[Halld-cal] newest revision of FCAL beam test NIM article

Matthew Shepherd mashephe at indiana.edu
Tue Apr 9 11:45:53 EDT 2013


Richard,

We attempted to study this in detail several years ago as we were designing the light guides.  

You may examine:

GlueX-doc 850

Table 1 is the table of interest.  There we indicate that a cylindrical light guide and cookie provides a factor of three improvement in photon collection efficiency over the air gap used in E852 and RadPhi.  When we wrote this, we consulted you on what the contribution of photo-statistics was to 7.3% RadPhi statistical term and then tried to assume that we just improve that component of uncertainty.

In my opinion we could remove all direct comparison with RadPhi from the NIM article and simply state that we are attempting to demonstrate that we can achieve the GlueX design resolution (which is really based on experience from E852 and RadPhi, but we don't need to make the connection so explicit in the paper).

Matt

------------------------------------------------------------------------
Matthew Shepherd, Associate Professor
Department of Physics, Indiana University, Swain West 265
727 East Third Street, Bloomington, IN 47405

Office Phone:  +1 812 856 5808

On Apr 9, 2013, at 11:33 AM, Richard Jones <richard.t.jones at uconn.edu> wrote:

> Kei,
> 
> In the abstract it states:
> Due to the addition of light guides, improvements are seen when these results
> are compared to extrapolations of the results of a previous experiment that used the same lead glass blocks.
> 
> How do we justify this statement?  Is it simply by comparing the test beam measurements with the final resolution from Radphi?  I would counter that the light guides are not the only difference there.  Back when we were setting up Radphi, Scott Teige reported on a test that he did in the context of the old E852 calorimeter where he directly compared the resolution of blocks with and without the "cookie" and (according to him) the difference was negligible.  I never saw data demonstrating this point.  You might ask Scott, since he is still right there in Bloomington.
> 
> Apart from that claim, I am skeptical that the addition of the cookie explains the big improvement shown in Fig. 8 over what we measured in Radphi.  I suspect that radiation damage in the blocks in Radphi is more important than the small improvement (maybe 25% more photoelectrons?) in photostatistics in determining the resolution at low energy.
> 
> The improvement is measured, so that is fine.  Attributing it all to the cookie needs to be proved, I think.  Here are some other effects that probably contribute at a similar level.
> 	• poor online gain balance in Radphi -- done using the flasher and plexi sheet, which was itself rad damaged and very non-uniform in illumination.  You can correct this to some extent offline, but not completely.
> 	• weaknesses in the gain calibration -- we had no clusters of "known" energy to use for this calibration, so we had to bootstrap the calibration off the pi0 mass peak.  This technique works, but it produces biases because of edges and dead blocks, whose effects propagate out into the intervening volume as you iterate the calibration procedure.
> 	• radiation damage -- a lot of this was already there before we started the main physics run.  We had a few beam tuning accidents during commissioning of the experiment.  Hence the lead wall, which was added before we started the main physics run in 2000.
> I don't mean to say that the data we took with Radphi was junk.  We achieved 7.2%/sqrt(E) in spite of all of these effects.  But we should not neglect them all and say that the final resolution       would have been what you saw in test beam, had we not left out the cookie.  That seems to be what is implied in the abstract.
> -Richard Jones 