Hi guys:
I have been thinking again about the perceived need for linearity
measurements capability in the calibration system. Let's look at the
numbers a bit more carefully and, please, correct me if I am wrong:
Nocc = M x [1-exp(-PDE x Nph/M)]
Where: Nocc is the number of pixels occupied by processing incident
photons
M is the number of pixels in the array
Nph is the number of incident photons
The ratio, then, of Nocc/M is the degree of non-linearity. For
example, if M= 56,000 pixels, PDE is 0.20 and Nph ~ 50,000 per side
per GeV, we get Nocc/M ~ 8.5%, assuming that ~20,000 photons go into
only one array. So, the number of ~10% non-linearity that Elton often
quotes is in the right ball park. So, the calculations are correct
but is the input correct?
Let's start with the 56,000 ph/side/GeV. This number comes out of
Andrei's calculations based on the number of p.e.'s we get from the
fibers using the Sr90 source and Irina's simulations of the energy
deposition in the matrix. However, these numbers were obtained with
naked fibers, where the full effect of the outer cladding comes into
play. In the BCAL, the fibers trap the light in the region between
the core and the first to second cladding volume, so the trapping
efficiency will be reduced. We will know the number of effective
photons trapped in the BCAL when we get the cosmic ray measurements
done. In any case, the number will be less.
Now come the really significant assumptions:
Even at oblique angles (less than 20 degrees) is it reasonable to
expect that all the photons on one side will go into a volume confined
in two dimensions within 2 cm x 2 cm? I have a difficult time
accepting this and perhaps a real simulation of the shower generated
by 1 GeV photon randomly distributed across the 2 cm width of a read-
out cell will give us better answers. The Moliere radius alone is
over 3 cm so spill over will occur even if the photon was incident
dead center on a read-out cell. So, how many photons will be
generated in a read-out cell as a fraction of the total energy
deposited in the BCAL? We need a better result to draw conclusions on
this.
Even if we take the 56,000 photons per side per GeV, this number
corresponds to the number of photons arriving at the interface of the
BCAL with the light guide. Here one has to start counting losses:
losses crossing the glued interface from fibers to light guide. Then,
and this is a significant number, losses from end of light guide to
the SiPM. Unless I recall wrong, our measurements here between
physical contact with Si grease between light guide and PMT, on one
hand, and air gap on the other, the latter was around 50% of the
former. In other words, just by the air gap we lose about 50% of the
light. So, the realistic maximum number of photons reaching
photosensors per side per GeV is much closer to 20,000 than to
56,000. If half of them are confined within the volume of one read-
out cell (one SiPM), then Nocc/M is ~ 3.5%.
Is such non-linearity an issue to try and monitor if it means more
complicated calibration system?
George