Hi Orlando et al.,
I have taken a look at the Hamamatsu data to try to identify and select 
the "noisy" sensors. Attached is a text file of the Hamamatsu data for 
all 4000 sensors, a script to read in the data and output plots. Take a 
look at plot_Hamamatsu_c2.png. Histogramed are the dark rate, dark rate 
RMS, and correlation between the two. The dark rate has a tight 
distribution between 10 and 30 MHz, with a small tail to 35 MHz. There 
are no sensors with a DR>40MHz (our spec is 100 MHz). The RMS 
distribution is not as concentrated, with a small tail extending out 
past 20MHz (max on the histogram). In the scatter plot one can see that 
there is very little correlation between large DR and high RMS. If 
anything there is a small anti-correlation.
One can also take a look specifically at the sensors that Orlando 
identified on Tuesday:
id         DR(MHz)          RMS DR (MHz)
1503    14.0                43.5
1673    10.2                67
1763    30.9                25.6
1772    16.3                56.4
Indeed, these sensors have a very large RMS, indicating that a single 
cell is very different than the rest. However, the overall dark rate is 
quite in line with other units. In fact many of them have relatively low 
(average) dark rate to compensate for the outlying cells.
My conclusion is that these sensors, while in some ways anomalous, are 
no worse than most of the other sensors. Therefore, I conclude that NO 
SPECIAL SELECTION SHOULD BE MADE BASED ON THE DARK RATE OR DARK RATE RMS.
I would appreciate prompt feedback if anyone has looked into this further.
Thanks, Elton.
-- 
Elton Smith
Jefferson Lab MS 12H5
12000 Jefferson Ave STE 16
Newport News, VA 23606
(757) 269-7625
(757) 269-6331 fax
-------------- next part --------------
3200	72.17	46.1	10.2	38.8	4.3	  
3164	72.4	39.9	10.2	38	8.3	  
3606	72.603	48.6	13.9	36.9	2.9	  
3587	72.41	49.4	10.9	36.9	5.3	  
3609	72.72	32.2	8	36.1	3.6	  
3195	72.416	38	10.2	36	8.1	  
3206	72.206	38.4	11.7	35.8	3.9	  
3602	72.607	47.8	11.7	35.7	5.3	  
3181	72.389	40.5	9.5	35.5	7	  
3172	72.186	39.1	10.9	35.5	5.2	  
3585	72.789	25.4	6.6	34.9	14.9	  
3178	72.405	23.6	6.6	34.9	6.9	  
4386	72.404	29.8	8.8	34.8	3.1	  
3540	72.23	42.2	10.2	34.8	7.6	  
3161	72.551	45	12.4	34.7	5.7	  
4384	72.34	51.8	12.4	34.6	3.3	  
3611	72.806	40.4	9.5	34.6	4	  
4010	72.541	36.3	10.2	34.5	4.4	  
4387	72.548	36.3	8.8	34.3	5.1	  
4380	72.14	48	11.7	34.2	3.7	  
3584	72.718	40.6	9.5	34.1	13.6	  
4377	72.156	52.3	12.4	33.9	4.2	  
3171	72.245	50.9	10.9	33.9	6.2	  
4383	72.23	43.5	10.9	33.8	3.8	  
4155	72.385	29.6	6.6	33.8	2.3	  
3126	71.955	20.6	5.1	33.8	4.1	  
4156	72.285	32.6	8	33.7	3.6	  
1992	73.062	20.4	5	33.7	4	  
4159	72.38	46.6	11.7	33.5	2.7	  
4003	72.186	29.1	5.8	33.5	2.7	  
3598	72.289	24	6.6	33.5	7.6	  
3538	72.124	29.6	8	33.4	9.3	  
3125	72.057	44	10.9	33.4	3.4	  
1989	72.966	44.2	11.5	33.4	2.3	  
4006	72.331	39.3	8.8	33.3	5.8	  
3536	72.188	36.4	8.8	33.3	13.3	  
4154	72.51	38.8	9.5	32.9	3.9	  
3599	72.501	42.3	10.9	32.9	17.3	  
4008	72.368	30.8	7.3	32.8	7.5	  
3555	72.813	43.1	10.9	32.8	14.6	  
3311	72.19	30.2	8	32.8	3.9	  
3169	72.086	27	7.3	32.8	4.3	  
4165	72.461	21.6	5.8	32.7	4.7	  
3537	72.105	22.7	7.3	32.7	7.7	  
3168	72.049	33.9	10.2	32.6	3.9	  
4091	72.315	38.1	10.2	32.5	5.8	  
3156	72.94	35	8.8	32.5	7.1	  
3123	72.168	36.1	9.5	32.5	8.7	  
4160	72.384	40.9	10.2	32.4	2.8	  
3549	72.583	34.9	8	32.4	19.4	  
2852	72.24	41.7	9.5	32.4	4.9	  
4167	72.596	31.3	8	32.3	5.7	  
3544	72.31	46.4	11.7	32.3	7	  
3179	72.161	25.3	6.6	32.3	7.5	  
4114	72.866	30.7	7.3	32.2	3	  
3183	72.219	55.3	12.4	32.2	7.6	  
4389	72.671	35.8	7.3	32.1	4.2	  
3545	72.388	27.3	6.6	32.1	11.9	  
3232	72.851	32	7.3	32.1	3.8	  
3208	72.22	27.8	7.3	32.1	3.8	  
3175	72.266	22.7	4.4	32.1	13.1	  
1872	72.747	24.8	5.8	32.1	3.8	  
4163	72.598	34.8	8	31.9	5	  
3542	72.368	31.5	8	31.9	9.7	  
3310	72.263	30.4	7.3	31.9	5	  
4115	72.922	46.4	12.4	31.8	7	  
4112	72.81	32.9	8.8	31.8	5.7	  
4016	72.611	35.8	8	31.8	9.1	  
3804	71.778	36.7	8	31.8	3.1	  
2026	72.98	27.5	6.5	31.8	4	  
2021	72.682	43.5	12.3	31.8	3.1	  
4095	72.405	29.2	7.3	31.7	4.5	  
3174	72.288	25.8	6.6	31.7	7.1	  
2850	72.307	31.9	7.3	31.7	5.2	  
4101	72.489	29.5	7.3	31.6	4.3	  
3579	72.63	30.8	8	31.6	13.5	  
3550	72.626	32.8	9.5	31.6	6.2	  
3661	72.298	24.9	6.6	31.5	8	  
1969	72.551	37.5	9.4	31.5	2.4	  
1874	72.629	42.6	9.4	31.5	5.1	  
4177	72.773	38.7	8.8	31.4	4.1	  
4169	72.698	21.6	5.8	31.4	4.1	  
4107	72.705	39.6	8.8	31.4	2.5	  
2006	72.501	42.3	9.4	31.4	3.1	  
4248	72.661	36	9.5	31.2	4.5	  
4170	72.658	34.4	9.5	31.2	6.2	  
4168	72.556	30.5	7.3	31.2	6.4	  
4161	72.457	36.6	10.2	31.2	4.5	  
3594	72.303	49.2	9.5	31.2	10.9	  
3547	72.49	40.5	10.9	31.2	6.2	  
3196	72.835	40.4	10.2	31.2	9.5	  
3180	72.546	39.1	9.5	31.2	11.6	  
4193	72.581	60.4	12.4	31.1	7.5	  
4025	72.811	36.6	10.9	31.1	4.1	  
4098	72.453	23	5.1	31	3.1	  
3541	72.29	44.3	11.7	31	11.9	  
3166	72.11	38.5	10.9	30.9	5.2	  
2854	72.627	33.7	8	30.9	7.2	  
2012	72.576	37.8	8.7	30.9	4.9	  
1763	72.715	21.2	5.8	30.9	25.6	  
4113	72.733	47.1	10.9	30.8	18.9	  
3543	72.4	47.1	13.1	30.8	7.2	  
2149	72.478	53.3	10.8	30.8	5.5	  
4180	72.866	34.3	7.3	30.7	23.7	  
4166	72.58	32.2	7.3	30.7	3.4	  
3955	72.823	28.2	6.6	30.7	3.4	  
3868	72.013	38.6	8.8	30.7	11.1	  
3588	72.21	24.3	5.8	30.7	8.7	  
3571	72.332	26.9	6.6	30.7	13.4	  
3226	72.878	33.1	7.3	30.7	2.8	  
2013	72.601	44.2	11.5	30.7	3.4	  
4178	73.125	30.1	8	30.6	9.8	  
4116	72.807	29.7	8	30.6	5.9	  
4111	72.803	45.8	11.7	30.6	4.6	  
2017	72.68	36.1	7.2	30.6	4.2	  
4120	72.99	37.3	8	30.5	4.8	  
3551	72.701	28	7.3	30.5	8	  
3314	72.163	30	7.3	30.5	4.4	  
3124	72.081	31.8	8.8	30.5	10.3	  
4110	72.685	43.3	11.7	30.4	3.3	  
3589	72.213	32.6	7.3	30.4	25.2	  
3213	72.275	52.1	11.7	30.4	3.8	  
3192	72.49	21.7	5.8	30.4	10.1	  
3151	72.744	33.6	10.2	30.4	9.9	  
4254	72.408	43.3	9.5	30.3	6.2	  
4162	72.494	26.3	5.8	30.3	7	  
3574	72.343	38.7	9.5	30.3	10.8	  
3473	72.245	52.5	13.1	30.3	3.5	  
3470	72.203	49.6	11.7	30.3	3.5	  
3469	72.658	48.1	12.4	30.3	4.5	  
2849	72.286	15.3	4.4	30.3	17.5	  
2005	72.361	47.7	11.5	30.3	2.3	  
4247	72.712	43.8	9.5	30.2	3.2	  
3233	72.705	51.7	11.7	30.2	3.8	  
3204	72.775	54.2	12.4	30.2	11.5	  
2001	72.245	34	7.9	30.2	3.8	  
4250	72.585	48.9	10.2	30.1	8.2	  
4179	72.936	48.4	10.9	30.1	7.8	  
4171	72.663	50.8	12.4	30.1	4.8	  
4030	73.245	31.8	8	30.1	4.4	  
4027	72.96	29.5	8.8	30.1	6.5	  
3566	71.723	43	9.5	30.1	6.7	  
3231	72.906	53.2	12.4	30.1	2.8	  
3159	72.966	38.7	10.2	30.1	23	  
3129	72.203	32.6	7.3	30.1	3.4	  
2025	72.881	48.7	10.1	30.1	6.7	  
1971	72.698	29.8	7.2	30.1	10.8	  
1985	72.886	39.4	9.4	30	10.2	  
1863	72.409	36	7.9	30	7.1	  
1437	72.736	50.7	11.6	30	10.2	  
3773	72.522	41.5	10.2	29.9	6	  
3564	72.18	43.1	10.9	29.9	7.4	  
3212	72.243	43	10.9	29.9	3.2	  
3184	72.261	52.8	11.7	29.9	4.6	  
4253	72.495	55	13.1	29.8	5.5	  
4197	72.681	46.6	12.4	29.8	6.4	  
4106	72.701	34.9	9.5	29.8	5.5	  
3658	72.201	37.4	8	29.8	8	  
3167	72.063	25.7	7.3	29.8	3.3	  
3157	72.614	36.6	8	29.8	7	  
2864	72.133	14.5	4.4	29.8	7.8	  
2335	72.795	26.8	6.5	29.8	4.3	  
2023	72.766	24.6	6.5	29.8	5.5	  
1871	72.527	47.8	12.3	29.8	12.6	  
1869	72.531	44.1	10.1	29.8	7.3	  
1857	72.318	27.1	6.5	29.8	6.1	  
4158	72.438	34.8	8.8	29.7	2.7	  
3595	72.341	59.4	14.6	29.7	10.4	  
3582	72.408	20.6	5.8	29.7	3.3	  
3257	72.458	38.3	8.8	29.7	5.8	  
2369	72.948	53.8	11.5	29.7	6.8	  
2150	72.492	55.3	12.3	29.7	9	  
2144	72.36	47.2	10.8	29.7	5.8	  
1988	72.945	37	8.7	29.7	13.9	  
4119	72.873	33.2	8.8	29.6	4.8	  
4029	73.022	43.8	10.2	29.6	2.7	  
3960	72.915	22.8	5.8	29.6	3.9	  
3957	72.88	20.3	5.8	29.6	2.7	  
3807	72.22	44.4	10.2	29.6	12.3	  
3577	72.419	49.9	12.4	29.6	6.8	  
3227	73.014	38.2	8.8	29.6	2.7	  
3170	72.291	24.5	5.8	29.6	7.3	  
3163	72.758	47.1	10.9	29.6	7.8	  
2019	72.776	31.1	7.2	29.6	7.3	  
4105	72.625	46.9	11.7	29.5	5.2	  
4022	72.9	29.2	6.6	29.5	6.5	  
3769	72.523	48.6	10.9	29.5	4.4	  
3764	72.121	27.7	7.3	29.5	3.9	  
3548	72.615	40.1	8.8	29.5	6.8	  
3165	72.692	49.1	11.7	29.5	7.6	  
2863	72.21	25.6	6.6	29.5	6.2	  
2853	72.386	42.3	10.9	29.5	6.5	  
4323	72.827	38.5	8	29.4	4.8	  
4103	72.608	42.3	11.7	29.4	5.5	  
4028	73.026	47.8	10.2	29.4	3.9	  
4004	72.216	35.5	10.9	29.4	19.3	  
3762	72.428	36.1	8	29.4	5.9	  
3580	72.512	30	7.3	29.4	9.5	  
3565	72.135	37.4	10.9	29.4	18.3	  
3155	72.921	32.2	7.3	29.4	11.5	  
3149	72.541	39.5	9.5	29.4	10.6	  
4099	72.465	37.4	9.5	29.3	10.1	  
3614	72.902	31.9	8.8	29.3	5.1	  
3586	72.402	37.6	8	29.3	3.2	  
3576	72.387	39.5	10.2	29.3	12.8	  
4373	72.487	37.1	8.8	29.2	5.4	  
4251	72.593	66.2	13.9	29.2	4.6	  
4102	72.583	44.3	10.2	29.2	7.8	  
4021	72.763	37.1	9.5	29.2	9	  
3780	72.696	42.6	10.9	29.2	5.1	  
3763	72.213	56.6	13.9	29.2	4.2	  
3632	72.801	45	12.4	29.2	3.1	  
3590	72.4	36.4	9.5	29.2	4.2	  
3173	72.89	48.2	11.7	29.2	7.6	  
3143	72.44	37	8.8	29.2	3.7	  
2427	72.54	32.7	7.2	29.2	6.1	  
1999	72.312	33.1	7.9	29.2	7.6	  
1873	72.616	42.2	10.8	29.2	13	  
4172	72.691	27.5	7.3	29.1	9.8	  
4031	73.203	31.3	8	29.1	6	  
4009	72.41	21.4	5.8	29.1	4.5	  
3876	72.013	44.2	12.4	29.1	9.4	  
3596	72.278	34.2	8	29.1	2.2	  
3230	72.888	29.9	8	29.1	4.1	  
3223	72.749	48.5	10.9	29.1	3.6	  
3211	72.467	55.6	10.9	29.1	4.5	  
3162	72.747	54.4	13.1	29.1	4.1	  
4181	72.855	39.3	8.8	29	6.6	  
4173	72.7	24	6.6	29	2.7	  
4164	72.625	34.6	8.8	29	3.9	  
3813	72.339	25.6	7.3	29	14	  
3383	72.813	38.4	8.8	29	15.5	  
2027	72.088	29.9	7.2	29	5.2	  
4135	72.653	33.6	8	28.9	9.7	  
3319	72.313	52.8	11.7	28.9	10.5	  
3234	72.756	41.7	8.8	28.9	2.4	  
2368	72.956	38.4	8.7	28.9	6.2	  
1972	72.658	40.3	9.4	28.9	8.6	  
4244	72.731	45.5	10.9	28.8	3.5	  
4225	71.756	37.7	8.8	28.8	4	  
4132	72.562	42.8	9.5	28.8	4	  
3822	72.568	23.3	5.8	28.8	8.6	  
3608	72.698	38.5	8.8	28.8	5.7	  
3556	72.778	46.6	11.7	28.8	8.3	  
1994	72.2	22	5	28.8	5.7	  
1868	72.495	39.4	10.1	28.8	4.9	  
4121	72.849	40.5	9.5	28.7	3.7	  
4109	72.604	38.8	10.9	28.7	3.1	  
3382	72.517	51.5	13.1	28.7	4.7	  
3235	72.82	38.6	9.5	28.7	4.7	  
3205	72.714	57.2	12.4	28.7	10.4	  
3186	72.093	51.4	10.2	28.7	7.4	  
2429	72.64	25.8	5.8	28.7	4.3	  
2376	72.867	53.3	10.8	28.7	7.1	  
1968	72.536	36.6	9.4	28.7	11.5	  
4032	73.15	39.7	11.7	28.6	7.3	  
3592	72.407	50.9	13.9	28.6	5.3	  
3569	72.305	28.5	6.6	28.6	5.3	  
2378	72.295	46	10.1	28.6	3.4	  
2377	72.119	49.3	10.8	28.6	3.4	  
2134	72.73	33.9	9.4	28.6	5.7	  
1979	72.761	36.1	10.1	28.6	7.6	  
4472	72.435	33.8	8.8	28.5	5.1	  
4469	72.534	58.5	12.4	28.5	3.6	  
4257	72.369	32.7	8	28.5	5.8	  
4104	72.62	37.4	9.5	28.5	3.6	  
4100	72.499	35.8	8	28.5	4.6	  
4005	72.13	27	6.6	28.5	12.8	  
3820	72.478	28	6.6	28.5	11.2	  
3805	71.689	35.3	10.9	28.5	10.2	  
3570	72.278	15.5	4.4	28.5	5.5	  
3477	72.315	38.9	9.5	28.5	4.2	  
3384	72.82	45.2	10.9	28.5	6.8	  
3153	72.57	36.2	8	28.5	3.6	  
2786	72.311	48.7	10.9	28.5	3	  
2423	72.176	40.6	10.8	28.5	6.5	  
2374	72.902	53.1	10.1	28.5	6.8	  
2031	72.205	47.1	11.5	28.5	3.6	  
4390	72.791	27.7	5.8	28.4	3.8	  
4175	72.745	29.4	7.3	28.4	5.6	  
3770	72.604	36.5	8	28.4	9.9	  
3633	72.685	38.4	8.8	28.4	6.9	  
3581	72.504	39.9	10.9	28.4	3.2	  
3160	72.984	55.1	10.9	28.4	20.1	  
2373	72.167	46.5	10.8	28.4	2.5	  
1862	72.331	47.6	10.8	28.4	10.3	  
4385	72.474	39.6	10.2	28.3	6.4	  
4147	72.9	24.7	6.6	28.3	5.7	  
4097	72.521	35.4	9.5	28.3	3.9	  
4096	72.488	38.8	8	28.3	5.7	  
4023	72.858	26.6	6.6	28.3	4.9	  
4017	72.606	40.8	9.5	28.3	9.1	  
3893	72.516	42.7	10.2	28.3	11.4	  
3668	72.437	51.5	12.4	28.3	2.7	  
3603	72.588	31.2	8.8	28.3	22.1	  
3237	72.479	58.1	11.7	28.3	4.9	  
2379	72.826	47.8	10.1	28.3	8.6	  
2370	72.522	41	9.4	28.3	2.7	  
2135	72.656	53.6	11.5	28.3	3.4	  
2133	72.618	34.4	7.9	28.3	3.9	  
4392	72.736	21.8	5.1	28.2	4	  
4195	72.628	35.4	8	28.2	4.5	  
4174	72.731	28.8	6.6	28.2	5	  
4126	72.461	30.5	8	28.2	6.1	  
4117	72.803	35.1	9.5	28.2	4	  
4015	72.575	25.8	6.6	28.2	5.8	  
3934	72.328	29.6	7.3	28.2	16.2	  
3734	72.691	32	8.8	28.2	12.2	  
3216	72.584	37	9.5	28.2	4	  
3145	72.566	21.2	5.1	28.2	5.4	  
3139	72.417	40.4	9.5	28.2	7.1	  
2851	72.355	13.6	3.6	28.2	5.4	  
2018	72.786	49.1	12.3	28.2	5.8	  
2016	72.675	52.2	11.5	28.2	10.7	  
2004	72.386	21.5	4.3	28.2	10.7	  
1986	72.841	43.3	10.1	28.2	14.4	  
1917	72.466	52.8	13	28.2	2.8	  
1870	72.559	45.5	12.3	28.2	20.9	  
4328	72.874	54.5	11.7	28.1	5	  
4198	72.705	48.7	10.9	28.1	5	  
4192	72.595	42.7	8.8	28.1	9.7	  
4011	72.48	35.3	6.6	28.1	18.7	  
3961	72.922	38.5	10.2	28.1	5.4	  
3954	72.833	28.4	6.6	28.1	4.6	  
3674	72.548	27.2	6.6	28.1	7.7	  
3147	72.627	26.9	7.3	28.1	6.8	  
2381	72.249	25.1	6.5	28.1	3.5	  
4090	72.297	30.6	7.3	28	10.4	  
3840	71.829	34.9	8.8	28	9.5	  
3719	72.249	40.9	10.9	28	11	  
3694	72.308	33.9	8.8	28	4.1	  
3669	72.42	51.5	13.1	28	7.2	  
3567	72.24	26.1	6.6	28	4.1	  
3214	72.519	34.7	8.8	28	6.9	  
3134	72.258	41.3	9.5	28	6.5	  
2040	72.42	49.3	11.5	28	7.2	  
4465	72.936	33	7.3	27.9	2.9	  
4128	72.493	29.6	6.6	27.9	11.2	  
4019	72.704	37.2	8	27.9	4.6	  
3944	72.574	36.6	8	27.9	8.3	  
3940	72.475	38.4	9.5	27.9	8.3	  
3666	72.433	48.5	11.7	27.9	10.2	  
3591	72.425	32	7.3	27.9	6.2	  
3539	72.205	44.4	10.2	27.9	7.2	  
3191	72.51	41.8	10.2	27.9	5.9	  
2163	72.398	35.9	9.4	27.9	2.9	  
4365	72.081	34.8	8.8	27.8	4.1	  
4336	72.55	27.8	8	27.8	2.8	  
4108	72.713	39.4	10.2	27.8	4.1	  
4024	72.797	41.2	10.9	27.8	3.5	  
3605	72.507	38.9	8.8	27.8	4.6	  
3308	72.723	49	10.2	27.8	5.5	  
3228	72.903	47.3	10.9	27.8	5	  
3158	72.703	39.9	8.8	27.8	7.5	  
2426	72.308	29.4	7.9	27.8	4.6	  
2372	72.366	47.5	11.5	27.8	2.8	  
4374	72.488	45.9	10.9	27.7	6.9	  
4144	72.803	28.6	6.6	27.7	9.3	  
3317	72.232	50.9	12.4	27.7	3.4	  
3150	72.816	26.2	5.8	27.7	5.4	  
2428	72.423	10.1	2.9	27.7	3.4	  
2352	72.826	39.6	8.7	27.7	6.5	  
2022	72.773	39.7	10.8	27.7	5.4	  
1970	72.471	36	9.4	27.7	8.6	  
4131	72.525	37.9	10.2	27.6	7.7	  
4118	72.843	42.8	11.7	27.6	14.2	  
4078	72.705	41.3	10.9	27.6	4.9	  
2857	72.491	42.4	10.2	27.6	7.4	  
2848	72.265	36.8	9.5	27.6	6.8	  
2432	72.677	13.4	3.6	27.6	5.3	  
4391	72.748	23.7	5.1	27.5	3.1	  
4370	73.025	52.6	13.9	27.5	2.3	  
4270	72.391	52.6	12.4	27.5	2.3	  
4176	72.785	59.5	13.1	27.5	6.4	  
3948	72.68	30.7	8	27.5	12.6	  
3765	72.108	26.3	7.3	27.5	3.8	  
3726	72.534	56.4	13.1	27.5	10.2	  
3680	72.616	35.1	10.2	27.5	15.2	  
3659	72.232	44.6	11.7	27.5	6	  
3635	72.888	40.2	10.2	27.5	14.7	  
3610	72.791	25.6	7.3	27.5	6.7	  
3154	72.917	19.8	4.4	27.5	10	  
2344	72.868	33.8	9.4	27.5	6	  
4376	72.46	33	8	27.4	6.3	  
4129	72.526	34.4	8.8	27.4	6.3	  
4077	72.609	47.9	13.1	27.4	5.5	  
4055	72.623	38.7	8	27.4	4.7	  
4052	72.54	30.4	5.8	27.4	4.7	  
4002	72.288	39.9	9.5	27.4	16.2	  
3935	72.196	38.2	10.2	27.4	12.2	  
3675	72.499	22.9	6.6	27.4	17	  
3601	72.52	38.3	10.9	27.4	12.4	  
3285	72.506	57.7	12.4	27.4	5.5	  
3255	72.528	40.2	10.2	27.4	5.9	  
2964	72.423	34	8.8	27.4	5.9	  
2862	72.346	35.9	8	27.4	4.1	  
2782	72.547	38.7	9.5	27.4	3.6	  
2331	72.933	25.7	5	27.4	7.3	  
2028	72.024	50.8	12.3	27.4	5.5	  
1906	72.634	24.2	5.8	27.4	2.9	  
4423	72.504	47.7	10.2	27.3	2.5	  
4371	72.59	56.6	12.4	27.3	3.9	  
4081	72.804	31.4	7.3	27.3	3.9	  
3803	71.914	42.2	9.5	27.3	20.2	  
3607	72.593	49.1	11.7	27.3	2.5	  
3563	72.21	38.4	9.5	27.3	3.3	  
3459	72.807	41.7	10.9	27.3	5.8	  
3401	72.355	34.8	9.5	27.3	3.3	  
3290	72.568	52.5	11.7	27.3	4.5	  
2380	72.196	21.2	5	27.3	4.5	  
2046	72.511	57	13.7	27.3	2.5	  
1990	72.945	21.5	5.8	27.3	4.5	  
959	72.65	37.6	9.4	27.3	11.4	  
4461	72.964	49.2	10.2	27.2	4.2	  
4196	72.551	41.9	10.9	27.2	2.1	  
4133	72.556	37.7	9.5	27.2	11.9	  
4130	72.626	29.8	6.6	27.2	10.3	  
3970	72.038	39.3	10.2	27.2	4.2	  
3947	72.669	34.9	8.8	27.2	4.2	  
3766	72.423	41.4	8.8	27.2	4.8	  
3289	72.111	43.1	10.2	27.2	9.3	  
2788	72.538	38.3	8.8	27.2	2.1	  
2367	72.72	35.7	7.9	27.2	3	  
2366	72.716	20.9	5.8	27.2	5.2	  
2038	72.305	41.9	10.1	27.2	11.7	  
4471	72.787	41	9.5	27.1	4	  
4364	72.453	49.5	11.7	27.1	2.6	  
4338	72.445	50.5	10.2	27.1	3.3	  
4337	72.44	29	7.3	27.1	4	  
4141	72.748	25.8	5.8	27.1	6.2	  
4085	72.853	43.6	11.7	27.1	4	  
4037	71.735	53.2	13.1	27.1	3.3	  
3974	72.24	44.7	10.2	27.1	3.3	  
3894	72.351	51.1	13.1	27.1	10.6	  
3885	72.29	30.7	7.3	27.1	20	  
3309	72.575	41.4	9.5	27.1	5.4	  
3144	72.555	32.8	8	27.1	4	  
3137	72.315	20.6	5.1	27.1	5	  
2332	72.967	44.3	10.8	27.1	5.8	  
2158	72.599	15.2	3.6	27.1	3.3	  
1932	72.667	41.3	9.4	27.1	1.4	  
1926	72.56	37.1	9.4	27.1	1.4	  
1900	72.516	36.8	8.7	27.1	4	  
1843	72.625	30.9	7.9	27.1	5	  
4467	72.554	33	7.3	27	3.6	  
4395	72.685	28.7	6.6	27	2	  
4335	72.603	30.7	7.3	27	2.9	  
4088	72.994	20.6	5.1	27	4.2	  
4056	72.532	49.7	10.9	27	2.9	  
3972	72.144	46.1	11.7	27	7.1	  
3946	72.614	22.5	5.8	27	4.2	  
3941	72.521	32.2	6.6	27	13.3	  
3508	72.449	60.3	13.1	27	4.2	  
3476	72.716	34.7	8.8	27	6	  
3474	72.725	38.1	8.8	27	6	  
3259	72.487	45.8	10.2	27	7.1	  
3199	72.411	40.4	8.8	27	3.6	  
2799	72.61	32.6	8	27	2	  
2787	72.608	50.7	10.9	27	3.6	  
2784	72.305	35.9	9.5	27	3.6	  
2382	72.325	34	7.2	27	5.6	  
2030	71.903	32.1	8.7	27	2	  
1925	72.573	34.4	7.9	27	5.6	  
1440	72.124	33.8	9.4	27	5.6	  
4466	72.885	16.7	3.6	26.9	3.2	  
4333	72.626	20.2	6.6	26.9	2.3	  
4329	72.691	45.1	10.9	26.9	3.8	  
4271	72.271	33.6	7.3	26.9	3.2	  
4151	72.99	31	8	26.9	8.7	  
3952	72.737	42	10.2	26.9	3.8	  
3936	72.225	47.8	10.9	26.9	12.5	  
3777	72.816	33.4	10.2	26.9	3.8	  
3623	72.603	42.5	10.2	26.9	9.2	  
3583	72.64	40.6	9.5	26.9	6.9	  
3249	72.513	59.2	11.7	26.9	3.8	  
3187	72.41	31.9	7.3	26.9	6.5	  
3035	72.407	37.6	10.2	26.9	2.3	  
2790	72.733	48.2	10.9	26.9	4.9	  
2781	72.506	28.9	6.6	26.9	3.2	  
2730	72.872	42.8	10.9	26.9	4.9	  
2351	72.551	52.3	10.8	26.9	6.2	  
4457	72.641	25.5	6.6	26.8	5.9	  
4442	72.648	21.8	5.1	26.8	2.6	  
4366	72.18	32.4	8.8	26.8	4	  
4355	71.875	47.1	10.9	26.8	3.4	  
4137	72.721	46.9	10.9	26.8	7	  
4127	72.446	29.4	7.3	26.8	8	  
4094	72.351	29.4	6.6	26.8	4	  
4062	72.31	34.4	8.8	26.8	8	  
4053	72.424	31.8	8	26.8	4	  
4013	72.407	33.9	8	26.8	13.1	  
3462	72.797	22.9	7.3	26.8	7	  
3458	72.629	52.9	11.7	26.8	4	  
3403	72.236	45.2	10.9	26.8	4	  
3349	72.137	45.8	11.7	26.8	4	  
3321	72.526	37.9	8	26.8	6.3	  
3254	72.692	49.3	12.4	26.8	3.4	  
3176	72.839	50.5	10.9	26.8	5.1	  
3142	72.366	37.5	8.8	26.8	8.3	  
3032	72.42	31	8.8	26.8	2.6	  
2861	72.325	22.5	5.1	26.8	4	  
2433	72.506	40.9	9.4	26.8	5.1	  
2131	72.491	24.7	6.5	26.8	4	  
2044	72.428	37.6	7.9	26.8	9.1	  
2029	71.984	49.3	10.8	26.8	2.6	  
2020	72.706	24.6	5.8	26.8	4.6	  
1984	72.87	35.5	9.4	26.8	15.6	  
1953	72.643	32.4	6.5	26.8	4	  
4443	72.641	14.5	3.6	26.7	2.8	  
4422	72.493	56.8	12.4	26.7	2.8	  
4369	72.705	25	8	26.7	2.8	  
4134	72.621	40.5	9.5	26.7	4.2	  
4058	72.32	33.8	8.8	26.7	6	  
3968	71.911	40	9.5	26.7	6.4	  
3879	72.208	36.1	10.2	26.7	16.7	  
3649	72.313	26.5	6.6	26.7	4.2	  
3619	72.323	48.1	10.9	26.7	8.6	  
3612	72.698	27.6	5.8	26.7	2.8	  
3347	72.173	39.3	8.8	26.7	4.2	  
3297	72.041	40.2	8.8	26.7	5.2	  
3247	72.371	46.3	10.2	26.7	5.2	  
3225	72.826	44.5	10.9	26.7	5.2	  
2371	72.801	20	5	26.7	7.4	  
1895	72.376	47.5	11.5	26.7	6.4	  
1841	72.535	29.8	8.7	26.7	2.8	  
4456	72.633	55	13.1	26.6	5.7	  
4317	71.997	47.2	10.9	26.6	6.9	  
4273	72.027	43.1	9.5	26.6	5.3	  
4243	72.723	20.2	5.1	26.6	3	  
4148	72.937	27.4	5.8	26.6	6.1	  
4140	72.761	28.9	6.6	26.6	8.7	  
4069	72.498	36.9	8.8	26.6	4.8	  
3976	72.201	35.6	8	26.6	3	  
3964	72.734	54.2	13.9	26.6	10.7	  
3877	72.106	34.1	7.3	26.6	14.5	  
3679	72.643	38.9	10.2	26.6	3.7	  
3575	72.319	31	8	26.6	11.1	  
3424	72.166	25	5.8	26.6	4.8	  
3329	72.537	35.8	10.2	26.6	4.3	  
3286	71.925	45.6	11.7	26.6	11.3	  
3194	72.386	9.4	2.2	26.6	3.7	  
3193	72.39	24.7	5.8	26.6	4.8	  
2743	72.868	15.4	3.6	26.6	4.3	  
2348	72.935	37.9	8.7	26.6	9.5	  
2219	72.586	32.4	7.2	26.6	6.1	  
4331	72.586	24.1	6.6	26.5	3	  
3881	72.151	37	9.5	26.5	6.9	  
3771	72.62	40.9	10.2	26.5	6.9	  
3198	72.406	25.5	5.1	26.5	3.7	  
3130	72.263	48.9	12.4	26.5	15.7	  
3054	72.669	36.7	8.8	26.5	18.8	  
3045	72.613	38.4	10.9	26.5	4.3	  
2968	72.538	37.9	10.2	26.5	3.7	  
2961	72.32	31.6	8.8	26.5	5.8	  
2858	72.182	39.4	8.8	26.5	3.7	  
2774	72.507	24.8	6.6	26.5	3.7	  
4368	72.491	28	6.6	26.4	3.8	  
4367	72.651	27.7	7.3	26.4	3.1	  
4139	72.795	35.5	8	26.4	3.1	  
4001	72.686	20.8	5.1	26.4	3.1	  
3953	72.74	34.4	8.8	26.4	7.6	  
3951	72.777	34.1	7.3	26.4	3.8	  
3921	72.581	40.1	8	26.4	3.1	  
3867	72.095	43.6	10.9	26.4	10.8	  
3616	72.216	48.5	12.4	26.4	5.4	  
3475	72.599	39.7	10.2	26.4	4.4	  
3328	72.288	29	5.8	26.4	6.9	  
3298	71.908	45.7	10.2	26.4	4.4	  
3296	72.197	32.3	7.3	26.4	5.4	  
2920	72.541	27	5.8	26.4	3.8	  
2891	72.656	33.4	8	26.4	3.8	  
2860	72.341	33.3	7.3	26.4	3.1	  
2574	72.505	24.2	6.6	26.4	3.8	  
2268	72.35	28.5	7.9	26.4	3.8	  
1976	72.765	38.9	9.4	26.4	4.4	  
4470	72.841	29.2	6.6	26.3	3.1	  
4421	71.628	45.5	10.2	26.3	3.8	  
4419	72.495	54	13.9	26.3	3.1	  
4316	72.099	34.9	10.2	26.3	7.9	  
4228	71.983	35.9	8.8	26.3	9.9	  
4145	72.885	24.2	5.8	26.3	4.9	  
3982	72.428	29.2	7.3	26.3	6.6	  
3971	72.012	40.2	8.8	26.3	7.3	  
3943	72.608	36.1	9.5	26.3	12.1	  
3939	72.386	39.4	9.5	26.3	3.1	  
3870	72.605	45.7	12.4	26.3	15.7	  
3767	72.395	31.4	8.8	26.3	16.4	  
3672	72.511	31.2	9.5	26.3	6.2	  
3631	72.798	34.8	8.8	26.3	12.9	  
3490	72.381	18.9	5.1	26.3	3.8	  
3264	72.413	19.9	5.1	26.3	3.1	  
3148	72.656	37.7	8.8	26.3	11.5	  
2888	72.639	31	7.3	26.3	3.1	  
2789	72.53	32	9.5	26.3	4.9	  
2422	72.486	20.9	4.3	26.3	3.1	  
2421	72.523	28.9	6.5	26.3	3.8	  
2036	72.365	49.1	12.3	26.3	4.4	  
2034	72.267	47.9	12.3	26.3	4.4	  
1914	72.358	36.9	9.4	26.3	8.3	  
4394	72.706	43.3	10.9	26.2	4.3	  
4330	72.836	40.8	10.2	26.2	3	  
4136	72.698	30.5	7.3	26.2	7.3	  
4087	72.99	25	6.6	26.2	4.9	  
4075	72.642	26.9	6.6	26.2	3.7	  
4020	72.711	29.4	8.8	26.2	16	  
3786	72.11	31.8	7.3	26.2	7	  
3744	72.315	43.6	10.9	26.2	7	  
3697	72.4	34	8.8	26.2	3.7	  
3422	73.041	58.6	12.4	26.2	3	  
3380	72.624	43	9.5	26.2	7.6	  
3353	72.363	47.8	11.7	26.2	3	  
2889	72.778	38.8	10.2	26.2	3	  
2746	72.725	27.3	6.6	26.2	5.4	  
2434	72.604	22.7	5	26.2	3.7	  
2353	72.811	32.9	7.9	26.2	5.4	  
1859	72.246	26.2	5.8	26.2	7.6	  
4446	72.673	37.2	9.5	26.1	3.6	  
4445	72.678	18.5	5.1	26.1	2.9	  
4432	72.402	43.2	10.9	26.1	2.9	  
4418	72.499	37.3	8.8	26.1	2.9	  
4407	72.583	43	10.9	26.1	2.9	  
4298	72.299	53.8	12.4	26.1	4.8	  
4199	72.615	40.8	8.8	26.1	4.3	  
4138	72.805	29.8	7.3	26.1	4.8	  
4076	72.58	20.9	5.1	26.1	6.9	  
4067	72.505	31.1	8	26.1	3.6	  
4047	72.374	25.2	5.1	26.1	2.9	  
3969	71.923	28.2	7.3	26.1	5.3	  
3889	72.285	32.4	8	26.1	7.3	  
3887	72.382	28.1	6.6	26.1	15.4	  
3644	72.213	46.1	11.7	26.1	3.6	  
3622	72.516	37.1	8	26.1	13.1	  
3507	72.706	53.4	11.7	26.1	2.9	  
3481	72.741	41.2	8.8	26.1	5.8	  
3334	72.736	34.7	9.5	26.1	4.3	  
3140	72.476	43.6	10.9	26.1	7.9	  
3138	72.33	23.2	5.8	26.1	2.9	  
3132	72.253	13.1	2.9	26.1	3.6	  
2823	72.928	52.3	11.7	26.1	5.3	  
2745	72.799	20.4	5.1	26.1	4.3	  
2334	73.065	40.1	8.7	26.1	9.6	  
2129	72.49	18.7	4.3	26.1	3.6	  
1910	72.631	37.3	10.1	26.1	4.3	  
1896	72.418	32.6	9.4	26.1	4.8	  
4433	72.534	33	6.6	26	2.7	  
4429	72.393	30.3	8	26	4.2	  
4406	72.539	26.4	6.6	26	2.7	  
4372	72.55	21.9	5.1	26	2.7	  
4359	72.057	44.6	9.5	26	4.7	  
4352	71.874	46.8	10.9	26	3.5	  
4349	72.77	43.7	9.5	26	3.5	  
4083	72.899	30.4	6.6	26	2.7	  
4082	72.898	36.3	9.5	26	2.7	  
4079	72.741	41.1	10.9	26	2.7	  
4014	72.536	37.3	8.8	26	2.7	  
3997	72.721	25.2	6.6	26	3.5	  
3869	72.12	38	10.2	26	8.5	  
3779	72.681	32.8	8	26	9.1	  
3624	72.489	44.9	10.9	26	8.2	  
3573	72.285	35	8.8	26	6.1	  
3480	72.444	34	6.6	26	6.5	  
3239	72.726	49	10.9	26	6.9	  
2868	72.278	35.7	9.5	26	3.5	  
2791	72.503	51.7	11.7	26	3.5	  
2663	72.647	23.5	5.8	26	4.2	  
2625	72.473	28.4	7.3	26	4.2	  
2621	72.561	32.2	6.6	26	4.2	  
2542	72.79	41.2	9.5	26	2.7	  
2436	72.306	25.7	7.2	26	2.7	  
2132	72.549	29.9	7.2	26	4.2	  
4358	72.486	58	12.4	25.9	2.4	  
4322	72.206	36.8	9.5	25.9	3.3	  
4301	72.16	34.6	8.8	25.9	4	  
4272	72.146	45.4	11.7	25.9	4.6	  
4152	73.025	34.8	10.2	25.9	4	  
4124	72.285	38.9	9.5	25.9	5.6	  
4086	72.961	20.6	5.1	25.9	2.4	  
4071	72.457	31.7	6.6	25.9	5.6	  
3686	72.717	28.8	6.6	25.9	14.1	  
3625	72.505	32.2	8	25.9	5.1	  
3546	72.507	42.3	10.9	25.9	24.1	  
3464	72.781	33.9	8	25.9	7.5	  
3418	73.098	55.9	12.4	25.9	3.3	  
3332	72.809	56	13.9	25.9	3.3	  
3291	72.359	34.7	8.8	25.9	2.4	  
3253	72.725	53.5	12.4	25.9	4	  
3141	72.433	44.3	9.5	25.9	3.3	  
2731	72.924	45.7	10.9	25.9	4.6	  
2049	72.685	51.8	13.7	25.9	4.6	  
2047	72.591	46	10.8	25.9	8.2	  
4454	72.614	39.6	8.8	25.8	4.9	  
4413	72.58	44.1	9.5	25.8	2.1	  
4327	72.506	33.8	8.8	25.8	3.8	  
4269	72.494	48	10.9	25.8	3.1	  
4268	72.313	47.7	10.2	25.8	2.1	  
4123	72.353	33.6	8	25.8	3.1	  
3994	72.647	30.8	7.3	25.8	4.4	  
3993	72.526	47.2	11.7	25.8	4.4	  
3600	72.487	44.1	9.5	25.8	7.4	  
3572	72.283	36.6	8.8	25.8	5.4	  
3468	72.728	44.9	11.7	25.8	3.8	  
3423	72.983	63.5	13.1	25.8	3.1	  
3417	73.006	57.1	13.1	25.8	3.8	  
3352	72.266	34.6	8	25.8	3.1	  
3236	72.711	50.7	11.7	25.8	7.4	  
3201	72.465	60.7	14.6	25.8	4.9	  
3189	72.429	53.8	10.9	25.8	3.8	  
2506	72.466	22.7	5.1	25.8	7.8	  
2430	72.266	31.3	7.9	25.8	3.8	  
2273	72.446	36	9.4	25.8	12.5	  
2266	72.31	33.6	7.2	25.8	3.8	  
4463	72.471	38.9	10.9	25.7	5.3	  
4444	72.74	33.9	8	25.7	2.7	  
4379	72.58	39.5	8.8	25.7	5.3	  
4146	72.898	35.1	8.8	25.7	4.8	  
4080	72.706	24.4	5.8	25.7	6.2	  
4054	72.51	38.2	8.8	25.7	2.7	  
3959	73.003	26.2	5.1	25.7	9.7	  
3706	72.583	45.7	9.5	25.7	6.6	  
3467	72.79	35.5	7.3	25.7	6.6	  
3404	72.168	57.7	13.1	25.7	8	  
3222	72.078	53.4	11.7	25.7	6.9	  
3210	72.481	44.4	9.5	25.7	7.3	  
3040	72.51	39.4	10.9	25.7	8.6	  
2946	72.098	33.9	8	25.7	4.2	  
2716	72.637	42.8	10.9	25.7	2.7	  
2683	72.665	44.5	10.2	25.7	4.8	  
2619	72.583	33.2	8	25.7	3.5	  
2539	72.633	17	4.4	25.7	2.7	  
2271	72.435	29.8	7.2	25.7	6.9	  
2246	72.637	43.8	10.1	25.7	1.5	  
2162	72.615	23.9	5.8	25.7	5.3	  
2009	72.461	32.6	9.4	25.7	4.2	  
1851	71.937	52.9	12.3	25.7	3.5	  
4417	71.725	48.9	13.1	25.6	4.5	  
4299	72.31	29.5	7.3	25.6	3.9	  
4295	72.728	46.6	11.7	25.6	4.5	  
4125	72.223	52.1	12.4	25.6	5.5	  
4073	72.615	42.2	11.7	25.6	3.2	  
4070	72.452	30.4	8	25.6	3.9	  
4065	72.41	32.6	8	25.6	7.2	  
4012	72.512	27.9	8	25.6	3.9	  
3989	72.544	50.8	12.4	25.6	6.8	  
3980	72.31	36.2	8.8	25.6	7.2	  
3912	72.391	42.7	10.9	25.6	6.4	  
3711	72.693	30.4	7.3	25.6	3.9	  
3695	72.213	38.1	11.7	25.6	3.9	  
3629	72.703	30.4	6.6	25.6	6.4	  
3053	72.686	31.5	7.3	25.6	6	  
2886	72.54	35.6	8.8	25.6	5.1	  
2859	72.301	14.2	3.6	25.6	2.2	  
2837	72.313	39.9	8.8	25.6	3.2	  
2780	72.383	18.5	4.4	25.6	0	  
2768	72.604	28.2	6.6	25.6	3.2	  
2748	72.749	41.3	10.9	25.6	4.5	  
2742	72.919	42.5	10.9	25.6	4.5	  
2728	72.646	42.2	10.2	25.6	7.2	  
2666	72.69	42.5	8.8	25.6	0	  
2541	72.88	41.9	10.2	25.6	3.9	  
2470	72.757	25.6	6.6	25.6	6.4	  
2215	72.46	43.7	10.8	25.6	5.5	  
2138	72.542	34.5	8.7	25.6	5.5	  
2039	72.42	36.1	8.7	25.6	5.1	  
