[Halld-cal] FCAL CW bases
Fernando J. Barbosa
barbosa at jlab.org
Mon Jun 21 15:52:18 EDT 2010
Hi Paul,
That's really strange that you are getting sparks at 1,500V on these
parts across the surface. You can try cleaning with Flux-off, followed
by a rinse in distilled water and thorough drying. I completely agree
that it is important to find out as much as possible before committing
to the re-design.
Yes, I got the pdf but getting dimensions from that and considering the
component dimensions is a bit inaccurate. You could send me your file
and I can try importing it.
Thanks and best regards,
Fernando
Paul Smith wrote:
> Hi Fernando,
>
> I have no objection to a board redesign which moves the diodes further
> apart.
>
> I tested one of the 2225 AVX capacitors with small wires soldered to
> the ends and just sitting in the air with no board or coating. It was
> soldered with water soluble flux which was then cleaned with distilled
> hot water under pressure and then ethyl alcohol. There was never any
> flux on the faces or sides, only a little on the end caps where the
> wires were soldered. So, it should be extremely clean. It sparks
> somewhere between 1500 and 2000 volts. Sparks are across the sides
> between corners. If you look at the end cap metallization, it is not
> exactly smooth at the boundary where the metallization meets the
> ceramic. After a spark, evidently a sharp point gets burned off since
> it will then withstand a slightly higher voltage before sparking
> again. Humidity was ~55%.
>
> The next thing I plan to try is epoxy-ing the diodes and humisealing
> the capacitors on one of the existing boards. I think it is important
> to learn as much as possible about the problem areas before redoing
> the board layout. If this works we can consider avoiding the use of
> epoxy.
>
> I sent a pdf of the present board layout a few weeks ago. I can send
> it again if you don't have it. I can send the new layout when I have
> it done.
>
> Paul
>
>
>
> On Jun 21, 2010, at 1:54 AM, Fernando J. Barbosa wrote:
>
>> Hi Paul,
>>
>> As we have discussed in the past, placing the diodes farther apart
>> (larger than the diode's pad-to-pad length) will prevent arcing on
>> this part of the board (the limitation will then be within each
>> component rather than component-to-component).
>>
>> There are a number of different types of epoxies and for different
>> applications, as you know. The problem with epoxy is that it does not
>> allow for re-working the circuitry, over time it de-laminates or
>> cracks (especially with high levels of radiation but this is not the
>> case here) and it is impossible to match the CTE of epoxy to that of
>> ceramic. Because epoxy is also slightly hygroscopic, arcing will
>> occur shortly after de-lamination. You are correct that thick coats
>> of epoxy are used on some components for protection from mechanical
>> stresses and environmental factors but its purpose is not to prevent
>> arcing, even though it may be rated for HV operation.
>>
>> The proper coating for HV should be acrylic-based (like the Humiseal
>> you have used) or Parelene. I prefer Humiseal because it cures
>> quickly, has very little out-gassing, is excellent in high humidity
>> environments and it allows for re-working of the circuitry.
>>
>> The problems I have encountered that lead to arcing have usually been
>> due to surface residues (cleanliness is the most important issue),
>> improper pad spacing (the pad-to-pad spacing[edges] under the
>> component should be about the same as the component length, sharp
>> edges, component spacing and defective components, in this order. Of
>> course, humidity is a big factor. IPC does not address HV conditions
>> properly, in my opinion, as it is concerned primarily with assembly
>> of PCBs for most commercial uses.
>>
>> We had some problems with HV in the FDC and the vast majority of
>> these were due to cleanliness. It turned out that the epoxy that held
>> the wires in place was contaminated from cleaning the solder residues
>> which also softened the epoxy; one component was placed too close to
>> another; residues on the PCB also caused quite a lot of problems.
>> These problems have been resolved by proper cleaning and we have
>> re-designed the boards more conservatively. We have also applied a
>> conformal coating of Humiseal to HV capacitors outside of the gas
>> volume because of humidity.
>>
>> The HV capacitors are 1825 from AVX rated at 4kV and operated at over
>> 2kV (1825JA221KAT1A). These are smaller than the ones you mention and
>> so I would look very carefully at the cleanliness of your assembled
>> PCB. These SMT components should operate at the specified HV provided
>> by the manufacturer for SMT assemblies - the disclaimer is just an
>> acknowledgment that operation of the component may be compromised if
>> humidity is high, for example.
