Performance testing of the FDC F1 TDC crates - 100 kHz is the maximum expected L1 trigger rate for the luminosity corresponding to 5x10^7 gamma/sec (average flux 
of the approved experiment). It would require significant efforts to run at higher flux because of the detector  (TOF, 
presumably FDC) and reconstruction limitations.

--

Logentry Text:
--
<ul><li>Note that the specification for high rate running is 200 kHz. (100 kHz is the expected average rate over the run).</li></ul>

---