Thought I would give everyone a heads up on this just in case you
have to deal with it. With the beam line configuration that we have now
there is a possibility that you may have the valves 2H00 and 2H00A close
without apparent reason. This valve set will not function without a
coincidence downstream of them. Usually this would come from the Saclay
targets vacuum gauges. Since the Saclay target is not installed and the
valves are I had to provide a signal for these valves. One of the
gauges is located on the DVCS vacuum jacket and the other is on the
transfer line from the buffer dewar. So far there have been two
instances where the valves closed. There is a GFCI outlet mounted to
the side of the turbo pump unit located next to the buffer dewar which
has tripped once. If this is the problem just reset the GFCI and start
the pump back up. You should be able to open the valve in just a couple
of minutes after the pump is restarted. The other problem occurred
during a cryo recovery. Temperature changes in the transfer line cause
changes in the vacuum jacket therefore the valves close. Once recovery
is finished you should be able to hit the reset button on the vacuum
panel and open the valve. However I would suggest a little caution
there. I don't know how vacuum tight the downstream beam pipe is and we
don't have anyway of knowing what kind of vacuum is there. So if it has
been an extended recovery we may need to consider bleeding up that
section of beam line before opening the valve. If anyone runs into this
problem and needs a hand feel free to call me and I will come into help.
Thanks
Denny