[Halld-jlab] Equipment Power OFF
Beni Zihlmann
zihlmann at jlab.org
Mon Jun 1 07:54:30 EDT 2015
Hi David,
the F125 can be powered off. In that state the crate consumes about 15
to 20 Amps on the 5 volt
line if the modules are powered (like now) the power goes up to 112 Amps
or more on the 5 volt line
depending on how many modules are in the crate.
The current status is that once the DAQ starts a run the modules are
powered up and never powered down
any more unless you reboot the crate. After reboot the modules come up
in power-off mode.
cheers,
Beni
> Hi David,
>
> Power-saving modes are not available in these modules. Given your
> need to use all of these modules on a weekly basis, there is no
> apparent benefit in turning power off to the VXS and VME crates. There
> is, however, still the possibility of turning off all the LV, bias and
> HV chassis but I have received no feedback on these yet.
>
> Best regards,
> Fernando
>
>
> On 5/29/2015 8:58 AM, David Lawrence wrote:
>> Hi Fernando,
>>
>> I seem to recall that our fADC modules could be put into a low
>> power consumption mode when not in use.
>> Do your numbers assume this? I would like to spend some time each
>> week on the development of the
>> event-tagging and L3 systems. Much of this will require all crates
>> with digitization modules. Even without
>> my testing I can imagine most of these crates will get turned on at
>> least once a week for some test or acquisition
>> of cosmic data. Is it worth us exploring a mechanism to basically put
>> crates into sleep mode either by
>> hand (via script) or when a period of inactivity is detected? I’m
>> just a little worried that nothing will really be
>> off for 2 weeks at a time so your basic concern won’t be addressed.
>>
>> Regards,
>> -David
>>
>> On May 27, 2015, at 2:01 PM, Fernando J Barbosa <barbosa at jlab.org>
>> wrote:
>>
>>> Hi all,
>>>
>>> Based on previous discussions, I would like for you to consider
>>> turning off the electronics (crates, LV and HV chassis) during this
>>> beam off period. Saving energy (money and less CO2 in the
>>> atmosphere) and possibly extended longevity of the electronics are
>>> appropriate to consider. For reference, power consumption is
>>> nominally 1.5kW-hr per chassis or ~ 90kW-hr total for Hall D.
>>>
>>> Although there is reluctance in switching equipment on and off, I
>>> would recommend switching off any equipment that is not in use or
>>> that will not be used for a period of at least 2 weeks. The power
>>> supplies in use have soft start attributes which limit in-rush
>>> currents to safe operating levels if performed in a controlled
>>> manner, i.e. remote or front panel.
>>>
>>> A power outage period is anticipated for Hall D, probably in June
>>> according to Tim, but the actual schedule is not known at this time.
>>> Tim will send an email at the appropriate time.
>>>
>>> Please, send me an estimate of expected usage (schedule would be
>>> grate) for each sub-system and we will start shutting off equipment
>>> in 2 weeks. I would also suggest considering this as a normal part
>>> of hall operations in the future.
>>>
>>> Thanks for your input in advance and best regards,
>>> Fernando
>>> <barbosa.vcf>_______________________________________________
>>> Halld-jlab mailing list
>>> Halld-jlab at jlab.org
>>> https://mailman.jlab.org/mailman/listinfo/halld-jlab
>
>
>
> _______________________________________________