[Halld-cal] [Halld-level1] L1 trigger meeting reminder
Mark Ito
marki at jlab.org
Mon Aug 10 12:08:36 EDT 2015
Alex, et al.,
One thing I have not seen in the email discussions is the idea of a mode
where both mode 7 and 8 data come out in the same event. That would
allow verification of the algorithm (and of any emulation, though
perhaps not necessary with this mode available), on a pulse-by-pulse
basis. I would think in such a mode would actually eliminate the need
for pure mode 8. Reading out both should not increase the data volume
over mode 8 by a large percentage and if we are reading out raw
waveforms at all we would not be in a mode where we are trying to push
the event read-out rate so doing the processing for both modes should
not bother us. I have mentioned this to Hai before in a hallway
conversation some weeks ago.
Also, in a previous experiment I was on, we always considered samples
both before the rising edge of the pulse and after the trailing edge (i.
e., before high-going threshold crossing and after low-going threshold
crossing). That ensured capture of the entire pulse, independent of
pulse amplitude. Also it gave more complete information about
close-in-time pulses in a single channel that appear as merged pulses;
not complete, but more complete. It is analogous to having the entire
pulse visible on an oscilloscope screen. Admittedly, the trailing edge
is less well defined since the slope of V vs. t (a negative slope if we
say all pulses are positive going) is much smaller in amplitude, but
since V itself is small, the effect is not so bad.
At a more advanced level one could think of having more complicated
processing of pulses where a double pulse is detected according to some
algorithm. In that case, it might be simpler to pass along the raw
waveform for such pulses and only such pulses, i. e., go ahead and read
them out. Such pulses should be a small fraction for the majority of
channels and thus not create a huge data volume load.
-- Mark
On 08/09/2015 11:24 PM, Alexander Somov wrote:
>
> Hi,
>
> We are going to discuss with Hai Dong the feasibility to
> implement some features in the new fadc250 firmware during
> the meeting tomorrow (Aug 10th). The meeting will be in
> our usual place, A110 at 9:30.
>
> Cheers,
> Alex
>
> P.S. I've updated GlueX-doc-2776, which containes some proposed
> changes.
> _______________________________________________
> Halld-level1 mailing list
> Halld-level1 at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-level1
More information about the Halld-cal
mailing list