This missing channel corresponds to crate N1-3-MID slot 4 channel 13. There are zero counts in the readout occupancy and zero counts on the scalers.
I went in to the hall and checked the LED signals on cables arriving at this ADC. They all have good signals. I swapped the cables of channels 13 and 14 and found that on the scalers channel 13 remained the one without counts. This means that the FADC channel is to blame.
---