[Halld-cal] [Revised Logentry] BCAL time-walk corrections
elton at jlab.org
elton at jlab.org
Tue Apr 14 15:55:02 EDT 2015
Logentry Text:
--
I have started thinking again about functional forms for the time-walk, in particular comparing your functional form from Shaun Krueger's thesis (6.2) with my eq. 5 in GlueX-doc-2618.
>From GlueX-doc-2618, Eq, 5 and 6:
t^w = t - f^w(P/T) + f^w(P_0/T), where
t^w is the walk-corrected time,
t is the time-over threshold for a fixed discriminator threshold T (in units of ADC counts)
P is the pulse height
P_0 is the reference pulse height that determines the time reference.
f^w(P/T) = c1/(P/T)^c2
Note that I am using the pulse height P instead of area A. If the pulse height is available, we should use it, as we know reflected pulses distort the pulse shape as a function of the position z along the module. Of course, we can always use A~15P, but this is an approximation which could be avoided.
>From Shaun Krueger's thesis, we have
F(P) = a + (b/(P+c))^d.
If we make the following identifications:
a = -(b/P_0+c))^d
c1 = -b^d
c2 = d
With these definitions, we see that the two formulas are the same, except for the following:
1. The function f^w is a function of f^w(P/T), and F is a function of F(P+c) -- that is scale vs offset adjustments. f^w gives the explicit dependence of the time-walk relative to thresholds, so it can easily be applied to different threshold conditions. This function is also only valid for P/T > 1, so the range of validity is clear. The parameter T has a well-defined physical interpretation, which c does not.
2. We also need to define the time reference, i.e. the time when the time-walk correction is zero. The reference chosen for the FADCs is the time at half-height. This corresponds to setting P_0=2T in t^w. It is not so clear to me how to set this offset in F(P).
It is likely that in practice the two forms will yield similar performance, but for the reasons given I think we should use Eqs 5 and 6 from GlueX-doc-2618 as we proceed to determine these constants.
---
This is a plain text email for clients that cannot display HTML. The full logentry can be found online at https://logbooks.jlab.org/entry/3332229
-------------- next part --------------
An HTML attachment was scrubbed...
URL: https://mailman.jlab.org/pipermail/halld-cal/attachments/20150414/2824c863/attachment.html
More information about the Halld-cal
mailing list