[Halld-cal] posted a GlueX doc to compare JLab and UTFSM SiPM measurements

Elton Smith elton at jlab.org
Tue Jan 19 15:52:36 EST 2016


Dear colleagues,

In preparation for writing a NIM article on the BCAL, I have tried to 
summarize the test results of the SiPMs. To my knowledge, we have not 
done this systematically before.
I have posted GlueX-doc-2910, which compares the measurements of SiPMs 
by JLab and UTFSM. See
http://argus.phys.uregina.ca/cgi-bin/private/DocDB/ShowDocument?docid=2910

Major conclusions of this study are
o measurements are generally consistent, but differences between various 
quantities are discussed
o entries of the UTFSM data into the JLab halld_sipm database are tagged 
as being at an over-bias of 1.2 V, but they are consistent with the NIM 
paper data at 0.9 V.
o dark rates reported in the UTFSM NIM paper are internally inconsistent 
and only one seems to be in line with JLab results.

Comments and feedback are welcome.

Cheers, Elton.

-- 
Elton Smith
Jefferson Lab MS 12H3
12000 Jefferson Ave STE 4
Newport News, VA 23606
(757)269-7625
(757)269-6331 fax



More information about the Halld-cal mailing list