Hi,
please find the transmittance plots updated using the new dark current, for all cookies, here:
https://www.dropbox.com/s/hddfavyk5xiphs7/cookies_transmission_slides_newDC1.pdf?dl=0
The summary of the measurements is here:
https://gluexdirc.mit.edu/transmission-measurements
Conclusions in short are:
- dark current taken as measurement at 200 nm provides consistent results between two days, also the transmittance does not drop for wavelengths < 350 nm as it used to
- transmittance for quartz windows is ~93% and flat, which is consistent with the Fresnel limit for this case (93%)
- FCAL cookie is very good
- Eljen cookies have lower transmittance than ours for wavelength < 400 nm
- our 2 mm thick cookies ensure relative transmittance > 97% (consistent with Belle II data) with both RTV615 and TSE3032 materials
- transmittance is consistent with data from manufacturers, Belle II data, and JLab data [1407.3258v1]
- cleaning/storing procedures should be refined, they affect measurements at ~ 4-5% level
Best regards,
Maria
On Dec 15, 2016, at 12:44 PM, Justin Stevens <jrsteven at jlab.org<mailto:jrsteven at jlab.org>> wrote:
Hi All,
We'll have a DIRC meeting tomorrow at 11 am. You can find the agenda at https://halldweb.jlab.org/wiki/index.php/December_16,_2016_PID_Upgrade.
-Justin
_______________________________________________
Halld-pid-upgrade mailing list
Halld-pid-upgrade at jlab.org<mailto:Halld-pid-upgrade at jlab.org>
https://mailman.jlab.org/mailman/listinfo/halld-pid-upgrade
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <https://mailman.jlab.org/pipermail/halld-pid-upgrade/attachments/20161216/cf527a41/attachment.html>
More information about the Halld-pid-upgrade
mailing list