[Halld-controls] Update for FCAL CAN System talk
Daniel Bennett
dwbennet at jlab.org
Fri Sep 21 14:07:39 EDT 2012
Hi Everyone,
I received some feedback from IU on some of the issues that were brought
up in yesterday's control meeting.
The first issue was the instability of the ADC Cathode reading, how base
responses could vary in the hundreds of volts some percentage of the
time. Paul tells me that this instability is due to the switching noise
of the capacitors in the Cockcroft Walton chain, and is not a physical
effect on the PMT Cathode. So this is something that I will just need
to work around. I have seen this noise on the scope when monitoring the
test pulser, but the capacitors can apparently affect ADC measurements
as well.
The other issue was setting a Trip Current on the firmware. The ADC
Current only monitors Base current, while the PMT will only pull in the
microAmp range. I had shown a distribution of ADC currents over a long
period of time for many bases, and there was a large peak from 2 to 4
mA, and a long tail extending out to 15mA, which the last 15mA bin
peaking due to the ADC maxing out. Paul says that this tail is likely
caused by extra current drawn by the ADC to take the actual ADC Current
measurement, and is also unavoidable.
I'll try to focus on solutions to both of these issues on the Firmware
side. One possibility for ADC Cathode would be to take several
measurements on the firmware during an ADC Cathode request, and if the
RMS is over a certain amount, throw that measurement out and re-measure.
The switching noise is in the nanosecond range, so multiple
measurements would be much slower than that scale. This would only
allow real effects on the ADC Cathode to show up back in Epics, but you
would risk a base not responding indefinitely if the base had an error
which caused every set of ADC Cathode measurements to produce an RMS
which was not below that threshold I just mentioned.
For ADC Current, this same solution would not be as ideal, since the
effect is a consistent high-side tail in the distribution. I will still
look into averaging these Firmware side, but Paul had suggested that
using the Power Supply current monitor, which will average the current
drawn by 100 bases, would be a better indicator for Base current draw.
Either way, the Bases are individually fused, and the entire PSU is
fused, so I do not know how critical a fast threshold current would be.
I'll keep working on all of this, and keep everyone updated.
-Dan
-------------- next part --------------
An HTML attachment was scrubbed...
URL: https://mailman.jlab.org/pipermail/halld-controls/attachments/20120921/27285a38/attachment.html
More information about the Halld-controls
mailing list