[Halld-pid] Hamamatsu PMT test

Beni Zihlmann zihlmann at jlab.org
Fri Feb 4 17:44:25 EST 2011


Hi All,
I had a first look at the Hamamatsu PMT 10534 looking at the
single photo electron peak to determine the gain:
Attached is a file showing the response at 1900V.
I used a x10 amplifier to get large enough signals into the ADC.
The resolution of the ADC is 0.1pC / channel
I tested 3 high voltages. Here the results. I have not estimated
the uncertainty but it is for sure better than 10%

    HV           Gain
----------------------
  1800          2.83 *10^6
  1900          4.12 *10^6
  2000          5.86 *10^6

have a nice week end
Beni
-------------- next part --------------
A non-text attachment was scrubbed...
Name: hamamatsu_1900volts.pdf
Type: application/pdf
Size: 16334 bytes
Desc: not available
URL: <https://mailman.jlab.org/pipermail/halld-pid/attachments/20110204/f05f1c27/attachment-0002.pdf>


More information about the Halld-pid mailing list