
>>>>>>> Thanks,
>>>>>>> Aaron
>>>>>>> ------------------------------------------------------------------------
>>>>>>> *From:* Greg Kibilko <greg at setechsales.com> 
>>>>>>> <mailto:greg at setechsales.com>
>>>>>>> *Sent:* Tuesday, February 18, 2020 10:32 AM
>>>>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>>>>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>; 
>>>>>>> Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Alberto Lucchesi 
>>>>>>> <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>> Pin Question
>>>>>>> Hi Aaron,
>>>>>>>
>>>>>>> I am sure Gianni will agree that a visit is in order.
>>>>>>>
>>>>>>> In the next couple of months are there any specific days/weeks 
>>>>>>> we should target to visit?  Or days/weeks we should try to avoid?
>>>>>>>
>>>>>>> Thanks!
>>>>>>> Greg
>>>>>>>
>>>>>>> On 2/18/2020 10:30 AM, Aaron Brown wrote:
>>>>>>>> Hello Gianni,
>>>>>>>>
>>>>>>>> We currently have 33 A7030TN modules and 2 SY4527 crates on hand.
>>>>>>>> I think that best idea would be to have someone from CAEN make 
>>>>>>>> a visit to the lab and take a look/fix the connector pins.
>>>>>>>>
>>>>>>>> Best regards,
>>>>>>>> Aaron Brown
>>>>>>>> ------------------------------------------------------------------------
>>>>>>>> *From:* Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>>>>>>>> *Sent:* Monday, February 17, 2020 12:12 PM
>>>>>>>> *To:* Aaron Brown <ambrown at jlab.org> <mailto:ambrown at jlab.org>; 
>>>>>>>> Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Greg Kibilko 
>>>>>>>> <greg at setechsales.com> <mailto:greg at setechsales.com>; Alberto 
>>>>>>>> Lucchesi <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>>> Pin Question
>>>>>>>> Hello Aaron,
>>>>>>>> thank you very much for the details.
>>>>>>>> Really strange a so high number of problems on the panel 
>>>>>>>> connector, as told from our experience it is more problematic 
>>>>>>>> the cable connector, this is the reason that we add the hot glue.
>>>>>>>> At this point let me know how you wish proceed.
>>>>>>>> To rework/repair the boards with pins pushed out we have two 
>>>>>>>> solutions, you ship the boards to our office in NY, or I could 
>>>>>>>> make an intervention in JLAB in spring.
>>>>>>>> How many boards you have in total?
>>>>>>>> May be it's worth to have a look to al the boards.
>>>>>>>>
>>>>>>>> Waiting your comments.
>>>>>>>> Best regards.
>>>>>>>> Gianni
>>>>>>>>
>>>>>>>> Il 14/02/2020 16:48, Aaron Brown ha scritto:
>>>>>>>>> Hello Gianni,
>>>>>>>>>
>>>>>>>>> We have experienced the issue of pins getting pushed back 12 
>>>>>>>>> separate times;
>>>>>>>>> 8 times it was the pins for the safety loops, and 4 times is 
>>>>>>>>> was an arbitrary pin.
>>>>>>>>> All occurrences were on the board R52 connector (male).
>>>>>>>>> I have attached a photo to highlight exactly what we're 
>>>>>>>>> describing.
>>>>>>>>> Also attached is a photo of the inside of one of the boards 
>>>>>>>>> showing how the channel pins
>>>>>>>>> are strain relieved, but the safety loop pins are not.
>>>>>>>>>
>>>>>>>>> If you have any questions, please do not hesitate to send an 
>>>>>>>>> email to the mailing list.