1907	72.7	33.2	7.9	25.6	5.1	  
1850	71.937	31.5	7.9	25.6	23.6	  
761	72.548	35.2	8.7	25.6	7.2	  
4319	72.088	37	10.2	25.5	2.7	  
4149	72.941	31.6	8	25.5	4.8	  
4142	72.848	34.8	8.8	25.5	4.8	  
3922	72.729	32.3	7.3	25.5	2.7	  
3915	72.458	38.4	10.2	25.5	6.2	  
3819	72.503	44.2	12.4	25.5	4.8	  
3798	72.606	33.4	9.5	25.5	1.5	  
3730	72.566	63.3	14.6	25.5	10.3	  
3712	72.791	31	7.3	25.5	2.7	  
3455	72.456	42.5	10.2	25.5	5.3	  
3419	72.321	34.2	9.5	25.5	5.8	  
3330	72.649	44	10.2	25.5	6.2	  
3292	72.251	36.3	9.5	25.5	4.8	  
3251	72.484	27	6.6	25.5	4.8	  
3220	72.423	38.5	9.5	25.5	4.8	  
3086	72.798	32.4	8	25.5	2.7	  
3038	72.485	37.9	8.8	25.5	5.8	  
3027	72.805	30.7	8	25.5	4.8	  
2836	72.298	43.3	10.9	25.5	2.7	  
2779	72.351	23.6	6.6	25.5	2.7	  
2775	72.414	29.8	7.3	25.5	1.5	  
2772	72.513	24.4	6.6	25.5	3.5	  
2763	72.576	23.8	5.1	25.5	4.2	  
2613	72.485	38.4	8.8	25.5	4.2	  
2545	72.685	38.6	10.9	25.5	4.2	  
2045	72.458	48	10.8	25.5	4.2	  
2041	72.405	48.4	10.8	25.5	4.2	  
1983	72.859	39.9	10.8	25.5	4.2	  
1982	72.856	38.1	9.4	25.5	23.9	  
1962	72.961	34.6	7.9	25.5	4.2	  
1960	72.844	28.5	7.2	25.5	4.2	  
1761	72.491	44	10.9	25.5	17.9	  
4434	72.346	43.6	9.5	25.4	3.1	  
4382	72.516	33.2	8.8	25.4	8.2	  
4381	72.582	45.3	10.9	25.4	7.2	  
4348	72.775	29.2	7.3	25.4	2.1	  
4345	72.745	26.3	6.6	25.4	2.1	  
4339	72.377	25.4	5.8	25.4	5	  
4326	72.29	55.1	12.4	25.4	5	  
4324	72.348	36.1	8	25.4	2.1	  
4066	72.544	27	6.6	25.4	3.9	  
3808	72.22	30.8	8	25.4	10.7	  
3717	72.853	21.5	5.8	25.4	6.4	  
3426	72.225	50.7	11.7	25.4	6	  
3322	72.697	34.9	8.8	25.4	3.9	  
3271	72.068	53.3	12.4	25.4	2.1	  
3265	72.548	44.4	10.2	25.4	5	  
3258	72.43	42.6	10.2	25.4	4.5	  
3146	72.56	47.1	11.7	25.4	7.5	  
2877	72.441	41	8.8	25.4	11.2	  
2767	72.653	27.5	6.6	25.4	3.1	  
2247	72.745	46.4	10.8	25.4	6	  
2214	72.406	54.3	10.8	25.4	4.5	  
1867	72.423	51.2	13	25.4	3.9	  
4448	72.621	43.6	10.2	25.3	2.5	  
4402	72.542	31.3	8.8	25.3	2.5	  
4401	72.466	29.6	6.6	25.3	2.5	  
4332	72.211	26	5.8	25.3	4.7	  
3931	72.803	46.8	12.4	25.3	8	  
3874	71.73	43.7	11.7	25.3	6.6	  
3836	72.508	36.6	9.5	25.3	19.2	  
3778	72.817	34.3	7.3	25.3	4.1	  
3621	72.425	42.8	11.7	25.3	5.7	  
3489	72.474	39.3	9.5	25.3	3.4	  
3338	72.52	36.7	8.8	25.3	2.5	  
3248	72.766	48.1	11.7	25.3	4.1	  
3221	72.291	42.7	10.2	25.3	6.2	  
3188	72.422	28.4	8.8	25.3	3.4	  
3057	72.783	23	5.8	25.3	5.2	  
3044	72.607	39.9	9.5	25.3	4.7	  
2883	72.571	28.3	7.3	25.3	4.7	  
2855	72.133	46.4	10.2	25.3	7.3	  
2793	72.562	26.4	5.8	25.3	3.4	  
2765	72.798	28.4	7.3	25.3	4.7	  
2761	72.571	34	8	25.3	10.6	  
2618	72.693	24.9	5.8	25.3	6.6	  
2576	72.379	44.1	8.8	25.3	4.1	  
2559	72.302	32.1	8	25.3	7	  
2431	72.245	35.2	7.9	25.3	5.2	  
2419	72.598	40.4	10.8	25.3	3.4	  
2257	72.4	39.2	10.8	25.3	7.7	  
957	72.728	16.6	4.3	25.3	8.7	  
3034	72.322	28.8	7.3	25.27	21.4	  
4450	72.361	41.2	10.9	25.2	3.6	  
4396	72.436	54.7	12.4	25.2	3.6	  
4378	72.508	26	7.3	25.2	4.3	  
4302	72.02	25.6	5.8	25.2	8.1	  
4290	72.759	49.7	10.9	25.2	2.8	  
4229	72.436	52.6	10.9	25.2	2.8	  
4068	72.518	36	9.5	25.2	5.4	  
4064	72.436	46.1	13.1	25.2	4.3	  
4046	72.328	42.7	10.2	25.2	9.4	  
3990	72.594	41.7	8.8	25.2	5.4	  
3886	72.396	44.5	10.2	25.2	14.5	  
3665	72.438	53.3	12.4	25.2	10.4	  
3652	72.401	44.8	10.9	25.2	11.7	  
3651	72.417	35.8	8.8	25.2	7.1	  
3640	72.041	37.9	9.5	25.2	3.6	  
3492	72.603	20.2	4.4	25.2	2.8	  
3482	72.505	43	9.5	25.2	4.3	  
3312	72.722	45.3	10.2	25.2	11.7	  
3135	72.34	39.5	10.2	25.2	7.5	  
2974	72.744	26.3	5.8	25.2	2.8	  
2511	72.7	43.8	10.9	25.2	4.3	  
2510	72.561	50.1	10.2	25.2	4.9	  
2509	72.587	23.5	7.3	25.2	7.1	  
2183	72.253	29.1	7.9	25.2	2.8	  
2179	72.303	31.5	8.7	25.2	4.3	  
2015	72.653	41.5	9.4	25.2	4.3	  
1866	72.418	35.4	8.7	25.2	12.1	  
1860	72.363	22.1	5	25.2	2.8	  
1853	72.125	53.9	12.3	25.2	2.8	  
4441	72.526	36.4	9.5	25.1	3	  
4440	72.532	23.5	5.1	25.1	4.4	  
4363	72.187	27.9	6.6	25.1	3	  
4296	72.647	36.9	10.2	25.1	3.8	  
4059	72.191	29.8	7.3	25.1	4.4	  
3986	72.424	37.5	9.5	25.1	8.3	  
3938	72.276	37.5	8.8	25.1	3.8	  
3848	72.198	32.8	9.5	25.1	5.5	  
3818	72.605	37.5	9.5	25.1	6	  
3776	72.72	35	8	25.1	5	  
3710	72.72	42.7	10.2	25.1	8.6	  
3685	72.749	33.7	9.5	25.1	3.8	  
3294	72.275	20	4.4	25.1	3	  
3250	72.809	22.9	5.8	25.1	3	  
3177	72.726	43.9	11.7	25.1	6.4	  
3133	72.211	26.3	5.8	25.1	3	  
2885	72.456	26.7	6.6	25.1	3.8	  
2769	72.743	37.3	9.5	25.1	22.7	  
2764	72.638	25.8	5.8	25.1	3	  
2620	72.736	45.7	10.2	25.1	5	  
2582	72.433	27.7	6.6	25.1	6.4	  
2254	72.478	32.5	8.7	25.1	5	  
1974	72.648	37.9	9.4	25.1	11.8	  
1883	72.242	43.7	11.5	25.1	3	  
1856	72.301	39.8	9.4	25.1	3	  
4435	72.29	27.3	5.8	25	3.2	  
4408	72.229	44.3	10.2	25	3.2	  
4313	72.27	48.4	12.4	25	3.9	  
3958	72.529	41.8	9.5	25	20.4	  
3872	72.03	48.9	12.4	25	17.3	  
3741	72.504	37.2	8.8	25	9.8	  
3689	72.806	22.7	5.8	25	5.1	  
3681	72.747	39.9	10.9	25	6.5	  
3627	72.616	40.9	10.2	25	6.1	  
3618	71.821	47.6	11.7	25	3.1	  
3502	72.724	56.5	13.1	25	3.2	  
3190	72.466	18.9	4.4	25	3.2	  
3089	72.905	33.2	8	25	3.2	  
2973	72.638	35	9.5	25	3.1	  
2807	72.998	54	13.1	25	4.6	  
2717	72.5	51.8	10.9	25	5.1	  
2535	72.831	18.6	4.4	25	3.2	  
2218	72.478	48	10.1	25	6.1	  
1980	72.736	38.6	9.4	25	3.9	  
1864	72.53	47.4	11.5	25	3.2	  
4414	72.391	44.8	9.5	24.9	3.2	  
4305	73.044	52.1	11.7	24.9	4	  
4049	72.411	20.4	5.1	24.9	4	  
4000	72.694	29.4	7.3	24.9	3.2	  
3995	72.656	22.7	5.1	24.9	5.2	  
3913	72.398	42.9	10.2	24.9	4	  
3906	72.306	28.9	8.8	24.9	3.2	  
3718	72.89	44.7	10.2	24.9	4.6	  
3713	72.786	43	10.9	24.9	7.4	  
3617	72.145	43.5	10.2	24.9	4.6	  
3615	72.2	45.9	10.9	24.9	4	  
3449	72.885	40	8	24.9	3.2	  
3327	72.371	51.2	11.7	24.9	4.6	  
3268	72.418	22.7	5.1	24.9	5.7	  
3263	72.593	58.2	13.1	24.9	4.6	  
3217	72.683	53.3	10.9	24.9	7.4	  
3207	72.631	25.7	6.6	24.9	5.2	  
3048	72.596	35.5	8	24.9	3.2	  
3017	72.713	47.8	10.9	24.9	9.9	  
2941	72.756	19.9	4.4	24.9	4.6	  
2802	72.883	20.9	5.1	24.9	3.2	  
2771	72.547	32.9	9.5	24.9	10.2	  
2735	72.943	44.6	10.2	24.9	3.2	  
2614	72.578	46.7	9.5	24.9	4	  
2465	72.797	32.3	7.9	24.9	5.2	  
2416	72.879	45.6	10.1	24.9	6.1	  
1993	72.956	32.8	7.9	24.9	17.3	  
1894	72.323	39.1	10.1	24.9	7	  
1762	72.651	25.3	5.8	24.9	16	  
4464	72.45	35.5	8	24.8	3.3	  
4460	72.243	54.2	12.4	24.8	3.3	  
4449	72.473	49.7	13.1	24.8	3.3	  
4320	72.398	39.8	9.5	24.8	4.7	  
4293	72.305	20.9	6.6	24.8	5.7	  
4200	72.211	36.5	8.8	24.8	4	  
4026	72.81	53.4	10.9	24.8	12	  
3843	72.131	35.1	9.5	24.8	9.4	  
3720	72.133	63.2	13.9	24.8	12.6	  
3709	72.72	34.1	8	24.8	4.7	  
3698	72.455	37.5	9.5	24.8	14.7	  
3690	72.783	18.5	5.1	24.8	4	  
3495	72.611	47.7	11.7	24.8	3.3	  
3491	72.578	28.2	6.6	24.8	3.3	  
3450	72.856	51.8	12.4	24.8	3.3	  
3331	72.629	30.4	7.3	24.8	5.2	  
3283	71.809	44.6	10.2	24.8	9.1	  
3270	72.206	45.1	12.4	24.8	4.7	  
3087	72.897	25.4	5.8	24.8	8.4	  
3051	72.703	42.2	10.9	24.8	7.4	  
2812	72.754	47.3	11.7	24.8	3.3	  
2811	72.851	52.7	10.9	24.8	3.3	  
2776	72.45	28.7	7.3	24.8	3.3	  
2773	72.485	18.6	5.1	24.8	3.3	  
2729	72.715	31.6	8	24.8	7	  
2627	72.353	16.2	4.4	24.8	4	  
2612	72.516	30.3	8	24.8	9.7	  
2581	72.388	38.1	9.5	24.8	7	  
2577	72.675	23.3	5.1	24.8	4.7	  
2466	72.775	27.8	7.2	24.8	3.3	  
2417	72.822	48.9	11.5	24.8	4	  
2407	72.533	35.9	9.4	24.8	4.7	  
2336	72.992	38.9	8.7	24.8	3.3	  
2033	72.302	35.6	9.4	24.8	8.1	  
2032	72.284	39.4	9.4	24.8	3.3	  
1865	72.445	29.4	7.2	24.8	4	  
1837	72.544	41.3	10.8	24.8	3.3	  
1834	72.441	46.4	11.5	24.8	5.7	  
1760	72.555	44.7	10.2	24.8	17.7	  
1444	72.031	37.6	8.7	24.8	7.4	  
284	72.247	33.3	8	24.8	15.4	  
4439	72.476	49.7	10.2	24.7	3.3	  
4400	72.612	38.9	9.5	24.7	3.3	  
4063	72.219	39.7	9.5	24.7	11.8	  
4043	72.238	55.4	13.9	24.7	4	  
3996	72.643	28.9	8	24.7	4.7	  
3920	72.568	35.4	8.8	24.7	9.7	  
3903	72.106	52.8	12.4	24.7	7	  
3883	72.293	36.2	9.5	24.7	10.3	  
3655	72.5	40	9.5	24.7	5.7	  
3646	72.398	39.6	10.2	24.7	5.7	  
3628	72.619	42.1	10.2	24.7	10.8	  
3620	72.313	50.8	13.1	24.7	8.1	  
3446	72.714	27.5	7.3	24.7	4.7	  
3256	72.327	39.2	10.9	24.7	9.1	  
3246	72.45	33.2	8.8	24.7	11	  
3238	72.718	57	11.7	24.7	3.3	  
3121	73.255	48.5	11.7	24.7	4	  
3120	73.155	38.1	10.2	24.7	3.3	  
3090	72.995	27.8	7.3	24.7	3.3	  
2803	72.96	30.5	8	24.7	3.3	  
2453	72.971	48.5	11.5	24.7	5.2	  
2354	72.838	14.7	2.9	24.7	5.2	  
1435	72.636	28.4	6.5	24.7	8.1	  
4452	72.22	46.8	10.9	24.6	3.2	  
4344	72.696	28	6.6	24.6	3.2	  
4343	72.624	47	10.9	24.6	3.2	  
4258	72.389	27.4	7.3	24.6	6.2	  
4074	72.538	51.2	14.6	24.6	15.7	  
3978	72.277	33.3	6.6	24.6	6.6	  
3963	72.991	48.4	11.7	24.6	11.3	  
3917	72.361	21.5	5.1	24.6	5.7	  
3904	72.196	38.4	8.8	24.6	12.7	  
3781	72.856	20.2	5.1	24.6	4	  
3647	72.292	38.9	9.5	24.6	3.2	  
3478	72.35	22.6	5.1	24.6	4.6	  
3400	72.391	49	10.9	24.6	5.2	  
3343	72.254	46	11.7	24.6	4	  
3293	71.931	20	5.1	24.6	4	  
3288	71.993	24.9	7.3	24.6	3.2	  
3284	72.328	46	11.7	24.6	6.2	  
3128	72.138	33.9	8	24.6	5.7	  
2935	72.667	33.9	8.8	24.6	5.2	  
2766	72.814	33.2	9.5	24.6	3.2	  
2547	72.598	31.5	8	24.6	4.6	  
2505	72.693	35.7	8	24.6	4.6	  
2448	72.871	38.9	10.8	24.6	4.6	  
2413	72.871	37.9	8.7	24.6	5.2	  
2188	72.303	48.9	10.8	24.6	3.2	  
2011	72.61	26.3	7.2	24.6	4	  
1959	72.853	38.9	10.1	24.6	3.2	  
1933	72.633	26	5.8	24.6	8.1	  
4447	72.408	29.4	6.6	24.5	3.1	  
4431	72.409	22.6	5.1	24.5	3.9	  
4397	72.666	35.9	7.3	24.5	3.9	  
4357	72.473	51.2	11.7	24.5	3.1	  
4325	72.307	41.7	9.5	24.5	7	  
4304	73.032	51.7	12.4	24.5	3.1	  
4224	71.718	50.7	10.9	24.5	4.5	  
4041	72.268	41.4	9.5	24.5	11.6	  
4007	72.319	39.2	9.5	24.5	5.1	  
3966	72.036	30.3	5.8	24.5	3.1	  
3789	72.204	36.5	8	24.5	8.8	  
3501	72.721	30	7.3	24.5	3.1	  
3483	72.605	25	5.8	24.5	4.5	  
3463	72.755	24.4	6.6	24.5	3.9	  
3427	72.306	44.2	10.2	24.5	4.5	  
3372	72.892	23.5	5.1	24.5	3.9	  
3335	72.648	40.9	10.2	24.5	5.1	  
3116	73.035	45.7	10.2	24.5	3.1	  
3088	72.881	41.9	10.9	24.5	6.6	  
3001	72.807	37.6	8	24.5	8.1	  
2934	72.59	28.6	8	24.5	5.7	  
2839	72.335	33	9.5	24.5	3.1	  
2736	72.964	37.7	9.5	24.5	5.1	  
2700	72.775	35.7	9.5	24.5	3.1	  
2330	73.007	36.9	7.9	24.5	3.1	  
2253	72.78	47.1	10.8	24.5	3.1	  
2136	72.575	17.8	3.6	24.5	5.7	  
1918	72.501	34.6	9.4	24.5	15	  
1905	72.592	39.9	10.8	24.5	3.9	  
1765	72.779	27.6	8	24.5	7.4	  
4437	72.164	35.2	7.3	24.4	2.9	  
4341	72.64	29.5	7.3	24.4	2.9	  
4321	72.366	30.4	8	24.4	6.5	  
4034	71.804	53.4	13.1	24.4	11.3	  
3979	72.32	37.9	8.8	24.4	4.4	  
3937	72.106	45.1	10.9	24.4	2.9	  
3809	72.306	38.9	10.9	24.4	3.7	  
3802	72.023	35.1	9.5	24.4	15.7	  
3783	72.027	41.2	9.5	24.4	5.6	  
3722	71.919	40	10.2	24.4	3.7	  
3707	72.593	35.5	8	24.4	4.4	  
3678	72.63	37.6	8.8	24.4	6.9	  
3613	72.69	29.3	5.8	24.4	8.4	  
3554	72.812	43.5	10.9	24.4	18.3	  
3466	72.558	21.2	5.1	24.4	6.1	  
3325	72.565	40.9	10.2	24.4	5	  
3278	71.824	33	8	24.4	5.6	  
3122	73.341	32	7.3	24.4	2.9	  
3118	72.981	49.2	12.4	24.4	2.9	  
3104	72.645	36.1	8	24.4	2.9	  
3095	72.378	40.3	10.2	24.4	5	  
3084	72.793	17	4.4	24.4	4.4	  
3030	72.821	33.5	8.8	24.4	3.7	  
2739	72.973	39.6	9.5	24.4	4.4	  
2704	72.805	38.6	8.8	24.4	5.6	  
2699	72.706	24.1	6.6	24.4	2.9	  
2697	72.83	24	5.8	24.4	4.4	  
2469	72.759	31.9	8	24.4	3.7	  
2425	72.331	28	6.5	24.4	4.4	  
2383	72.316	29.6	6.5	24.4	7.3	  
2191	72.396	25.7	7.2	24.4	2.9	  
2050	72.92	23.2	5.8	24.4	2.9	  
2000	72.088	38.2	10.1	24.4	22.7	  
1923	72.575	37.7	7.9	24.4	9.7	  
1922	72.477	28.1	7.2	24.4	16.1	  
730	72.27	30.5	8	24.4	8.1	  
4436	72.319	25.4	5.8	24.3	3.5	  
4035	71.808	40.3	10.9	24.3	8.4	  
3985	72.395	32.2	7.3	24.3	9.9	  
3902	72.096	32	8	24.3	6.4	  
3898	72.007	46.2	12.4	24.3	5.4	  
3496	72.708	29.2	8	24.3	2.6	  
3471	72.507	26.4	8	24.3	7.6	  
3378	72.93	27.1	6.6	24.3	3.5	  
3115	72.958	52.6	12.4	24.3	3.5	  
3050	72.668	32.2	8.8	24.3	7.6	  
3016	72.624	30.7	7.3	24.3	5.4	  
3002	72.911	21.5	5.1	24.3	2.6	  
3000	72.817	44.6	10.2	24.3	3.5	  
2967	72.558	29.8	8	24.3	2.6	  
2878	72.483	38.1	9.5	24.3	3.5	  
2624	72.445	30.5	8	24.3	4.3	  
2262	72.29	16.7	4.3	24.3	6.4	  
2252	72.788	48.9	10.8	24.3	5.4	  
2062	72.621	27.6	7.2	24.3	6.4	  
1854	72.175	34	8.7	24.3	8	  
4405	72.4	38.4	8.8	24.2	4.7	  
4375	72.345	49.1	10.9	24.2	4	  
4297	72.398	49.1	10.2	24.2	2.2	  
4203	72.364	64	13.1	24.2	4	  
4051	72.523	40.3	10.2	24.2	4	  
4038	71.936	30.2	6.6	24.2	2.2	  
3882	72.258	29.6	7.3	24.2	12.5	  
3801	72.471	25.3	6.6	24.2	8.3	  
3797	72.538	41.7	10.2	24.2	5.3	  
3784	71.915	35.9	9.5	24.2	4.7	  
3692	72.298	25.5	5.8	24.2	5.3	  
3494	72.658	18.2	5.1	24.2	2.2	  
3465	72.583	23.6	5.1	24.2	5.8	  
3267	72.463	18.9	4.4	24.2	8.6	  
2983	72.398	37.9	6.6	24.2	6.7	  
2833	72.295	49.2	10.9	24.2	2.2	  
2830	72.946	51.2	13.1	24.2	4.7	  
2698	72.862	36.7	8.8	24.2	6.7	  
2622	72.577	31.5	8.8	24.2	4.7	  
2549	72.566	36.2	8	24.2	5.3	  
2540	72.956	32.1	7.3	24.2	5.3	  
2450	72.902	30	7.2	24.2	4	  
2164	72.607	42	9.4	24.2	6.3	  
2157	72.588	47.3	11.5	24.2	7.9	  
1940	72.316	44.1	10.1	24.2	4	  
4453	72.103	22.7	5.1	24.1	2.9	  
4291	72.168	51.6	11.7	24.1	1.6	  
4240	72.694	19.6	5.8	24.1	4.5	  
4072	72.097	32.9	8.8	24.1	2.9	  
4061	72.111	42.3	8.8	24.1	3.8	  
3975	72.219	37.1	8.8	24.1	3.8	  
3814	72.415	35.4	9.5	24.1	5.6	  
3756	72.9	52	9.5	24.1	10.9	  
3662	72.32	52.1	12.4	24.1	9.8	  
3472	72.437	25.9	6.6	24.1	6.6	  
3295	72.43	53.7	13.1	24.1	2.9	  
3110	72.833	35.7	8	24.1	4.5	  
3022	72.711	39.7	10.9	24.1	4.5	  
3009	72.513	40.4	10.9	24.1	6.1	  
3003	72.902	28.1	7.3	24.1	2.9	  
2840	72.391	39.1	10.9	24.1	4.5	  
2815	72.685	54.5	13.9	24.1	3.8	  
2615	72.725	45.6	10.2	24.1	5.1	  
2587	72.481	36.3	8	24.1	7	  
2536	72.873	34.6	7.3	24.1	5.1	  
2279	72.74	39	9.4	24.1	7	  
2258	72.74	36.5	10.1	24.1	2.9	  
2095	73.116	28.9	6.5	24.1	7	  
1954	72.744	32.6	7.9	24.1	2.9	  
1890	72.311	42.8	11.5	24.1	1.6	  
1445	72.263	29.8	8	24.1	5.1	  
1442	72.272	163.6	32.8	24.1	9.2
4231	72.073	19.2	5.1	24	7.6	  
4202	72.496	57.2	13.1	24	2.4	  
4060	72.2	35.2	8.8	24	0	  
3999	72.776	22.4	5.1	24	12.8	  
3945	72.571	27.2	7.3	24	9.1	  
3676	72.535	35.5	8.8	24	4.8	  
3664	72.461	46.7	11.7	24	13.9	  
3630	72.708	19.9	5.8	24	0	  
3553	72.727	29.7	6.6	24	23.7	  
3479	72.285	43.9	11.7	24	5.9	  
3399	72.4	18.6	5.1	24	4.2	  
3119	72.933	41.9	9.5	24	0	  
3047	72.613	29.1	8	24	10.3	  
2972	72.626	34.6	7.3	24	7.6	  
2922	72.793	48.1	11.7	24	3.4	  
2562	72.693	32.6	6.6	24	5.4	  
2548	72.117	37.5	8.8	24	9.1	  
2507	72.66	26.7	5.8	24	5.9	  
2341	73.038	38.8	7.9	24	0	  
2171	72.666	22.7	5	24	4.8	  
2139	72.41	49.1	11.5	24	4.8	  
2093	72.926	40.4	10.8	24	2.4	  
1849	72.023	27.7	8.7	24	6.4	  
759	72.496	24.9	5.8	24	6.4	  
4455	72.101	27.2	8	23.9	2.9	  
4451	72.458	41.6	8.8	23.9	3.8	  
4430	72.436	24.1	6.6	23.9	1.6	  
4266	72.337	17.7	4.4	23.9	4.5	  
3992	72.535	26.8	5.1	23.9	4.5	  
3991	72.496	30	8.8	23.9	5.1	  
3956	72.913	36.3	8.8	23.9	3.8	  
3704	72.588	28	7.3	23.9	10.5	  
3650	72.375	39.6	10.9	23.9	3.8	  
3460	72.533	39.9	10.2	23.9	7.1	  
3368	72.857	19.8	3.6	23.9	3.8	  
3260	72.584	40.1	10.2	23.9	3.8	  
3252	72.344	53.9	12.4	23.9	7.5	  
3240	72.705	42.2	10.2	23.9	3.8	  
3219	72.268	42.8	11.7	23.9	6.6	  
3108	72.848	34.4	7.3	23.9	3.8	  
3096	72.476	42.3	10.2	23.9	4.5	  
3085	72.822	25.6	5.8	23.9	3.8	  
2999	72.839	41.3	10.9	23.9	5.1	  
2921	72.605	30.1	8	23.9	7.5	  
2856	72.542	52.8	10.2	23.9	3.8	  
2794	72.74	49.6	10.9	23.9	4.5	  
2703	72.791	24.2	5.1	23.9	2.9	  
2583	72.537	33.9	8.8	23.9	5.1	  
2526	72.58	38.6	8.8	23.9	3.8	  
2391	72.551	41.1	9.4	23.9	5.1	  
2362	72.899	27.6	6.5	23.9	2.9	  
2259	72.493	40.2	9.4	23.9	3.8	  
2126	72.661	33.1	7.2	23.9	4.5	  
2054	72.799	29.9	7.2	23.9	5.1	  
2007	72.511	25.6	5.8	23.9	7.5	  
1930	72.568	37	8.7	23.9	7.9	  
1911	72.511	39.5	10.8	23.9	2.9	  
1891	72.365	38.6	8.7	23.9	3.8	  
1852	72.046	37.1	10.1	23.9	2.9	  
281	72.12	19.1	4.3	23.9	5.7	  
1434	72.536	29.1	7.2	23.88	18.8	  
4403	72.516	24.6	5.8	23.8	2.2	  
4360	72.07	28.5	6.6	23.8	4.8	  
4354	72.516	50	12.4	23.8	4.1	  
4340	72.673	38.2	9.5	23.8	3.3	  
4315	72.321	26.1	7.3	23.8	4.1	  
4194	72.644	23.3	5.8	23.8	4.1	  
4050	72.431	37.2	8.8	23.8	5.4	  
3973	72.133	38	8.8	23.8	8.8	  
3942	72.616	24.6	5.1	23.8	3.3	  
3925	72.609	32.9	8	23.8	5.9	  
3909	72.253	28.4	5.8	23.8	10	  
3897	72.586	49.4	12.4	23.8	13.4	  
3775	72.692	40	9.5	23.8	5.4	  
3757	73.212	40.4	10.2	23.8	6.4	  
3558	72.902	43.1	10.9	23.8	12.4	  
3509	72.034	44.7	10.2	23.8	5.4	  
3377	72.666	27.9	7.3	23.8	4.1	  
3371	72.896	33.4	7.3	23.8	2.2	  
3364	72.364	38.6	10.2	23.8	4.1	  
3339	72.585	58.6	12.4	23.8	5.4	  
3215	72.295	39.6	9.5	23.8	4.8	  
3107	72.767	37.1	7.3	23.8	2.2	  
3082	72.716	38.1	10.2	23.8	6.4	  
2994	72.719	24.3	6.6	23.8	4.1	  
2932	72.48	30.5	8	23.8	7.3	  
2825	73.085	39.4	10.2	23.8	6.4	  
2288	72.708	16	4.3	23.8	4.8	  
2182	72.473	40.6	10.8	23.8	3.3	  
4427	72.399	29.5	8	23.7	2.7	  
4426	72.337	43.7	9.5	23.7	2.7	  
4424	72.199	37.1	10.9	23.7	2.7	  
4303	71.973	36.6	7.3	23.7	6.1	  
4267	72.271	30.8	8.8	23.7	5.6	  
3950	72.834	46	12.4	23.7	2.7	  
3880	72.193	30.7	8	23.7	11.6	  
3855	72.466	31.5	8	23.7	5	  
3821	72.502	34.9	10.2	23.7	28	  
3811	72.383	43.4	10.2	23.7	23.5	  
3799	72.602	47.3	12.4	23.7	4.4	  
3748	72.73	32.5	8.8	23.7	10.5	  
3732	72.73	40.1	10.2	23.7	5.6	  
3488	72.642	38.5	8.8	23.7	3.6	  
3487	72.58	29.5	7.3	23.7	3.6	  
3485	72.655	24.7	6.6	23.7	2.7	  
3373	72.915	52.4	11.7	23.7	2.7	  
3300	72.586	49.6	10.9	23.7	3.6	  
3209	72.57	32.8	8	23.7	4.4	  
3152	72.868	18.5	4.4	23.7	7.4	  
3117	73.038	32.6	7.3	23.7	2.7	  
2955	72.225	30.1	8	23.7	4.4	  
2933	72.594	37.2	9.5	23.7	3.6	  
2749	72.779	44.1	9.5	23.7	3.6	  
2709	72.813	45.1	10.9	23.7	6.6	  
2628	72.397	22.9	6.6	23.7	8.2	  
2560	72.707	16.5	4.4	23.7	2.7	  
2342	73.053	33	6.5	23.7	2.7	  
2276	72.49	35.4	9.4	23.7	13.3	  
1950	72.624	32	6.5	23.7	4.4	  
4399	72.553	15.8	4.4	23.6	3	  
4356	72.521	39.3	8.8	23.6	5.8	  
3998	72.713	45.2	12.4	23.6	20.1	  
3983	72.427	31.3	8	23.6	4.6	  
3908	72.307	51.5	12.4	23.6	11.7	  
3884	72.289	26.4	5.8	23.6	3	  
3854	72.447	30.4	8.8	23.6	12.7	  
3838	71.92	42.1	10.9	23.6	8.3	  
3806	72.096	46.1	11.7	23.6	5.2	  
3755	72.916	42	11.7	23.6	7.6	  
3684	72.643	44.3	9.5	23.6	5.8	  
3638	72.002	36.6	9.5	23.6	10.6	  
3559	72.98	11.5	2.9	23.6	5.8	  
3513	71.824	49.1	10.2	23.6	3	  
3461	72.565	30.9	7.3	23.6	8	  
3363	72.33	30.2	7.3	23.6	3	  
3261	72.368	46	10.9	23.6	3	  
3109	72.89	39.7	10.2	23.6	3	  
2993	72.724	30.9	6.6	23.6	5.2	  
2975	72.336	49.2	12.4	23.6	4.6	  
2892	72.838	44.2	8.8	23.6	4.6	  
2867	72.177	47.9	11.7	23.6	7.6	  
2843	72.525	33	8	23.6	3.9	  
2824	72.958	31.2	8.8	23.6	7.2	  
2777	72.563	32.8	8	23.6	3.9	  
2770	72.655	26.3	5.8	23.6	3	  
2707	72.701	29.8	6.6	23.6	7.6	  
2702	72.979	47.5	10.9	23.6	3	  
2575	72.726	23.3	5.8	23.6	3.9	  
2474	72.772	31	7.3	23.6	5.8	  
2437	72.517	42	10.1	23.6	3.9	  
2406	72.763	36.6	8.7	23.6	5.2	  
2173	72.662	32.1	7.9	23.6	3	  
2115	72.758	17.9	4.3	23.6	5.2	  
1975	72.741	44.1	10.1	23.6	9.4	  
1955	72.727	26.9	6.5	23.6	5.8	  
1912	72.353	39.4	11.5	23.6	20	  
1838	72.511	48.4	10.8	23.6	8.3	  
4409	72.272	32.5	8	23.5	3.2	  
4256	71.921	55.4	12.4	23.5	4	  
4239	72.645	33.6	9.5	23.5	3.2	  
4236	72.444	44.7	9.5	23.5	3.2	  
4233	72.058	24.4	6.6	23.5	5.4	  
4040	72.127	36.4	9.5	23.5	6.4	  
3774	72.606	30.5	7.3	23.5	3.2	  
3454	72.6	40.5	11.7	23.5	4.7	  
3448	72.692	16.9	3.6	23.5	3.2	  
3245	72.706	33.2	8	23.5	3.2	  
3218	72.662	32.1	8	23.5	7.3	  
3025	72.723	40.6	9.5	23.5	11.3	  
2903	72.541	42.6	10.2	23.5	4.7	  
2826	73.116	36.4	8.8	23.5	5.4	  
2411	72.823	43.9	9.4	23.5	5.4	  
2189	72.402	41.2	10.8	23.5	4.7	  
2174	72.638	33.8	7.9	23.5	3.2	  
1949	72.558	19.7	5	23.5	3.2	  
1947	72.499	32.5	8.7	23.5	8.5	  
1845	72.223	28.9	6.5	23.5	4	  
726	72.054	52.6	11.6	23.5	10.1	  
4438	72.069	27.1	7.3	23.4	3.4	  
4404	72.397	38.3	9.5	23.4	3.4	  
4314	72.396	30.7	8	23.4	3.4	  
4310	72.461	37.8	9.5	23.4	3.4	  
3984	72.408	34.4	8	23.4	3.4	  
3949	72.721	37	9.5	23.4	4.2	  
3929	72.793	29.6	8	23.4	9.9	  
3895	72.68	33.2	7.3	23.4	10.5	  
3772	72.523	39.4	10.2	23.4	4.2	  
3715	72.87	22.5	4.4	23.4	3.4	  
3708	72.71	45.3	10.2	23.4	4.2	  
3497	72.612	33.1	7.3	23.4	3.4	  
3299	72.672	47.6	9.5	23.4	5.5	  
3280	72.547	46.6	12.4	23.4	8.2	  
3020	72.706	44.5	10.2	23.4	3.4	  
3004	73.006	29.8	7.3	23.4	3.4	  
2890	72.67	29.2	8	23.4	8.2	  
2586	72.467	39	8.8	23.4	4.2	  
2553	72.171	28.9	8	23.4	10.5	  
2538	72.949	34.5	8	23.4	3.4	  
2424	72.438	28	5.8	23.4	4.2	  
2361	72.863	39.3	9.4	23.4	3.4	  
2328	73.026	18.5	4.3	23.4	5.5	  
2260	72.675	26.3	6.5	23.4	3.4	  
2251	72.539	25.4	5	23.4	4.2	  
2243	72.47	36.4	7.9	23.4	4.2	  
2168	72.632	11.8	2.9	23.4	4.9	  
2068	72.422	44	10.1	23.4	4.2	  
1991	73.029	31	8.7	23.4	4.2	  
1987	72.955	33.6	7.9	23.4	4.2	  
1981	72.861	28.4	6.5	23.4	4.9	  
4306	72.568	44.2	10.2	23.3	4.9	  
4255	72.307	36.2	9.5	23.3	4.3	  
4245	72.456	47.7	10.9	23.3	4.3	  
4222	72.026	48.1	11.7	23.3	3.5	  
3891	72.348	14.7	3.6	23.3	3.5	  
3871	71.992	33.3	8.8	23.3	17.9	  
3845	72.121	38.5	10.2	23.3	12.7	  
3785	71.925	25	5.8	23.3	10.6	  
3759	73.008	28.5	7.3	23.3	9.7	  
3716	72.975	25	5.8	23.3	6.6	  
3714	72.804	49.9	11.7	23.3	6.6	  
3451	72.68	45.8	10.9	23.3	3.5	  
3402	72.295	31.8	8.8	23.3	8.6	  
3374	72.652	45.5	10.2	23.3	3.5	  
3366	72.215	50.1	12.4	23.3	4.3	  
3342	72.831	61.4	13.1	23.3	6.1	  
3307	72.773	44	10.2	23.3	4.9	  
3242	72.665	17.5	4.4	23.3	6.6	  
2537	72.731	37	8	23.3	3.5	  
2443	72.931	40.6	9.4	23.3	4.3	  
2263	72.654	42.8	10.1	23.3	4.9	  
2175	72.542	45	11.5	23.3	4.3	  
1978	72.73	36.6	9.4	23.3	22.7	  
1920	72.569	32.5	8.7	23.3	10.6	  
1876	71.944	34.4	9.4	23.3	4.3	  
1467	72.717	27.2	5.8	23.3	9	  
4353	72.49	63	12.4	23.2	5	  
4351	72.476	51.8	10.9	23.2	3.5	  
4346	72.386	35.1	8.8	23.2	4.3	  
3988	72.506	33.8	8.8	23.2	5	  
3967	72.027	47.2	10.9	23.2	11.8	  
3900	71.821	40.5	9.5	23.2	9.7	  
3864	72.555	25.8	6.6	23.2	3.5	  
3839	71.831	34.1	7.3	23.2	3.5	  
3796	72.495	36.6	8.8	23.2	3.5	  
3782	72.034	28.9	7.3	23.2	7.1	  
3760	73.024	14.5	3.6	23.2	7.1	  
3724	72.39	36.6	9.5	23.2	16.3	  
3512	72.574	50.1	10.9	23.2	3.5	  
3320	72.621	38.7	8.8	23.2	4.3	  
3262	72.4	46.6	12.4	23.2	8.7	  
3091	72.986	43.1	9.5	23.2	3.5	  
3081	72.695	43.5	9.5	23.2	12.8	  
3042	72.531	36.3	11.7	23.2	5	  
2710	72.831	48.4	11.7	23.2	5	  
2662	72.122	18	4.4	23.2	3.5	  
2551	72.148	32.6	7.3	23.2	7.1	  
2546	72.338	43	10.2	23.2	3.5	  
2170	72.636	20.2	4.3	23.2	3.5	  
2154	71.862	45	9.4	23.2	4.3	  
2088	72.749	39.5	9.4	23.2	3.5	  
2079	72.538	38.6	10.1	23.2	3.5	  
1967	72.568	37.7	11.5	23.2	15.5	  
4232	72.345	54.8	12.4	23.1	3.5	  
3919	72.491	49.2	11.7	23.1	7.9	  
3896	72.594	40.8	11.7	23.1	4.3	  
3833	72.214	32	7.3	23.1	8.3	  
3747	72.701	41	9.5	23.1	3.5	  
3657	72.341	40.3	10.2	23.1	15.3	  
3626	72.575	43	8.8	23.1	3.5	  
3604	72.597	39.5	10.2	23.1	7.1	  
3333	72.461	35.6	7.3	23.1	3.5	  
3112	72.851	38.2	8.8	23.1	5	  
2965	72.415	36.6	10.2	23.1	5.6	  
2958	72.341	30.9	8.8	23.1	3.5	  
2844	72.635	42.1	8.8	23.1	4.3	  
2801	72.868	33.5	7.3	23.1	6.6	  
2732	72.726	34.9	8	23.1	6.6	  
2711	72.515	43.5	10.9	23.1	4.3	  
2554	72.606	34.6	7.3	23.1	6.6	  
2375	72.711	40.3	8.7	23.1	5	  
2316	72.732	47.5	11.5	23.1	8.3	  
2270	72.411	49.3	10.1	23.1	17.5	  
2141	72.211	36.7	8.7	23.1	3.5	  
4044	72.188	26.8	6.6	23	4.3	  
3987	72.455	22.8	5.8	23	6.1	  
3910	72.361	25.8	5.8	23	22.7	  
3890	72.365	31.6	8.8	23	10.7	  
3865	72.543	28.2	6.6	23	7.9	  
3746	72.615	35.9	8.8	23	10.1	  
3728	72.636	38.2	10.9	23	15	  
3484	72.174	41.1	10.2	23	5	  
3103	72.646	40.9	9.5	23	3.4	  
3058	72.766	30.2	7.3	23	5.6	  
2987	72.626	27.7	7.3	23	6.1	  
2795	72.717	26.9	6.6	23	3.4	  
2785	72.645	38.9	8.8	23	7.1	  
2778	72.555	26.3	6.6	23	8.7	  
2508	72.185	34.2	9.5	23	4.3	  
2475	72.846	28	7.3	23	6.1	  
2467	72.685	46.6	10.1	23	3.4	  
2290	72.998	37	7.9	23	5	  
2151	71.893	65.4	12.3	23	4.3	  
2140	72.375	27.3	5.8	23	3.4	  
2010	72.568	49.5	10.8	23	11.3	  
2008	72.464	40.1	12.3	23	6.1	  
1846	72.708	34.8	7.9	23	7.1	  
729	72.11	32	8	23	6.6	  
4411	72.168	29.4	8	22.9	4.2	  
4342	72.408	47.6	11.7	22.9	4.9	  
4042	72.314	31.8	6.6	22.9	5.5	  
4039	72.014	27	5.8	22.9	4.9	  
3965	72.828	57.5	13.1	22.9	10.1	  
3899	71.94	40.2	10.9	22.9	6.6	  
3727	72.632	48.5	12.4	22.9	21.7	  
3691	72.811	43	10.2	22.9	14.5	  
3519	72.042	30.4	8	22.9	4.9	  
3367	72.104	53	11.7	22.9	4.9	  
3273	72.3	25.1	6.6	22.9	6.1	  
3106	72.79	43.6	10.9	22.9	3.3	  
2989	72.646	37.7	10.2	22.9	7.5	  
2980	72.438	35.5	8.8	22.9	7.9	  
2940	72.786	18.9	5.1	22.9	6.1	  
2810	72.165	32.4	7.3	22.9	3.3	  
2718	72.415	40.1	8	22.9	5.5	  
2468	72.455	32.4	7.3	22.9	3.3	  
2449	72.813	34.3	7.9	22.9	3.3	  
2301	72.491	31.5	8.7	22.9	4.2	  
2137	72.485	35.4	9.4	22.9	3.3	  
1973	72.646	26.8	6.5	22.9	11	  
1884	72.236	46.2	10.1	22.9	3.3	  
1433	72.541	25.6	7.2	22.9	8.3	  
4309	72.519	47.6	10.9	22.8	5.4	  
3977	72.226	39.9	8.8	22.8	7.8	  
3742	72.411	40.6	9.5	22.8	3.1	  
3562	72.97	24.2	5.1	22.8	4.7	  
3511	72.189	39.5	11.7	22.8	6	  
3370	72.588	38.4	10.2	22.8	4	  
3357	72.421	36.8	9.5	22.8	7.8	  
3354	72.385	29.8	5.8	22.8	7.8	  
3324	72.371	43.8	10.9	22.8	4	  
3316	72.675	32.2	7.3	22.8	20.4	  
3275	71.89	45.6	11.7	22.8	6	  
3243	72.863	48.4	11.7	22.8	4.7	  
2988	72.635	32.8	8	22.8	3.1	  
2937	72.581	31.8	8.8	22.8	7.4	  
2926	72.405	22.8	5.1	22.8	6	  
2914	72.231	45.7	10.2	22.8	3.1	  
2827	73.001	43.5	10.2	22.8	6	  
2708	72.718	35.8	8.8	22.8	3.1	  
2676	72.713	39.1	10.2	22.8	4.7	  
2630	72.361	37.4	9.5	22.8	4.7	  
2617	72.55	41.1	9.5	22.8	3.1	  
2472	72.403	43.9	10.9	22.8	7	  
2403	72.623	47.4	10.1	22.8	4.7	  
2360	72.847	33.7	7.9	22.8	3.1	  
2223	72.232	33.9	7.2	22.8	3.1	  
2216	72.631	51	12.3	22.8	5.4	  
2167	72.673	31.1	6.5	22.8	6	  
2130	72.512	37.3	10.1	22.8	14.1	  
2094	72.93	42.7	10.1	22.8	7	  
1919	72.533	28.2	5.8	22.8	4	  
1836	72.411	31.3	7.2	22.8	4	  
960	72.523	39.3	9.4	22.8	7.4	  
4425	72.206	47.5	10.9	22.7	4.6	  
4350	72.492	37.3	8.8	22.7	3.8	  
4265	72.51	49.5	11.7	22.7	4.6	  
3878	72.186	30.5	8.8	22.7	16.7	  
3721	72.007	46.6	12.4	22.7	11.8	  
3702	72.456	22.7	5.8	22.7	5.8	  
3425	72.703	35.3	7.3	22.7	3.8	  
3416	72.51	50	11.7	22.7	6.9	  
3318	72.578	47.7	11.7	22.7	6.4	  
3315	72.721	49.5	10.2	22.7	7.8	  
3056	72.87	21.6	5.1	22.7	3.8	  
2966	72.524	36.3	9.5	22.7	10.6	  
2753	72.787	40.4	10.2	22.7	4.6	  
2557	72.199	21.1	4.4	22.7	9.7	  
2556	72.683	19.5	5.1	22.7	3.8	  
2516	72.829	31.7	8	22.7	6.4	  
2444	72.717	29.5	7.2	22.7	2.8	  
2392	72.621	27	7.2	22.7	5.8	  
2248	72.575	19.3	5	22.7	5.2	  
2228	72.13	42.1	9.4	22.7	5.8	  
2222	72.404	55	13	22.7	2.8	  
2172	72.85	35	8.7	22.7	4.6	  
2159	72.087	53.3	11.5	22.7	3.8	  
2084	72.62	34.6	7.9	22.7	4.6	  
4246	72.351	40.5	10.9	22.6	3.5	  
4204	72.176	53.5	13.1	22.6	6.7	  
3928	72.71	41.3	8.8	22.6	9.6	  
3905	72.203	38.1	10.9	22.6	9.2	  
3853	72.312	45	11.7	22.6	5	  
3844	72.145	48.4	12.4	22.6	3.5	  
3841	71.732	42.3	11.7	22.6	3.5	  
3753	72.785	22.5	5.8	22.6	2.4	  
3653	72.461	28	6.6	22.6	6.7	  
3506	72.233	45.1	11.7	22.6	3.5	  
3392	72.401	53.6	11.7	22.6	19.1	  
3323	72.496	30.4	8.8	22.6	3.5	  
3023	72.7	27.6	7.3	22.6	4.3	  