>>
>> In my opinion, the CW bases should be robust by design as we consider
>> at least 10 years of operation and the cost of re-design should not
>> be that significant compared to production and maintenance costs.
>>
>> I will be glad to check your layout if I can import your file.
>>
>> Best regards,
>> Fernando
>>
>>
>> Paul Smith wrote:
>>> Hi Fernando,
>>>
>>> I'm not on the calorimetry email list. Please feel free to forward
>>> to the list if appropriate.
>>>
>>> Comments in-line below:
>>>
>>> On Jun 18, 2010, at 11:28 AM, Matthew Shepherd wrote:
>>>
>>>>
>>>> Hi Paul,
>>>>
>>>> I'm not sure you are on the calorimetry list -- see the note below
>>>> from Fernando.
>>>>
>>>> -Matt
>>>>
>>>> Begin forwarded message:
>>>>
>>>>> From: "Fernando J. Barbosa" <barbosa at jlab.org>
>>>>> Date: June 18, 2010 11:24:28 AM EDT
>>>>> To: Matthew Shepherd <mashephe at indiana.edu>
>>>>> Cc: Hall-D Calorimetry <halld-cal at jlab.org>, "Fernando J. Barbosa"
>>>>> <barbosa at jlab.org>
>>>>> Subject: FCAL CW bases
>>>>>
>>>>> Hi Matt,
>>>>>
>>>>> I have a few concerns about the plan for the CW bases you
>>>>> suggested at the GlueX meeting and regarding the HV sparks
>>>>> observed on the present design:
>>>>>
>>>>> 1. HV sparks across components - Due to close proximity of
>>>>> components, HV sparks or leakage are observed. The present
>>>>> proposed solution is to pot with epoxy. This seems to prevent HV
>>>>> sparks.
>>>>>
>>>>> My concern is that this is temporary solution but not a cure for
>>>>> the inherent problem. I really don't want to rely on potting with
>>>>> epoxy (long term issues) and a conformal coating should only be
>>>>> used as a preventive measure.
>>>
>>> This problem is between adjacent diodes in the CW multiplier chain.
>>> If one follows the IPC "rule of thumb" of 500 volts for a 0.1 inch
>>> spacing on uncoated FR4, there shouldn't be a problem. However,
>>> even though the pads have round corners, the diode leads have square
>>> corners which aren't always completely covered by solder. Even
>>> though conformal coating is supposedly good for >1500 volts/mil,
>>> this seems to be the case only if the electric field is transverse
>>> to the coating. As we reported a few weeks ago, sparks will form
>>> under the coating in a plane parallel to the board surface,
>>> presumably between the board and the coating. Evidently the coating
>>> doesn't completely bond to the FR4. Slots, etc are possible, but at
>>> extra cost. This problem could be solved by moving the diodes
>>> further apart, resulting in a longer board.
>>>
>>> What are the long term issues with epoxy you refer to? Radiation
>>> damage to the epoxy? Isn't epoxy used all the time in chamber
>>> construction, etc? Aren't through hole HV capacitors and resistors
>>> potted/molded in epoxy by the manufacturer?
>>>
>>>
>>>>>
>>>>> 2. HV sparks are observed across one HV capacitor - the capacitor
>>>>> is rated for operation at 2,500V but it cannot hold 1,500V. The
>>>>> proposed solution is to use a low viscosity epoxy to under fill
>>>>> the capacitor and prevent sparking.
>>>>>
>>>>> A capacitor that is sold and rated at 2,500V should withstand that
>>>>> same voltage or the manufacturer has a big problem. Most likely
>>>>> the PCB pads to which the cap is soldered to are too close or have
>>>>> sharp edges - this limits the maximum operating voltage.
>>>>> Alternatively, you could choose a larger capacitor with the pads
>>>>> spaced farther apart but this may not be necessary.
>>>
>>> This capacitor is an AVX # 2225WC103KAT1A The data sheet states:
>>> "may require protective surface coating to prevent external
>>> arcing". The length is 0.22", so the clearance is way less than the
>>> 500 volts/0.1 inch rule of thumb. The arcing is observed NOT
>>> between the pads (which do have round corners) , but most definitely
>>> between the metallization of the capacitor terminals along the sides
>>> of the capacitor (between the ends). Epoxy underneath the capacitor
>>> doesn't solve the problem; it also needs the sides and top covered
>>> with epoxy as well in order to withstand the rated 2500 volts.