>>>>>>>>>
>>>>>>>>> Warm regards,
>>>>>>>>> Aaron Brown
>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>> *From:* Dsg-hallc_controls 
>>>>>>>>> <dsg-hallc_controls-bounces at jlab.org> 
>>>>>>>>> <mailto:dsg-hallc_controls-bounces at jlab.org> on behalf of 
>>>>>>>>> Gianni Di Maio <g.dimaio at caen.it> <mailto:g.dimaio at caen.it>
>>>>>>>>> *Sent:* Friday, February 14, 2020 3:26 AM
>>>>>>>>> *To:* Amrit Yegneswaran <yeg at jlab.org> <mailto:yeg at jlab.org>
>>>>>>>>> *Cc:* Patrizia Rossi <rossi at jlab.org> <mailto:rossi at jlab.org>; 
>>>>>>>>> dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>; Greg Kibilko 
>>>>>>>>> <greg at setechsales.com> <mailto:greg at setechsales.com>; Alberto 
>>>>>>>>> Lucchesi <a.lucchesi at caen.it> <mailto:a.lucchesi at caen.it>
>>>>>>>>> *Subject:* Re: [Dsg-hallc_controls] [EXTERNAL] CAEN Connector 
>>>>>>>>> Pin Question
>>>>>>>>> Dear Amrit,
>>>>>>>>>
>>>>>>>>> thank you for your email, I apologize I was not aware about 
>>>>>>>>> the dsg mailing list (dsg-hallc_controls at jlab.org 
>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>).
>>>>>>>>> I full agree with you the pushed pins issue will be surely 
>>>>>>>>> addressed by CAEN, we are considering to do a maintenance 
>>>>>>>>> intervention at JLAB in the next future.
>>>>>>>>> What I would like to know  to have a clear idea of the 
>>>>>>>>> situation is
>>>>>>>>> How many boards with pushed pin/s you have?
>>>>>>>>> How many pins per board?
>>>>>>>>> If I well understood ( but may be I'm wrong) there are 4 
>>>>>>>>> boards with this issue, that is one pin pushed back,and every 
>>>>>>>>> time is one of the two safety pin loop.
>>>>>>>>>
>>>>>>>>> Could you or some one of the group confirm me these numbers?
>>>>>>>>>
>>>>>>>>> Or in case I'm wrong could you provide me a report updated to 
>>>>>>>>> today?
>>>>>>>>>
>>>>>>>>>
>>>>>>>>> Best regards.
>>>>>>>>>
>>>>>>>>> Gianni
>>>>>>>>>
>>>>>>>>>
>>>>>>>>> Il 13/02/2020 19:16, Amrit Yegneswaran ha scritto:
>>>>>>>>>> hello,
>>>>>>>>>> this issue of pushed pins needs to be addressed by CAEN.
>>>>>>>>>> dsg is not in this business.
>>>>>>>>>> also, i am very disappointed that i'm receiving the gianni's 
>>>>>>>>>> e-mail only today now.
>>>>>>>>>> i urge all of you, for the sake of transparency, particularly 
>>>>>>>>>> from the caen side to use the mailing list, like we do here 
>>>>>>>>>> at jlab.
>>>>>>>>>> amrit
>>>>>>>>>> if anyone in the dsg group receives a "singular" e-mail make 
>>>>>>>>>> sure you broadcast it to the mailing list.
>>>>>>>>>> ------------------------------------------------------------------------
>>>>>>>>>> *From:* Dsg-hallc_controls 
>>>>>>>>>> <dsg-hallc_controls-bounces at jlab.org> 
>>>>>>>>>> <mailto:dsg-hallc_controls-bounces at jlab.org> on behalf of 
>>>>>>>>>> Brian Eng <beng at jlab.org> <mailto:beng at jlab.org>
>>>>>>>>>> *Sent:* Thursday, February 13, 2020 12:32 PM
>>>>>>>>>> *To:* dsg-hallc controls <dsg-hallc_controls at jlab.org> 
>>>>>>>>>> <mailto:dsg-hallc_controls at jlab.org>
>>>>>>>>>> *Subject:* [Dsg-hallc_controls] Fwd: [EXTERNAL] CAEN 
>>>>>>>>>> Connector Pin Question
>>>>>>>>>> In regards to Brad's previous email about the potting 
>>>>>>>>>> material here's one of the emails CAEN sent.