2998	72.84	34.2	9.5	22.6	5.7	  
2976	72.361	44	10.9	22.6	3.5	  
2960	72.33	41.6	10.2	22.6	5.7	  
2846	72.792	35.3	8	22.6	5.7	  
2462	72.66	40	9.4	22.6	2.4	  
2245	72.555	13.1	2.9	22.6	5.7	  
2148	71.923	32.6	7.9	22.6	3.5	  
2055	72.623	33.2	8.7	22.6	3.5	  
1882	72.084	22.2	5.8	22.6	3.5	  
4252	72.406	51.4	13.1	22.5	5.4	  
4226	72.231	43.5	10.2	22.5	4.8	  
3888	72.349	29.9	7.3	22.5	8.4	  
3856	72.412	31.5	8	22.5	3.1	  
3828	72.688	38.1	10.2	22.5	9.8	  
3790	72.417	31.7	9.5	22.5	12.3	  
3735	72.896	31.1	8	22.5	9.5	  
3561	72.987	35.6	8.8	22.5	5.4	  
3500	72.543	29.8	6.6	22.5	1.7	  
3493	72.396	23.3	5.8	22.5	1.7	  
3486	72.689	40	10.2	22.5	3.1	  
3341	72.491	37.2	10.2	22.5	9.8	  
3287	72.019	55	13.1	22.5	4	  
3113	72.961	39.3	10.9	22.5	1.7	  
3063	72.296	34.9	8.8	22.5	4.8	  
2829	72.938	44.3	10.9	22.5	7.5	  
2762	72.644	38.6	10.2	22.5	4	  
2721	72.676	45.2	10.9	22.5	5.4	  
2550	72.583	32.1	8	22.5	4	  
2471	72.679	32.5	7.3	22.5	4.8	  
2412	72.697	43.4	9.4	22.5	1.7	  
2207	72.565	31.6	7.9	22.5	4	  
2187	72.275	42.8	11.5	22.5	7.9	  
2176	72.591	34.9	7.9	22.5	4.8	  
2165	72.175	38.2	7.9	22.5	3.1	  
2096	72.94	43.4	10.1	22.5	7.5	  
2092	72.847	30.8	7.9	22.5	4	  
2014	72.576	39.4	10.1	22.5	11.4	  
1847	72.63	45.2	10.8	22.5	1.7	  
767	73.071	38.5	8	22.5	13.6	    
4415	72.123	27.5	7.3	22.4	2.6	  
3901	71.739	34.7	8	22.4	5.8	  
3842	72.037	42	9.5	22.4	8.2	  
3705	72.621	40.1	10.2	22.4	15.8	  
3693	72.306	41.7	10.9	22.4	14	  
3637	72.998	29.9	8.8	22.4	0	  
3498	72.498	32	6.6	22.4	0	  
3452	72.661	32.7	7.3	22.4	0	  
3447	72.665	38.4	8	22.4	0	  
3428	72.646	34.8	8	22.4	2.6	  
3421	72.775	52.2	10.9	22.4	3.6	  
3301	72.716	49.3	11.7	22.4	5.2	  
3244	72.591	31.6	8	22.4	12.2	  
3101	72.64	37.6	8.8	22.4	6.9	  
3067	72.499	36	10.2	22.4	5.8	  
3031	72.844	33.2	8	22.4	9.7	  
2818	72.803	53.1	11.7	22.4	5.2	  
2813	72.162	38.3	10.9	22.4	4.5	  
2754	72.798	33	8	22.4	3.6	  
2664	72.206	23.3	5.8	22.4	3.6	  
2578	72.671	19	4.4	22.4	0	  
2504	72.205	35.5	8.8	22.4	7.3	  
2501	72.391	36.1	8	22.4	9.4	  
2464	72.835	39.3	9.4	22.4	5.2	  
2196	72.423	24.1	6.5	22.4	9	  
2080	72.538	47.6	11.5	22.4	4.5	  
2035	72.383	43.3	10.1	22.4	6.3	  
285	72.301	26.8	6.5	22.4	5.2	  
4217	72.354	34.6	8	22.3	3.1	  
3933	72.695	36.8	8	22.3	11.5	  
3670	72.493	30.3	8	22.3	12.4	  
3514	72.135	35	8.8	22.3	4.8	  
3453	72.625	36.1	7.3	22.3	1.7	  
3420	72.775	45	9.5	22.3	5.5	  
3389	72.481	55.2	13.1	22.3	7.1	  
3029	72.917	42	10.9	22.3	12.1	  
2669	72.466	39.2	8.8	22.3	5.5	  
2544	72.422	27.4	5.8	22.3	1.7	  
2515	72.776	32.6	9.5	22.3	6.6	  
2178	72.545	43.8	10.8	22.3	1.7	  
1913	72.391	47.3	11.5	22.3	3.1	  
1903	72.563	39.7	10.8	22.3	10.9	  
768	72.377	42	10.8	22.3	9.2	  
4261	72.328	37.2	8.8	22.2	3.6	  
4249	72.321	28.8	7.3	22.2	4.4	  
4234	72.138	54.4	12.4	22.2	2.4	  
3787	72.118	45	10.9	22.2	5.8	  
3671	72.525	50.1	12.4	22.2	7.8	  
3515	71.866	40.8	9.5	22.2	5.1	  
3503	72.413	35.7	8.8	22.2	2.4	  
3456	72.536	34.7	8	22.2	3.6	  
3406	72.92	40.4	10.9	22.2	4.4	  
3313	72.723	56.5	11.7	22.2	5.8	  
3277	72.208	31.2	8	22.2	5.8	  
3099	72.559	25.4	6.6	22.2	10.8	  
3098	72.541	33	8.8	22.2	2.4	  
2884	72.545	37	7.3	22.2	2.4	  
2623	72.343	28.9	7.3	22.2	2.4	  
2440	72.677	25.9	6.5	22.2	3.6	  
2420	72.515	25.8	5.8	22.2	3.6	  
2414	72.533	30.2	8.7	22.2	5.8	  
2405	72.622	44.9	9.4	22.2	2.4	  
2333	73.043	45.7	11.5	22.2	5.1	  
2296	72.536	47.3	11.5	22.2	4.4	  
2161	72.225	21.2	5.8	22.2	2.4	  
1877	71.843	34.6	7.9	22.2	4.4	  
1431	72.392	18.4	4.3	22.2	5.1	  
737	72.194	42.4	9.4	22.2	5.1	  
4289	72.51	41.3	9.5	22.1	7.1	  
4274	72.583	46.8	12.4	22.1	2.9	  
4241	72.743	22.4	5.1	22.1	2.9	  
4207	72.694	17.8	4.4	22.1	4.7	  
4045	72.203	45.2	11.7	22.1	3.9	  
3847	72.221	28.8	8	22.1	2.9	  
3788	72.09	31	8	22.1	9.2	  
3673	72.614	29.8	7.3	22.1	7.5	  
3648	72.298	39.9	8.8	22.1	6.5	  
3276	72.214	25	7.3	22.1	11.5	  
3102	72.65	42.7	10.9	22.1	2.9	  
3052	72.698	29.6	8.8	22.1	7.1	  
2543	72.404	43.8	10.2	22.1	2.9	  
2404	72.701	34.6	9.4	22.1	5.4	  
2090	72.705	41.6	10.1	22.1	5.4	  
2067	72.357	33.3	7.9	22.04	19.5	  
4361	72.101	35.1	8.8	22	4.1	  
3926	72.691	36.9	9.5	22	4.1	  
3907	72.327	32.5	10.2	22	7.2	  
3892	72.48	37.1	8.8	22	4.9	  
3725	72.491	37.3	8	22	5.6	  
3634	72.718	38.8	8	22	3.2	  
3534	72.503	42.6	10.9	22	3.2	  
3517	72.558	50.5	10.9	22	3.2	  
3396	72.361	34.6	6.6	22	6.7	  
3365	72.102	42.5	11.7	22	8.1	  
3100	72.644	43.2	9.5	22	12.3	  
3077	72.6	56.6	13.9	22	4.9	  
3036	72.42	33.6	8.8	22	8.1	  
2995	72.753	28	7.3	22	9.3	  
2918	72.426	45.2	11.7	22	4.1	  
2796	72.876	26.2	5.8	22	3.2	  
2734	72.711	39.5	8.8	22	6.7	  
2272	72.568	33.6	7.2	22	4.9	  
2208	72.468	35.2	7.9	22	9	  
2155	72.202	27.6	7.2	22	14.9	  
2091	72.743	37.5	8.7	22	8.6	  
2043	72.492	34.1	7.2	22	4.1	  
1899	72.536	25.7	6.5	22	19.6	  
1881	72.037	47.6	10.8	22	3.2	  
1848	72.053	25.7	7.2	22	3.2	  
1084	72.301	32.2	7.2	22	3.2	  
4242	72.636	30.7	7.3	21.9	3.4	  
4237	72.651	44.6	10.9	21.9	4.3	  
4188	72.158	47.3	12.4	21.9	3.4	  
3918	72.478	16.8	4.4	21.9	3.4	  
3849	72.11	50.8	10.2	21.9	5.7	  
3846	72.228	41.9	10.2	21.9	5.1	  
3687	72.693	34.4	8	21.9	4.3	  
3654	72.806	46.8	10.9	21.9	8.3	  
3557	72.86	39.3	8	21.9	9.1	  
3510	72.29	36.1	8.8	21.9	6.3	  
3429	72.88	52.8	10.9	21.9	6.9	  
3340	72.425	13.1	3.6	21.9	3.4	  
3114	72.963	18.5	5.1	21.9	7.4	  
2982	72.516	35.3	8	21.9	5.1	  
2959	72.333	32.8	8.8	21.9	6.9	  
2939	72.685	36.6	8	21.9	9.8	  
2887	72.555	27.8	5.8	21.9	5.1	  
2838	72.856	50.3	12.4	21.9	8.7	  
2835	72.943	45.5	9.5	21.9	9.8	  
2345	73.025	32.4	7.2	21.9	3.4	  
2323	72.964	40.8	8.7	21.9	3.4	  
2217	72.184	22.2	6.5	21.9	3.4	  
2202	72.586	14.9	3.6	21.9	6.3	  
2169	72.743	40.1	8.7	21.9	3.4	  
2160	72.237	20.4	5	21.9	3.4	  
748	72.156	36.6	9.4	21.9	13.2	  
734	72.389	42.6	10.1	21.9	5.1	    
727	72.006	27.2	5.8	21.9	7.8	  
4416	72.116	56.7	12.4	21.8	7	  
4276	72.825	32.2	8.8	21.8	3.6	  
4208	72.59	37.3	8.8	21.8	3.6	  
3962	72.9	28.8	8	21.8	8.4	  
3924	72.564	26	6.6	21.8	13.6	  
3696	72.32	26.9	6.6	21.8	7.9	  
3660	72.242	25.9	6.6	21.8	6.4	  
3445	72.768	55.7	12.4	21.8	3.6	  
3415	72.745	27.5	7.3	21.8	4.5	  
3326	72.424	46.4	10.2	21.8	8.4	  
3075	72.718	35.4	9.5	21.8	9.9	  
2997	72.805	22.2	5.1	21.8	5.2	  
2797	72.598	41.9	10.2	21.8	3.6	  
2633	72.314	32.4	7.3	21.8	5.2	  
2579	72.578	28.9	8	21.8	3.6	  
2512	72.385	45.6	10.2	21.8	5.9	  
2410	72.609	38.2	10.8	21.8	5.2	  
2408	72.651	35.5	9.4	21.8	5.2	  
2350	72.986	43.4	10.8	21.8	3.6	  
2220	72.24	41.8	10.1	21.8	3.6	  
2156	72.236	28.4	6.5	21.8	13.3	  
2087	72.763	36.1	7.9	21.8	5.9	  
2082	72.411	29.4	7.2	21.8	6.4	  
1909	72.639	35.6	7.9	21.8	10.6	  
4347	72.519	25.1	7.3	21.7	6.5	  
4263	72.453	40.8	9.5	21.7	3.7	  
3911	72.395	36.5	9.5	21.7	6	  
3740	73.003	37	8	21.7	8.5	  
3682	72.616	36.3	8.8	21.7	4.6	  
3663	72.427	45.2	10.9	21.7	7.1	  
3499	72.526	34.5	8	21.7	4.6	  
3346	72.476	47	10.9	21.7	3.7	  
3302	72.843	32.4	8	21.7	3.7	  
3272	72.392	21.1	5.1	21.7	8	  
3083	72.779	29.5	8	21.7	10	  
3072	72.568	19.3	4.4	21.7	6	  
2996	72.748	40.8	10.9	21.7	4.6	  
2991	72.641	39.5	9.5	21.7	3.7	  
2981	72.541	43.6	11.7	21.7	6.5	  
2969	72.438	34.8	8	21.7	11.7	  
2962	72.335	52.6	11.7	21.7	7.1	  
2945	72.054	23	7.3	21.7	4.6	  
2757	72.783	39.2	10.2	21.7	4.6	  
2641	72.41	19.6	5.1	21.7	4.6	  
2629	72.478	30.9	8	21.7	6.5	  
2513	72.358	35.9	9.5	21.7	4.6	  
2441	72.863	31.7	7.2	21.7	3.7	  
2355	72.971	44.1	10.8	21.7	3.7	  
2302	72.425	24.2	6.5	21.7	7.1	  
2289	72.568	41.5	9.4	21.7	4.6	  
2249	72.577	16.1	3.6	21.7	3.7	  
2244	72.521	26.3	6.5	21.7	3.7	  
2142	72.256	33.2	9.4	21.7	3.7	  
764	72.891	40.6	8.7	21.7	5.3	    
4410	72.153	36.4	8.8	21.6	3.8	  
4275	72.755	39.6	10.2	21.6	3.8	  
3768	72.503	42.7	10.2	21.6	7.1	  
3750	72.82	39.4	10.2	21.6	7.6	  
3700	72.5	31.4	8	21.6	10.8	  
3639	72.02	58.5	13.9	21.6	10.4	  
3369	72.547	34.9	7.3	21.6	4.6	  
3033	72.389	53.7	12.4	21.6	11.1	  
2948	72.158	25.2	5.8	21.6	3.8	  
2919	72.423	33	8	21.6	3.8	  
2847	72.732	31.9	7.3	21.6	3.8	  
2832	72.984	14.5	3.6	21.6	8.5	  
2758	72.739	38.2	9.5	21.6	4.6	  
2473	72.772	27.4	5.1	21.6	3.8	  
2386	72.448	27.1	7.2	21.6	11.7	  
2294	72.48	41.2	10.8	21.6	4.6	  
2264	72.646	39.8	10.1	21.6	10.4	  
2192	72.313	30	6.5	21.6	6.6	  
2143	72.203	47.3	11.5	21.6	3.8	  
1942	72.405	39.9	10.8	21.6	5.4	  
1829	72.223	37.5	9.4	21.6	3.8	  
1080	72.201	35.1	8.7	21.6	3.8	  
742	72.093	27.9	6.5	21.6	6	  
287	72	42	9.4	21.6	15.5	  
4264	72.197	55.3	12.4	21.5	4.6	  
3834	72.771	42.7	10.2	21.5	10.5	  
3739	72.917	49.8	10.9	21.5	11.1	  
3701	72.466	19.9	5.1	21.5	7.1	  
3405	72.868	24.2	6.6	21.5	3.8	  
3376	72.651	61.5	12.4	21.5	3.8	  
3097	72.571	54.9	13.1	21.5	4.6	  
3062	72.197	34.1	8	21.5	4.6	  
2876	72.36	39.1	8.8	21.5	10.1	  
2682	72.339	38.7	8	21.5	6	  
2631	72.786	13.6	3.6	21.5	5.4	  
2390	72.666	24.4	5.8	21.5	6	  
2327	73.001	21.9	5.8	21.5	6.6	  
2295	72.575	37.2	7.2	21.5	3.8	  
2250	72.616	40.1	8.7	21.5	3.8	  
2224	72.222	24.8	5.8	21.5	3.8	  
2190	72.323	37.1	8.7	21.5	10.8	  
1977	72.753	33.2	7.9	21.5	6.6	  
1828	72.207	22.3	5	21.5	3.8	  
4280	72.82	26.7	5.8	21.4	3.7	  
3932	72.851	31	6.6	21.4	11.2	  
3835	72.707	44.3	10.9	21.4	14.4	  
3793	72.22	30.9	6.6	21.4	13	  
3683	72.601	37.5	10.2	21.4	3.7	  
3520	72.055	42.5	10.9	21.4	4.6	  
3504	72.513	30.4	7.3	21.4	4.6	  
3356	72.3	51.7	11.7	21.4	4.6	  
3046	72.608	38.1	10.2	21.4	5.3	  
2956	72.236	24.7	6.6	21.4	4.6	  
2929	72.483	28	6.6	21.4	8.1	  
2925	72.289	29.5	6.6	21.4	9.4	  
2667	72.268	25	5.8	21.4	3.7	  
2645	72.503	30.4	7.3	21.4	4.6	  
2640	72.459	39	9.5	21.4	4.6	  
2637	72.395	21.8	5.8	21.4	4.6	  
2479	72.852	41.3	10.2	21.4	3.7	  
2053	72.725	34.8	7.2	21.4	5.3	  
1465	72.63	31.1	7.2	21.4	14.9	  
282	72.145	29.8	7.2	21.4	5.3	  
4238	72.729	30.2	6.6	21.3	5.2	  
4205	72.681	47.8	10.2	21.3	5.9	  
3837	72.317	35.6	9.5	21.3	7.6	  
3792	72.352	33.7	9.5	21.3	11.8	  
3761	73.208	33.7	8	21.3	4.5	  
3723	72.326	43.7	11.7	21.3	7.6	  
3688	72.827	37.8	8.8	21.3	3.5	  
3636	72.897	37.3	8	21.3	3.5	  
3521	72.13	26.3	5.1	21.3	8.9	  
3505	72.415	47.7	12.4	21.3	5.2	  
3281	72.208	42.4	11.7	21.3	4.5	  
3037	72.501	48.7	13.1	21.3	4.5	  
3026	72.717	22.6	6.6	21.3	5.9	  
2675	72.575	40.9	9.5	21.3	4.5	  
2644	72.435	18.9	5.1	21.3	3.5	  
2564	72.279	18.4	4.4	21.3	4.5	  
2527	72.496	27.7	7.3	21.3	10.1	  
2415	72.523	30.7	7.9	21.3	5.9	  
2291	72.59	27.4	6.5	21.3	3.5	  
1844	72.603	23.3	5	21.3	7.6	  
1835	72.431	41.1	8.7	21.3	13.6	  
762	72.765	49.7	12.3	21.3	4.5	  
745	72.168	29.9	7.2	21.3	7.6	  
1432	72.475	27.5	6.5	21.27	18.3	  
4286	72.541	49.1	10.2	21.2	3.3	  
3927	72.773	26.7	6.6	21.2	3.3	  
3533	72.323	35.3	8.8	21.2	3.3	  
3407	72.868	18.5	4.4	21.2	3.3	  
3381	72.68	56	11.7	21.2	6.4	  
3305	72.451	52.2	10.2	21.2	5.1	  
2970	72.51	35.7	8	21.2	10.4	  
2954	72.228	38.7	8.8	21.2	12.7	  
2834	72.974	23	5.1	21.2	8.9	  
2657	72.539	9.2	2.2	21.2	5.1	  
2555	72.591	42.4	10.2	21.2	3.3	  
2514	72.347	38.7	10.2	21.2	7	  
2500	72.489	31.2	9.5	21.2	3.3	  
2461	72.571	28	6.5	21.2	8.4	  
2293	72.618	39.3	8.7	21.2	6.4	  
2225	72.225	54.5	11.5	21.2	3.3	  
2212	72.217	38.3	9.4	21.2	4.3	  
2184	72.522	47.9	10.1	21.2	6.4	  
1908	72.697	33.5	7.9	21.2	8	  
1747	72.138	43.3	10.2	21.2	8.9	  
3857	72.352	32.7	8.8	21.1	10	  
3791	72.352	41.7	9.5	21.1	4.1	  
3752	72.71	44.7	9.5	21.1	4.9	  
3516	72.583	51.8	11.7	21.1	4.1	  
3386	72.605	50.5	12.4	21.1	7.4	  
3028	72.782	24.6	5.1	21.1	4.9	  
2947	72.13	45.2	10.9	21.1	5.6	  
2881	72.43	22.2	5.1	21.1	4.1	  
2792	72.426	44.7	10.2	21.1	9.2	  
2688	72.079	21.1	5.1	21.1	4.1	  
2593	72.567	45.9	10.2	21.1	4.9	  
2566	72.646	28.4	6.6	21.1	4.1	  
2563	72.315	38.1	10.2	21.1	6.3	  
2558	72.34	40.8	10.2	21.1	3	  
2532	72.458	25.6	5.8	21.1	4.9	  
2438	72.645	41.4	9.4	21.1	6.9	  
2227	72.11	38.3	10.1	21.1	3	  
2221	72.263	14	3.6	21.1	3	  
2209	72.371	45.9	11.5	21.1	4.9	  
2197	72.293	29.1	6.5	21.1	6.9	  
2070	72.13	32.4	7.9	21.1	5.6	  
784	71.88	38.5	9.4	21.1	8.4	  
771	72.56	53.6	13	21.1	3	    
765	72.983	45.5	10.1	21.1	4.9	  
4259	72.095	36.5	8	21	4.7	  
4209	72.675	26.8	6.6	21	2.6	  
3800	72.475	24.4	5.8	21	11.4	  
3745	72.63	51.9	10.9	21	9.9	  
3345	72.463	34	8	21	6.1	  
3105	72.741	27.1	8	21	8.2	  
3018	72.603	31.8	7.3	21	11.4	  
3014	72.612	37.8	10.2	21	3.8	  
2882	72.458	34.6	8.8	21	5.4	  
2800	72.516	53.5	11.7	21	2.6	  
2686	72.087	39	9.5	21	4.7	  
2646	72.641	32.4	8	21	3.8	  
2632	72.361	26.8	5.8	21	3.8	  
2585	72.345	24.2	5.8	21	3.8	  
2584	72.457	31	7.3	21	2.6	  
2255	72.588	52	10.8	21	2.6	  
2145	72.031	36.8	9.4	21	2.6	  
2099	72.342	24	6.5	21	3.8	  
2074	71.923	36.1	7.9	21	3.8	  
1825	72.09	35.1	7.9	21	4.7	  
1755	72.235	47.8	12.4	21	8.2	  
1429	72.351	25.6	5.8	21	4.7	  
4183	72.735	27.5	6.6	20.9	1.9	  
4089	72.625	40.4	10.2	20.9	11.6	  
3916	72.448	33.4	7.3	20.9	9.4	  
3795	72.412	29.5	8	20.9	7.6	  
3758	73.12	50.1	12.4	20.9	5.9	  
3743	72.395	36.5	7.3	20.9	1.9	  
3523	72.52	32.8	8	20.9	3.3	  
3443	72.912	27.2	7.3	20.9	3.3	  
3390	72.436	44.8	10.9	20.9	7.1	  
2949	72.159	29.9	6.6	20.9	5.2	  
2642	72.458	28	6.6	20.9	5.9	  
2518	72.34	34.9	8	20.9	6.5	  
2402	72.663	31.1	7.2	20.9	5.2	  
2322	73.01	16.5	4.3	20.9	4.3	  
2312	72.424	18.6	4.3	20.9	3.3	  
2292	72.566	38.7	7.9	20.9	1.9	  
2267	72.575	18.6	5	20.9	1.9	  
2226	72.221	27.1	6.5	20.9	1.9	  
1878	71.866	31.3	7.9	20.9	9	  
1827	72.146	36.4	10.1	20.9	9	  
777	72.308	49.2	11.6	20.9	10.2	  
749	72.196	43.4	9.4	20.9	3.3	  
1754	72.453	35.5	8	20.86	16.7	  
4182	72.626	23.5	6.6	20.8	0	  
3703	72.405	36.8	8.8	20.8	5.6	  
3699	72.419	53.7	11.7	20.8	10.1	  
3522	72.214	40.1	8	20.8	5.6	  
3442	73.006	39.9	10.9	20.8	2.8	  
3344	72.448	45.5	11.7	20.8	8.8	  
3337	72.87	39.3	9.5	20.8	0	  
3304	72.898	26.8	8	20.8	0	  
3069	72.516	43.7	10.9	20.8	4.8	  
3055	72.816	33.4	8	20.8	3.9	  
2910	72.327	39.4	8.8	20.8	2.8	  
2733	72.815	36.5	8.8	20.8	2.8	  
2701	72.576	42.6	10.9	20.8	2.8	  
2687	72.02	40.1	10.2	20.8	6.2	  
2671	72.249	25.1	6.6	20.8	5.6	  
2658	72.467	43.1	10.2	20.8	6.2	  
2590	72.137	33.5	8.8	20.8	2.8	  
2534	72.326	19.9	4.4	20.8	4.8	  
2499	72.468	39.8	8	20.8	3.9	  
2409	72.581	19.3	4.3	20.8	2.8	  
2398	72.648	36.5	7.9	20.8	2.8	  
1855	72.128	38.5	8.7	20.8	6.2	  
1832	72.235	45.7	10.8	20.8	6.2	  
1181	72.137	49.9	10.9	20.8	6.8	  
1079	72.092	33.7	8	20.8	4.8	  
753	72.016	38.7	10.1	20.8	2.8	  
752	72.011	45.3	11.6	20.8	3.9	  
747	72.388	47.3	10.1	20.8	2.8	  
731	72.196	35.7	7.2	20.8	3.9	  
1957	72.751	30.7	7.9	20.76	20.5	  
4283	72.643	52.6	12.4	20.7	1.9	  
4235	72.122	47.3	10.2	20.7	1.9	  
4210	72.615	35.4	8.8	20.7	1.9	  
4184	72.817	35.6	8.8	20.7	1.9	  
3923	72.663	48.8	11.7	20.7	9.9	  
3829	72.666	27.2	8	20.7	4.4	  
3751	72.803	27.2	5.8	20.7	7.1	  
3457	72.46	30.8	6.6	20.7	5.9	  
3361	72.453	46.4	12.4	20.7	5.2	  
3306	72.796	19.9	5.1	20.7	3.4	  
3064	72.421	31.6	8	20.7	5.2	  
3060	72.345	39.6	9.5	20.7	5.2	  
3021	72.742	34.9	8	20.7	8.6	  
3019	72.622	47.2	10.9	20.7	5.2	  
2805	72.368	52	12.4	20.7	1.9	  
2720	72.575	31.6	7.3	20.7	6.6	  
2668	72.313	28	7.3	20.7	1.9	  
2636	72.348	33.1	7.3	20.7	3.4	  
2439	72.761	24.1	5	20.7	5.2	  
2357	72.758	27.1	5.8	20.7	6.6	  
2231	72.191	37.7	7.9	20.7	1.9	  
2230	72.208	32.2	7.2	20.7	1.9	  
2083	72.55	32.4	7.2	20.7	3.4	  
1840	72.421	46.6	10.8	20.7	1.9	  
1046	72.047	48	12.3	20.7	3.4	  
785	71.823	31.7	8	20.7	5.9	  
758	71.799	36	10.1	20.7	3.4	  
754	71.914	25.5	6.5	20.7	3.4	  
1916	72.452	37.5	7.9	20.64	27	  
4279	72.801	48.1	12.4	20.6	2.6	  
4277	72.883	53.2	12.4	20.6	2.6	  
4262	72.244	31.1	8	20.6	2.6	  
4260	72.326	51.6	11.7	20.6	3.8	  
3348	72.537	37.3	10.2	20.6	3.8	  
2911	72.299	32.1	8	20.6	2.6	  
2865	72.274	37	9.5	20.6	8.4	  
2750	72.48	47.4	11.7	20.6	4.8	  
2677	72.366	38.6	8.8	20.6	4.8	  
2529	72.058	39.6	9.5	20.6	2.6	  
2496	72.465	26	6.6	20.6	3.8	  
2494	72.393	18.8	5.1	20.6	4.8	  
2455	72.392	42.3	9.4	20.6	3.8	  
2454	72.426	48.5	10.8	20.6	3.8	  
2297	72.383	27	6.5	20.6	3.8	  
2229	72.215	27.3	7.2	20.6	2.6	  
2181	72.392	42.9	10.8	20.6	6.2	  
2114	72.786	41.1	12.3	20.6	9.7	  
2072	72.138	41.6	12.3	20.6	7.4	  
1963	72.932	37.6	8.7	20.6	16.2	  
1958	72.843	32.6	7.9	20.6	9.3	  
1886	72.128	28.3	7.2	20.6	10.1	  
1875	72.015	26	6.5	20.6	2.6	  
1821	72.056	44.5	11.5	20.6	11.3	  
1075	71.991	30.3	8.7	20.6	8.4	  
817	71.946	26.5	7.2	20.6	4.8	    
757	71.757	41.7	9.4	20.6	3.8	  
277	71.843	37.7	8.7	20.6	25.5	  
4281	72.728	41.7	9.5	20.5	3.1	  
4190	72.504	44.8	10.9	20.5	5.8	  
3914	72.489	35.4	8	20.5	7.1	  
3861	72.543	47.4	11.7	20.5	9.5	  
3830	72.888	30.3	8.8	20.5	4.2	  
3824	72.595	46.7	11.7	20.5	5.1	  
3749	72.814	25.2	7.3	20.5	9.5	  
3535	72.188	35.3	7.3	20.5	6.5	  
3282	72.183	47.9	12.4	20.5	3.1	  
2985	72.54	34.3	8	20.5	9.1	  
2831	72.951	21.9	5.1	20.5	7.6	  
2819	72.341	23.1	4.4	20.5	5.1	  
2816	72.346	35.5	9.5	20.5	3.1	  
2808	72.27	34.3	8.8	20.5	6.5	  
2804	72.565	22.7	5.8	20.5	3.1	  
2752	72.585	46.1	11.7	20.5	4.2	  
2714	72.719	55.4	12.4	20.5	10.7	  
2589	72.278	32.2	6.6	20.5	5.1	  
2394	72.483	35.7	8.7	20.5	4.2	  
2349	72.978	40.8	9.4	20.5	15.4	  
2256	72.544	34.2	8.7	20.5	4.2	  
2147	72.248	28	7.2	20.5	7.6	  
2146	72.256	20.6	5	20.5	9.5	  
2063	72.623	29.1	7.2	20.5	3.1	  
2051	72.603	46	10.1	20.5	8.6	  
1823	72.022	29.7	8.7	20.5	8.6	  
781	71.991	29.2	5.8	20.5	7.1	  
780	72.136	31.7	7.2	20.5	8.6	  
4206	72.546	31.5	7.3	20.4	3.5	  
3860	72.5	31.2	6.6	20.4	5.3	  
3754	72.91	35.9	9.5	20.4	3.5	  
3641	71.916	36.4	9.5	20.4	6.7	  
3431	72.686	50.1	12.4	20.4	3.5	  
3409	73.046	42.2	8.8	20.4	3.5	  
3007	72.126	39.6	10.2	20.4	7.8	  
2952	72.228	36.5	10.2	20.4	11.6	  
2817	72.415	33.4	8.8	20.4	6	  
2659	72.371	62	13.1	20.4	4.5	  
2650	72.533	21.5	5.1	20.4	4.5	  
2604	72.375	31.8	8	20.4	6.7	  
2498	72.377	25.9	5.8	20.4	3.5	  
2493	72.418	29.6	7.3	20.4	8.8	  
2401	72.58	24	6.5	20.4	7.8	  
2399	72.614	43.3	8.7	20.4	4.5	  
2347	72.991	29.4	6.5	20.4	3.5	  
2265	72.736	34.9	7.9	20.4	6.7	  
2071	72.218	35.8	9.4	20.4	6.7	  
2002	72.365	35.5	8.7	20.4	15.5	  
1952	72.611	33.1	8.7	20.4	9.2	  
1887	72.228	44.2	10.8	20.4	18.2	  
1407	72.376	35.5	9.4	20.4	11.6	  
996	72.035	45	9.4	20.4	5.3	  
735	72.623	54.7	11.6	20.4	4.5	  
118	72.359	52.5	13.7	20.4	17.9	  
4285	72.624	45.6	11.7	20.3	3.7	  
4282	72.636	37.3	9.5	20.3	3.7	  
3866	72.663	27.4	7.3	20.3	5.5	  
3827	72.795	44.7	10.9	20.3	11.7	  
3826	72.815	40.4	10.2	20.3	8.9	  
3408	72.92	21.9	5.1	20.3	3.7	  
3355	72.403	24.9	6.6	20.3	6.8	  
3303	72.733	30.3	7.3	20.3	3.7	  
3092	72.46	34	8	20.3	6.2	  
3013	72.501	35	9.5	20.3	3.7	  
2909	72.324	41.7	10.2	20.3	3.7	  
2871	72.353	45	12.4	20.3	11.4	  
2738	72.5	47.1	10.9	20.3	4.7	  
2737	72.818	49.7	10.9	20.3	5.5	  
2672	72.241	21.5	5.8	20.3	5.5	  
2649	72.475	18.6	4.4	20.3	3.7	  
2605	72.287	34.9	7.3	20.3	6.2	  
2525	71.915	24.2	6.6	20.3	3.7	  
2476	72.753	45.2	10.9	20.3	3.7	  
1133	71.963	37.5	10.2	20.3	4.7	  
963	72.381	40.6	10.9	20.3	3.7	  
850	72.631	47.2	10.8	20.3	12.1	  
770	73.168	40	8.7	20.3	6.2	  
4191	72.678	27	6.6	20.2	10.3	  
3851	72.335	45.4	10.9	20.2	9	  
3432	72.62	57.6	12.4	20.2	3.9	  
3411	72.7	30.6	8	20.2	3.9	  
3395	72.201	49.4	13.1	20.2	8.1	  
3066	72.42	42.9	12.4	20.2	6.3	  
2979	72.409	51	13.1	20.2	6.3	  
2906	72.208	23.4	5.8	20.2	4.9	  
2902	72.13	34.7	8.8	20.2	7	  
2684	72.024	35	9.5	20.2	5.6	  
2670	72.231	21.3	5.1	20.2	3.9	  
2660	72.48	43.1	10.2	20.2	6.3	  
2654	72.531	14	3.6	20.2	5.6	  
2502	72.462	38.2	9.5	20.2	3.9	  
2477	72.455	17.8	4.4	20.2	3.9	  
2397	72.599	21.4	5	20.2	4.9	  
2396	72.518	46.7	10.8	20.2	5.6	  
2153	72.186	27.2	5.8	20.2	7	  
2152	72.221	30.7	7.9	20.2	5.6	  
2104	72.478	38.6	10.1	20.2	9	  
2089	72.715	47.1	10.8	20.2	5.6	  
1427	72.331	29.9	7.2	20.2	3.9	  
1105	71.886	41.6	10.2	20.2	5.6	    
123	72.999	55.6	13	20.2	8.1	    
3831	72.991	23.3	6.6	20.1	4	  
3816	72.51	31.7	8	20.1	4	  
3815	72.511	40.3	9.5	20.1	4	  
3526	72.277	27.9	7.3	20.1	4	  
3518	71.96	33.8	6.6	20.1	4	  
3351	72.448	48.5	10.9	20.1	4	  
3059	72.311	29.9	5.8	20.1	4	  
2821	72.218	35.5	7.3	20.1	4	  
2719	72.633	30.3	7.3	20.1	4	  
2603	72.378	20.9	5.1	20.1	4	  
2533	72.387	10.6	2.2	20.1	5.7	  
2519	72.355	22.2	5.1	20.1	4	  
2451	72.382	45.5	10.8	20.1	5.7	  
2387	72.517	13.4	3.6	20.1	5	  
1944	72.344	45.7	12.3	20.1	7.6	  
1880	72.028	42.6	10.1	20.1	9.1	  
1042	71.95	45.4	10.2	20.1	7.6	  
769	73.13	44.1	11.6	20.1	6.4	  
4213	72.526	33.8	8	20	5	  
3858	72.436	38.2	10.2	20	9.2	  
3794	72.413	34	8.8	20	5.8	  
3061	72.202	42.6	10.2	20	9.2	  
3049	72.703	33.2	8.8	20	4.1	  
2944	72.02	38	9.5	20	5	  
2755	72.63	24.6	6.6	20	4.1	  
2705	72.621	41	9.5	20	4.1	  
2647	72.548	46.6	10.9	20	5	  
2634	72.496	36.4	9.5	20	9.2	  
2497	72.411	15.1	3.6	20	5	  
2489	72.645	35.9	8.8	20	4.1	  
2356	72.865	26.3	6.5	20	5	  
2280	72.548	21.2	5	20	10.9	  
2085	72.623	35.9	9.4	20	5	  
2073	72.029	38.5	9.4	20	4.1	  
803	72.691	48.4	10.8	20	12.7	  
774	73.13	42.8	10.1	20	10.5	    
733	72.393	54.1	13	20	4.1	  
98	72.124	50.7	12.3	19.95	11.3	    
4214	72.656	43.7	10.9	19.9	4.1	  
4211	72.559	29.3	7.3	19.9	4.1	  
3738	72.726	34.3	8.8	19.9	7.7	  
3736	72.815	27.3	8	19.9	7.1	  
3436	73.055	31.6	7.3	19.9	4.1	  
3362	72.447	31	7.3	19.9	4.1	  
3070	72.414	35.3	9.5	19.9	5	  
3065	72.411	36.1	9.5	19.9	5	  
2899	72.116	37.3	9.5	19.9	9.7	  
2828	72.858	24.4	5.8	19.9	4.1	  
2726	72.45	49.5	10.9	19.9	4.1	  
2722	72.655	23	5.8	19.9	5	  
2643	72.492	33.1	8.8	19.9	5	  
2594	72.56	47.7	9.5	19.9	4.1	  
2318	72.971	12.2	2.9	19.9	4.1	  
2119	72.651	39.9	8.7	19.9	11.7	  
1938	72.238	31.4	7.9	19.9	9.2	  
1921	72.538	24.1	6.5	19.9	7.7	  
1100	71.858	39.8	8.7	19.9	7.1	  
1065	72.307	35.3	8.7	19.9	12.7	  
755	71.766	63.4	14.5	19.9	6.5	  
290	72.745	32.6	8	19.9	19.6	    
286	72.695	50.1	11.6	19.9	5.8	    
114	72.168	54.7	12.3	19.9	4.1	  
1071	72.001	27	6.5	19.88	18.2	  
95	72.192	52.8	-5.5	19.83	11	    
4288	72.381	40.8	10.9	19.8	5.8	  
3817	72.523	49.7	11.7	19.8	5.8	  
3656	72.525	27.5	8	19.8	5.8	  
3552	72.69	23.9	5.8	19.8	6.5	  
3444	72.558	29.4	8	19.8	5.8	  
3439	72.579	37.8	8	19.8	4	  
3435	73.035	52.2	11.7	19.8	4	  
3430	72.543	45.7	12.4	19.8	5	  
3111	72.843	32.8	8	19.8	4	  
3074	72.608	43.2	10.9	19.8	4	  
3011	72.48	30.3	6.6	19.8	5	  
3006	72.38	34.9	8.8	19.8	4	  
2943	72.038	17	3.6	19.8	6.5	  
2917	72.507	36	8.8	19.8	5	  
2873	72.348	34.4	8	19.8	4	  
2747	72.573	29.4	7.3	19.8	5.8	  
2725	72.585	26.8	6.6	19.8	5	  
2639	72.655	33.4	8	19.8	7.1	  
2610	71.985	36.1	9.5	19.8	4	  
2607	71.874	34.4	8	19.8	4	  
2591	72.274	34.2	7.3	19.8	5.8	  
2517	72.098	41.7	10.2	19.8	4	  
2495	72.3	37.3	9.5	19.8	5.8	  
2452	72.431	53.9	11.5	19.8	4	  
2304	72.385	23.3	5	19.8	12.4	  
2106	72.572	51	11.5	19.8	4	  
2076	72.325	47.8	10.8	19.8	5	  
2037	72.363	30	7.9	19.8	5	  
1936	72.056	34.5	7.9	19.8	5	  
1831	72.23	42.3	10.8	19.8	9.2	  
961	71.877	42.8	9.4	19.8	8.2	  
848	72.535	37.7	8.6	19.8	10.5	  
839	71.558	53	11.5	19.8	10.5	    
793	72.241	54.2	13	19.8	11.7	  
779	72.19	52.4	10.1	19.8	7.7	  
763	71.751	42.1	10.8	19.8	4	  
117	72.323	31.7	8.7	19.8	5.8	  
3438	72.696	29.4	7.3	19.7	3.8	  
3413	72.749	37.3	9.5	19.7	3.8	  
3269	72.443	35.9	8	19.7	4.8	  
3080	72.686	27.7	6.6	19.7	3.8	  
3010	72.511	44	10.9	19.7	3.8	  
2896	71.999	46.9	11.7	19.7	5.7	  
2880	72.442	22.3	6.6	19.7	5.7	  
2706	72.696	34.9	8.8	19.7	6.4	  
2446	72.261	25.3	5.8	19.7	4.8	  
2384	72.655	41.6	9.4	19.7	6.4	  
2117	72.678	39.8	10.1	19.7	4.8	  
1764	72.793	47.9	10.9	19.7	3.8	  
1092	72.215	32	7.2	19.7	3.8	  
819	72.106	38.7	10.8	19.7	5.7	    
772	73.16	40	9.4	19.7	9.7	  
1034	71.924	21.5	5.1	19.69	24.2	  
2060	72.438	47.5	12.3	19.64	21.9	  
4215	72.566	50.8	12.4	19.6	4.7	  
4185	72.544	42.1	10.9	19.6	3.6	  
3643	72.104	36.8	9.5	19.6	9.6	  
3440	72.653	41.2	10.2	19.6	3.6	  
3359	72.472	27.2	6.6	19.6	4.7	  
3350	72.531	45.1	10.9	19.6	5.5	  
2984	72.533	47.5	10.2	19.6	4.7	  
2916	72.405	36.6	8	19.6	6.9	  
2809	72.447	55.3	12.4	19.6	3.6	  
2651	72.511	22.1	4.4	19.6	4.7	  
2638	72.705	25.8	5.8	19.6	4.7	  
2567	72.468	27.2	7.3	19.6	4.7	  
2460	72.561	33.6	6.5	19.6	6.9	  
2198	72.505	22.2	5.8	19.6	5.5	  
2077	72.42	56.5	14.4	19.6	8.1	  
2056	72.51	36.7	8.7	19.6	5.5	  
1995	72.257	43.5	10.1	19.6	6.3	  
1965	73.001	21.3	4.3	19.6	13.8	  
1945	72.309	29.9	7.2	19.6	3.6	  
1935	72.154	48.3	11.5	19.6	3.6	  
1839	72.505	41.9	8.7	19.6	11.3	  
1533	72.016	25.5	6.6	19.6	6.9	  
1050	72.039	36	7.2	19.6	7.5	  
981	72.406	33.4	9.4	19.6	4.7	  
818	72.011	31.8	7.9	19.6	4.7	  
775	73.069	43.2	11.6	19.6	6.9	    
133	72.61	50.7	12.3	19.6	4.7	  
86	72.083	34.9	-0.5	19.51	14.3	    
4287	72.301	55.2	11.7	19.5	6.1	  
4221	72.648	53.8	12.4	19.5	3.3	  
3930	72.699	31	6.6	19.5	7.4	  
3859	72.539	43.8	10.2	19.5	10.4	  
3531	72.521	16.6	3.6	19.5	3.3	  
3441	72.97	32.2	9.5	19.5	3.3	  
3414	72.703	20.5	6.6	19.5	5.3	  
3073	72.603	21.2	5.8	19.5	3.3	  
2957	72.238	42.1	10.9	19.5	8.5	  
2898	72.138	41.7	11.7	19.5	5.3	  
2814	72.603	60.8	13.1	19.5	4.4	  
2806	72.516	59.3	13.9	19.5	3.3	  
2798	72.405	43.3	10.2	19.5	3.3	  
2751	72.687	46.4	10.2	19.5	3.3	  
2653	72.527	23.5	6.6	19.5	3.3	  
2573	72.28	42.9	9.5	19.5	3.3	  
2325	73.052	9.3	2.2	19.5	5.3	  
1897	72.445	47.7	10.8	19.5	6.8	  
1824	71.936	36.4	8.7	19.5	5.3	  
1415	72.068	38.4	8.7	19.5	6.8	  
85	72.147	41.5	-4.1	19.5	17	    
4218	72.525	22.5	5.1	19.4	4.1	  
4212	72.517	27.2	7.3	19.4	5.1	  
3850	72.245	44.7	11.7	19.4	9.9	  
3532	72.378	31.2	7.3	19.4	2.8	  
3266	72.733	30.7	7.3	19.4	2.8	  
3068	72.483	30.5	6.6	19.4	2.8	  
2915	72.428	33.9	8.8	19.4	2.8	  
2866	72.158	40.6	10.9	19.4	5.1	  
2842	72.671	29.1	7.3	19.4	7.2	  
2724	72.482	57.9	10.9	19.4	2.8	  
2648	72.56	41.2	9.5	19.4	2.8	  
2608	72.518	51.2	10.9	19.4	2.8	  
2565	72.433	22.4	5.8	19.4	5.1	  
2447	72.345	33.6	7.9	19.4	4.1	  
2100	72.49	24.6	5.8	19.4	9.4	  
2065	72.423	37.2	7.9	19.4	7.2	  
1885	72.11	34.9	7.9	19.4	6.6	  
838	71.511	15.4	3.6	19.4	6.6	  
783	71.803	70.3	13.7	19.4	4.1	    
778	72.403	56.7	13.7	19.4	8.9	  
526	71.801	33.6	9.4	19.4	8.4	  
132	72.373	38.7	9.4	19.4	9.9	  
1807	72.603	60.4	13.1	19.35	12.7	  
3412	72.896	33.2	7.3	19.3	2	  
3410	72.954	27.8	7.3	19.3	2	  
3015	72.613	40.9	10.2	19.3	2	  
2936	72.564	33	7.3	19.3	3.6	  
2869	72.222	25.6	5.8	19.3	7	  
2841	72.759	48	10.9	19.3	5.6	  
2820	72.318	38.4	10.2	19.3	5.6	  
2756	72.611	38.2	9.5	19.3	2	  
2689	72.083	38.2	9.5	19.3	5.6	  
2478	72.504	31.4	8.8	19.3	2	  
2445	72.441	38.7	7.9	19.3	3.6	  
2442	72.721	52.8	10.8	19.3	3.6	  
2358	72.675	30.9	7.9	19.3	24.4	  
2239	71.935	28.9	6.5	19.3	9.3	  
2206	72.67	25.9	5.8	19.3	6.3	  
1961	72.703	29.1	7.2	19.3	10.2	  
1929	72.562	26.2	6.5	19.3	12.6	  
1596	72.158	40.6	9.5	19.3	11.9	  
1463	72.491	37.4	8.7	19.3	4.7	  
1190	72.033	24.9	6.5	19.3	7	  
1177	72.134	46.4	10.2	19.3	7	  
797	72.43	41	10.1	19.3	8.2	  
4216	72.654	41.1	8.8	19.2	0	  
3527	72.377	28.8	7.3	19.2	6	  
3525	72.336	38.2	8	19.2	4.3	  
3524	72.463	19.9	5.1	19.2	4.3	  
3434	72.836	52.7	12.4	19.2	0	  
3078	72.594	31.1	7.3	19.2	3	  
3071	72.491	30	8	19.2	5.2	  
3008	72.418	31.3	8.8	19.2	6.8	  
2759	72.496	21.8	5.8	19.2	3	  
2723	72.631	16.4	3.6	19.2	4.3	  
2588	72.187	37.6	8.8	19.2	0	  
2528	71.961	32	7.3	19.2	5.2	  
2400	72.606	24.9	5.8	19.2	5.2	  
2395	72.492	30.6	7.2	19.2	4.3	  
