Eugene, Magnet Report
* current was measured at 1300 A
* central line matched with scaling of 1500-A map to better than a
few percent
* during ramping to 1500 A, magnet quenched
+ helium dumped
+ rupture disk broken
* magnet designed to stand full quench with all power going into
magnet
+ there is a 20-to-1 ratio of copper to superconductor
* quench started in sub-coil 1 of coil 1 moved initially to coil 4
+ temperature went up by 30 K
+ can measure resistance and infer temperature
* no indication of overheating inside magnet
+ most of energy went to magnet, very little in dump resister
+ about 4 times more resistance in dump resister
* everything went normal, perhaps with exception of one sub-coil
* pressure could cause leaks, but no indications of that so far
* vacuum problems seen, but known from before, there was a plan to
fix anyway
* temperature of He depends on pressure, pressure was too high,
crossed critical current
* coil resistance looks nominal
* magnet has never has a quench before, this was a first
* quenches have been captured before
* If cryo plant shuts down, pressure increases at full current, ramp
down speed may not be fast enough to keep magnet sub-critical
+ perhaps add pressure relief chamber
* current cryo plant is weak
* plans are still developing
* root cause: possibly coil moved
+ on the other hand no voltage spike in record, and resistance
rose in two seconds then took off
* path forward
1. restore cryoplant, clean out carbon
o magnet could be full in two weeks
2. low current test with magnet cold
o requires review and TOSP
3. then warm up, work on cryoplant, prepare for full power test
* definition of full current may have to be adjusted
* quench detector needs to be changed to allow lower threshold
* more interlocks on pressure system will have to be installed
* Claus assumes that no re-baselining will be approved until magnet
fully tested
* magnet now at 30 K, perhaps 40 K, will be kept below 80 K
* physics and 12 GeV will jointly conduct over post-incident
investigation
* TOSP goals will have to be developed
* there were no personnel safety issues associated with the quench
Bert, Safety
* in conjunction with Lehman review, CDC will have two people doing
electrical safety review tomorrow
+ interviewees have been notified
+ will look at Hall A, experimental review process
o how we bring in electrical equipment from outside Lab
Retrieved from
"https://halldweb1.jlab.org/wiki/index.php/Private:Hall_D_Staff_Meeting,_May_6,_2013"
--
Mark M. Ito, Jefferson Lab, marki at jlab.org, (757)269-5295
More information about the Halld-jlab
mailing list