>>>>>>>>>>
>>>>>>>>>> Still waiting on more feedback since we've definitely seen it 
>>>>>>>>>> on the board side multiple times.
>>>>>>>>>>
>>>>>>>>>>> Begin forwarded message:
>>>>>>>>>>>
>>>>>>>>>>> *From: *Gianni Di Maio <g.dimaio at caen.it 
>>>>>>>>>>> <mailto:g.dimaio at caen.it>>
>>>>>>>>>>> *Subject: **Re: [EXTERNAL] CAEN Connector Pin Question*
>>>>>>>>>>> *Date: *February 6, 2020 at 5:42:31 PM EST
>>>>>>>>>>> *To: *Brian Eng <beng at jlab.org <mailto:beng at jlab.org>>
>>>>>>>>>>> *Cc: *Greg Kibilko <greg at setechsales.com 
>>>>>>>>>>> <mailto:greg at setechsales.com>>
>>>>>>>>>>>
>>>>>>>>>>> Dear Brian,
>>>>>>>>>>>
>>>>>>>>>>> please could you tell me if the pins pushed out are present 
>>>>>>>>>>> only on the RADIALL connector installed on the boards 
>>>>>>>>>>> (female) or even on the RADIALL connectors installed on the 
>>>>>>>>>>> cables ( male).
>>>>>>>>>>> >From my experience usually in the past the problem appeared 
>>>>>>>>>>> only on cable side, this is the reason that we decided to 
>>>>>>>>>>> add the hot glue, in order to strengthen the tightness of 
>>>>>>>>>>> the pin and prevent a possible escape from the correct 
>>>>>>>>>>> position, but this as said only for the cable connector.
>>>>>>>>>>> The problem on the panel connector is very rare, how many 
>>>>>>>>>>> problematic boards you have with pins pushed out?
>>>>>>>>>>> Could you send me some pictures?
>>>>>>>>>>>
>>>>>>>>>>> Best regards.
>>>>>>>>>>> Gianni
>>>>>>>>>>
>>>>>>>>>
>>>>>>>>> -- 
>>>>>>>>>
>>>>>>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>>>>>>
>>>>>>>>> *Maintenance Division Manager*
>>>>>>>>>
>>>>>>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>>>>>>
>>>>>>>>> Skype: gdimaio_caen
>>>>>>>>>
>>>>>>>>> 	Cell:+393482556525 	
>>>>>>>>>
>>>>>>>>> *CAEN S.p.A.*
>>>>>>>>>
>>>>>>>>> Tel.: +39 0584 388 398
>>>>>>>>> Fax: +39 0584 388 959
>>>>>>>>> Email: info at caen.it <mailto:info at caen.it>
>>>>>>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>>>>>>> Cod.Fisc. e p.IVA IT00864500467
>>>>>>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=9Z9mjiHXVYe4UyLKO3S5MyBF6a-z0Gk2u-VphJNkCR4&e=  
>>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=NkBWvHI5Zs-9oXlFaDnzpnLIvyQVuV_qo4jAGvl6JM8&s=HXuJxESZtIm45hzrZH6q0YrYEYtV0J5kDumRkl8BGW4&e=> 
>>>>>>>>>
>>>>>>>>>
>>>>>>>>> Le informazioni trasmesse attraverso la presente comunicazione 
>>>>>>>>> via mail ed eventuali allegati sono da intendersi di esclusiva 
>>>>>>>>> spettanza del destinatario. Nel caso in cui le stesse 
>>>>>>>>> raggiungessero, per qualunque motivo, soggetti diversi dai 
>>>>>>>>> destinatari, questi dovrebbero darne immediata notizia al 
>>>>>>>>> mittente. In ogni caso, eventuali soggetti diversi dai 
>>>>>>>>> legittimi destinatari della presente comunicazione sono 
>>>>>>>>> esplicitamente diffidati da ogni utilizzazione, sia ai sensi 
>>>>>>>>> del D.Lgs. n. 196/2003 "Codice in materia di protezione dei 
>>>>>>>>> dati personali" sia ai sensi degli art. 616 e ss. Codice 
>>>>>>>>> Penale che sanziona la violazione del segreto sulla 
>>>>>>>>> corrispondenza.