2078	72.434	41.6	10.1	19.2	4.3	  
1750	72.327	40.4	10.2	19.2	4.3	  
858	72.399	35.1	8.6	19.2	9.1	  
857	72.514	30.1	7.2	19.2	8	  
851	72.691	53.4	10.1	19.2	6	  
801	72.703	50	13.7	19.2	3	  
800	72.655	24.7	5	19.2	8.6	  
786	71.844	30.3	7.2	19.2	5.2	    
347	72.396	28.4	7.2	19.2	0	  
310	71.756	32	8	19.2	8.6	  
292	72.998	41.6	11.6	19.2	8.6	  
2057	72.441	32.2	9.4	19.19	13.6	  
42	72.023	10.7	5.3	19.17	9.5
1094	72.156	30.5	6.5	19.13	15.4	  
4220	72.66	48.3	10.2	19.1	2	  
3737	72.814	31.4	8.8	19.1	7.7	  
3437	72.916	26	6.6	19.1	2	  
3433	72.668	27.2	5.8	19.1	4.8	  
3397	72.401	20.7	5.8	19.1	7.1	  
3360	72.166	24.4	7.3	19.1	3.7	  
3076	72.698	43.8	9.5	19.1	3.7	  
2963	72.313	29.4	8.8	19.1	4.8	  
2895	71.938	46.5	10.9	19.1	2	  
2635	72.469	25.1	6.6	19.1	9.4	  
2326	72.936	39.9	9.4	19.1	4.8	  
1458	72.35	51.1	11.6	19.1	8.3	  
1171	72.132	20.4	5.1	19.1	11.2	  
965	71.895	18.6	5.1	19.1	6.4	  
853	72.67	44.4	10.1	19.1	6.4	  
840	72.106	47.4	11.5	19.1	5.6	  
808	72.551	41.1	10.8	19.1	4.8	  
751	71.83	55.9	13	19.1	4.8	  
280	71.876	43.9	10.8	19.1	2	  
40	72.001	28.7	0.3	19.1	12	  
4219	72.61	42.8	10.9	19	5.2	  
3863	72.655	31.1	7.3	19	2.8	  
3012	72.503	39.4	10.9	19	4.2	  
2894	71.916	37.7	10.2	19	5.2	  
2599	72.463	25.5	5.1	19	8	  
2598	72.523	38.6	10.2	19	4.2	  
2520	72.78	40	9.5	19	5.2	  
2203	72.646	22.7	5	19	6.7	  
1532	72.325	29.6	6.6	19	6.7	  
361	72.118	37.2	9.4	19	6	  
226	72.108	49.2	11.6	19	9.1	  
81	72.045	48.5	-4.8	19	14	    
2127	72.375	23.6	5.8	18.96	9	  
4278	72.523	41.5	8.8	18.9	5.5	  
3825	72.673	33.8	7.3	18.9	4.6	  
3093	72.35	30.9	7.3	18.9	5.5	  
3043	72.583	49.9	13.1	18.9	7	  
3005	72.406	34.3	7.3	18.9	3.4	  
2913	72.299	33.3	7.3	18.9	3.4	  
2822	72.283	26.2	6.6	18.9	5.5	  
2694	72.281	21.6	5.1	18.9	5.5	  
2674	72.324	29.6	7.3	18.9	3.4	  
2600	72.311	39.1	9.5	18.9	7	  
2482	72.509	28.6	8	18.9	4.6	  
2385	72.452	40.2	10.1	18.9	4.6	  
2111	72.571	35.9	8.7	18.9	5.5	  
2086	72.608	37.2	8.7	18.9	3.4	  
1939	72.257	32.3	9.4	18.9	3.4	  
1937	72.06	44.4	10.8	18.9	3.4	  
1412	72.068	45.9	10.9	18.9	10.6	  
1128	72.148	24.1	5.8	18.9	7.7	  
1029	72.106	41.4	9.4	18.9	8.8	  
980	72.316	29.4	7.2	18.9	7	  
910	72.05	27.1	6.5	18.9	7.7	  
871	71.879	29.7	7.9	18.9	4.6	  
846	72.402	37.6	7.9	18.9	6.3	  
836	71.627	56.5	13	18.9	7	  
822	72.161	52.2	12.2	18.9	4.6	  
766	71.689	36.7	9.4	18.9	3.4	  
756	71.701	29.7	6.5	18.9	5.5	  
1448	72.026	28.9	6.5	18.88	12.1	  
44	72.073	27.9	1	18.85	13.9	
3862	72.664	42.8	10.2	18.8	3.8	  
3530	72.291	34.6	8.8	18.8	3.8	  
3394	72.388	10.8	2.9	18.8	3.8	  
2744	72.563	30.4	7.3	18.8	3.8	  
2652	72.585	31.1	6.6	18.8	6.5	  
2531	72.366	38.2	10.2	18.8	4.9	  
2480	72.766	50.5	13.9	18.8	3.8	  
2393	72.52	35.8	8.7	18.8	6.5	  
2103	72.485	46.8	12.3	18.8	10.9	  
1123	71.864	44.1	9.4	18.8	12.2	  
962	71.879	37.8	9.4	18.8	4.9	  
925	72.653	50.1	12.2	18.8	19.2	  
820	72.115	37.9	8.6	18.8	4.9	  
782	71.953	30.9	5.8	18.8	5.8	  
732	71.841	40.5	9.4	18.8	6.5	  
231	72.023	32.8	8.7	18.8	4.9	  
222	71.726	27.7	6.5	18.8	3.8	  
120	72.343	34	8.7	18.8	8.5	  
1534	72.096	41.4	9.5	18.73	6.8	  
4223	72.72	27	7.3	18.7	4	  
2953	72.221	28	7.3	18.7	4	  
2727	72.351	61.8	10.9	18.7	6.7	  
2685	72.173	42.6	10.9	18.7	5.1	  
2568	72.6	33.4	7.3	18.7	4	  
2486	72.634	44.4	9.5	18.7	5.1	  
2232	71.973	52.2	11.5	18.7	4	  
2186	72.478	55.6	11.5	18.7	7.4	  
1927	72.63	21.2	5	18.7	4	  
888	72.437	49.1	11.5	18.7	6.7	    
874	71.977	29.9	7.2	18.7	6	  
809	72.421	39.3	10.1	18.7	4	  
370	72.255	33.8	7.2	18.7	6	  
358	71.973	37.3	8.7	18.7	5.1	  
303	72.09	29.4	8.7	18.7	4	  
279	71.963	35.1	8.7	18.7	4	  
265	71.786	35.7	8.7	18.7	4	  
227	72.14	39.6	9.4	18.7	6	  
225	71.77	46.3	11.6	18.7	4	  
2311	72.31	37.1	9.4	18.62	10.9	  
3823	72.631	33.4	7.3	18.6	7.6	  
3812	72.375	45	10.2	18.6	6.1	  
3398	72.352	19.1	5.1	18.6	8.8	  
3393	72.383	21.5	5.1	18.6	5.3	  
3391	72.385	28.9	6.6	18.6	4.3	  
2740	72.563	45.4	11.7	18.6	5.3	  
2696	72.175	24.4	5.8	18.6	4.3	  
2693	72.395	19.6	4.4	18.6	4.3	  
2552	72.485	17.1	4.4	18.6	5.3	  
2205	72.546	34	6.5	18.6	6.1	  
2118	72.677	55.1	12.3	18.6	5.3	  
1423	72.363	27.5	6.5	18.6	4.3	  
914	72.387	24.6	6.5	18.6	5.3	  
875	72.075	36.3	8.6	18.6	9.3	  
870	71.885	38.1	7.9	18.6	5.3	  
828	72.058	18.3	4.3	18.6	5.3	    
746	71.849	44.7	10.8	18.6	4.3	  
583	71.864	32.2	8	18.6	6.9	  
368	72.171	20.9	5.1	18.6	5.3	  
365	72.15	24.2	6.5	18.6	5.3	  
359	72.186	34.1	8	18.6	4.3	  
334	72.505	34	8	18.6	6.9	  
330	72.05	32.7	8	18.6	6.9	  
289	72.795	29.2	7.2	18.6	11.7	    
141	71.913	32.1	8.7	18.6	8.8	  
2483	72.56	41.7	8.8	18.58	6.4	  
2052	72.604	27	7.2	18.55	16.2	  
2101	72.298	45	10.1	18.54	10.4	  
3529	72.444	32.6	8	18.5	4.4	  
2741	72.486	28.2	8.8	18.5	4.4	  
2388	72.648	29	7.2	18.5	5.4	  
1951	72.654	40.6	10.1	18.5	4.4	  
1948	72.542	29.3	7.9	18.5	4.4	  
1766	72.843	34	8.8	18.5	5.4	  
1748	72.116	45.5	11.7	18.5	5.4	  
1093	72.111	30.1	8	18.5	5.4	  
1039	71.857	35.1	8.7	18.5	7	  
975	72.344	29.6	7.2	18.5	10.9	  
937	72.33	47.1	12.2	18.5	4.4	  
916	72.463	31.7	7.9	18.5	4.4	  
909	71.969	29.5	7.9	18.5	8.9	  
883	72.348	30.5	8.6	18.5	4.4	  
834	71.801	52.9	13	18.5	11.8	  
799	72.567	20.1	5	18.5	4.4	  
736	71.758	28.7	6.5	18.5	4.4	  
693	71.573	42.3	9.4	18.5	4.4	  
527	71.755	24.2	6.5	18.5	5.4	  
362	72.289	34.3	8	18.5	8.3	    
360	72.078	30.3	9.4	18.5	4.4	  
276	71.897	38.2	9.4	18.5	5.4	  
91	72.253	31.7	-1.2	18.5	10.9	    
1785	72.09	48.7	9.5	18.48	53	  
103	72.083	36.4	9.4	18.46	11.9	  
1563	72.095	36.6	10.2	18.45	7.6	  
1726	72.038	30.5	8.8	18.41	12	  
2950	72.25	16.9	3.6	18.4	4.4	  
2681	72.081	34.4	7.3	18.4	4.4	  
2081	72.523	21.5	5.8	18.4	4.4	  
1956	72.71	34.5	7.9	18.4	4.4	  
1759	72.671	39	10.2	18.4	6.3	  
1419	72.205	23.9	5.1	18.4	6.3	  
1213	72.287	21.7	5.8	18.4	4.4	  
1057	72.025	32	7.2	18.4	8.9	  
938	72.098	40.6	9.3	18.4	10.9	  
554	72.013	32.4	7.2	18.4	11.4	  
228	72.472	55.1	12.3	18.4	6.3	    
77	71.965	41.6	-1.2	18.4	10	    
2116	72.695	43.5	10.1	18.39	10.1	  
2059	72.317	44.3	10.8	18.38	11.1	  
1537	72.208	23.3	5.8	18.33	11.1	  
3358	72.569	20.8	5.1	18.3	4.4	  
2990	72.628	34.1	8	18.3	6.3	  
2673	72.176	18.5	4.4	18.3	4.4	  
2456	72.201	47.3	10.1	18.3	6.3	  
2320	72.868	53.4	12.3	18.3	4.4	  
2235	72.078	48	10.8	18.3	5.5	  
1879	72.041	48	12.3	18.3	4.4	  
1453	71.779	27.1	7.2	18.3	5.5	  
1155	71.883	47.5	10.2	18.3	5.5	  
841	72.078	35.3	10.1	18.3	9	  
823	72.336	54.1	13.7	18.3	7.7	  
528	71.821	40.8	9.4	18.3	8.4	  
364	72.099	56.6	12.3	18.3	6.3	  
327	71.91	41.1	10.1	18.3	7.7	  
230	72.083	27.2	7.2	18.3	4.4	  
189	72.928	22.5	6.5	18.3	5.5	    
128	72.322	21.1	5.1	18.3	4.4	  
87	72.22	43.8	-4.1	18.3	10.5	    
78	71.952	39.4	-1.9	18.3	7.7	    
2102	72.457	38.5	9.4	18.29	6.4	  
51	71.952	38.3	-2.6	18.29	11.4	  
950	72.394	25.8	7.2	18.27	10.8	  
82	72.089	21.7	2.4	18.23	6.8	    
307	72.308	44.3	11.6	18.22	12.1	  
2923	72.284	34.6	8.8	18.2	5.4	  
2679	72.276	35.7	8.8	18.2	4.3	  
1924	72.585	30.3	8.7	18.2	12.6	  
917	72.565	33	7.9	18.2	5.4	  
912	72.319	48.6	12.2	18.2	7	  
791	72.194	31.1	5.8	18.2	4.3	  
313	72.115	28.7	6.5	18.2	6.3	  
300	71.994	28.5	8	18.2	6.3	  
220	71.803	43.1	10.8	18.2	4.3	  
173	72.316	41.4	10.1	18.2	6.3	  
145	72.264	35.3	8.7	18.2	4.3	  
127	72.324	26	5.8	18.2	4.3	  
124	72.285	41.7	9.4	18.2	4.3	  
2978	72.329	55.3	12.4	18.17	8	  
2107	72.478	38.1	9.4	18.12	11.1	  
4189	72.671	34.4	9.5	18.1	7.7	  
2907	72.24	28.6	7.3	18.1	4.2	  
2905	72.206	32.8	8.8	18.1	4.2	  
2690	72.432	48.2	10.9	18.1	4.2	  
1468	72.729	31.2	8	18.1	6.2	  
1424	72.42	20.4	5.1	18.1	6.2	  
1064	72.278	32.2	8.7	18.1	5.3	  
1024	72.005	47	11.6	18.1	6.2	  
976	72.295	31.1	7.2	18.1	7	  
942	72.116	51.2	12.2	18.1	5.3	  
844	72.29	18.6	4.3	18.1	8.9	  
821	72.251	30.1	7.9	18.1	4.2	  
796	72.42	32.1	7.2	18.1	10.5	  
715	72.49	27.9	6.5	18.1	5.3	  
309	72.601	33.4	7.2	18.1	4.2	  
268	71.86	31.6	8.7	18.1	4.2	  
143	72.38	36.4	8.7	18.1	4.2	  
125	72.31	30.3	7.2	18.1	4.2	  
750	71.684	62	13	18.09	5.4	  
1773	72.851	44.3	10.2	18.03	70.4	  
1888	72.238	35	8.7	18.02	9	  
1597	72.165	44.7	10.9	18.02	23	  
738	71.709	29.3	6.5	18.01	8	  
3528	72.47	25.2	5.8	18	3.9	  
2879	72.446	30.7	8	18	5.1	  
2530	72.055	26.5	6.6	18	3.9	  
2310	72.443	27	7.2	18	5.1	  
2300	72.251	59.6	13	18	3.9	  
2298	72.818	29.4	7.2	18	5.1	  
2204	72.667	24.6	6.5	18	5.1	  
1571	72.13	19.8	4.4	18	6.8	  
1520	72.007	37.3	8	18	5.1	  
1478	72.366	36.4	8.7	18	3.9	  
1426	72.419	31.7	7.2	18	6.8	  
1244	72.143	27.7	5.8	18	5.1	  
1021	72.002	36.2	8	18	11.3	  
879	72.245	27.3	5.7	18	5.1	    
872	71.935	39.3	9.3	18	7.6	  
787	71.881	41.8	10.8	18	6.8	  
773	71.781	34.2	9.4	18	5.1	  
718	72.371	24	5.8	18	3.9	  
384	72.12	18.7	5.1	18	5.1	  
357	71.982	41.8	10.1	18	6.8	  
314	72.121	35.6	8.7	18	3.9	  
266	71.791	49.9	11.6	18	5.1	  
89	72.208	45.9	-4.8	18	12.7	    
1226	72.026	38	10.9	17.99	8.3	  
147	71.848	56.7	13.7	17.99	8.9	  
1090	72.353	30.5	7.2	17.98	8.4	  
1130	72.233	44.7	9.4	17.96	8	  
2346	72.875	49.7	11.5	17.95	14.1	  
2121	72.37	42.1	9.4	17.9	7.4	  
2109	72.577	49.3	11.5	17.9	3.6	  
1892	72.33	31.2	5.8	17.9	3.6	  
941	72.17	35.8	9.3	17.9	4.8	  
896	72.437	25.1	5.7	17.9	7.4	  
889	72.47	42.1	9.3	17.9	6.7	  
865	72.031	55.9	14.4	17.9	3.6	  
795	72.344	39.8	10.8	17.9	4.8	  
790	72.248	48.6	11.6	17.9	3.6	  
789	72.05	46.1	12.3	17.9	5.8	  
654	71.9	23.5	5.1	17.9	4.8	  
366	72.046	39.9	8.7	17.9	8.1	  
348	71.864	35.5	8.7	17.9	3.6	  
216	72.309	43.5	11.6	17.9	3.6	  
43	72.07	32.7	-1.2	17.9	11.4	
878	72.163	34	7.9	17.89	6.8	  
2875	72.364	24.4	6.6	17.88	9.5	  
1724	72.148	23.9	5.1	17.86	8.4	  
99	72.075	25.8	7.2	17.86	15.9	  
306	72.419	40.4	9.4	17.84	12.2	  
2942	72.037	28.8	7.3	17.8	3	  
2904	72.214	40.6	11.7	17.8	5.5	  
2485	72.686	38.9	10.9	17.8	3	  
1898	72.423	51.3	13	17.8	7.2	  
1461	72.376	23.8	5.8	17.8	4.4	  
1428	72.388	30.7	8	17.8	8.4	  
1028	71.964	33.4	7.2	17.8	6.4	  
893	72.463	24.4	5.7	17.8	6.4	    
882	72.323	40.4	9.3	17.8	4.4	  
862	72.265	39.1	9.3	17.8	10.8	  
845	72.11	34.4	7.9	17.8	5.5	  
842	72.202	28.8	6.5	17.8	4.4	  
589	71.806	47.5	11.6	17.8	7.9	  
565	72.152	38.5	8.7	17.8	5.5	  
560	71.745	38.4	9.4	17.8	4.4	  
529	71.85	54.2	13	17.8	4.4	  
335	72.421	37.4	8.7	17.8	4.4	  
293	71.791	37.2	10.1	17.8	14.6	  
275	71.698	35.2	8.7	17.8	3	  
210	72.509	49.4	10.8	17.8	7.2	  
207	72.208	34.4	8	17.8	3	  
192	72.189	39.7	10.1	17.8	3	  
188	72.816	29.1	6.5	17.8	5.5	    
144	72.123	41.1	10.1	17.8	7.2	  
126	72.183	66.8	13.7	17.8	4.4	  
1002	72.26	50.5	11.6	17.78	6.6	  
940	72.028	54.4	13.7	17.78	7.4	  
1399	72.241	42.1	10.2	17.77	9.4	  
495	72.119	43.2	10.1	17.76	8.2	  
2870	72.265	33.8	9.5	17.7	4	  
2692	72.141	32.7	8	17.7	2.2	  
2569	72.493	27.7	7.3	17.7	4	  
2359	72.719	38.3	9.4	17.7	4	  
2284	72.545	37.5	8.7	17.7	4	  
1416	72.363	39.2	7.2	17.7	6.9	  
936	72.407	36.6	10.1	17.7	2.2	  
897	72.422	26.9	6.5	17.7	4	  
557	71.936	49.7	10.8	17.7	6.9	    
375	72.198	33.7	8.7	17.7	2.2	  
369	72.22	28.6	6.5	17.7	2.2	  
350	72.305	42.7	11.6	17.7	4	  
329	72.005	38.1	9.4	17.7	10.1	  
326	71.908	37	8.7	17.7	6.1	  
264	71.619	41.3	10.8	17.7	4	  
223	71.669	39.5	9.4	17.7	4	  
219	71.622	38.9	10.8	17.7	2.2	  
214	71.976	50	13	17.7	2.2	  
194	71.959	20.1	5.1	17.7	2.2	  
1466	72.636	40.7	10.2	17.69	4.1	  
743	71.755	33	6.5	17.69	9.5	  
1016	72.01	38.6	8.7	17.68	7.8	    
2488	72.631	23.4	6.6	17.65	8.7	  
2283	72.761	40.3	10.1	17.64	11.5	  
2112	72.577	39.7	9.4	17.64	13.2	  
301	71.988	32	8.7	17.63	9.6	  
296	71.89	35.3	8	17.62	7	  
157	72.184	29.2	6.5	17.62	7.7	    
2951	72.195	22.4	5.8	17.6	10.4	  
2678	72.127	26.4	5.8	17.6	0	  
2592	72.123	26.7	6.6	17.6	5.7	  
2309	72.841	50.8	12.3	17.6	0	  
2210	72.626	54.5	13	17.6	4.6	  
1795	72.226	28.4	6.6	17.6	19.5	  
1460	72.303	40.9	10.9	17.6	4.6	  
1218	72.225	43.8	9.4	17.6	5.7	  
924	72.712	37.8	9.3	17.6	0	  
922	72.708	30.5	7.2	17.6	0	    
826	72.535	30.5	7.9	17.6	5.7	  
824	71.671	36	8.6	17.6	3.3	  
713	72.591	44.7	10.1	17.6	0	  
702	72.463	53.5	12.3	17.6	4.6	  
342	71.973	46.1	12.3	17.6	0	  
325	71.794	38.6	10.1	17.6	0	  
206	72.36	37.2	8.7	17.6	8.1	  
191	72.301	32.4	7.2	17.6	6.6	    
190	72.619	25.6	6.5	17.6	4.6	  
171	72.237	32.9	7.2	17.6	8.7	  
971	72.412	58.2	13.8	17.59	13.3	  
2305	72.398	24.7	7.2	17.58	12	  
1209	72.26	27.9	6.5	17.58	6.8	  
906	72.146	15.7	4.3	17.58	12	  
48	72.042	38.3	-2.6	17.58	10
1788	72.138	26	7.3	17.57	16	  
1577	72.015	18.2	4.4	17.57	9	  
80	72.018	38.4	-1.9	17.57	10.5	    
403	72.741	23.9	6.5	17.56	8.4	  
2120	72.39	36.3	8.7	17.52	8.7	  
2680	72.273	34.3	8	17.5	2.2	  
2606	72.183	28.9	7.3	17.5	2.2	  
2596	72.293	47.1	10.2	17.5	4	  
1964	72.95	26.8	7.2	17.5	2.2	  
929	72.405	42.7	10.1	17.5	5.2	  
927	72.538	45.5	10.8	17.5	4	  
920	72.648	21.3	5	17.5	2.2	  
837	71.495	40.1	8.6	17.5	2.2	  
697	71.735	56.2	10.8	17.5	4	  
657	71.979	22.9	5.8	17.5	4	  
534	72.062	29.5	8	17.5	7	  
481	72.791	50.4	13.7	17.5	4	  
462	72.024	46.7	10.8	17.5	7	  
332	72.287	35.1	8	17.5	4	  
317	72.245	35	8	17.5	6.2	  
316	72.458	37.2	10.1	17.5	7.8	  
312	71.826	21.8	4.3	17.5	6.2	  
205	72.307	42.6	10.1	17.5	2.2	  
186	72.444	27	6.5	17.5	4	  
112	71.723	38.4	8	17.5	7.8	  
1833	72.419	18	4.3	17.49	7.8	  
860	72.29	35.3	7.2	17.49	8.5	  
58	72.691	33.8	-1.2	17.49	16	    
2285	72.488	40	8.7	17.48	9.8	  
2066	72.243	31.3	7.2	17.48	12.4	  
105	71.99	39	8.7	17.47	12.7	  
943	71.934	39.8	8.6	17.46	8.1	  
270	72.164	32.8	7.2	17.45	22.7	  
1446	72.075	47.7	13.8	17.44	7.6	  
933	71.919	33.7	7.9	17.44	8.2	  
57	72.136	43.6	-1.9	17.42	11.2	    
2695	72.098	32.4	7.3	17.4	3.1	  
2124	72.887	33.1	7.9	17.4	4.5	  
1931	72.674	43	10.1	17.4	4.5	  
1222	72.13	33.8	8.7	17.4	6.6	  
1142	71.996	28.2	7.2	17.4	5.6	    
945	72.115	37.5	9.3	17.4	5.6	  
931	72.071	45.6	11.5	17.4	9.4	  
911	72.296	50.3	9.3	17.4	6.6	  
704	72.523	36.4	9.4	17.4	4.5	    
698	72.436	62.5	13.7	17.4	5.6	  
311	71.606	42	10.8	17.4	9.4	  
247	72.011	29.7	7.2	17.4	5.6	  
215	72.221	40.5	9.4	17.4	4.5	  
212	71.891	36.3	8.7	17.4	3.1	  
182	72.091	42.4	11.6	17.4	3.1	  
2521	72.659	36	9.5	17.39	3.3	  
1521	72.132	36.2	7.3	17.38	9.9	  
489	71.921	28	6.5	17.38	10.1	  
349	72.576	29.6	6.5	17.38	5.8	  
187	72.693	28	6.5	17.38	3.4	  
324	71.704	31.4	7.2	17.37	6.8	  
100	72.141	42.7	13	17.37	13.8	  
470	72.379	41	10.1	17.36	9	  
1208	72.226	24.9	5.8	17.35	9.7	  
1418	72.173	35.9	8	17.34	6.2	  
2274	72.456	36.3	8.7	17.31	14.1	  
1826	72.12	30.1	7.9	17.31	11.9	  
62	72.587	31.7	-0.5	17.31	10.6	    
2108	72.255	37.2	10.1	17.3	5	  
892	72.476	34	8.6	17.3	6.9	  
873	72.075	39.3	10.1	17.3	8.4	    
832	72.004	38.4	8.6	17.3	6	  
538	72.115	32.2	8	17.3	6	  
218	71.893	45.2	10.1	17.3	6	  
202	72.083	32.6	8.7	17.3	3.7	  
198	71.836	37.1	8.7	17.3	3.7	  
185	72.407	42.6	9.4	17.3	5	  
1801	72.441	38.7	8.8	17.29	21.9	  
2105	72.546	27.2	6.5	17.28	10.1	  
1562	71.905	26.8	6.6	17.28	10.2	  
1101	72.384	35.2	8	17.28	6.2	  
493	72.094	28	6.5	17.28	7	  
328	72.124	31.4	7.2	17.28	6.2	  
54	72.555	34	-1.2	17.28	12.3	    
389	71.87	36.9	8.7	17.27	7.1	  
39	72.063	40.9	-1.9	17.27	21.5	    
798	72.154	40.6	10.8	17.26	8	  
487	72.059	59	13.7	17.26	7.2	  
66	72.394	28	0.3	17.26	8.5	    
1349	72.393	38.2	9.4	17.25	74.8	  
2233	72.645	36.6	9.4	17.24	5.6	  
2928	72.394	31.4	8.8	17.21	8.4	  
2458	72.455	39.1	10.1	17.2	7.9	  
2307	72.45	44.4	9.4	17.2	4.1	  
2306	72.416	41.4	10.1	17.2	5.3	  
1623	72.262	21.4	5.1	17.2	4.1	  
1369	72.233	40.1	10.2	17.2	5.3	  
918	72.424	49.7	10.1	17.2	7.9	  
849	72.12	48.6	10.1	17.2	4.1	  
709	72.636	38.7	9.4	17.2	4.1	  
707	72.585	53.7	10.8	17.2	5.3	    
553	72.042	22.3	6.5	17.2	9.3	  
469	71.955	26	7.2	17.2	5.3	  
461	72.027	59.4	13.7	17.2	5.3	  
340	71.859	26.4	5.8	17.2	4.1	  
208	72.31	43.7	12.3	17.2	5.3	  
204	72.188	29.9	8	17.2	4.1	  
193	72.008	32	8	17.2	4.1	  
1229	72.149	30.2	7.2	17.19	6.4	  
829	72.735	57.2	12.2	17.19	5.4	  
262	71.912	29.7	7.2	17.19	8	  
158	72.183	52.3	11.6	17.19	5.4	  
524	71.948	21	5.1	17.18	5.5	  
274	71.982	38.9	9.4	17.18	7.3	  
159	72.3	48.6	13	17.18	6.5	  
905	72.255	34.4	7.9	17.17	4.4	  
712	72.505	13.1	2.9	17.17	4.4	  
2269	72.726	45.8	10.1	17.16	5.7	  
47	72.078	54.6	-4.8	17.16	12.5	  
2691	72.121	31.1	7.3	17.13	6.8	  
2389	72.41	32.9	7.2	17.12	5	  
831	72.161	43.6	9.3	17.12	10.3	  
723	72.115	28.7	7.2	17.11	12.4	  
2912	72.311	35.5	10.2	17.1	5.6	  
2571	72.558	40	8.8	17.1	4.4	  
2481	72.491	38.4	9.5	17.1	4.4	  
2113	72.767	40.9	10.8	17.1	6.5	  
1673	72.205	44.8	10.2	17.1	67	  
966	72.148	19.6	4.3	17.1	6.5	  
852	72.049	22.6	5.7	17.1	5.6	  
620	72.083	35.1	8.7	17.1	5.6	  
531	72.006	58.4	13.7	17.1	5.6	  
473	71.915	17.1	4.3	17.1	4.4	  
341	72.073	45.7	10.8	17.1	4.4	  
181	72.231	35.4	9.4	17.1	5.6	    
162	72.307	35.1	8.7	17.1	6.5	  
119	71.736	41.2	10.8	17.1	4.4	    
930	72.231	38.5	9.3	17.09	5.7	  
1242	72.217	36.6	9.4	17.08	5.7	  
652	71.616	30.2	8.7	17.08	21.7	  
497	72.19	48.7	13	17.07	8.3	  
465	72.146	35.9	8.7	17.06	7.6	  
1275	72.165	52.7	12.3	17.05	5	  
1462	72.498	29.6	6.5	17.03	8.5	  
1195	72.198	39.8	8	17.02	6.3	  
2897	72.024	34.6	9.5	17.01	5.3	  
1449	72.021	17	5.1	17.01	6.4	  
1134	72.143	26	7.2	17.01	5.4	  
2572	72.463	38.1	9.5	17	4.7	  
2125	72.873	43.1	10.8	17	4.7	  
1095	72.112	30.2	8	17	13.5	  
825	72.455	21.2	5	17	4.7	  
543	72.149	17.6	5.1	17	5.8	  
448	72.153	48.1	11.6	17	10	  
211	72.593	29.1	5.8	17	4.7	  
151	71.913	29.8	8	17	7.3	  
142	71.929	22	5.1	17	4.7	  
2524	72.548	32.4	8	16.99	12.2	  
1254	72.192	38.2	8.7	16.99	4.7	  
856	72.05	23.5	5.7	16.99	4.7	  
272	72.404	44.1	10.8	16.99	10.6	  
172	72.323	40.6	9.4	16.99	4.7	    
160	72.474	35	10.1	16.99	4.7	  
2299	72.891	51.1	12.3	16.98	8.2	  
2058	72.337	46.8	10.8	16.98	5.9	  
242	72.006	25.2	5.1	16.97	4.9	  
932	72.016	33.4	7.9	16.96	7.8	  
496	72.235	48.5	12.3	16.96	7	  
490	71.903	22.7	6.5	16.96	7	  
422	72.189	37.3	9.4	16.96	6.1	  
390	71.958	67.4	14.5	16.96	7.8	  
1012	72.086	49.5	10.9	16.95	5.1	  
901	72.408	21.2	5.7	16.95	5.1	  
474	71.911	13.7	3.6	16.95	6	    
109	71.959	24	5.8	16.95	13.6	  
1132	72.147	25.4	5.8	16.93	10.9	  
22	72.38	39.6	-2.6	16.93	11.1
111	71.73	64.8	14.5	16.92	8.5	  
1966	72.848	35.9	9.4	16.9	5.9	  
954	72.596	23.6	5.8	16.9	5.9	  
811	72.2	42.2	11.5	16.9	5.5	  
720	72.245	28.7	6.5	16.9	6.8	  
424	72.1	40	9.4	16.9	7.7	  
336	72.385	30.1	7.2	16.9	4.8	  
217	72.671	35.8	8.7	16.9	4.8	  
2185	72.429	24.8	5.8	16.89	6	  
180	72.001	41.8	10.8	16.89	4.9	  
1915	72.47	24.3	5	16.88	5	  
498	72.288	25.5	6.5	16.88	8.5	  
148	71.839	34.3	9.4	16.88	7.8	  
1758	72.554	30.1	8.8	16.87	5	  
486	72.011	29.1	8	16.87	6.1	  
522	72.178	35.9	8	16.86	6.2	  
139	71.741	51.3	13	16.86	6.8	  
2313	72.782	42.3	9.4	16.85	5.2	  
346	72.616	38.7	8	16.85	7.2	    
1527	72.203	32.8	8	16.84	12.1	  
688	71.821	52.6	11.6	16.84	7.2	  
485	72.115	36.6	8.7	16.84	7.2	  
256	71.903	47.4	10.1	16.84	7	  
92	72.914	45.1	-2.6	16.84	15.9	    
52	72.388	40.6	-1.9	16.84	11.5	    
701	72.481	15.8	4.3	16.83	10.1	  
315	72.715	22.2	5.8	16.83	8.8	    
1733	72.22	48.5	10.2	16.82	8.6	  
1160	72.068	34.2	8.7	16.81	8.1	  
2211	72.521	55.8	12.3	16.8	4.9	  
1822	72.013	43	10.8	16.8	8.9	  
915	72.241	39.1	8.6	16.8	4.9	  
352	71.591	30.9	7.2	16.8	4.9	  
140	71.933	43.3	11.6	16.8	6	  
2602	72.153	27.2	5.1	16.79	4.9	  
200	71.939	25.1	5.8	16.79	5.8	  
155	72.075	37.5	10.1	16.79	4.9	  
49	72.041	40.6	-2.6	16.79	4.9	    
792	71.718	31.8	8	16.78	6.1	  
523	72.049	45.3	10.1	16.78	5	  
337	71.972	46.6	11.6	16.78	5.8	  
1451	72.086	28.7	8	16.77	8.2	  
1420	72.026	29.3	7.2	16.77	6.8	  
710	72.523	18.2	5.1	16.77	8.4	  
321	71.687	40	9.4	16.77	5.1	    
294	71.691	49.2	11.6	16.76	6.1	  
221	71.492	30.6	7.2	16.76	5.2	  
1438	72.096	23.3	5.8	16.75	5.2	  
388	71.868	56.8	12.3	16.74	7.2	  
356	71.658	26	6.5	16.73	5.4	  
64	72.471	48.4	-4.1	16.73	10.1	    
45	71.707	43.5	-2.6	16.73	11.5	  
1893	72.312	44.1	8.7	16.71	5.5	  
1745	72.713	24.1	5.8	16.71	10.5	  
1223	72.053	42.6	10.2	16.71	6.6	  
655	71.951	50.4	10.8	16.71	5.5	  
2122	72.985	20.6	5.8	16.7	4.9	  
923	72.053	45.4	12.2	16.7	6	  
699	72.448	20.2	5.8	16.7	6.9	  
550	72.073	30	7.2	16.7	6	  
53	72.061	58.6	-5.5	16.7	7.6	    
926	71.994	38.2	10.1	16.69	4.9	  
913	71.587	47.2	11.5	16.69	4.9	  
788	71.655	32.6	8.7	16.69	4.9	  
163	72.551	32.2	7.2	16.69	8.5	  
68	72.301	47.9	-4.1	16.69	10.2	    
1408	72.698	41.2	10.2	16.68	7.8	  
1230	72.161	28.5	8.7	16.68	12.4	    
696	72.446	14.4	5.1	16.68	5	  
525	71.965	51.8	11.6	16.68	7	  
440	72.121	28.4	6.5	16.67	6.1	  
184	72.355	32.2	8	16.67	5.1	  
65	72.045	21.9	2.4	16.66	10.3	    
453	71.905	24.4	5.1	16.65	6.3	  
79	71.868	40.6	-2.6	16.64	10.3	    
1749	72.199	67.3	13.9	16.63	7	  
2597	72.087	45.5	10.2	16.62	8.1	  
419	72.01	31.9	7.2	16.62	7.9	  
685	71.845	21.5	5.1	16.61	7.2	  
689	72.113	60.3	14.5	16.6	4.8	  
195	71.984	26.5	7.2	16.6	4.8	  
1365	72.124	44.5	10.9	16.59	5.6	  
632	72.571	51.5	10.8	16.59	6	  
624	72.117	30.2	8	16.59	6	  
494	72.114	56.2	13	16.59	4.8	  
161	72.357	35.8	8	16.59	4.8	  
102	72.176	36	8.7	16.59	10.3	  
692	72.345	29.8	6.5	16.58	6	    
558	71.835	28	6.5	16.58	8.3	  
443	72.119	51.4	11.6	16.58	4.9	  
304	72.191	23.9	6.5	16.58	6.4	  
666	72.625	42.4	10.1	16.57	13.3	  
260	71.802	43.2	10.1	16.57	4.9	  
245	71.893	41.5	10.8	16.57	4.9	  
69	72.105	46.7	-4.1	16.57	10.3	    
2110	72.469	24.6	6.5	16.56	6.1	  
1751	72.358	22.2	5.1	16.56	13.4	  
1070	72.076	23.8	5.1	16.56	6	  
695	72.236	28.2	7.2	16.56	5	  
243	71.728	28.7	7.2	16.56	6.2	  
56	72.639	51.3	-4.1	16.56	7.7	    
458	71.761	48.4	13	16.55	7	  
457	71.863	39.9	10.1	16.55	6.2	  
2194	72.462	41.3	10.8	16.54	5.1	  
1787	71.965	29.6	7.3	16.54	19.7	  
466	72.208	32.4	8	16.54	8	  
273	72.282	27.9	7.2	16.54	6.8	  
1000	72.39	41.6	10.2	16.53	9.1	  
94	72.855	53.2	-5.5	16.53	9.4	    
1496	72.005	24.7	5.8	16.52	7.8	  
299	72.113	47.7	12.3	16.52	11.6	  
2908	72.222	36.7	8	16.51	5.4	  
592	71.813	29.6	7.2	16.51	5.3	  
1406	72.76	26.1	7.2	16.5	4.6	  
1287	71.92	32.7	8	16.5	7.9	  
919	72.173	35.5	8.6	16.5	4.6	  
908	72.131	29.9	7.9	16.5	6.8	  
600	72.031	29.2	8	16.5	7.2	  
271	72.466	21.8	5.8	16.5	6.4	    
257	71.958	47	10.8	16.5	8.9	  
1422	72.413	28.4	8.7	16.49	6.3	  
1395	72.417	33.7	7.2	16.49	4.6	  
1333	72.041	40.6	10.2	16.49	6.5	  
863	72.025	28.5	6.5	16.49	4.6	  
716	72.43	43.2	9.4	16.49	9.5	  
60	72.638	47.7	-3.3	16.49	6.5	    
1191	72.379	26.1	5.8	16.48	4.7	  
953	72.511	31.3	8	16.48	4.7	  
416	72.33	51.7	12.3	16.48	11.3	  
249	72.495	28.7	5.8	16.48	5.9	  
248	72.37	29.7	8	16.48	7.4	  
213	71.783	38.2	10.1	16.48	4.7	  
1528	72.238	19	5.8	16.47	22.5	  
2200	72.644	46.6	10.1	16.46	4.9	  
1259	72.339	15.2	4.3	16.46	5.9	  
843	71.616	34.8	7.9	16.46	4.8	  
694	72.406	27.5	6.5	16.46	4.8	  
510	72.413	19.5	5.1	16.46	6.4	  
38	71.988	30.3	0.3	16.46	16.3	    
2490	72.208	27.5	6.6	16.45	7.6	  
968	72.086	32	7.2	16.45	9.5	  
588	71.861	28.5	6.5	16.45	4.9	  
451	72.027	29.5	6.5	16.45	7.6	  
295	71.59	43.2	10.8	16.45	9.5	  
61	72.075	45.8	-4.8	16.45	7.4	    
2314	72.743	31.1	7.2	16.44	4.9	  
2275	72.475	13.6	3.6	16.44	11.4	  
430	72.235	54.5	11.6	16.44	5.7	  
333	72.305	37	8.7	16.44	4.9	  
251	72.292	25.1	7.2	16.44	5	  
2601	72.227	45	10.9	16.43	5	  
618	72.086	42.8	9.4	16.43	5	    
96	72.74	51.8	-4.8	16.43	8.8	    
724	72.019	23.5	5.8	16.42	6.2	  
1455	72.223	33.6	8	16.41	5.2	  
2570	72.483	17	4.4	16.4	4.3	  
1421	72.481	36.9	8.7	16.4	7.3	  
1270	71.913	41.2	9.4	16.4	6.9	  
859	72.089	37.1	8.6	16.4	4.3	  
598	72.305	47.7	9.4	16.4	7.9	  
178	72.136	50.3	12.3	16.4	4.3	  
154	72.102	58.4	12.3	16.4	5.6	  
121	71.772	52.7	12.3	16.4	4.3	  
1170	72.158	34.2	8	16.39	7.2	  
1139	71.851	25.3	5.8	16.39	5.3	  
827	71.648	38.3	9.3	16.39	4.4	  
323	71.79	34.3	8.7	16.39	8.2	  
690	72.178	48.1	13	16.38	5.3	  
1371	72.199	42.9	10.2	16.37	8.4	  
2286	72.683	23.8	5.8	16.36	9	  
705	72.513	13	4.3	16.36	5.6	  
343	72.524	22.5	5.8	16.36	12.5	  
46	71.811	36.1	-0.5	16.36	8.3	    
2061	72.538	38.2	10.8	16.35	4.6	  
505	72.329	53.2	12.3	16.35	5.4	  
170	72.065	41.4	10.1	16.35	5.8	  
1626	72.353	19.5	4.4	16.34	4.7	  
1414	72.535	40.9	8.7	16.34	4.7	  
1062	72.444	17.8	5.1	16.34	5.8	  
261	71.589	18.7	5.1	16.34	4.6	  
2128	72.362	36	7.9	16.33	7.4	  
1540	72.036	30.3	8	16.31	10.1	  
1410	72.72	41.8	10.2	16.3	3.9	  
663	72.108	29	8	16.3	4.8	  
477	71.934	15	3.6	16.3	4.9	  
164	72.196	52.8	11.6	16.3	3.9	  
866	72.051	38.5	10.1	16.29	4.9	  
2484	72.425	32.4	7.3	16.28	5	  
2213	72.332	33.9	8.7	16.28	4	  
1394	72.31	44.6	10.2	16.28	4	  
864	72.163	34.6	9.3	16.27	4.1	  
423	72.105	31.6	8	16.27	6.4	  
196	71.628	20	5.1	16.27	12	  
627	72.465	46.7	10.8	16.26	4.1	  
393	71.932	48.5	10.8	16.26	5	  
240	72.985	23.6	5.8	16.26	4.1	  
175	72.28	39.1	8.7	16.25	4.1	  
72	72.093	57.9	-6.2	16.25	7.7	    
5	72.229	27.9	1	16.24	9.3	  
2874	72.37	24.4	5.8	16.23	4.2	  
1363	72.094	17.5	4.3	16.23	6	  
156	71.989	28.6	5.8	16.23	5.2	  
1500	72.008	29.8	7.3	16.22	5.4	  
661	72.048	25.2	5.8	16.22	4.3	  
153	72.04	29.8	8	16.22	5.6	  
660	72.011	31.1	6.5	16.21	4.3	  
374	71.648	48.1	11.6	16.21	7.1	    
18	72.336	38.9	-1.2	16.21	10.4	  
2	72.12	32.4	-1.2	16.21	6.5	  
1282	71.895	30.1	7.2	16.2	6.6	  
951	72.401	32.6	8.7	16.2	4.4	  
420	72.038	36.4	10.8	16.2	5.3	  
201	71.841	27.1	6.5	16.2	5.4	  
1800	72.318	27.3	6.6	16.19	10.5	  
1780	71.93	29.2	7.3	16.19	10.9	  
621	72.304	49.6	12.3	16.19	4.4	  
305	72.509	23.5	6.5	16.19	5.7	  
267	71.971	30.1	8	16.19	4.4	  
259	72.061	30.1	6.5	16.19	3.4	  
244	71.811	46.5	10.1	16.19	4.5	  
2308	72.891	43.3	9.4	16.18	6.9	  
970	72.576	32.3	8	16.17	12.8	  
519	72.27	45.2	11.6	16.17	4.4	    
446	72.101	47.5	10.1	16.17	3.4	  
717	72.424	26.3	6.5	16.16	5.9	  
169	71.91	38.4	9.4	16.16	5.9	  
73	71.993	33.6	-1.2	16.16	6.5	    
1566	72.392	38.9	10.2	16.15	8.8	  
616	71.962	44.3	10.8	16.15	6.8	  
475	72.361	44.4	10.1	16.15	3.6	  
263	71.68	38.3	8	16.15	3.6	  
605	72.11	30.8	7.2	16.14	5.9	  
168	72.011	51.5	12.3	16.14	4.9	  
106	72.049	32.3	7.2	16.14	7.8	  
1772	72.858	65	13.9	16.13	56.4	  
1737	72.364	37.5	8.8	16.13	9.6	  
876	72.214	37.4	8.6	16.13	6	  
658	72.117	46.8	11.6	16.13	4.7	  
70	72.071	44.4	-4.8	16.13	4.8	    
6	72.322	43.5	-2.6	16.13	9.4	  
1257	72.149	56	10.9	16.12	5.1	  
703	71.673	28.4	7.2	16.12	6.2	  
648	71.905	39.6	9.4	16.12	6.9	  
593	72.005	41.7	10.8	16.12	4.8	  
629	72.445	32.2	8.7	16.11	4.8	  
250	72.01	41.6	10.1	16.11	3.7	  
108	71.871	49.9	13	16.11	10.5	  
1164	71.901	44.2	10.9	16.1	4.8	  
672	72.144	40.8	10.1	16.09	3.9	  
630	72.545	41.1	10.1	16.09	2.5	  
682	71.936	31.7	8	16.08	7.2	  
1192	72.316	30.9	7.2	16.07	21.2	  
1096	72.12	30.6	8	16.07	7.3	  
935	71.815	30.5	7.9	16.07	8.8	  
609	72.146	38.2	8.7	16.07	3.9	  
395	72.128	58.6	13.7	16.07	5	  
26	72.423	43.3	-3.3	16.07	11.4	  
1513	72.399	41.3	8.8	16.06	10.3	  
1251	72.218	29.9	8	16.06	6	  
1165	72.009	38.3	10.2	16.05	6.2	  
722	72.213	19.2	3.6	16.05	3.9	  
2900	72.119	39	8.8	16.04	4.1	  
483	72.221	26.1	7.2	16.04	2.6	  
344	72.487	51	13	16.03	4	  
237	71.976	41.9	9.4	16.03	6.2	  
3	72.105	62.8	-6.2	16.03	8.8	
673	72.546	34.2	8	16.02	8.8	  
619	72.206	43.6	11.6	16.02	4.1	  
590	71.79	14.6	3.6	16.02	5.2	  
427	72.26	37.7	8.7	16.02	2.8	 
