[dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Xiangdong Wei
xwei at jlab.org
Mon Sep 21 16:45:33 EDT 2020
Hi Tyler,
Thank you for fast responding. I didn’t reconfigure the rack2 connections, only used a GPIB cable to connect both power supplies. In principle, the connections were the same as your drawing (the adapter is on the lock-in).
I did use the front panel to set the GPIB address, following the Oxford manual.
What I didn’t do was checking if the cable was good or not. I’ll look for another one and check again. (The cave is locked out now)
Thanks,
Xiangdong
________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Monday, September 21, 2020 3:55 PM
To: Xiangdong Wei <xwei at jlab.org>; thomaso at jlab.org <thomaso at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Xiangdong,
Just to make sure everything is hooked up correctly and since I can’t see it in person, attached is a hand drawn sketch for how the set up needs to be.
There needs to a a GPIB cable between the two power supplies. The USB-GPIB adapter needs to go on to the connector of one of the power supplies. In the sketch, it’s on P.S. # 1 but it could also be on P.S. # 2.
If everything is already set up as in the sketch, it could be that the second power supply didn’t actually get assigned the new GPIB address. How did you set its address?
You would need to set it on the front panel of the power supply. In the attached manual, page 48 has the steps to get the power supply into the mode to set the address. Page 49 has the info on what to do to set the GPIB address.
Tyler
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Monday, September 21, 2020 3:33:03 PM
To: Tyler Lemon <tlemon at jlab.org>; thomaso at jlab.org <thomaso at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Tyler,
I set the GPIB address of the 2nd power supply to 29, and tried to connect to the 1st power supply via a glib cable between them. None of them work with this configuration—-both displayed errors. Either one works after disconnecting glib cable from another, after selecting the correct GPIB address. How do I connect 2 supplies (with same bus address)? Do we need another GPIB-USB adapter?
Thanks,
Xiangdong
________________________________
From: Tyler Lemon <tlemon at jlab.org>
Sent: Friday, September 18, 2020 9:05 AM
To: Xiangdong Wei <xwei at jlab.org>; thomaso at jlab.org <thomaso at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Xiangdong,
Another point about controlling two power supplies from the same computer, the power supplies each need to have a unique GPIB address.
If the power supplies have previously been used for the rotation of target polarization program, they should already be set up this way.
-Tyler
________________________________
From: dsg-hdice <dsg-hdice-bounces at jlab.org> on behalf of Tyler Lemon <tlemon at jlab.org>
Sent: Friday, September 18, 2020 8:51 AM
To: Xiangdong Wei <xwei at jlab.org>; thomaso at jlab.org <thomaso at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Xiangdong,
I added a second copy of the Oxford power supply VI to the fsNMR LabVIEW project.
Both copies are in a new folder in the project (see attached screenshot).
You only have to put the correct GPIB address into each of the programs and run each program to control both power supplies.
-Tyler
________________________________
From: Xiangdong Wei <xwei at jlab.org>
Sent: Friday, September 18, 2020 8:31 AM
To: thomaso at jlab.org <thomaso at jlab.org>; Tyler Lemon <tlemon at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Tyler,
After reviewing the UITF run plans, I just realized that we need to remotely drive 2 independent Oxford superconducting magnet power supplies at the same time, not just 1.
Would you please check to see if it is possible to add another VI to Rack 2 computer to realize this mode?
Thanks,
Xiangdong
________________________________
From: dsg-hdice <dsg-hdice-bounces at jlab.org> on behalf of thomaso at jlab.org <thomaso at jlab.org>
Sent: Thursday, September 17, 2020 3:54 PM
To: Tyler Lemon <tlemon at jlab.org>
Cc: dsg-hdice at jlab.org <dsg-hdice at jlab.org>
Subject: Re: [dsg-hdice] Found a built in averaging and referencing feature on sweep tab of Zurich lock-in amplifier
Hi Tyler,
Thanks for sharing this with us! I will check it out shortly.
-Tom
> Hi Xiangdong and Tom,
>
> After our meeting, I ran a test sweep with the Zurich lock-in amplifier.
> While it was running I was looking around the sweep tab's options and
> found a sweep mode that lets you set a number of samples to take at each
> frequency during a sweep. I also saw on the "History" tab in the sweep tab
> there is a "Reference" option. These two options seem to result in what we
> are trying to do with the fsNMR program.
>
> So you all can look at it, attached are three screenshots with what to
> click along with the steps below.
>
> screenshot1.jpg
> 1. Go to settings tab on right of sweep tab
> 2. use drop down menu to change application to "Parameter Sweep Averaged"
>
> screenshot2.jpg
> 3. Click "Advanced" button on settings tab
> 4. Enter averaging settings
>
> At this point you'd run a sweep as normal. For the reference feature:
>
> screenshot3.jpg
> 5. Go to history tab on sweep tab
> 6. select result to use as reference
> 7. click "Reference" button
> 8. Click button to toggle reference "on"
>
> The next sweep you'd run will be referenced to whatever you selected in
> step 6. An example result after averaging is in the amplitude plot for
> screenshot1.jpg and in the attached CSV file.
>
> I look forward to hearing what you think about and find with these
> options.
>
> -Tyler
>
>
>
> _______________________________________________
> dsg-hdice mailing list
> dsg-hdice at jlab.org
> https://mailman.jlab.org/mailman/listinfo/dsg-hdice
>
_______________________________________________
dsg-hdice mailing list
dsg-hdice at jlab.org
https://mailman.jlab.org/mailman/listinfo/dsg-hdice
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/dsg-hdice/attachments/20200921/76fb28bd/attachment.html>
More information about the dsg-hdice
mailing list