>>>>>>>>> The information contained in this transmission may contain 
>>>>>>>>> privileged and confidential information and is intended only 
>>>>>>>>> for the use of the person(s)named above. If you are not the 
>>>>>>>>> intended recipient, or an employee or agent responsible for 
>>>>>>>>> delivering this message to the intended recipient, any review, 
>>>>>>>>> dissemination, distribution or duplication of this 
>>>>>>>>> communication is strictly prohibited. If you are not the 
>>>>>>>>> intended recipient, please contact the sender immediately and 
>>>>>>>>> destroy all copies of the original message.
>>>>>>>>>
>>>>>>>>
>>>>>>>> -- 
>>>>>>>>
>>>>>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>>>>>
>>>>>>>> *Maintenance Division Manager*
>>>>>>>>
>>>>>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>>>>>
>>>>>>>> Skype: gdimaio_caen
>>>>>>>>
>>>>>>>> 	Cell:+393482556525 	
>>>>>>>>
>>>>>>>> *CAEN S.p.A.*
>>>>>>>>
>>>>>>>> Tel.: +39 0584 388 398
>>>>>>>> Fax: +39 0584 388 959
>>>>>>>> Email: info at caen.it <mailto:info at caen.it>
>>>>>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>>>>>> Cod.Fisc. e p.IVA IT00864500467
>>>>>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=9Z9mjiHXVYe4UyLKO3S5MyBF6a-z0Gk2u-VphJNkCR4&e=  
>>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=CK4IA3vcXcfmGx6m3v3SLbVbALDNmIRVqwR42dIp0vQ&s=3g8vV8eb0W81I5E9dK0wOFl7hjtwyjGD3KSLHzdqEd4&e=> 
>>>>>>>>
>>>>>>>>
>>>>>>>> Le informazioni trasmesse attraverso la presente comunicazione 
>>>>>>>> via mail ed eventuali allegati sono da intendersi di esclusiva 
>>>>>>>> spettanza del destinatario. Nel caso in cui le stesse 
>>>>>>>> raggiungessero, per qualunque motivo, soggetti diversi dai 
>>>>>>>> destinatari, questi dovrebbero darne immediata notizia al 
>>>>>>>> mittente. In ogni caso, eventuali soggetti diversi dai 
>>>>>>>> legittimi destinatari della presente comunicazione sono 
>>>>>>>> esplicitamente diffidati da ogni utilizzazione, sia ai sensi 
>>>>>>>> del D.Lgs. n. 196/2003 "Codice in materia di protezione dei 
>>>>>>>> dati personali" sia ai sensi degli art. 616 e ss. Codice Penale 
>>>>>>>> che sanziona la violazione del segreto sulla corrispondenza.
>>>>>>>> The information contained in this transmission may contain 
>>>>>>>> privileged and confidential information and is intended only 
>>>>>>>> for the use of the person(s)named above. If you are not the 
>>>>>>>> intended recipient, or an employee or agent responsible for 
>>>>>>>> delivering this message to the intended recipient, any review, 
>>>>>>>> dissemination, distribution or duplication of this 
>>>>>>>> communication is strictly prohibited. If you are not the 
>>>>>>>> intended recipient, please contact the sender immediately and 
>>>>>>>> destroy all copies of the original message.