14	72.256	51.3	-4.8	16.02	9.9
1	72.146	31.3	-0.5	16.02	4.1
1732	72.236	33.6	7.3	16.01	14.6	  
802	71.517	46.8	10.8	16.01	4.1	  
610	72.153	27.7	6.5	16.01	2.8	  
331	72.225	22.7	5.1	16.01	2.8	  
1457	72.293	32.6	8.7	16	4.1	  
719	71.604	35.3	8.7	16	2.8	  
667	72.126	38.1	8.7	16	2.8	  
515	72.382	42	10.8	16	4.1	  
353	71.586	37.7	10.1	16	4.1	  
258	72.187	31.5	8	16	5.5	  
1411	72.711	34.9	8	15.99	4.2	  
1359	72.11	32.2	8	15.99	13.9	  
447	72.103	28.9	7.2	15.99	2.9	  
2287	72.381	36.6	10.1	15.98	4.2	  
1744	72.508	21.8	5.1	15.98	9.4	  
816	71.613	14	3.6	15.98	5.7	    
683	71.92	20.3	5.1	15.98	4.3	  
656	72.011	58.6	12.3	15.98	2.9	  
2595	72.08	46.5	11.7	15.97	0.5	  
886	71.825	59.8	12.2	15.97	7.9	  
706	72.513	31.1	7.2	15.97	3	  
431	72.305	53.2	12.3	15.97	4.4	  
320	71.736	28.4	6.5	15.97	8.3	  
318	71.52	39.7	8.7	15.97	2.9	    
1375	72.076	40.1	9.4	15.96	6.8	  
405	72.458	31.4	6.5	15.96	5.6	  
199	71.886	27.7	6.5	15.96	3	  
27	72.363	19.9	3.2	15.96	13.4	    
847	71.591	21.3	5	15.95	3.2	    
625	72.426	31.5	8.7	15.95	3	    
664	72.149	20.1	5.1	15.94	3.1	  
417	72.111	44.3	10.1	15.94	4.4	  
2491	72.146	26	5.8	15.93	3.2	  
1404	72.781	17.4	5.1	15.93	3.3	  
1345	71.999	33.3	7.2	15.93	4.6	  
1158	72.202	37.6	9.4	15.93	6.1	  
19	72.605	32.8	-0.5	15.93	12.1	  
967	72.075	26	7.2	15.91	4.9	  
255	71.767	41.3	10.1	15.91	5.9	  
1066	71.981	18.9	3.6	15.9	1	  
721	72.323	33.4	8.7	15.9	3.2	  
902	72.348	25.7	5.7	15.89	5	  
885	71.808	46.1	11.5	15.89	8.2	  
2492	72.181	29.4	8	15.88	1.2	  
2317	72.46	47.2	10.8	15.88	8.5	  
1756	72.417	32.1	7.3	15.88	13.5	  
1471	72.131	19.6	5.1	15.88	3.5	  
1194	72.211	23.3	5.8	15.88	5	  
944	71.903	24.4	6.5	15.88	4.8	  
596	72.11	42.1	9.4	15.88	9.3	  
982	72.464	35.7	7.2	15.87	11.8	  
674	72.194	35.3	9.4	15.87	1.3	  
607	72.089	15.6	3.6	15.87	3.3	  
668	72.685	33.4	8	15.86	4.7	  
152	72.008	34.2	8.7	15.86	1.3	  
1390	72.271	19.6	5.1	15.85	3.4	  
869	71.61	61.9	13.7	15.84	9.5	  
659	71.596	32.6	8	15.84	1.8	  
628	72.241	37.6	9.4	15.84	1.5	  
252	71.646	36.2	9.4	15.84	1.2	  
241	71.788	45.2	10.1	15.84	1.4	  
149	71.887	33.7	9.4	15.84	3.5	  
471	72.38	49.1	11.6	15.83	1.4	  
1538	72.2	25.6	5.8	15.82	3.7	  
1203	72.138	38.9	9.4	15.82	3.9	  
714	71.626	23.3	5.8	15.82	1.5	  
613	71.936	51.7	10.8	15.82	3.3	  
604	72.991	48.7	10.1	15.82	4.7	  
854	71.726	41.4	9.3	15.81	3.8	  
1502	72.1	40.2	9.5	15.8	7.5	  
921	71.703	41.2	8.6	15.8	1	  
830	71.323	33.6	7.2	15.8	3.7	  
491	71.847	61.5	13	15.8	8.4	  
2236	72.033	18.1	5	15.79	1.6	  
904	72.2	54.4	13	15.79	5	  
662	71.481	35.3	8.7	15.79	3.5	    
232	72.872	30.6	7.2	15.79	3.4	  
602	72.053	39.3	10.1	15.78	3.7	  
351	71.522	29.3	7.2	15.78	6.8	  
174	72.398	43.1	10.8	15.78	3.7	  
1413	72.636	40.6	10.9	15.77	1.5	  
1403	72.751	20.7	5.1	15.77	3.8	  
623	72.365	41.4	10.1	15.77	1.6	  
472	72.41	20.1	5.1	15.77	1.6	  
2238	72.655	37.4	10.1	15.76	3.4	  
1743	72.643	47	12.4	15.76	14.8	  
1235	72.288	39.7	9.4	15.76	7.8	  
1127	72.268	33.4	7.2	15.76	4.5	  
1091	71.986	34	7.2	15.76	5.4	  
617	71.78	26.4	5.1	15.76	3.8	  
355	71.749	26.1	6.5	15.76	4.5	  
1499	71.973	26.2	6.6	15.75	6.5	  
1032	72.125	39.8	8.7	15.75	5.4	  
833	71.433	54.1	12.2	15.75	2.4	  
700	72.199	46.1	10.1	15.75	5.2	  
1791	72.116	44.5	10.2	15.74	18.4	  
708	71.652	40.7	8.7	15.74	1.6	  
1400	72.673	39.1	10.2	15.73	3.9	  
684	71.858	17.5	4.3	15.73	6.3	  
880	72.483	31.5	7.9	15.72	4.1	  
615	71.888	12.2	2.9	15.72	6.6	  
254	71.541	41.4	10.1	15.72	5.2	  
675	72.158	42.9	9.4	15.71	1.7	  
1456	72.25	29.2	5.8	15.7	5.7	  
30	72.267	25.1	1.7	15.7	10.9	    
2237	72.676	39.9	10.1	15.69	4	  
1799	72.314	27.3	8	15.68	17.3	  
595	71.757	24.6	5.1	15.68	2.3	  
76	71.916	39.4	-3.3	15.68	12.4	    
35	72.009	33.9	-1.9	15.68	10.8	    
1121	72.32	24.4	6.5	15.67	9.7	  
877	72.35	46.2	11.5	15.67	4.3	  
2522	71.94	41.2	10.9	15.66	2.1	  
671	72.681	42.1	10.8	15.66	3.8	  
345	72.373	36.4	9.4	15.66	2.3	  
322	71.775	39.1	10.1	15.66	6.4	  
1781	71.918	38.8	8.8	15.65	9.5	  
1113	72.48	32.8	8.7	15.65	3.8	  
1112	72.521	41.6	9.4	15.65	4.3	  
884	72.658	47.3	12.2	15.65	1.9	  
597	72.093	53.5	12.3	15.65	1.9	  
84	71.706	46.3	-3.3	15.65	9	    
2097	72.278	42.2	10.1	15.64	4.5	  
1479	72.449	26.7	5.8	15.63	9.8	  
409	72.425	14.1	3.6	15.63	1.8	  
1291	72.203	27.2	7.2	15.62	2.9	  
958	72.148	43	10.9	15.62	2.8	  
88	71.595	45.4	-4.1	15.62	9.8	    
1385	72.404	22.2	5.1	15.61	4	  
1099	72.229	34.5	8	15.61	4.8	  
993	72.736	41.7	10.2	15.61	6.2	  
990	72.791	36.3	10.2	15.61	9	  
813	72.065	39.3	8.6	15.61	3.5	  
594	71.968	50.7	12.3	15.61	3.3	  
591	71.834	48.5	12.3	15.61	4.1	    
2098	72.178	42.7	11.5	15.6	2	  
1104	72.471	38.7	10.9	15.6	2.9	  
319	71.613	24.7	5.8	15.6	6.7	  
1102	72.367	46.4	10.2	15.59	6.7	  
949	72.378	48.8	11.6	15.59	4.7	  
665	71.725	41.4	9.4	15.59	3.3	  
653	71.643	28	8	15.59	7.6	  
253	71.656	38.7	8.7	15.59	4.3	  
2242	72.47	29	6.5	15.58	2.6	  
1089	72.471	37.8	10.2	15.58	10	  
1476	72.213	17.8	4.3	15.57	16.1	  
807	71.48	50.5	10.8	15.57	5	    
404	72.401	22.5	5.1	15.56	2	  
1739	72.444	47.8	10.2	15.54	6.2	  
1131	72.175	19.9	5.8	15.54	5	  
679	71.966	32.2	8.7	15.54	4.3	  
599	71.701	43.8	10.1	15.53	1.7	  
1647	72.001	43.3	9.5	15.52	7.9	  
891	72.868	50.2	11.5	15.5	2.6	  
678	72.055	40.1	9.4	15.49	2.3	    
235	71.888	36.3	8.7	15.49	6.1	  
1779	72.061	25.3	6.6	15.48	20.3	  
867	72.033	37.3	10.8	15.48	2.4	  
1117	72.418	25	6.5	15.47	5.5	  
2303	72.971	43.3	8.7	15.45	5.2	  
1393	72.303	42.3	9.4	15.45	5.1	  
999	72.495	33.4	9.4	15.45	5.5	  
488	71.765	61	13.7	15.45	7.7	  
1746	72.76	35.4	8	15.44	3.5	  
1401	72.363	39.1	9.4	15.44	2.8	  
1398	72.342	21.4	5.8	15.44	4	  
1396	72.307	33.5	9.4	15.42	3.7	  
601	71.678	23.3	4.3	15.42	2.7	  
236	71.869	36.2	8	15.42	2.8	  
55	71.757	39.5	-1.2	15.42	5.1	    
1734	72.329	38.2	8	15.4	15.4	  
1098	72.131	25.8	5.8	15.4	8.5	  
2234	72.678	38.4	10.1	15.39	3.6	  
1572	72.103	43.9	10.9	15.39	17	  
1353	72.113	30.5	6.5	15.39	2.9	  
414	72.318	70.3	14.5	15.39	2.1	  
1706	72.14	38.6	9.5	15.38	6	  
1384	72.439	28.8	8	15.38	2.2	  
669	71.77	31.8	5.8	15.38	3.6	  
338	71.6	40.8	10.1	15.38	3.6	    
868	71.668	30.5	7.2	15.37	10.8	  
1736	72.423	49.4	10.9	15.36	6.2	  
9	72.423	63.6	-6.2	15.36	13.6	  
1576	72.11	31.9	7.3	15.35	11.2	  
614	71.91	45.1	10.8	15.35	8.4	  
59	71.665	34.8	-1.2	15.35	5.3	    
1147	72.639	43.2	10.9	15.34	4.9	  
2240	72.583	30.4	7.2	15.33	5.1	  
1741	72.532	41.3	10.2	15.33	6.6	  
1293	72.086	28.9	7.2	15.32	8.8	  
1115	72.271	22.4	5.8	15.32	4.8	  
626	71.692	33.1	8	15.32	3.8	  
1063	72.211	28.5	6.5	15.29	5.4	  
956	72.07	35.3	8.7	15.29	5.6	  
233	71.68	29	6.5	15.29	1.2	  
28	72.402	24.3	1.7	15.29	10.2	    
1368	72.448	23.6	6.5	15.28	9.6	  
1519	72.031	30	6.6	15.27	8.6	  
1292	72.106	27.5	7.2	15.27	3.5	  
1119	72.393	18.9	5.1	15.27	3.3	    
815	71.687	38.2	9.3	15.27	7.8	  
1753	72.419	34.1	8	15.26	12.6	  
1454	72.221	19	5.1	15.25	3.2	  
1145	72.618	41.9	10.9	15.23	6.4	  
1103	72.445	37.2	8	15.23	3.1	  
810	71.364	34.8	8.6	15.23	6.1	    
456	71.772	47.3	10.8	15.23	9.8	  
75	71.539	50.3	-3.3	15.22	8.1	    
1789	72.145	19.9	5.1	15.21	12	  
1189	72.453	32	8	15.21	6	  
302	72.136	17.7	4.3	15.21	4.4	  
150	71.785	42.1	10.8	15.21	3.2	  
898	72.614	46.1	10.1	15.2	2.1	  
812	71.498	58.7	11.5	15.19	5.2	  
90	71.599	40.5	-4.1	15.19	5.8	    
1450	72.094	42.7	10.9	15.18	6.3	  
1381	72.491	27.3	5.8	15.18	2.3	  
1156	72.328	41.5	9.4	15.18	4.8	  
167	71.787	38.7	8.7	15.18	2.8	  
1615	72.679	34.9	9.5	15.17	14.2	  
1258	72.336	21.8	5.1	15.15	4.8	  
36	71.977	22.6	1.7	15.15	8.6	    
2193	72.328	38.5	10.1	15.14	2.8	  
1256	72.41	26.5	5.8	15.13	5.4	  
1354	72.147	28.4	8	15.12	3.1	  
1350	72.199	24.3	4.3	15.12	3.6	  
1106	72.526	30	6.5	15.12	3.9	  
13	72.559	61.4	-5.5	15.12	4.9	  
1707	72.356	22.7	5.1	15.11	6.2	  
603	71.621	35.1	8.7	15.11	2.8	  
1125	72.246	25.7	5.8	15.1	5.2	  
670	71.726	43.9	10.1	15.1	3.9	  
239	71.715	35.4	8	15.1	2.4	  
985	72.574	38.2	11.6	15.09	7.1	  
972	72.617	13.4	2.9	15.09	4.9	    
964	72.15	35.6	8.7	15.08	7.1	  
900	72.453	42	12.2	15.07	3.8	  
392	71.853	52.5	12.3	15.06	3.9	  
234	71.765	38.9	9.4	15.06	4.2	  
1738	72.35	41.8	10.2	15.05	10.3	  
1735	72.355	39.6	9.5	15.05	5.2	  
1425	72.366	35.1	9.4	15.05	4.3	  
1138	72.048	34.1	8.7	15.04	3	  
197	71.703	22.1	5.1	15.04	6.4	    
1124	72.288	23.3	5.1	15.03	3.6	    
1110	72.241	22.7	5.1	15.03	4.6	  
681	71.503	52.1	11.6	15.03	5.3	  
1622	72.231	32.9	8.8	15.02	4.4	  
1231	72.466	44.1	12.3	15.02	5.7	  
1472	72.173	22.7	5.8	15.01	3.5	  
1136	72.02	32.9	8	15.01	3.3	  
903	72.265	51.2	12.2	15.01	5	  
2523	71.884	46	10.2	14.99	3.2	  
1581	72.018	33	6.6	14.98	5.5	  
1567	72.343	39.1	10.2	14.98	9.1	  
1475	72.286	27.4	7.2	14.98	3.4	  
1311	72.16	43	10.9	14.98	7	  
1005	72.362	41.8	10.2	14.98	5.4	  
1742	72.623	30.3	7.3	14.97	6.1	  
890	71.745	21.2	4.3	14.97	3.9	  
32	72.174	38.2	-1.9	14.97	19.1	    
1389	72.191	24.7	6.5	14.96	3.6	  
1524	72.123	31.8	7.3	14.95	7.7	  
1594	72.23	30.5	7.3	14.93	15.9	  
1498	72.115	39.3	9.5	14.93	5.3	  
1314	72.15	35	8.7	14.93	4.6	  
1109	72.44	26.8	7.2	14.92	3.9	  
34	72.085	33.2	0.3	14.92	11.6	    
907	71.676	24.7	6.5	14.88	5.8	  
1193	72.248	25.5	7.2	14.87	3.2	  
1319	72.182	33.7	7.2	14.84	5.7	  
1380	72.507	28.4	7.2	14.82	3.1	  
1108	72.467	21.4	5.8	14.82	4.8	  
676	71.403	35.7	7.2	14.81	2	    
1163	72.125	38.8	10.2	14.8	4.7	  
1069	72.155	26.5	7.2	14.79	5.1	  
928	71.417	46.6	11.5	14.78	5	    
1120	72.286	34.2	8	14.77	4.7	  
1729	72.272	23.5	5.8	14.76	3.1	  
1417	72.143	33	8.7	14.76	4.6	  
1178	72.666	42.6	10.9	14.76	3.3	  
1122	72.339	38.8	10.9	14.76	2.5	  
97	71.647	53.1	13	14.76	6.1	  
1559	72.327	25.6	5.8	14.75	7.9	  
1116	72.419	15.6	4.3	14.75	5.2	  
1704	72.256	43.1	10.2	14.74	15	  
1126	72.21	26.9	7.2	14.74	6.4	  
1161	72.015	23.6	5.8	14.73	8.8	  
1033	72.243	32.6	9.4	14.72	5	  
1784	72.023	24.7	6.6	14.68	17.9	  
1561	72.216	38.4	9.5	14.68	15.2	  
974	72.669	24.8	6.5	14.67	5.1	  
1692	72.18	35.2	8	14.66	5.2	  
1778	71.945	46.8	11.7	14.64	4.1	  
1006	72.406	20.6	5.1	14.64	6.9	  
4	72.181	45.7	-2.6	14.64	5.4	  
1793	72.183	43.3	10.2	14.63	19.3	  
1584	72.479	22.9	5.8	14.63	8.9	  
1031	72.12	31.8	8.7	14.63	9.1	  
1740	72.452	33.9	8.8	14.62	3.7	  
1629	72.45	28.3	6.6	14.62	5.2	  
1646	72.093	54.2	11.7	14.61	9.7	  
1447	72.128	32.8	7.2	14.61	4.2	  
1564	72.135	37	8.8	14.59	3.4	  
1383	72.335	15	3.6	14.59	3.7	  
977	72.681	33.3	8	14.59	5	  
421	71.911	57	11.6	14.58	5.6	  
1320	72.041	24.2	5.1	14.57	6.7	  
606	71.526	41.5	8.7	14.57	3.4	  
1352	72.163	15.4	3.6	14.56	3	  
21	72.575	67.2	-6.2	14.56	9.2	  
1409	72.106	15.4	3.6	14.55	2.3	  
1693	72.118	30.8	7.3	14.54	4.8	  
1107	72.478	12.2	2.9	14.54	3.1	    
608	71.592	50.6	11.6	14.53	3.2	  
1595	72.253	42.2	10.9	14.52	9.6	  
1388	72.234	18.2	4.3	14.52	5.1	  
1812	72.641	36.8	8.8	14.51	9.7	  
1044	72.318	27.6	6.5	14.51	4.9	  
1556	72.387	20.4	4.4	14.5	11.9	  
1166	72.205	30.3	8	14.5	3.6	  
1560	72.332	36	8	14.49	4	  
1111	72.446	18.9	4.3	14.48	1.7	  
1770	72.78	22.3	5.1	14.47	23.4	  
1035	72.37	32.2	8	14.47	3.8	  
1298	71.967	48.3	10.2	14.45	4.5	  
1086	72.205	25.8	7.2	14.45	4.4	  
894	71.758	31.8	6.5	14.45	3.1	  
1489	72.495	35.2	10.2	14.43	5.2	  
23	72.562	59.4	-6.2	14.43	6.9	  
1376	72.533	29.6	8	14.42	3.7	  
1048	72.351	29.1	7.2	14.42	3.9	  
997	72.608	30.3	7.2	14.42	3.1	  
1355	72.462	45.6	10.9	14.41	3.3	  
1148	72.671	27.3	6.5	14.41	3.7	  
2281	72.375	20.9	4.3	14.4	6.6	  
1027	72.332	39.5	9.4	14.4	4.1	  
1058	72.535	30.3	8	14.38	2.7	  
17	72.601	33	0.3	14.38	7.3	  
166	71.668	23	5.1	14.37	4.5	  
1082	72.2	18.7	4.3	14.36	5.4	  
1047	72.893	30.5	7.2	14.36	6.3	    
1339	72.704	35.9	8.7	14.34	5.6	  
1072	72.269	42.3	10.9	14.34	4.5	  
8	71.698	42.4	-3.3	14.34	9.1	  
1366	72.456	12	3.6	14.33	5.1	  
37	71.998	63.4	-6.2	14.33	11.4	    
1627	72.344	21.8	5.1	14.32	8.6	  
1114	72.451	19.3	4.3	14.32	5.3	  
1056	72.618	28.3	7.2	14.32	2.5	  
1172	72.385	38.8	9.4	14.3	4.5	  
1806	72.522	39.9	12.4	14.29	24.2	  
1233	72.527	46.6	11.6	14.29	3.7	  
1169	72.275	38.1	9.4	14.29	2.4	  
1067	72.093	23	5.1	14.29	4.9	  
1607	72.594	26.8	7.3	14.28	11.9	  
1536	72.229	24.3	6.6	14.28	5.1	  
521	71.765	46.6	10.1	14.28	5.2	    
1731	72.177	14.8	3.6	14.27	3.9	  
445	71.724	45.8	10.1	14.27	6	  
1061	72.314	27.8	6.5	14.26	5.2	  
459	71.69	49.5	10.8	14.26	5	  
1351	72.25	33.1	7.2	14.25	3	  
442	71.766	56.1	13	14.25	7.4	  
25	72.513	47.4	-4.8	14.25	8.2	  
1730	72.234	44.5	9.5	14.24	2.7	  
2241	72.432	28.6	7.2	14.23	3.5	  
1382	72.363	30	7.2	14.23	2.6	  
979	72.686	48.1	12.3	14.21	5.8	  
1358	72.528	38.8	9.4	14.19	2.4	  
995	72.681	34.4	8	14.19	4	    
1051	72.721	35.2	8	14.18	4.7	  
1001	72.191	33	8.7	14.18	5.9	  
1391	72.24	15.6	4.3	14.17	6.4	  
1184	72.707	54	12.3	14.17	2.8	  
1052	72.618	40.8	8.7	14.17	4	  
101	71.542	33.5	9.4	14.17	6.9	  
1775	72.808	24.8	5.1	14.16	21.9	  
1228	72.378	27.3	6.5	14.16	5	  
1783	72.061	35	8.8	14.15	8.9	  
1405	72.088	19	5.1	14.15	3.3	  
1501	72.213	30	8	14.14	12.2	  
899	71.766	24.1	6.5	14.14	3.5	  
1088	72.691	34.4	8.7	14.13	4.3	  
1045	72.892	35.3	8	14.12	5	  
339	71.673	29.8	7.2	14.11	3.4	  
1776	71.853	38.1	8.8	14.09	5.2	  
1232	72.285	55.1	10.9	14.09	5	  
1546	72.283	43.7	12.4	14.07	13.6	  
1318	72.108	16.6	4.3	14.06	5.6	  
1197	72.321	37.2	8.7	14.06	3.9	  
1059	72.438	29	7.2	14.05	3.9	  
1625	72.335	31.4	6.6	14.04	5.7	  
1503	72.125	36.3	9.5	14.03	43.5	  
1377	72.596	39.7	8.7	14.02	3.3	  
1323	72.132	19.8	5.1	14.01	2.8	  
1081	72.637	30.4	8	14.01	8.4	  
1601	72.453	22.7	5.1	13.99	23.9	  
1290	72.237	41.2	10.2	13.99	4.2	  
1180	72.722	32.1	8.7	13.98	2.4	  
1043	72.326	26.2	6.5	13.98	4.8	  
24	71.707	40.5	-2.6	13.97	43.6	  
994	72.768	23.7	5.1	13.96	2.5	  
71	71.533	52.1	-4.8	13.96	8.1	    
1188	72.535	32.8	7.2	13.95	2.3	  
991	72.863	31.7	7.2	13.95	5.7	  
1182	72.706	33.4	8	13.94	4.5	  
1038	72.5	39.9	10.9	13.94	3.8	  
983	72.694	44.8	10.2	13.94	4.7	  
1669	71.916	38.2	10.2	13.93	2.1	  
1610	72.675	37.5	8.8	13.92	9.2	  
1771	72.872	46.2	10.9	13.91	26.4	  
1379	72.449	30.6	6.5	13.91	3	  
1338	72.678	33.5	6.5	13.91	3.5	  
1162	72.052	24.8	6.5	13.9	4.8	  
1068	72.091	27	7.2	13.9	7.5	  
1187	72.62	43.8	10.2	13.89	2.9	  
1196	72.275	45.3	8.7	13.88	6.7	  
1137	72.241	46.7	10.2	13.88	3.5	  
1185	72.668	45.9	10.2	13.87	2	    
1183	72.177	42.1	10.2	13.87	3.2	  
1288	72.37	34.6	6.5	13.86	8.9	  
1236	72.691	25.5	6.5	13.86	2.4	  
1555	72.578	14.2	3.6	13.85	5.9	  
1250	72.742	32.9	8.7	13.84	3.1	  
1346	72.565	37.5	8.7	13.81	3.1	  
1152	72.633	31.5	8	13.81	2.9	  
1477	72.343	63.7	14.5	13.8	7.6	  
1055	72.55	23.2	5.1	13.8	3.4	  
992	72.82	27.9	7.2	13.79	4.2	  
1551	72.575	30.3	7.3	13.78	7.4	  
1708	72.323	26	5.8	13.77	10.1	  
1486	72.51	40.4	8.8	13.77	4.5	  
1374	72.754	33	8	13.77	3.4	  
1040	72.622	40	8.7	13.77	5.1	  
1023	72.661	51.4	13.1	13.77	3.6	  
1017	72.866	26.5	7.2	13.77	4.9	    
1255	72.408	33.4	8.7	13.76	4.2	  
1246	72.831	29	6.5	13.76	2.9	  
1672	72.118	30.8	8	13.75	7.4	  
1583	72.416	16.2	5.1	13.75	5.8	  
1154	72.535	38.1	8	13.74	3.5	  
1053	72.445	38.6	8.7	13.74	2.8	  
1526	72.097	32.9	8.8	13.73	8.9	  
1482	72.366	31.7	8	13.73	4.6	  
1373	72.494	46.1	11.6	13.72	2.5	  
1782	72.028	47.9	11.7	13.71	6.9	  
1645	72.155	60.4	13.1	13.71	14.2	  
1473	72.331	43.4	10.9	13.7	4	  
1186	72.231	36.3	8.7	13.7	3.9	  
1348	72.362	23.5	6.5	13.68	2.8	  
1326	72.286	39.9	9.4	13.68	3.3	  
1143	72.73	41.7	10.9	13.68	2.8	  
1026	72.233	30.4	8.7	13.68	3.5	    
1344	72.37	38.6	8.7	13.66	1.8	    
1008	72.477	21.1	5.1	13.66	4	  
1261	72.247	29.7	7.2	13.65	6.7	  
1025	72.443	47.5	10.9	13.65	4.4	  
1515	72.513	30.9	7.3	13.64	4	  
1332	72.678	32	8	13.62	5.2	  
1484	72.446	28.6	8	13.61	3.8	  
1248	72.785	21.6	5.8	13.6	2.7	    
1796	72.305	28.9	7.3	13.59	15.5	  
1599	72.351	34	9.5	13.58	6.8	  
1241	72.894	27.3	7.2	13.58	3.6	  
1020	72.285	32.4	8.7	13.58	4.5	  
1176	72.455	32.2	8	13.57	2.5	  
1144	72.71	54	13.1	13.57	4.4	  
1613	72.784	17.5	4.4	13.56	8	  
1245	72.88	30.1	6.5	13.56	3.2	  
1087	72.766	31.3	7.2	13.56	3.1	    
1085	72.738	28.5	8	13.56	4.1	  
1009	72.551	16	3.6	13.56	3.7	  
1078	72.592	32.7	8	13.55	2.3	    
1624	72.327	34.1	6.6	13.54	8.7	  
1378	72.434	29.4	6.5	13.54	1.9	  
1249	72.468	32.2	8	13.54	4.3	  
1168	72.262	20.4	5.1	13.54	4.9	  
1341	72.362	25.1	5.8	13.53	2	  
1239	72.754	18.9	4.3	13.53	2.5	  
1542	72.135	45.6	10.9	13.52	18.6	  
1317	72.161	28.9	6.5	13.51	5.6	  
1146	72.711	25.3	6.5	13.51	2.5	  
16	71.537	22.9	1.7	13.51	9.5	  
1541	72.175	37.2	8.8	13.5	16.8	  
1372	72.863	35.3	8.7	13.5	5.6	  
1321	72.045	32.8	8	13.5	3.2	  
1286	72.493	38.4	10.9	13.5	5.5	  
1173	72.365	33.2	9.4	13.5	3.4	  
988	72.573	42.2	9.4	13.5	4.7	  
1151	72.686	19.2	5.1	13.49	2.7	  
998	72.158	30.3	8	13.49	7.3	  
1786	72.134	47.7	10.2	13.48	18.3	  
1705	72.102	34.5	8.8	13.48	7.2	  
1630	72.42	24	5.8	13.47	7.4	  
1018	72.38	30.1	8	13.47	3.9	  
1334	72.676	30.7	8	13.46	2.2	  
1243	72.546	19.9	4.3	13.46	3.9	    
1179	72.396	50.5	10.9	13.46	2.7	  
1362	72.593	28.9	6.5	13.45	2.1	  
1347	72.366	25	8	13.45	1.8	  
1279	72.714	25.5	5.1	13.45	8.4	  
1019	72.91	43.2	9.4	13.44	4.3	    
1360	72.5	15	3.6	13.42	4.9	  
1141	72.73	45.3	9.4	13.42	2.7	  
1315	72.266	44.9	10.9	13.41	3.3	  
1364	72.631	20.9	5.1	13.39	2.2	  
1356	72.371	33.8	8	13.39	4.4	  
1252	72.66	38	9.4	13.39	3.5	  
1712	72.399	21.4	5.8	13.38	3.2	  
1535	72.24	20.9	5.8	13.38	6.5	  
1234	72.335	47.3	10.2	13.38	8.2	  
1310	72.783	26.5	6.5	13.37	2.6	  
1276	72.723	35.3	7.2	13.37	5.9	  
1370	72.935	49.2	10.2	13.36	2.8	  
1217	72.631	38.2	8.7	13.36	2.9	  
1545	72.215	29.4	7.3	13.35	13.9	  
1074	72.382	34.5	8.7	13.35	3.1	  
1224	72.545	17.5	4.3	13.34	5.2	  
1221	72.583	44.4	11.6	13.34	2.3	  
1329	72.553	29.4	7.2	13.32	3	  
1325	72.321	24.8	5.8	13.32	3.6	    
1313	72.441	49.7	11.6	13.32	12.2	  
1149	72.69	22.6	5.1	13.32	2.7	  
1022	72.288	36.9	9.4	13.32	4.5	  
1504	72.291	47.7	13.1	13.29	10.9	  
1367	72.768	29.2	6.5	13.29	3.5	  
1240	72.753	55.8	13.8	13.29	3.1	  
107	71.533	33.4	8.7	13.29	5.7	  
1327	72.449	33.1	8	13.26	1.6	  
1711	72.43	23.6	5.1	13.25	6.4	  
1340	72.582	46.5	10.2	13.24	3.2	  
1167	72.198	28.7	6.5	13.24	3.9	    
1725	72.141	15	4.4	13.23	4.8	  
1073	72.121	31.2	8	13.23	6.5	  
1670	71.941	32.2	8	13.22	4.4	  
1470	72.665	36.6	8.7	13.22	3.3	  
1076	72.495	29.8	6.5	13.21	2.5	  
1702	72.638	27.6	7.3	13.2	4.9	  
1307	72.803	34.9	8.7	13.19	5.1	  
1054	72.453	34.4	7.2	13.19	6	  
1591	72.51	45	10.9	13.17	9.3	  
1265	72.313	34.7	8.7	13.17	3.1	  
1337	72.57	30.6	7.2	13.16	2.1	  
1083	72.678	44.2	11.6	13.16	3.2	  
1036	72.207	50.9	11.6	13.16	7.2	  
1728	72.233	28.9	6.6	13.14	6.7	  
1281	72.697	34.7	9.4	13.14	3.1	  
1805	72.487	63.5	13.9	13.13	6.6	  
1273	72.663	37.3	9.4	13.13	3.6	  
1013	72.733	40.9	10.2	13.13	3.6	  
1777	71.91	49.8	10.2	13.12	4.9	  
1677	72.269	17.3	3.6	13.12	48.1	  
1216	72.521	28	7.2	13.12	3.6	  
1636	72.533	34.7	8.8	13.11	5.9	  
984	72.512	16.5	3.6	13.11	6.7	  
455	71.647	34.9	9.4	13.11	3.8	  
1487	72.476	29.8	6.6	13.1	3	  
1060	72.408	33.1	8	13.1	6.4	  
1361	72.545	26.8	6.5	13.09	2.7	  
1175	72.378	31.8	6.5	13.09	2.4	  
1309	72.813	25.4	5.8	13.07	3.5	    
1283	72.622	27.9	6.5	13.06	2.9	  
1011	72.642	30.2	7.2	13.06	2.4	  
1214	72.691	25	5.1	13.05	3.5	  
1481	72.471	49.9	13.1	13.04	2.4	  
1817	72.704	33.2	8.8	13.02	17.7	  
1331	72.296	40.6	9.4	13.02	2.3	  
1007	72.37	27.9	5.8	13.02	5.3	  
1357	72.438	39.4	8.7	13	4.6	  
1303	72.616	41.9	10.9	13	3.5	  
1335	72.55	31.4	7.2	12.99	2	  
1324	72.258	26.6	6.5	12.99	2.3	  
1474	72.291	38.1	10.9	12.98	3.8	  
1819	72.724	48.3	10.2	12.97	18	  
1278	72.71	32.2	8.7	12.97	2.7	  
1494	72.487	28.8	6.6	12.96	5.2	  
1306	72.721	36.4	10.2	12.96	2.6	  
1553	72.655	19.9	4.4	12.95	12.5	  
1633	72.431	29.8	8	12.94	7.4	  
1547	72.371	43	12.4	12.94	10.8	  
1289	72.306	38.5	11.6	12.94	2.4	  
1260	72.232	35.1	8.7	12.94	4.7	  
1650	72.246	67.4	13.9	12.93	11.2	  
1543	72.115	56	12.4	12.93	3.8	  
1480	72.522	20.8	5.8	12.92	2.1	  
1316	72.283	45.5	10.2	12.92	4.6	  
1511	72.422	29.5	7.3	12.9	10.6	  
1718	72.549	32.7	8	12.89	3.7	  
1488	72.528	39.6	9.5	12.88	3.1	  
1299	72.512	45.2	10.9	12.88	3.5	  
1277	72.341	34.4	8.7	12.86	2.7	  
1219	72.644	26	6.5	12.86	1.8	  
1210	72.788	21.2	5.1	12.86	2.6	  
1517	71.993	25.2	6.6	12.85	5.7	  
1274	72.695	29.4	7.2	12.85	3.4	  
1077	72.21	35.2	8	12.85	4.3	  
1569	72.633	40.9	10.2	12.83	5.7	  
1199	72.555	39.4	8.7	12.83	3.7	  
1668	72.05	27.4	8	12.78	5.5	  
1211	72.539	27.1	6.5	12.78	4.3	  
1305	72.27	48.9	10.9	12.77	3.3	  
1797	72.403	42	9.5	12.75	14.5	  
1308	72.214	39.3	8.7	12.75	2.5	  
1140	72.534	47.1	12.3	12.75	3.2	  
1269	72.479	21.1	5.1	12.74	2.6	  
1212	72.762	20.1	5.1	12.74	3.5	  
1586	72.496	28.4	6.6	12.73	7	  
1302	72.53	52.6	11.6	12.73	3.1	  
978	72.416	34	8	12.73	5.3	  
1582	72.418	29.4	7.3	12.72	5.1	  
1469	72.749	23.5	6.5	12.72	2.9	  
1266	72.398	27.3	6.5	12.72	2.7	  
1790	72.146	39.1	10.2	12.71	9.1	  
1602	72.51	30.1	6.6	12.71	12	  
1525	72.119	26.4	6.6	12.71	23	  
1295	72.48	36.6	8	12.71	3.9	  
1014	72.465	35.2	8	12.71	4	  
1717	72.552	36.7	8.8	12.69	3.3	  
1557	72.405	30.3	7.3	12.69	8.2	  
1280	72.509	49.1	12.3	12.69	2.8	    
1037	72.251	42	9.4	12.69	4	  
1015	72.471	44.2	10.2	12.69	4.2	  
1603	72.397	27.9	7.3	12.68	13.3	  
1312	72.291	33.5	8	12.68	2.8	  
1174	72.351	28.5	8	12.65	3	  
1531	72.305	33	8	12.64	6.9	  
1204	72.676	47.1	10.2	12.62	2.7	  
1642	72.719	31.2	7.3	12.6	2.4	  
1237	72.355	43.8	10.9	12.6	5.2	  
1808	72.626	62.2	13.9	12.59	5.1	  
1284	72.458	48.1	12.3	12.58	1.7	  
1620	72.713	31.1	7.3	12.55	14.5	  
1635	72.51	29.5	7.3	12.53	3.7	  
1272	72.631	28	7.2	12.53	2.7	  
1215	72.694	35.5	9.4	12.52	1.9	  
1723	72.609	39.9	9.5	12.51	6.6	  
1579	72.337	44.3	10.2	12.51	3.1	  
1611	72.653	41.5	10.2	12.5	6.4	  
1304	72.33	52.5	12.3	12.49	4	  
1573	72.232	10.6	2.9	12.48	3.9	  
1202	72.643	22.1	5.1	12.47	2.8	  
1713	72.458	18.9	4.4	12.46	4.1	  
1639	72.628	32	7.3	12.46	4.8	  
1483	72.523	34	6.6	12.46	4.4	  
1714	72.388	35.6	8.8	12.45	6.6	  
1297	72.493	43.6	10.2	12.45	2.9	  
1719	72.516	32.6	8.8	12.44	5.5	  
1530	72.607	36.7	9.5	12.44	5	  
1300	72.363	28	7.2	12.43	3.9	  
1207	72.768	23.1	5.8	12.43	2.1	  
1804	72.426	41.2	9.5	12.38	8.7	  
1263	72.341	25.8	6.5	12.37	4.5	  
1616	72.821	32.4	8	12.36	4.6	  
1634	72.531	38.5	10.2	12.35	4.2	  
1238	72.481	45.7	11.6	12.35	3.8	  
1264	72.375	34.8	8	12.34	3	  
1716	72.545	38.4	8	12.33	4.5	  
1715	72.544	40.4	11.7	12.33	4.1	  
1574	72.231	43.6	10.2	12.33	3.3	  
1271	72.598	26.1	6.5	12.31	2.7	  
1198	72.503	24.4	5.8	12.31	2.8	  
1608	72.578	27	6.6	12.3	6.8	  
1294	72.481	31.5	8.7	12.29	2.9	  
1514	72.693	29.3	7.3	12.27	5.1	  
1701	72.601	25.2	6.6	12.26	4.5	  
1205	72.705	22.4	5.8	12.26	1.7	  
1810	72.542	50.6	12.4	12.24	8.2	  
1612	72.675	39.3	10.2	12.24	5.1	  
1663	72.56	26.9	7.3	12.23	6.1	  
1600	72.362	43.5	10.9	12.23	4.8	  
1614	72.759	30.8	7.3	12.22	5.6	  
1621	72.708	33.3	7.3	12.21	2.5	  
1587	72.636	30.7	8	12.21	3	  
1200	72.539	35.1	7.2	12.2	4	  
1710	72.444	50.7	11.7	12.18	4.6	  
1589	72.414	28.2	7.3	12.17	4.7	  
1575	72.241	36.7	9.5	12.13	3	  
1485	72.613	33.4	8	12.13	3.8	  
1722	72.739	37.1	10.2	12.12	2.3	  
1301	72.343	24.9	6.5	12.11	2.8	  
1803	72.53	35.3	8.8	12.09	9.2	  
1640	72.612	22.6	6.6	12.09	2.9	  
1267	72.322	41.8	10.2	12.09	2.6	  
1703	72.705	34	8	12.08	2.5	  
1588	72.713	46.5	12.4	12.08	3	  
1641	72.6	33.9	8	12.07	2.5	  
1631	72.517	31	7.3	12.07	4.1	  
1606	72.582	37.5	9.5	12.06	3.4	  
1604	72.467	33.9	9.5	12.04	2.4	  
1696	72.446	37.5	10.2	12.03	5.8	  
1691	72.316	28.6	7.3	12.02	2.4	  
1618	72.837	38.5	8.8	12.01	4.7	  
1697	72.501	31.8	7.3	12	4.7	  
1548	72.408	47.7	11.7	11.99	3.8	  
1813	72.636	33.4	9.5	11.98	7.7	  
1554	72.678	42.3	9.5	11.98	5.7	  
1590	72.536	37	9.5	11.96	4.5	  
1549	72.521	35.2	8.8	11.96	5.7	  
1720	72.631	34.8	10.2	11.95	3.2	  
1592	72.22	23.6	6.6	11.95	6.9	  
1617	72.823	27.7	5.8	11.94	2.3	  
1695	72.429	30.8	7.3	11.92	2.2	  
1296	72.376	49.3	10.9	11.91	3.4	  
1652	72.313	39.1	8.8	11.89	9	  
1578	72.348	39.5	9.5	11.88	2.5	  
1529	72.222	34.5	8.8	11.87	3.5	  
1619	72.734	43.2	11.7	11.86	4.2	  
1609	72.548	23.7	6.6	11.86	3.2	  
1767	72.311	45	9.5	11.85	5	  
1651	72.508	33.2	8	11.84	4	  
1598	72.258	26.2	5.8	11.84	3.1	  
1660	72.624	29.4	6.6	11.83	2.7	  
1688	72.42	27	7.3	11.8	5.4	  
1637	72.645	27	6.6	11.77	5.5	  
1721	72.608	46.1	11.7	11.76	4	  
1522	72.518	29.4	7.3	11.76	2.4	  
1671	72.157	34.5	7.3	11.73	3.6	  
1818	72.742	54.2	12.4	11.71	5.7	  
1700	72.515	25.8	5.8	11.69	4.4	  
1506	72.418	32	8	11.69	9.2	  
1552	72.608	38.1	10.9	11.67	2.1	  
1816	72.534	44.8	10.9	11.66	3.1	  
1664	72.66	29	6.6	11.65	3.8	  
1510	72.411	42	9.5	11.65	3.9	  
1698	72.416	35.1	8.8	11.63	3.5	  
1605	72.459	25.9	6.6	11.63	7.2	  
1507	73.016	23.6	6.6	11.61	4.1	  
1568	72.303	27.2	7.3	11.6	6.6	  
1509	72.61	45	10.2	11.59	4.9	  
1694	72.355	31.8	8	11.58	3.1	  
1585	72.529	35.3	9.5	11.58	1.7	  
1632	72.548	34.6	10.2	11.57	7.6	  
1681	72.436	39.6	9.5	11.55	2.9	  
1814	72.675	34.4	7.3	11.54	9.8	  
1558	72.443	20.8	5.1	11.52	6.2	  
1674	72.73	21.7	5.1	11.5	3.1	  
1815	72.788	58.6	12.4	11.48	4.2	  
1709	72.353	39.2	9.5	11.47	2.7	  
1667	72.73	35.9	10.2	11.46	4.3	  
1820	72.923	46	9.5	11.45	5.6	  
1505	72.438	43.8	10.9	11.44	8.1	  
1550	72.533	39.9	8.8	11.43	3.5	  
1684	72.411	39.3	10.2	11.42	4.9	  
1699	72.438	34.9	9.5	11.41	2.4	  
1570	72.114	22.2	5.8	11.4	4.5	  
1678	72.325	33	7.3	11.38	19.1	  
1662	72.49	22	5.8	11.37	3.4	  
1508	72.711	32	8.8	11.36	5.3	  
1676	72.233	30.7	7.3	11.33	2.2	  
1493	72.513	28	7.3	11.31	3.8	  
1687	72.523	35.1	8.8	11.3	1.9	  
1638	72.597	43.8	10.9	11.28	2.4	  
1675	72.632	34.9	8.8	11.27	2.9	  
1492	72.601	36.8	8.8	11.27	3.6	  
1682	72.628	28.5	7.3	11.25	3.6	  
1689	72.425	35.5	9.5	11.23	2	  
1679	72.333	17.4	5.1	11.23	5.4	  
1665	72.575	22.7	5.8	11.19	3.3	  
1648	72.373	23	5.1	11.19	1.6	  
1658	72.546	45.1	12.4	11.17	2.4	  
1655	72.315	26	5.8	11.16	4	  
1686	72.416	25.5	6.6	11.1	3.3	  
1666	72.623	33.6	8.8	11.1	2.6	  
1644	72.447	40.4	10.9	11.09	3.6	  
1544	72.34	42.5	9.5	11.05	3.5	  
1661	72.551	17.8	3.6	10.97	3.6	  
1690	72.546	39.4	10.9	10.96	4.5	  
1491	72.681	32.8	7.3	10.82	3.6	  
1653	72.306	20.6	5.1	10.8	6.4	  
1657	72.476	35.9	8.8	10.73	3.1	  
1654	72.285	29.2	6.6	10.68	3.4	  
1683	72.477	10.6	2.9	10.67	4.9	  
1685	72.454	31.1	7.3	10.55	5.7	  
1656	72.481	29.6	6.6	10.5	3.7	  
1659	72.407	43.1	10.2	10.4	2.3
-------------- next part --------------
A non-text attachment was scrubbed...
Name: plot_Hamamatsu_c1.png
Type: image/png
Size: 18258 bytes
Desc: not available
Url : https://mailman.jlab.org/pipermail/halld-cal/attachments/20121207/332faa7f/attachment-0002.png 
-------------- next part --------------
A non-text attachment was scrubbed...
Name: plot_Hamamatsu_c2.png
Type: image/png
Size: 16848 bytes
Desc: not available
Url : https://mailman.jlab.org/pipermail/halld-cal/attachments/20121207/332faa7f/attachment-0003.png 
-------------- next part --------------
void plot_Hamamatsu() {
  gROOT->Reset();
  gStyle->SetPalette(1,0);
   char string[256];
   char filename[80];
   Int_t j,jj;
   
