    Hi Fernando,<br>
    <br>
    Thanks for pushing these procurements through. I am looking
    particularly at the Bcal and Fcal requirements, but some of my
    comments are more general:<br>
    <br>
    1. Belden 9907 is listed in the Belden catalog as RG58/U type (not
    RG58A/U).<br>
    2. Labels: At the time cables are ordered, we should also have them
    labeled. It is much easier to have quality labels attached at the
    time they are fabricated because the connector is not yet on, and
    this allows for sturdy labels to be placed without going over the
    connector. I believe you already have labeling scheme which has been
    reviewed and could be used for all cables, although it might be
    useful to check the labels one last time. It is also useful to
    include the length of the cable on the label.<br>
    2a. Labels for the Fcal cables should be coordinated with the Fcal
    cables and labeling which already exists at IU. <br>
    3. Discussion on spares: We should consider the strategy here, and
    comments/suggestions are welcome.&nbsp; In the final installation we will
    want all cables to be correctly labeled. Therefore, at the very
    least, when a "spare" is installed its label will need to be updated
    to match its function. We also have several lengths of cables of
    RG-58, although the RG-59 cables are all 55'. It might make sense
    for the RG-58 cables to simply order uncut cable,&nbsp; connectors and
    labels that can be installed as required. That allows the additional
    cable to be used for any need that arises (including ones/twos
    connections that are not on the list). For the RG-59 one could order
    spares of the 55' length, but uncut rolls again provides
    flexibility.&nbsp; <br>
    4. RG-59 type cable. Belden 8281F does not have a fire rating in my
    catalog (is it CL2?) This particular cable also has a fairly large
    diameter (0.305"), so the SHV connector will of course need to match
    it.<br>
    5. Quantities of Bcal and Fcal cables are correct. <br>
    5a. There are no HV cables for Bcal, but what about the required LV
    cables? Will they be purchased from another list? (We do not want
    them to fall between the cracks).<br>
    6. I would like Chuck (or designate) to review all the dimensions in
    the tables once again and check that nothing has changed. For
    example, it may be that the detailed design of the Bcal readout may
    have some effect on the actual lengths. (The lengths were of course
    determined long before we have detailed the geometry of the Bcal
    readout). There may have changes to the geometry for other systems
    as well.<br>
    <br>
    Cheers, Elton.<br>
    <pre class="moz-signature" cols="72">Elton Smith
Jefferson Lab MS 12H5
12000 Jefferson Ave
Suite #16
Newport News, VA 23606
(757) 269-7625
(757) 269-6331 fax </pre>
    <br>
    On 5/23/11 11:04 PM, Fernando J. Barbosa wrote:
    <blockquote cite="mid:4DDB204F.9030207@jlab.org" type="cite">Dear
      colleagues,
      <br>
      <br>
      We have already reviewed the cable specs over a year ago. As I am
      starting procurement of all these cables, I find it opportune to
      ask you to perform a final check.
      <br>
      <br>
      I have attached a file that shows the cable lengths (actual),
      connectors at each end and quantities for each detector subsystem.
      Please let me know if there are any discrepancies before I submit
      the PR by the end of this week.
      <br>
      <br>
      The procurements for the items in green are complete and the
      cables for the collimator are not defined at this time and will be
      procured at a later date.
      <br>
      <br>
      Thanks in advance and best regards,
      <br>
      Fernando
      <br>
      <pre wrap="">
<fieldset class="mimeAttachmentHeader"></fieldset>
_______________________________________________
Halld-jlab mailing list
<a class="moz-txt-link-abbreviated" href="mailto:Halld-jlab@jlab.org">Halld-jlab@jlab.org</a>
<a class="moz-txt-link-freetext" href="https://mailman.jlab.org/mailman/listinfo/halld-jlab">https://mailman.jlab.org/mailman/listinfo/halld-jlab</a></pre>
    </blockquote>
  </body>
</html>