>>>>>>>>
>>>>>>>
>>>>>>> -- 
>>>>>>> *Greg Kibilko*
>>>>>>> *SETS - SE Technical Sales, Inc.*
>>>>>>> 112 Tyra Lane
>>>>>>> Townville, SC 29689
>>>>>>> P: (864) 356-2606
>>>>>>> CAEN Products 
>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=3nJ2zMdxx2iYrQfVevkoQ136crvz5WPX-u9vLXbDKyc&e=>
>>>>>>> CAEN SyS Products & Solutions 
>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=skGqBl6x9y3h92KiLpckUYhVQAtLyVD90zuz4wljKO4&e=>
>>>>>>> IOxOs Products 
>>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=Ny4G7qu4sv_T0NtiToDEcRuwjKB8XwFEyYHHIK3W5Sg&s=OLAIrxGqeN6aRvKctgEs_ax0G-qYK9hWu6Do-hdBywM&e=>
>>>>>>
>>>>>> -- 
>>>>>> *Greg Kibilko*
>>>>>> *SETS - SE Technical Sales, Inc.*
>>>>>> 112 Tyra Lane
>>>>>> Townville, SC 29689
>>>>>> P: (864) 356-2606
>>>>>> CAEN Products 
>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=OnPfX_7lx0ht8FdYXuDrd7OFUQ58Ifa1FEn5qe97ndk&e=>
>>>>>> CAEN SyS Products & Solutions 
>>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=LjH-sZKE_oLptW3aG24eGtFrP1pCGT8A5-PZU5mC3xI&e=>
>>>>>> IOxOs Products 
>>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=RuKyQ246VCfLgDisBhNfXaVwXLPxS8snA1X68m1DNYk&s=cxNHULDWIflLFm2_PSTqcGcmRiirjlGJyPqqHNTYsB8&e=>
>>>>>
>>>>> -- 
>>>>> *Greg Kibilko*
>>>>> *SETS - SE Technical Sales, Inc.*
>>>>> 112 Tyra Lane
>>>>> Townville, SC 29689
>>>>> P: (864) 356-2606
>>>>> CAEN Products 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=wWIxhZ0sQFRqCM2a7qnyYSru8vh21i-yFlS9DiglFLE&e=>
>>>>> CAEN SyS Products & Solutions 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=zcy1CKSquMtAvMb6ArLv6fcDY94k9m3jXSVKz6kgGrQ&e=>
>>>>> IOxOs Products 
>>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=YJAckWJYg3im1sjAx40VNIKTFS7CLO9TnLUALivkW6M&e=>
>>>>
>>>>
>>>> -- 
>>>>
>>>>  CAEN S.p.A.  Gianni Di Maio
>>>>
>>>> *Maintenance Division Manager*
>>>>
>>>> Email: g.dimaio at caen.it <mailto:g.dimaio at caen.it>
>>>>
>>>> Skype: gdimaio_caen
>>>>
>>>> 	Cell:+393482556525 	
>>>>
>>>> *CAEN S.p.A.*
>>>>
>>>> Tel.: +39 0584 388 398
>>>> Fax: +39 0584 388 959
>>>> Email: info at caen.it <mailto:info at caen.it>
>>>> via Vetraia,11 • 55049 • Viareggio (LU) • Italy
>>>> Cod.Fisc. e p.IVA IT00864500467
>>>> https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=9Z9mjiHXVYe4UyLKO3S5MyBF6a-z0Gk2u-VphJNkCR4&e=  
>>>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it&d=DwMF-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=ksMkvqymFO1sFCHsTbtFDMSRRxzdO2ZNaUaJOP8I-po&s=m_dmwoWv0btIaKtYcLCUaVxX_F_Gy3P48QiYGL-j4Cw&e=> 
>>>>
>>>>
>>>> Le informazioni trasmesse attraverso la presente comunicazione via 
>>>> mail ed eventuali allegati sono da intendersi di esclusiva 
>>>> spettanza del destinatario. Nel caso in cui le stesse 
>>>> raggiungessero, per qualunque motivo, soggetti diversi dai 
>>>> destinatari, questi dovrebbero darne immediata notizia al mittente. 