   #define npts 4000;
  Float_t x1[npts];
  Float_t x2[npts];
  Float_t x3[npts];
  Float_t x4[npts];
  Float_t x5[npts];
  Float_t x6[npts];
  ifstream in;
  in.open("Hamamatsu_data.dat");
  
  Float_t y1;
  Float_t y2;
  Float_t y3;
  Float_t y4;
  Float_t y5;
  Float_t y6;
  Int_t i = 0; 
   Double_t xmin=0;
   Double_t xmax=0.5;
   Double_t ymin=0;
   Double_t ymax=100;
   Int_t nbins=50;
  TH1F *p1 = new TH1F("p1","delta",nbins,0,4000);
  TH1F *p2 = new TH1F("p2","delta",nbins,60,80);
  TH1F *p3 = new TH1F("p3","delta",nbins,0,100);
  TH1F *p4 = new TH1F("p4","delta",nbins,0,20);
  TH1F *p5 = new TH1F("p5","delta",nbins,0,50);
  TH1F *p6 = new TH1F("p6","delta",nbins,0,20);
  TH2F *RMS_vs_DR = new TH2F("RMS_vs_DR","DR RMS vs DR",nbins,0,50,nbins,0,20);
  while (i<npts) { // one iteration of this loop per set of points
    
    in >> y1 >> y2 >> y3 >> y4 >> y5 >> y6;    
    printf("%f %f %f %f %f %f\n",y1,y2,y3,y4,y5,y6);
    p1->Fill(y1);
    p2->Fill(y2);
    p3->Fill(y3);
    p4->Fill(y4);
    p5->Fill(y5);
    p6->Fill(y6);
   RMS_vs_DR->Fill(y5,y6);
    i++;
   