>>>
>>> The previous version of the CW base used a Johanson 302H51W103KY4
>>> for this component; this was a 3530 case size. It was rated for
>>> 3000 volts and would withstand 2500 volts if the humidity was 50% or
>>> lower. Humidity levels higher than 60% would cause sparks at 2000
>>> volts. Unfortunately, this capacitor is no longer made, and in fact
>>> Johanson now has capacitors rated up 6000 volts in the 2225 case
>>> size. They also include the disclaimer "may require a surface
>>> coating to prevent external arcing".
>>>
>>> AVX does make a 3640 case size capacitor; unfortunately it is 4X as
>>> expensive as a 2225. To my knowledge this is the only option for a
>>> physically larger surface mount HV cap. Several manufacturers make
>>> 2500 volt rated capacitors in the 2225 case size.
>>>
>>>>>
>>>>> I strongly suggest that the base be re-designed to space the
>>>>> components and pads farther apart to provide for reliable
>>>>> operation by design, without a conformal coating. Once that is
>>>>> verified, a conformal coating is highly recommended because of
>>>>> moisture and dust.
>>>
>>>
>>> Assuming we can find a surface mount capacitor physically large
>>> enough to withstand it's rated voltage without any sort of coating,
>>> the next issue is with the 1 Gohm feedback resistors. The previous
>>> version of the CW base used a Victoreen (now Ohmite) MC102821007F,
>>> in a 5025 case size. Unfortunately, this is no longer a standard
>>> product and is only available as a special order. The present
>>> version of the base uses an Ohmite HVF2512T1007FE. This is rated at
>>> 2500 volts in a 2512 case size, so has slightly better clearance
>>> than the 2225 capacitor, but still well in excess of the 500
>>> volt/0.1 inch rule. I haven't actually seen sparks across this
>>> resistor since they form across the capacitor first, but have also
>>> coated it with epoxy both underneath and top and sides in order to
>>> get it to work. Venkel does make a 1 Gohm resistor in a 5020 size,
>>> but it is 6X as expensive as the Ohmite 2512.
>>>
>>>
>>> To summarize, the diode clearance issue could be solved by a larger
>>> layout. The 2500 volt rated resistors and capacitors are difficult
>>> to get and expensive in larger case sizes and still won't work
>>> without a coating unless the humidity is very low. Conformal
>>> coating doesn't seem to be adequate for insulation parallel to the
>>> board surface and sparking can occur underneath. Epoxy does work,
>>> but if you think there are long term issues, the only alternative
>>> for producing a layout that will operate without any type of
>>> additional insulation would be to use non surface mount components
>>> for the parts that see the entire 2+ kV across them. I suppose
>>> another alternative might be to use 2 components in series so that
>>> each only sees half the voltage.
>>>
>>> Because of the block to block spacing, it is not possible to make
>>> the PCB any wider. It is possible to make it longer. This works
>>> out fine for increasing the diode spacing, but doesn't help much
>>> with the other parts. It might be necessary to go to more layers to
>>> use the longer board to advantage with larger, non surface mount
>>> parts or doubled parts.
>>>
>>> I would point out that CW bases made by Hamamatsu (for example) are
>>> completely potted. Obviously, I will do whatever is required to
>>> pass the JLab review, but changing the layout enough to allow it to
>>> operate without additional insulating material is going to be
>>> expensive in terms of engineering time and parts cost. Is this
>>> requirement going to apply to HV distribution boards on the
>>> chambers, for example? I know the ATLAS TRT drift tubes had to use
>>> epoxy on their HV distribution/blocking capacitors. The fundamental
>>> problem seems to be that HV surface mount components have extremely
>>> optimistic voltage ratings. Presumably the capacitor dielectric can
>>> withstand the rated voltage, but they must test them in an oil bath
>>> or something other than air.
>>>
>>> I am planning to visit a new contract manufacturer next week (our
>>> previous CM in Indianapolis went out of business). I will discuss
>>> the manufacturing issues involved in coating HV components with
>>> them, whether with epoxy or some other material.
>>>
>>> In any case, your comments/suggestions are extremely useful, and I
>>> will keep in touch,
>>>
>>> Paul
>>>
>> <barbosa.vcf>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: barbosa.vcf
Type: text/x-vcard
Size: 188 bytes
Desc: not available
Url : https://mailman.jlab.org/pipermail/halld-cal/attachments/20100621/497cf54b/attachment-0001.vcf
More information about the Halld-cal
mailing list