>>>> In ogni caso, eventuali soggetti diversi dai legittimi destinatari 
>>>> della presente comunicazione sono esplicitamente diffidati da ogni 
>>>> utilizzazione, sia ai sensi del D.Lgs. n. 196/2003 "Codice in 
>>>> materia di protezione dei dati personali" sia ai sensi degli art. 
>>>> 616 e ss. Codice Penale che sanziona la violazione del segreto 
>>>> sulla corrispondenza.
>>>> The information contained in this transmission may contain 
>>>> privileged and confidential information and is intended only for 
>>>> the use of the person(s)named above. If you are not the intended 
>>>> recipient, or an employee or agent responsible for delivering this 
>>>> message to the intended recipient, any review, dissemination, 
>>>> distribution or duplication of this communication is strictly 
>>>> prohibited. If you are not the intended recipient, please contact 
>>>> the sender immediately and destroy all copies of the original message.
>>>>
>>
>> -- 
>> *Greg Kibilko*
>> *SETS - SE Technical Sales, Inc.*
>> 112 Tyra Lane
>> Townville, SC 29689
>> P: (864) 356-2606
>> CAEN Products 
>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=S7h3K8fgiI00eZJ2kEf5yMIeKJgvTkyxBfOWhilcrqc&e=>
>> CAEN SyS Products & Solutions 
>> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=ejkMDj6QjXOr7xaA_U4V8SBEF3Ty-1iA6DRqqHJIPmU&e=>
>> IOxOs Products 
>> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=GtuKScKnnvQz5mqUxXps-eQDnWdyIH3rVmFwdtCRzW8&s=SK--V8r9rxg_h2_07vjN3jo9DU_UFOEVmwAT7ndFXdU&e=>
>
> -- 
> *Greg Kibilko*
> *SETS - SE Technical Sales, Inc.*
> 112 Tyra Lane
> Townville, SC 29689
> P: (864) 356-2606
> CAEN Products 
> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=lb66aikn92Ry8Wk9SaQ8vFQizV9CzQKqezE8YwZi0gg&e=>
> CAEN SyS Products & Solutions 
> <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=C0T3o7YcsYSeDSiX-EqAjmW1uJ_jYH_AlvzkicfbySA&e=>
> IOxOs Products 
> <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwMD-g&c=CJqEzB1piLOyyvZjb8YUQw&r=TkJUDpm1-rWCci4nhe47nw&m=8n85U3kspPySw_6XgdPf_7vG1P-esCOYDMN-HlhRyTc&s=NGpHRCiyokCFKbLwHfQR3HDyRk33sGD8AmL6QmU1MZQ&e=>
-- 
*Greg Kibilko*
*SETS - SE Technical Sales, Inc.*
112 Tyra Lane
Townville, SC 29689
P: (864) 356-2606
CAEN Products <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.caen.it_csite&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=aZli3FzdIezaEFEionestJK-qW0tpSrcEdw7kfTf4qQ&e= >
CAEN SyS Products & Solutions <https://urldefense.proofpoint.com/v2/url?u=https-3A__caensys.com_products-2Dand-2Dsolutions_&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=Da68Jxi7xWH-4D1mE2eB8CrhZvikIgLGQF3e4QwCPL8&e= >
IOxOs Products <https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ioxos.ch&d=DwID-g&c=CJqEzB1piLOyyvZjb8YUQw&r=0dv9yTu0KREdd74L0wgLztdO3ePmhsIUc7EqYumdPxc&m=T7Kr_Z3gRG4RC14MpIhi0OhX8Y8MzC-ogjqhplpl6Aw&s=BEqFRHfIC-KgY6_zuyP_aJQCTLp5I3oStgvrLgTCzdE&e= >
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hallc_controls/attachments/20200227/2cd7a6b9/attachment-0001.html>
    
    
More information about the Dsg-hallc_controls
mailing list