[Halld-cal] [Revised Logentry] FCAL HV scan
gasparan at jlab.org
gasparan at jlab.org
Mon Nov 19 12:23:37 EST 2018
Hi Sasha,
a very good test. It looks like we may have serious problems if we need
to set 1800 V range.
We will need to have more information (gains, electronic noise, ...)
before making some decision.
Thanks,
Ashot
>
> Logentry Text:
> --
> FCAL readout: RAW mode (mode 10)
>
> Use violet LED, 500 Hz
>
> 1500 V
> ----------
>
> Run 51604, 130 K events, 17 V
> Run 51605, 122 K events, 22 V
>
> 1550 V
> ---------
>
> Run 51606, 120 K events, 17 V
> Run 51607, 120 K events, 22 V
>
> 1600 V
> ----------
>
> Run 51608, 17 V
> Run 51609, 22 V
>
> 1650 V
> ----------
>
> Run 51610, 17 V
> Run 51611, 22 V
>
> 1700 V
> ----------
>
> Problems with setting HV on channel (22,22 ; -8,-8)
>
> Run 51612, 17 V
> Run 51613, 22 V
>
>
> 1750 V
> ----------
>
> Not able to set this voltage on the following channels:
> Channel, Read back voltage
>
> (1, 31) 1610 V
> (7, 46) 1601
> (13, 23) 1618
> (13, 47) 1651
> (20, 26) 1600
> (22, 22) 1607
> (25, 26) 1620
> (29,7) 1640
> (33,47) 1609
> (35, 37) 1630
> (44, 30) 1590
> (41, 39) 1660
> (44, 30) 1590
> (47, 38) 1625
> (49, 31) 1612
>
> Run 51614, 17 V
> Run 51615, 22 V
>
> 1800 V
> ---------
>
> Cannot set HVs on 41 channels
>
> Run 51616, 17 V
> Run 51617, 22 V
>
>
>
>
>
>
> ---
>
> This is a plain text email for clients that cannot display HTML. The full
> logentry can be found online at
> https://logbooks.jlab.org/entry/3628470_______________________________________________
> Halld-cal mailing list
> Halld-cal at jlab.org
> https://mailman.jlab.org/mailman/listinfo/halld-cal