  }
   c1 = new TCanvas("c1","c1 plot_Hamamatsu",200,10,700,500);
   c1->SetBorderMode(0);
   c1->SetFillColor(0);
   c1->Divide(2,2);
   c1->cd(1);
   c1_1->SetBorderMode(0);
   c1_1->SetFillColor(0);
   
   p1->SetTitle("");
   // p1->GetXaxis()->SetRangeUser(xmin,xmax);
   // p1->GetYaxis()->SetRangeUser(ymin,ymax);
   p1->GetXaxis()->SetLabelSize(0.05);
   p1->GetXaxis()->SetTitleSize(0.05);
   p1->GetYaxis()->SetLabelSize(0.05);
   p1->GetYaxis()->SetTitleSize(0.05);
   p1->GetYaxis()->SetTitleOffset(1.5);
   p1->GetXaxis()->SetTitle("Serial Number");
   // p1->GetXaxis()->SetNdivisions(505);
   p1->SetLineColor(1);
   p1->Draw();
   c1->cd(2);
   c1_2->SetBorderMode(0);
   c1_2->SetFillColor(0);
   
   p2->SetTitle("");
   // p2->GetXaxis()->SetRangeUser(xmin,xmax);
   // p2->GetYaxis()->SetRangeUser(ymin,ymax);
   p2->GetXaxis()->SetLabelSize(0.05);
   p2->GetXaxis()->SetTitleSize(0.05);
   p2->GetYaxis()->SetLabelSize(0.05);
   p2->GetYaxis()->SetTitleSize(0.05);
   p2->GetYaxis()->SetTitleOffset(1.5);
   p2->GetXaxis()->SetTitle("Operating Voltage at 25 degC (V)");
   // p2->GetXaxis()->SetNdivisions(505);
   p2->SetLineColor(1);
   p2->Draw();
 
