[Halld-cpp] TAC detector signal test
    Beni Zihlmann 
    zihlmann at jlab.org
       
    Mon Mar 21 10:16:23 EDT 2022
    
    
  
Hi All,
I had a look at the TAC signal right at the input to the fADC250.
This is located in channel 0, slot 20, D2-3-TOP
HV used is channel 20, module 14 (double slot size), D2-8-TOP. The 
channel in the module is labeled "TAC".
HV = -1700V drawing about 585uA. The trip limit is set at 1000uA.
attached you find 3 pictures of a scope shot for documentation.
The noise level is very low, smaller than 0.5mV. There is a clear but 
very small 50Hz baseline modulation visible
at the level of about 0.4mV peak to peak.->TAC_signal3.png
The signal itself has a rise time of about 3ns and a decay time of about 
8ns (both 10% to 90% and 90% to 10% respectively).
With a trigger threshold at 10mV the rate is very small way below 1Hz. 
->TAC_signal1.png
Higher trigger rates, 10Hz or more (my estimate), is seen with very low 
trigger threshold (less than 1mV). -> TAC_signal2.png
Based on these observation the detector looks operational working quite 
nice. These tests can however not assess the
functionality at high rates.
cheers,
Beni
-------------- next part --------------
A non-text attachment was scrubbed...
Name: TAC_signal1.png
Type: image/png
Size: 1037304 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-cpp/attachments/20220321/12fdffa5/attachment-0003.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: TAC_signal2.png
Type: image/png
Size: 959834 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-cpp/attachments/20220321/12fdffa5/attachment-0004.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: TAC_signal3.png
Type: image/png
Size: 1202212 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-cpp/attachments/20220321/12fdffa5/attachment-0005.png>
    
    
More information about the Halld-cpp
mailing list