Good morning Amanda,
The nmr control rack 2 is set to run 1754 sweeps to verify that it can record the parameters for more than 1024 sweeps. It is at the 508th sweep now. Two problems were found this morning: 1. Clicking "un-Pause" button caused power supply to ramp down to 0. I caught it at 0.5001Tesla (It was set to sweep between 0.5728T and 0.6028T, but the point was showing 0). 2. The sweeps end up at different values, sometimes at 0.6028T (correct), other times at 0.6027T (wrong). It looks like the hold button starts before the ramping finishes. Would you think about it offline and modify the program after the run finish? It would be Thursday morning.
Thanks,
Xiangdong
Xiangdong Wei
x-5266