   c1->cd(3);
   c1_3->SetBorderMode(0);
   c1_3->SetFillColor(0);
   
   p3->SetTitle("");
   // p3->GetXaxis()->SetRangeUser(xmin,xmax);
   // p3->GetYaxis()->SetRangeUser(ymin,ymax);
   p3->GetXaxis()->SetLabelSize(0.05);
   p3->GetXaxis()->SetTitleSize(0.05);
   p3->GetYaxis()->SetLabelSize(0.05);
   p3->GetYaxis()->SetTitleSize(0.05);
   p3->GetYaxis()->SetTitleOffset(1.5);
   p3->GetXaxis()->SetTitle("RMS Operating Voltage (mV)");
   // p3->GetXaxis()->SetNdivisions(505);
   p3->SetLineColor(1);
   p3->Draw();
   c1->cd(4);
   c1_4->SetBorderMode(0);
   c1_4->SetFillColor(0);
   
   p4->SetTitle("");
   // p4->GetXaxis()->SetRangeUser(xmin,xmax);
   // p4->GetYaxis()->SetRangeUser(ymin,ymax);
   p4->GetXaxis()->SetLabelSize(0.05);
   p4->GetXaxis()->SetTitleSize(0.05);
   p4->GetYaxis()->SetLabelSize(0.05);
   p4->GetYaxis()->SetTitleSize(0.05);
   p4->GetYaxis()->SetTitleOffset(1.5);
   p4->GetXaxis()->SetTitle("RMS Gain Variation / Gain (%)");
   // p4->GetXaxis()->SetNdivisions(505);
   p4->SetLineColor(1);
   p4->Draw();
 
  c2 = new TCanvas("c2","c2 plot_Hamamatsu",200,10,700,500);
   c2->SetBorderMode(0);
   c2->SetFillColor(0);
   c2->Divide(2,2);
   c2->cd(1);
   c2_1->SetBorderMode(0);
   c2_1->SetFillColor(0);
   
   p5->SetTitle("");
   // p5->GetXaxis()->SetRangeUser(xmin,xmax);
   // p5->GetYaxis()->SetRangeUser(ymin,ymax);
   p5->GetXaxis()->SetLabelSize(0.05);
   p5->GetXaxis()->SetTitleSize(0.05);
   p5->GetYaxis()->SetLabelSize(0.05);
   p5->GetYaxis()->SetTitleSize(0.05);
   p5->GetYaxis()->SetTitleOffset(1.5);
   p5->GetXaxis()->SetTitle("Array Dark Rate (MHz)");
   // p5->GetXaxis()->SetNdivisions(505);
   p5->SetLineColor(1);
   p5->Draw();
   c2->cd(2);
   c2_2->SetBorderMode(0);
   c2_2->SetFillColor(0);
   
   p6->SetTitle("");
   // p6->GetXaxis()->SetRangeUser(xmin,xmax);
   // p6->GetYaxis()->SetRangeUser(ymin,ymax);
   p6->GetXaxis()->SetLabelSize(0.05);
   p6->GetXaxis()->SetTitleSize(0.05);
   p6->GetYaxis()->SetLabelSize(0.05);
   p6->GetYaxis()->SetTitleSize(0.05);
   p6->GetYaxis()->SetTitleOffset(1.5);
   p6->GetXaxis()->SetTitle("(RMS Array Dark Rate) /Mean (%)");
   // p6->GetXaxis()->SetNdivisions(505);
   p6->SetLineColor(1);
   p6->Draw();
   c2->cd(3);
   c2_3->SetBorderMode(0);
   c2_3->SetFillColor(0);
   
   RMS_vs_DR->SetTitle("");
   // RMS_vs_DR->GetXaxis()->SetRangeUser(xmin,xmax);
   // RMS_vs_DR->GetYaxis()->SetRangeUser(ymin,ymax);
   RMS_vs_DR->GetXaxis()->SetLabelSize(0.05);
   RMS_vs_DR->GetXaxis()->SetTitleSize(0.05);
   RMS_vs_DR->GetYaxis()->SetLabelSize(0.05);
   RMS_vs_DR->GetYaxis()->SetTitleSize(0.05);
   RMS_vs_DR->GetYaxis()->SetTitleOffset(1.5);
   RMS_vs_DR->GetXaxis()->SetTitle("RMS DR vs DR (MHz)");
   // RMS_vs_DR->GetXaxis()->SetNdivisions(505);
   RMS_vs_DR->SetLineColor(1);
   RMS_vs_DR->Draw("colz");
  in.close();
 
   sprintf(filename,"plot_Hamamatsu_c1.png");
   c1->SaveAs(filename);
   sprintf(filename,"plot_Hamamatsu_c2.png");
   c2->SaveAs(filename);
  
}
    
    
More information about the Halld-cal
mailing list