Hi Xiangdong,

Is 1:30 a good time?

Regards,
Amanda

----- Original Message -----
From: "Xiangdong Wei" <xwei at jlab.org>
To: "Amanda Hoebel" <amandah at jlab.org>
Cc: dsg-hdice at jlab.org
Sent: Tuesday, August 29, 2017 8:50:25 AM
Subject: Re: [dsg-hdice] NMR power supply synchronization

Amanda, 

My schedule is open today, so we can meet in HDice lab anytime. Please pick a time and let me know. 

Thanks, 

Xiangdong 




Xiangdong Wei 

x-5266 


From: "Amanda Hoebel" <amandah at jlab.org> 
To: dsg-hdice at jlab.org 
Sent: Tuesday, August 29, 2017 8:33:19 AM 
Subject: Re: [dsg-hdice] NMR power supply synchronization 

Hi Xiangdong, 

I will add in a check for the signal generator upon startup. For the program failing to control the power supply, Pete thinks the USB might be hanging up. He suggested we switch to GPIB card. When we last tried to run with GPIB card the temperature and liquid level sensors would not work. I'm going to try fixing the program to allow the sensors to work with the GPIB card, though for now we can insert the card and see if the program runs continuously for 1500 sweeps without the sensors. Will you be in the lab this afternoon? 

Regards, 
Amanda 

----- Original Message ----- 
From: "Amanda Hoebel" <amandah at jlab.org> 
To: "Peter Bonneau" <bonneau at jlab.org> 
Sent: Monday, August 28, 2017 2:37:57 PM 
Subject: Fwd: [dsg-hdice] NMR power supply synchronization 

----- Forwarded Message ----- 
From: "Xiangdong Wei" <xwei at jlab.org> 
To: "Amanda Hoebel" <amandah at jlab.org> 
Sent: Monday, August 28, 2017 12:38:46 PM 
Subject: Fwd: [dsg-hdice] NMR power supply synchronization 

Hi Amanda, 

After we talked about the Rack 2 control at HDice lab last Thursday, I started NMR runs to measure the averaged signal sizes. What I found so far are listed Below: 

1. When starting the VI, the NMR program checked hardware status, but ignored the non-opertional condition (the signal generator output was off) and proceed after a few seconds. The VI even displayed a message "waiting for response". 

2. The VI failed to control magnet power supply within 1 hour of running. Unplug and re-plug the USB connector (for GPIB) didn't help. The computer was rebooted and all components on the control rack 2 were individually power-cycled to isolate the error source. It seems like NMR run can only be started after power-cycling the magnet power supply. (As we discussed before, this can not be done on any magnet without a persistence current switch, like IBC. Power-cycle magnet means loss of target polarization.) 

3. Finally a long run (average of 1500 sweeps, more that 60 hours) to study noise was started on Friday afternoon. 
The program started as normal but failed to control power I checked on Saturday morning (trace #477): while power supply stuck at the lower edge of the sweeping window as before, the program keep recording data and update the average as if everything was working. The power supply was power-cycled and NMR VI was restarted for 1000 sweeps. The problem reappeared after several hundred sweeps before Sunday noon (checked on trace #663). Attached pictures 1, 2 and 3 show the status. From the signal to noise ratio, one can conclude that there was at least a couple of hundred useful sweeps in the data stream. Would you please check if the VI can report the starting field before each sweep in a separate file (similar to recording temperature and helium level) so one can check it to find out the ramping status (if needed)? 

4. The program halt at the start of condition #3 (field reading reading jumped to the upper end at t=31s and jumped dwon to 0 at t=62s). Please see the attach picture 4. 

Please take a look at the issues. 

I started another run but you can stop it and start program debugging anytime. 

Thanks, 

Xiangdong 



Xiangdong Wei 


x-5266 


From: "Xiangdong Wei" <xwei at jlab.org> 
To: "Amanda Hoebel" <amandah at jlab.org> 
Sent: Wednesday, August 23, 2017 1:37:42 PM 
Subject: Re: [dsg-hdice] NMR power supply synchronization 

Hi Amanda, 

Thank you for the info, I'll continue to test the Rack 2 NMR program. 

Thanks, 

Xiangdong 




Xiangdong Wei 

x-5266 


From: "Amanda Hoebel" <amandah at jlab.org> 
To: dsg-hdice at jlab.org 
Sent: Wednesday, August 23, 2017 11:58:06 AM 
Subject: Re: [dsg-hdice] NMR power supply synchronization 

Hi Xiangdong, 

As mentioned in the lab, the malfunction might have occurred from not power-cycling all instrumentation after the isolation transformer box had been removed. Taking the box out of the rack required the removal of several cables from some of the instrumentation, which could have caused a communication failure. If anything is taken from the rack it is probably best to power-cycle everything together to avoid program failure. 

The program seems to be working now. If the problem persists, let me know and I will take a look at it again. 

Regards, 
Amanda 

----- Original Message ----- 
From: "Xiangdong Wei" <xwei at jlab.org> 
To: "Amanda Hoebel" <amandah at jlab.org> 
Cc: dsg-hdice at jlab.org 
Sent: Tuesday, August 22, 2017 5:24:15 PM 
Subject: Re: [dsg-hdice] NMR power supply synchronization 

Amanda, 

I tried to restart the Rack 2 test with PD1 this afternoon after cooling down the dewar over the weekend. The test run could not be carried out because the VI, especially the field plot malfunctioned: The VI failed to set the correct power supply parameters correctly, causing the ramping stopped at the low end of the sweeping window (Please see attached pictures). This happened in the past and was fixed. Please take a look. I'll be in HDice lab at 10:10am, after finishing an analysis meeting. 

Thanks, 

Xiangdong 



Xiangdong Wei 

x-5266 
_______________________________________________ 
dsg-hdice mailing list 
dsg-hdice at jlab.org 
https://mailman.jlab.org/mailman/listinfo/dsg-hdice 
_______________________________________________ 
dsg-hdice mailing list 
dsg-hdice at jlab.org 
https://mailman.jlab.org/mailman/listinfo/dsg-hdice



More information about the dsg-hdice mailing list